I work for a company that has a small CNC Dept and a large sheet metal Dept. I maintain the calibration for the entire shop. Lately I have failed several digital calipers for not being within .002 of an inch. This has caused problems with the sheetmetal side. I was told by their "Engineer" that I was being unreasonable. That there has to be two standards, one for CNC and one for Sheet metal. I'm at a loss.
Is there a double standard? Can someone shed some light. I was taught that calipers were calipers no mater what you are inspecting. Please help!:confused:
 
Elsmar Forum Sponsor
N

natecsss

#2
Re: Calibration standards

How did you come to the .002" standard? Are you calibrating to a standard (NIST, ANSI) or just using something random?

When calibration is done at our facility, it doesn't matter what department the calipers are used. Either they meet the standard we use or they don't and can not be approved.
 
#3
Re: Calibration standards

Sounds like a typical situation I've seen/heard before.

A caliper IS a caliper. Maybe it's the WRONG item - or maybe the TOLERANCES are the wrong ones for each type of product. Sheet metal isn't going to be held to tight tolerances, except for punched/blanked/sheared dimensions. IS this the REAL reason for the comment? Measuring over a bend with a caliper is fine - but the tol. isn't likely to be held if "close toleranced"
 
L

louc74

#4
Re: Calibration standards

Thanks, that's what I thought. Calibration standard is the same regardless. I will hold my ground and I will give him the bill when the calibration house fails them.
 
C

cbehrens

#5
I suspect that the part tolerances in the sheet metal side of the business are less restrictive that on the machining side. If that is the case the calibration tolerance may in fact be different from process to process. You really need to review the required accuracy of the gage against the tolerances of the items you are measuring. It may not be the same for both processes!
 
L

louc74

#6
I was instructed to use that tolerance when trained in the Calibration Lab at the last company I was employed. The ASQ's primer also uses that tolerance. I will not work to two standards in the shop. In the course of the day, I cant be sure the tool will only be used for what dept it was calibrated for. I was curious if the there was a standard that used a greater tolerance. I thank all for the input and look forward to more. When I get the owner of the company and this other person in the same room to discuss the issue, I want to be have all the info I can get.
 
#7
I was instructed to use that tolerance when trained in the Calibration Lab at the last company I was employed. The ASQ's primer also uses that tolerance. I will not work to two standards in the shop. In the course of the day, I cant be sure the tool will only be used for what dept it was calibrated for. I was curious if the there was a standard that used a greater tolerance. I thank all for the input and look forward to more. When I get the owner of the company and this other person in the same room to discuss the issue, I want to be have all the info I can get.
You also have to consider the "shape" of the feature being measured, when considering the type of equipment. A caliper is often the "fall back" equipment, but it has limitations, which aren't just accuracy/precision related. That may also help you make the case for getting/using a cmm, for example.
 

normzone

Trusted Information Resource
#8
Re: Calibration standards

Thanks, that's what I thought. Calibration standard is the same regardless. I will hold my ground and I will give him the bill when the calibration house fails them.
Since we left rack-and-gear equipment behind and went digital, the fail mode I see in rough use calipers is usually jaw wear.

But you are correct - the parts derived may have different tolerancing, but the tools remain the same.
 

Jen Kirley

Quality and Auditing Expert
Staff member
Admin
#9
I was instructed to use that tolerance when trained in the Calibration Lab at the last company I was employed. The ASQ's primer also uses that tolerance. I will not work to two standards in the shop. In the course of the day, I cant be sure the tool will only be used for what dept it was calibrated for. I was curious if the there was a standard that used a greater tolerance. I thank all for the input and look forward to more. When I get the owner of the company and this other person in the same room to discuss the issue, I want to be have all the info I can get.
With all due respect to your last employer and the ASQ Primer, they don't apply to instruments that they have not been specifically written to address. The only appropriate outside source for accuracy/tolerance standards are your customers. To my knowledge, not even NIST will specify specific acceptability tolerances.

Ideally the calipers' accuracy will be measurable in increments at least 10 times that of the part and/or material tolerances. That means for tolerances listed in thousandths, your calipers should be able to read in ten-thousandths. That is called the "rule of 10," but when asked to provide the source of that "rule" I was not able to find it. So it is an industry practice that just makes sense to help ensure accuracy.

I hope this helps!
 

BradM

Staff member
Admin
#10
I work for a company that has a small CNC Dept and a large sheet metal Dept. I maintain the calibration for the entire shop. Lately I have failed several digital calipers for not being within .002 of an inch. This has caused problems with the sheetmetal side. I was told by their "Engineer" that I was being unreasonable. That there has to be two standards, one for CNC and one for Sheet metal. I'm at a loss.
Is there a double standard? Can someone shed some light. I was taught that calipers were calipers no mater what you are inspecting. Please help!:confused:
Is there anything wrong with having a broader tolerance for the sheet metal side?

If the instrument satisfies their process requireme, what would be the logic in failing
Them?

