Definition Calibration vs. Verification of Calibration vs. Verification - Definitions
harry
Super Moderator
#31
Re: Calibration vs. Verification - Definitions
Yes, those were my interpretations based on reading in between the lines and my understanding of the subject.
In other words, you are looking for a specific document with all these words explained (when you talk about normative reference) so that others do not need to guess what it means. I am not sure if such a document exists?
What about definitions by and found in 'NIST' documents? Do you or can you accept them as 'the' interpretation?
Elsmar Forum Sponsor
Stijloor
Staff member
Super Moderator
#33
Hi Marc,
Thank you for the links to similar threads.
Proof that history repeats itself. For an excellent reason though...;)
But I'm concerned that some Covers get too carried away with semantics..:frust::frust:
Stijloor.
D
DsqrdDGD909
#34
Marc's posted reminded me of my Dad..."If I have to come upstairs one more time to quiet you kids down, I'll..."
Thanks Marc.
Jim Wynne
Staff member
Admin
#35
I think we could benefit from a review of the question at hand, and some explanation on my part. I know that many here probably think I'm being nit-picky and pedantic, but...
Here is the original statement from AndyN, which has now been split from this thread (thanks, Brad :agree1:):
... if anyone is making decisions about product or process 'acceptability' (does it meet a specification), then the equipment used should be calibrated or at least verified as operating correctly.
My response was
"Calibrated" and "verified as operating correctly" are the same thing.
ISO 9001-2000 (7.6)says that measurement devices used in making decisions must be
calibrated or verified at specified intervals, or prior to use, against measurement standards traceable to international or national measurement standards...
Note that (A) there is no explanation of the difference between "calibrated" and "verified" in the standard (or any reference I know of) and (B) in either case, it says that standards must be traceable.
Thence opened the floodgates. :mad::D
This all has to do with definition of terms and normative references; both ideas are fundamentals in our profession. Ironically though, they seem to be among the most widely misunderstood or overlooked concepts. It boils down to this: if you use terms in your documentation (and this applies to everyone, including ISO) that may be subject to more than one definition, you must provide a definition that becomes the "court of last resort." That's what a "normative reference" is. If I provide a definition of a term in my quality manual, and the manual is duly approved, everyone else in the company is enjoined from creating their own definitions or interpretations regardless of any commonly-accepted denotations. Dictionary denotations should be deviated from only when absolutely necessary, but when it's important that everyone singin' from the same hymnal, you do what you have to do to maintain clarity and order.
In the present case, insofar as the ISO-9001-2000 is concerned, "calibrated" and "verified" both unambiguously mean that applicable devices must be compared to a traceable standard. Anyone who disagrees with this is reading something that ain't there. Furthermore, ISO provides no normative reference which would lead us to believe that there is a substantive difference between calibrating a device against a traceable standard and verifying a device against a traceable standard. Furthermore, we know (or at least I hope we know) that calibration is nothing more than the act of comparing something (anything, not just a measurement device) to a standard. ISO-9001 expands the definition slightly to require traceability of the standard(s) used, but that expansion also applies to "verified."
Now--if someone can supply a traceable :cool: normative reference that contradicts what I've said, I will gladly concede the matter and feel better for having learned something. When I say a traceable normative standard, I mean one for which an unbroken trail exists between ISO-9001-2000 and the definitions. I'm not looking for someone's personal interpretation or opinion; this is a matter of black-and-white.
Finally, a word about why it's worth making an issue over something like this. This is a widely-referenced forum, and beyond terrific group of people we've come to know and respect, many people who never register and never post look to it for information. For all of us who post regularly, we need to do our best to make sure that the most accurate information possible is presented. Errors need to be corrected when they pop up in order to maintain the integrity of the system. There is nothing personal in this. I've made mistakes myself; we all do from time to time. But we owe it to everyone who comes here to do our best to make sure we provide good advice. :D
harry
Super Moderator
#36
Thanks for your post, Jim. We (or at least I) learned not to assume things.
Perhaps you would like to look at this link for a draft copy of 'International vocabulary of basic and general terms in metrology (VIM)'. Definitions of calibration, verification and validation (as in ISO 9001) are given therein (page 20-22). I do not think there are much changes in the final copy.
