Caliper Calibration Instruction - Example Caliper Calibration Procedure Wanted
janah214
I do not have much experience with internal calibration. Could someone please send me a sample of an instruction for caliper calibration that includes tolerance?
Hershal
Metrologist-Auditor
Welcome to the cove.
I posted a reply to your question in the ISO9001 list, and repeated here. Hope it helps.
The best work I can recommend is The Metrology Handbook, published by ASQ Press. It is for Metrology (aka calibration, the science of measurement) what the Juran reference work is for quality.
As for documentation, you need to keep the information on the gage blocks you use, along with the certificate number from the gage block's last calibration, and next due date. That provides the traceability information above your organization. You should become familiar with measurement uncertainty so that you know the errors involved with your calibration.
As for instructions, if you are going to do much calibration, join GIDEP www.gidep.org I think is the website, and you can access military (e.g. USN's NA 17-20 series) which are considered validated.
Another thought, if you do a lot of internal cal, is to consider seeking accreditation down the road for your internal lab. The accreditation is to ISO/IEC 17025. There are five accrediting bodies in the U.S. that are accepted currently, so talk to all, if that is an option for you.
Also, get a subscription to Cal Lab Magazine, the leading periodical in Metrology. www.callabmag.com
Hope this helps.
Hershal
Marc
The basics:
Both the inside jaws and the outside jaws need to be calibrated, as well as the depth rod and the step measurement, if these are used. Calipers should be frequently checked for accuracy. They are more susceptible to damage than other tools.
To check for wear in the jaws, do this: clean them and close them. Then hold them up to the light and if they're worn you'll see light shining through the gaps. You can continue to use the calipers if you measure with the unworn surfaces. For total reliability, however, you'll have to send the calipers for servicing. The surfaces can be ground flat again.
For the outside jaws it's a simple matter of inserting a series of gage blocks between them and recording the caliper readings. They must not deviate by more than one graduation (.001") over the first 4" of range. From 4" to 8" the error may be .0015" (one and one-half graduation). From 8" to 12" the error can be .002". Accuracy may vary among different models and the manufacturer's specs should be consulted for this information. Take readings at 1-inch intervals. Three gage blocks (see below) of 1", 2" and 3" sizes will be all you need.
To calibrate the inside jaws you may use a set of ring gages. Do not rely on very small ring gages because the inside jaws can not accurately measure small inside diameters. You may also set a calibrated .0001" micrometer to 1" (and higher, if possible) and then use the inside jaws to measure this distance. Since the micrometer has a discrimination ten times that of the calipers, you'll get an accurate reading.
Repeatability means the dial hand returns to the same position on different attempts to measure the same gage block. If this fails, then you'll have to have the calipers serviced.
Millerman
Hello Hershall, what are the implications of Joining "gidep" the concept is alluring, but worried about adversely implicating my organization. What has your experience(s) been?
Al Rosen
You can only join if you are involved with US Gov contracts and is usually required to join and submit applicable data.
DietCokeofEvil
There is an ASME standard for calipers- ASME B89.1.14-2018 as well. If you need a procedure, I'd be happy to send you ours if you need it. Message me.
Did they change this requirement? I have used and accessed GIDEP when in the automotive industry and all that they wanted was an idea of the cost savings provided by using their documentation. So figure the cost of finding a qualified (lab) person to write a cal procedure etc. Easily, $300 - which is used to justify to the bureaucrats why GIDEP exists to help American Manufacturing.
Al Rosen
No, the requirements haven't changed. I would think the government gets it's vehicles from one of the US auto manufacturers (are there any left?) so that may be how you qualified.
