Re: AMS2750 rev D - para 3.2.6.2 - "Signature of the calibration company representati
Hey guys, Someone have procedure explaining how to do a calibration and uniformity according to AMS2750, for relieve stress furnace (max 500 ?C)?
As Mikishots said, more details would be nice: range of the furnace, for what specifications the stress is done.
Would explain in Fahrenheit (cause it's easier to remember the numbers)
Say range is: 200C to 500C (392 F to 932 F)
Difference is less than 600F so you need to do any intermediate point at every survey and at least once a year the minimum and the maximum.
However if your parts are for Boeing and they are calling their own Pyrometry standard, the intervals between TUS points shall be no greater than 300 F in which case you should split your range in subranges no bigger than 300 F.
So, you would have one range from 392 to 600 (say) and one range from 600 to 932.
And now you shall cover each range at every TUS and at least once a year the maximum and the minimum.
That means you have 392-600-900. Assuming monthly TUS, in order to satisfy the requirements you shall perform TUS as follows:
January: 392 - 600 (you cover both ranges)
February: 600 - 932 (you cover both ranges)
After that you alternate every month. This way you cover both ranges every month and you also cover the minimum and maximum requirement.