D
DsqrdDGD909
#37
From Harry's Link above:
(From 2.27 verification)
"NOTE
Verification should not be confused with calibration of a measuring system, or vice versa." :confused:
Ok - how about real world examples:
1. I use the services of a 17025 lab to calibrate my analytical scales using NIST traceable weights.
Q. This can be referred to as either calibration or verification?
2. I buy a set of NIST traceable weights to "check" my scales in between calibrations/verifications from #1 above.
Q. What is this called?
3. I buy/use a set of weights that are not traceable to NIST for use as an operator check in between whatever we call #2 above.
Jim Wynne
Staff member
Admin
#38
There are no normative definitions given in this link, and its definition of "verification" seems to indicate that calibration is a form of verification.
This one is mostly a discussion of the meaning of "calibration," and whether or not "calibration" necessarily includes adjustment. I think we've agreed that it doesn't.
I didn't read the whole thread, but there doesn't seem to be anything that authoritatively addresses the question at hand. There's some interesting conversation, but no real meat. At one point, one person provides yet another unfounded definition for "verification": Verification is the act of using the data from calibration (or in-service check) to determine suitability for a task.
This conversation is mainly about validation vs. calibration.
More interesting discussion, with one Cover providing his own definition of verification, without normative reference.
This thread generally treats calibration and verification as synonymous.
I don't think the definition of "calibration" is in question.
The question here should be qualified to be:
Calibration vs. Verification of Calibration.
All due respect, but the question is, "For the purposes of ISO-9001-2000, is there a substantive, defined difference between "calibrated" and "verified?"
Jim Wynne
Staff member
Admin
#39
Thanks for the link, Harry. It seems to be a step in the right direction, but the definition given for "verification" which is
confirmation through examination of a given item and provision of objective evidence that it fulfils specified requirements
seems to aptly describe the calibration process, but then we're warned in a note that...
EXAMPLES
a) Demonstration that a given reference material as claimed is homogeneous down to samples having a mass of 10 mg for the quantity and measurement procedure concerned.
b) Demonstration that stated performance properties of a measuring system are achieved.
The examples given aren't particulary helpful; what they seem to be saying is that calibration is a form of verification, and vice-versa, but don't call calibration "verification," and vice-versa. :confused:
harry
Super Moderator
#40
The following is a word for word reproduction of the definition from VIM
2.22 Calibration – definition (a) – { definition (b) for IEC related matters}
Operation establishing the relation between quantity values provided by measurement standards and the corresponding indications of a measuring system, carried out under specified conditions and including evaluation of measurement uncertainty.
Notes 1: the relations referred to in definitions (a) and (b) can be expressed by calibration diagrams, calibration function, or calibration tables.
2.27 Verification
Confirmation through examination of a given item and provision of objective evidence that it fulfils specified requirements.
Note: verification should not be confused with calibration of a measuring system, or vice versa.
To me, the obvious difference is that calibration need to be carried out under specified conditions and includes evaluation of measurement uncertainty.
Uncertainty is in turn an essential part of traceability. Without the measurement uncertainty part, the traceability chain is broken (even if all the other essential components that contribute to traceability remain the same).
Can we now safely conclude that the key difference between calibration and verification is the need for determination of measurement uncertainty.
