Calipers found out of calibration after only 1 year!

#1
Dear all, I am handling a quality problem inditing 85000 parts involving 45 IP's, 45 part numbers.

The calipers failed the yearly calibration cycle. The calibration facility reported that the calipers were off by 0.003 on the I.d and 0.0055 on the O.D

Everything, ever since those calipers were last calibrated (a year back) is suspect. How do you do a risk analysis in this scenario ?

:(
senior members please advice !
 
Elsmar Forum Sponsor
V

Vimal Mahobia

#2
Re: calipers found out of cal's : after 1 year

Dear all, I am handling a quality problem inditing 85000 parts involving 45 IP's, 45 part numbers.

The calipers failed the yearly calibration cycle. The calibration facility reported that the calipers were off by 0.003 on the I.d and 0.0055 on the O.D

Everything, ever since those calipers were last calibrated (a year back) is suspect. How do you do a risk analysis in this scenario ?

:(
senior members please advice !
First and foremost a question, how do you decide on the calibration cycle? When high volume of parts are to be measured, then the calibration interval should commensurate with QUANTITY in stead of TIME. A caliper used by a Quality engineer has lesser usage vis-a-vis the caliper used by quality inspector, and the highest usage is by the patrolling inspection team. All the three calipers must have different calib. cycles. The popular mistake is in keeping all one-type-of-tools in a single calib.cycle.

:caution: The risk is great and IMO all components in question, must be quarantined and put to 100% inspection with a new caliper.
 
Likes: QE
P

prototyper

#4
First, I would go back through the measurement records. Are there any results within 0.003 of the tolerance. This will help to quantify the problem.
Second, assuming the calipers are worn, how would a part being oversize by 0.003 effect the fit and function?
Last, have the worst condition measured parts been used by the customer? Did they report a problem?

This will enable you to determine the level of risk and thereby the appropriate action to take.
 
U

Umang Vidyarthi

#5
Dear all, I am handling a quality problem inditing 85000 parts involving 45 IP's, 45 part numbers.

The calipers failed the yearly calibration cycle. The calibration facility reported that the calipers were off by 0.003 on the I.d and 0.0055 on the O.D

Everything, ever since those calipers were last calibrated (a year back) is suspect. How do you do a risk analysis in this scenario ?

:(
senior members please advice !
If the calipers were used day in and day out in one year duration, then hypothetically the wear on ID was @ 0.003/12 = 0.00025/per month. Similarly on the OD was @ 0.0055/12 = 0.000458/per month. w.r.t. Tolerance of the parts in question you may decide the safe months (starting from last calibration), as well as the border line cases. Beyond that, every thing is suspect. You may have to recall the supplies as well as quarantine the material in your facility.

Secondly, have another look at your calibration frequency, it appears out of sync.

Hope this helps

Umang :D
 

SteelMaiden

Super Moderator
Super Moderator
#6
we verify ours every day against known standards. there is no way we could afford to put a year's worth of product at risk. Wow, sorry to hear aboutthis problem. Good luck to you.
 
Likes: QE

bobdoering

Stop X-bar/R Madness!!
Trusted Information Resource
#7
This is also a good time to determine if standard calipers are the right tool for the job. Sure, Gage R&R will tell you if they function as an adequate measurement tool, but has no time function - calibration does. But, beyond that some knowledge of the gage is necessary. Calipers are handy because they measure many things, and do it fairly conveniently. But, unless you get special calipers, they generally do not have adequate measurement surfaces for high volume or high wear measurement (such as the carbide anvils on a micrometer). Not only that, when they do wear, there is little that can be done for them - whereas an indicator tip can be easily replaced a very low cost.

Although this is a very unfortunate problem (I have lived through them, too), it is a good learning opportunity for all of us. Do not rely on just gage R&R for gage choice!
 
Likes: QE

BradM

Staff member
Admin
#8
What does your procedure say?
This was short and sweet, but makes an excellent point. If you don't have remedial action in your procedures, you need to put it in there.

The question is not if a failure will occur, but when. There are a lot of golden gems in all of these posts on this thread, and need to be considered when going through this entire process. Part of this should be determining how failures will be evaluated and handled. You really should implement Remedial Action for when you have calibration failures.

Risk analysis is important, here, and SteelMaiden's post is one to not be taken flippantly. If your risk is high, then you need to take steps to mitigate that risk.

How different was this wear from the last time they were calibrated? The instrument could have been near tolerance the last time they were checked. Hopefully, your instruments were calibrated by a competent lab, and you purchased data with the calibrations. All calibration documentation needs to be analyzed, and used for effective management decisions.

Let us know how this progresses.
 
Likes: QE

CarolX

Super Moderator
Super Moderator
#9
Dear all, I am handling a quality problem inditing 85000 parts involving 45 IP's, 45 part numbers.

The calipers failed the yearly calibration cycle. The calibration facility reported that the calipers were off by 0.003 on the I.d and 0.0055 on the O.D

Everything, ever since those calipers were last calibrated (a year back) is suspect. How do you do a risk analysis in this scenario ?

:(
senior members please advice !
:topic:
In addition, you may want to look at the manufactures stated accuracy. 12" Mitutoyo dial calipers have a stated accuracy of .002".


As to your risk analysis - I think prototyper covered it pretty well. I really can't add to that advice.
 

BradM

Staff member
Admin
#10
:topic:
In addition, you may want to look at the manufactures stated accuracy. 12" Mitutoyo dial calipers have a stated accuracy of .002".