Put orange and blue dots on them respectively, proceduralize it, and move on. :)
 
Thread starter Similar threads Forum Replies Date
C Micrometer and Vernier Caliper simple calibration process General Measurement Device and Calibration Topics 8
Q Using Caliper A to verify Caliper B - Verification of Calibration ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
E Caliper Calibration on a Shoestring Calibration and Metrology Software and Hardware 11
Q Engineering purchased a new Caliper - 7.6 - Initial Calibration Required? General Measurement Device and Calibration Topics 6
B How do I calculate Correction to Caliper Calibration Data General Measurement Device and Calibration Topics 4
D Calibration of Pin Gages using a Calibrated Caliper that measures to 4 decimal places General Measurement Device and Calibration Topics 15
H Can you tell me a suitable "tooling" for Caliper Calibration - ISO 17025 ISO 17025 related Discussions 16
E Possible escapes due to caliper out of calibration Calibration Frequency (Interval) 4
J Caliper Calibration Instruction - Example Caliper Calibration Procedure Wanted General Measurement Device and Calibration Topics 22
J Digital Caliper Calibration vs. Verification - Full Scale vs. Used Range General Measurement Device and Calibration Topics 8
B Electric 0-8" Caliper Calibration Tolerance Question. General Measurement Device and Calibration Topics 9
O Mitutoyo Digital Caliper to PC USB question General Measurement Device and Calibration Topics 2
DietCokeofEvil What is the general consensus on Caliper tolerances? General Measurement Device and Calibration Topics 1
G How to compute or estimate a micrometer or caliper flatness using optical flat General Measurement Device and Calibration Topics 2
G Procedure Creation specifically for dimensional (caliper, micrometer, dti) Document Control Systems, Procedures, Forms and Templates 8
G Uncertainty Budget Examples for Caliper, Micrometer and Dial Gauge Measurement Uncertainty (MU) 3
G Caliper or Micrometer's Dimensional Flatness General Measurement Device and Calibration Topics 5
G Manual Procedure with Uncertainty Budgeting for either a Caliper or Micrometer General Measurement Device and Calibration Topics 6
T GR&R Data from Caliper Measurements APQP and PPAP 3
N Digital vs Dial Caliper Comparison Measurement Uncertainty (MU) 17
C Uncertainty measurements for a 6 inch caliper Measurement Uncertainty (MU) 11
P Strange Gage R&R results for a Caliper (Tolerance Method) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
B Please review my Vernier Caliper Gauge R&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 8
F What are the Advantages of a Digital Caliper Compared with a Dial Caliper? General Measurement Device and Calibration Topics 8
4 Federal Specification GGG-C-105C - Caliper, Micrometer, Gage, Depth Micrometer General Measurement Device and Calibration Topics 7
M Do we need a Gage R&R on Caliper used for Product Audits? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
B Uncertainty in Length Measurement using a Caliper Measurement Uncertainty (MU) 3
T Vernier Caliper Accuracy Issue - Inside measurements were slightly off General Measurement Device and Calibration Topics 7
J Advice for a Length Gage other than Caliper for 52"+ Aluminum Tube Manufacturing and Related Processes 13
J Easy Method to Test a Caliper - Good or bad? General Measurement Device and Calibration Topics 10
Jim Wynne Great Moments In Metrology History--Underwear Caliper Coffee Break and Water Cooler Discussions 6
A Usage of Calibrated Equipment - Vernier Caliper Misc. Quality Assurance and Business Systems Related Topics 10
B Must a vernier caliper and a rule be calibrated according to ISO 9001? General Measurement Device and Calibration Topics 50
apestate Etalon Dial Caliper Gage Information Needed General Measurement Device and Calibration Topics 2
P How to Interpret Caliper GR&R (Gage R&R) Graphical Result Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
L Best Caliper for Shop Use? Digital vs. Dial Caliper vs. Vernier-type General Measurement Device and Calibration Topics 53
J Dial Caliper or Digital Caliper - Which is the best? General Measurement Device and Calibration Topics 27
M Slide caliper gage R&R - What range and kind of samples Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
J Caliper repair procedure - I need a caliper repair procedure for dial calipers General Measurement Device and Calibration Topics 2
A How to Calibrate a Vernier Caliper - Detailed Procedure General Measurement Device and Calibration Topics 25
D Help - FDA Audit - Caliper Calibrated .750 Use - Calibrated @ 0 & 1" General Measurement Device and Calibration Topics 7
H Five Foot Caliper vs. Tape Measure - Tolerance issues General Measurement Device and Calibration Topics 5
D No In-house laboratory - Does this mean I have to send out simple things like caliper General Measurement Device and Calibration Topics 8
F ESD workbench "calibration" Manufacturing and Related Processes 2
C Correct Calibration Method for Dial Depth Gage General Measurement Device and Calibration Topics 6
F Nist traceable calibration certificates General Measurement Device and Calibration Topics 1
G Tool tracebility and First calibration requirements for aerospace (AS9100) organisation AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
D Calibration of Digital thermometer with surface probe General Measurement Device and Calibration Topics 1
T Calibration or Verification -> Cm and Cmk, etc. Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 3
Q Quality Technician (calibration / metrology) - Memphis, TN Job Openings, Consulting and Employment Opportunities 1

Similar threads

Top Bottom