Joining GIDEP: Membership Requirements
Participation Guidelines
Any organization that meets one of the following criteria may become a GIDEP member:
  • U.S. or Canadian industrial organization who supplies items or services (directly or indirectly) to the U.S. Government or to the Canadian Department of National Defence
  • U.S. Government department, agency, or activity
  • Canadian Department of National Defence or Canadian Space Agency
C Micrometer and Vernier Caliper simple calibration process General Measurement Device and Calibration Topics 8
Q Using Caliper A to verify Caliper B - Verification of Calibration ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
L Caliper Calibration Standards - One for CNC and one for Sheet Metal? General Measurement Device and Calibration Topics 9
E Caliper Calibration on a Shoestring Calibration and Metrology Software and Hardware 11
Q Engineering purchased a new Caliper - 7.6 - Initial Calibration Required? General Measurement Device and Calibration Topics 6
B How do I calculate Correction to Caliper Calibration Data General Measurement Device and Calibration Topics 4
D Calibration of Pin Gages using a Calibrated Caliper that measures to 4 decimal places General Measurement Device and Calibration Topics 15
H Can you tell me a suitable "tooling" for Caliper Calibration - ISO 17025 ISO 17025 related Discussions 16
E Possible escapes due to caliper out of calibration Calibration Frequency (Interval) 4
J Digital Caliper Calibration vs. Verification - Full Scale vs. Used Range General Measurement Device and Calibration Topics 8
B Electric 0-8" Caliper Calibration Tolerance Question. General Measurement Device and Calibration Topics 9
O Mitutoyo Digital Caliper to PC USB question General Measurement Device and Calibration Topics 2
DietCokeofEvil What is the general consensus on Caliper tolerances? General Measurement Device and Calibration Topics 1
G How to compute or estimate a micrometer or caliper flatness using optical flat General Measurement Device and Calibration Topics 2
G Procedure Creation specifically for dimensional (caliper, micrometer, dti) Document Control Systems, Procedures, Forms and Templates 8
G Uncertainty Budget Examples for Caliper, Micrometer and Dial Gauge Measurement Uncertainty (MU) 3
G Caliper or Micrometer's Dimensional Flatness General Measurement Device and Calibration Topics 5
G Manual Procedure with Uncertainty Budgeting for either a Caliper or Micrometer General Measurement Device and Calibration Topics 6
T GR&R Data from Caliper Measurements APQP and PPAP 3
N Digital vs Dial Caliper Comparison Measurement Uncertainty (MU) 17
C Uncertainty measurements for a 6 inch caliper Measurement Uncertainty (MU) 11
P Strange Gage R&R results for a Caliper (Tolerance Method) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
B Please review my Vernier Caliper Gauge R&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 8
F What are the Advantages of a Digital Caliper Compared with a Dial Caliper? General Measurement Device and Calibration Topics 8
4 Federal Specification GGG-C-105C - Caliper, Micrometer, Gage, Depth Micrometer General Measurement Device and Calibration Topics 7
M Do we need a Gage R&R on Caliper used for Product Audits? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
B Uncertainty in Length Measurement using a Caliper Measurement Uncertainty (MU) 3
T Vernier Caliper Accuracy Issue - Inside measurements were slightly off General Measurement Device and Calibration Topics 7
J Advice for a Length Gage other than Caliper for 52"+ Aluminum Tube Manufacturing and Related Processes 13
J Easy Method to Test a Caliper - Good or bad? General Measurement Device and Calibration Topics 10
Jim Wynne Great Moments In Metrology History--Underwear Caliper Coffee Break and Water Cooler Discussions 6
A Usage of Calibrated Equipment - Vernier Caliper Misc. Quality Assurance and Business Systems Related Topics 10
B Must a vernier caliper and a rule be calibrated according to ISO 9001? General Measurement Device and Calibration Topics 50
apestate Etalon Dial Caliper Gage Information Needed General Measurement Device and Calibration Topics 2
P How to Interpret Caliper GR&R (Gage R&R) Graphical Result Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
L Best Caliper for Shop Use? Digital vs. Dial Caliper vs. Vernier-type General Measurement Device and Calibration Topics 53
J Dial Caliper or Digital Caliper - Which is the best? General Measurement Device and Calibration Topics 27
M Slide caliper gage R&R - What range and kind of samples Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
J Caliper repair procedure - I need a caliper repair procedure for dial calipers General Measurement Device and Calibration Topics 2
A How to Calibrate a Vernier Caliper - Detailed Procedure General Measurement Device and Calibration Topics 25
D Help - FDA Audit - Caliper Calibrated .750 Use - Calibrated @ 0 & 1" General Measurement Device and Calibration Topics 7
H Five Foot Caliper vs. Tape Measure - Tolerance issues General Measurement Device and Calibration Topics 5
D No In-house laboratory - Does this mean I have to send out simple things like caliper General Measurement Device and Calibration Topics 8
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 0
N IATF Calibration Lean in Manufacturing and Service Industries 3
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
E Calibration Records needed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
D Limited Range Calibration - 5000 lb Industrial floor scale General Measurement Device and Calibration Topics 3