As to your risk analysis - I think prototyper covered it pretty well. I really can't add to that advice.
Good point, Carol. I didn't even think about going into that. You may need to purchase more accurate instruments, where small departures from the tolerance does not have an effect on the measurement process.

Also, buy a gauge block set. If you cannot afford a set, purchase two or three of them (across the range of use).
 
Thread starter Similar threads Forum Replies Date
S Calibrating our own equipment, can we? Micrometers to calibrate vernier calipers General Measurement Device and Calibration Topics 13
G How many Uncertainty Measurements for Micrometers and Calipers? Measurement Uncertainty (MU) 3
N International / Traceable Standards - Procedures for Calipers, Micrometers, Etc. ISO 17025 related Discussions 2
C ISO 9001:2008 Audit Major Finding - Dial Calipers not set to zero ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 16
J FDA 483 Warning Letter for use of Calipers 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 18
Q Number of Calibration Verification Points - Calipers, height gages, etc General Measurement Device and Calibration Topics 6
J What do you use to Calibrate 40" Calipers General Measurement Device and Calibration Topics 10
V Gauge R&R on Vernier Calipers used to Measure 5 CTQ Features Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
F Measuring Ceramic Tubes & Discs - any suggestions for replacement of calipers? Manufacturing and Related Processes 6
F Adjusting Digital and /or Vernier Calipers General Measurement Device and Calibration Topics 2
E Gage Repair Shops near Pheonix, AZ - Variable Gages - Micrometers, Calipers, etc. Calibration and Metrology Software and Hardware 3
A ID Calipers Accuracy and Pin Gages Manufacturing and Related Processes 13
J Performing Internal Calibrations - Dial gauges, micrometers, height gauges, calipers General Measurement Device and Calibration Topics 8
M Devices with Measuring Function - Do rulers and calipers count? EU Medical Device Regulations 7
C Seeking a holder/jig to calibrate large OD micrometer calipers (up to 23") General Measurement Device and Calibration Topics 8
E Calibration of Calipers, Mics, Height gauges - Recommended Gage Block Types General Measurement Device and Calibration Topics 7
B MSA (Measurement Systems Analysis) with One Operator using a CMM or Common Calipers Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
O Calibrating a Gage Block In-house with Calibrated Digital Calipers General Measurement Device and Calibration Topics 7
C Repair of Helios Calipers General Measurement Device and Calibration Topics 3
N Basic MSA - A simple way to complete an MSA study on Dial Calipers Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
W ASTM Spec or what standard is used to calibrate calipers? General Measurement Device and Calibration Topics 2
R Need of working standard for vernier and micrometer calipers General Measurement Device and Calibration Topics 2
G Should Serial # of all Measurement Instruments (Calipers,etc) be in Control Plans? FMEA and Control Plans 7
B Room temperature for calibrations on basic equipment like calipers, micrometers, etc. General Measurement Device and Calibration Topics 14
F Is there a standard that I should use to Calibrate Calipers and Micrometers? General Measurement Device and Calibration Topics 8
R Validation of A Procedure for Calibrating Dial Calipers using Traceable Gauge Blocks General Measurement Device and Calibration Topics 6
S Calibrated vs. Inspected - Calipers with a Certificate of Inspection AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
L Precision tool storage such as micrometers and calipers? General Measurement Device and Calibration Topics 5
C Does your Company Supply Calipers for Employees? Manufacturing and Related Processes 59
I What is Gauge Capability? Dial Thickness Gage and Digital Vernier Calipers Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 17
ScottK Automating variable inspection - Calipers, micrometers, and optical comparators Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
C Struggling to teach a co-worker how to read dial calipers, any advice? General Measurement Device and Calibration Topics 13
O Calibration tolerances needed - Calipers, digital indicators, feeler gages, etc General Measurement Device and Calibration Topics 4
C Control of Vernier Calipers - Calipers easily switched between operators and lost Quality Assurance and Compliance Software Tools and Solutions 16
CarolX Seeking training material on the use of micrometers and calipers Training - Internal, External, Online and Distance Learning 6
J Caliper repair procedure - I need a caliper repair procedure for dial calipers General Measurement Device and Calibration Topics 2
C What is the proper method for cleaning dial calipers? General Measurement Device and Calibration Topics 4
J In-house calibration - Resource for calibration, adjustment of calipers, indicators General Measurement Device and Calibration Topics 2
L Stability Testing - Calipers used to check lead thickness when setting dies Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
O In-House Laboratory Scope - Calipers, Pin Gages and Digital Indicators General Measurement Device and Calibration Topics 4
M Help DOD inspection - Standard temperature calipers should be calibrated at? General Measurement Device and Calibration Topics 7
apestate Calibrating .001" resolution Dial Calipers - Calibration and safeguarding procedure General Measurement Device and Calibration Topics 6
D Gage R&R Required on Micrometers and Calipers? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
J Dial Calipers vs. Digital Calipers - Has anyone replaced all their dial indicators? General Measurement Device and Calibration Topics 5
M Calibration Procedures for checking and certifying micrometers and calipers General Measurement Device and Calibration Topics 3
T What is the best procedure for calibrating / certifying our Calipers? General Measurement Device and Calibration Topics 8
M Test failure Root cause not found Customer Complaints 10
M Vigilance - incident found in an article EU Medical Device Regulations 3
Y We found out we have been using a equipment without validation for past 4 years Quality Manager and Management Related Issues 6
A Non-Conformances Found After 3rd Party Sorting Supplier Quality Assurance and other Supplier Issues 12

Similar threads

Top Bottom