Thread starter Similar threads Forum Replies Date
S Calibration/Verification of customer fixtures IATF 16949 - Automotive Quality Systems Standard 6
Q Calibration verification records 7.1.5.2.1 IATF 16949 - Automotive Quality Systems Standard 2
B AS9100D 7.1.5.2 Calibration or Verification Method using outside cal lab AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
T Calibration or Verification -> Cm and Cmk, etc. Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 3
M Calibration or Verification? What terminology to use ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
V Calibration certificate verification Qualification and Validation (including 21 CFR Part 11) 3
G Assigning a calibration tolerance - An x-y coordinate machine - Uncertainty as my verification tolerance General Measurement Device and Calibration Topics 4
J Calibration/Verification Records (IATF 16949 7.1.5.2.1) IATF 16949 - Automotive Quality Systems Standard 5
D How to set equipment calibration/verification criteria General Measurement Device and Calibration Topics 1
E IATF 16949 Cl. 7.1.5.2.1 - Calibration and Verification Records Requirements IATF 16949 - Automotive Quality Systems Standard 1
S Calibration and Verification requirement for foreign EASA Part 145 EASA and JAA Aviation Standards and Requirements 7
J IATF 16949 Cl. 7.1.5.2.1 - Gauges (Measuring Jigs) Calibration/Verification Records IATF 16949 - Automotive Quality Systems Standard 6
M Clarification on Calibration/Verification Records 7.1.5.2.1d (IATF 16949) General Measurement Device and Calibration Topics 11
B Calibration/Verification Records - IATF 16949 section 7.1.5.2.1 f) IATF 16949 - Automotive Quality Systems Standard 6
P IATF 16949 Cl. 7.1.5.2.1 "Calibration/Verification Records" Interpretation IATF 16949 - Automotive Quality Systems Standard 12
J Pneumatic & Electrical Torque Tool - Calibration/Verification of 'Power Tools' General Measurement Device and Calibration Topics 2
Q Using Caliper A to verify Caliper B - Verification of Calibration ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
K Verification instead of calibration?(ISO 9001:2008) Calibration Frequency (Interval) 10
K "Verification of Calibration" - CMM arm General Measurement Device and Calibration Topics 6
I Verification or calibration Fluke 8508a with 5720a? TUR not 4:1? Measurement Uncertainty (MU) 1
Q Number of Calibration Verification Points - Calipers, height gages, etc General Measurement Device and Calibration Topics 6
A Calibration/Verification of thermo-hygrometer in storage areas ISO 13485:2016 - Medical Device Quality Management Systems 9
G Is this Calibration or Verification ? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 23
D How Working Standard Calibration result use in Verification of Equipment General Measurement Device and Calibration Topics 7
M Calibration Certification Qualification vs. Verification by Uncertified Staff General Measurement Device and Calibration Topics 6
A Torque Wrench Calibration/Verification General Measurement Device and Calibration Topics 24
G BRC (British Retail Consortium) - Calibration vs. Verification Food Safety - ISO 22000, HACCP (21 CFR 120) 4
G TS-16949 7.6.2 ? Calibration/Verification Records - Assessment of the Impact IATF 16949 - Automotive Quality Systems Standard 5
J Balance Calibration Verification question General Measurement Device and Calibration Topics 9
J Log Book for Analytical Balance Calibration Verification General Measurement Device and Calibration Topics 4
Uriel Alejandro How to Record Evidence for Verification/Calibration of Equipment AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
M Establishing Ranges for Weights for Calibration Verification of a Micro-Balance Calibration Frequency (Interval) 1
H Analytical Balance Calibration and Verification General Measurement Device and Calibration Topics 9
D Precision Balance Verification and Calibration Process General Measurement Device and Calibration Topics 3
D Equipment Verification and Calibration Report Standard General Measurement Device and Calibration Topics 9
P Record Retention requirements of Calibration and Verification Service Providers ISO 13485:2016 - Medical Device Quality Management Systems 11
K Calibration and Verification Records - TS 16949 Clause 7.6.2 General Measurement Device and Calibration Topics 8
P Is it compulsory to send new measuring equipment for calibration / verification? Calibration Frequency (Interval) 3
D Calibration vs. Verification and Measurement Device Labeling General Measurement Device and Calibration Topics 10
M Basis used for Calibration or Verification of Measuring Equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
D For Reference Only or Calibration/Verification - Tape Measures for cutting bar stock ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 26
M Methods for Verification/Calibration of Video (Optical) Extensometer General Measurement Device and Calibration Topics 2
C Updating our Laboratory Calibration / Verification Procedures IATF 16949 - Automotive Quality Systems Standard 5
R Verification vs. Calibration and what to do with the results of each General Measurement Device and Calibration Topics 3
B Difference between Calibration & Verification and How to calibrate Go/No Go gauge? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
S Audit and Calibration question - does verification alone meet the requirements? General Auditing Discussions 5
D Calibration vs. Verification - Can we calibrate dial indicators in-house? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
P Internal Calibration or Verification - Personnel Requirements IATF 16949 - Automotive Quality Systems Standard 3
S Verification of calibration certificate Miscellaneous Environmental Standards and EMS Related Discussions 1
R Calibration (Verification) of RF cables General Measurement Device and Calibration Topics 2
Similar threads
Top Bottom