Can a Powerstrip be used in a Medical Device?
E
eephotoki
According to IEC 60601-1 subclause 57.2e:
Auxiliary mains socket outlets on non-permanently installed equipment, intended for the provision of mains supply to other equipment, or to separted parts of the Equipment shall be of a type that cannot except a MAINS PLUG.
The reason they give is because of "Extra lekage currents caused by other equipment".
Question:
So besides a MAINS PLUG what other plugs can the "Auxiliary mains socket outlets" be is 2 prong ok or a usb powered device. Also does this mean that a Powerstrip which is accepts a three prong plug(L,N,Gnd) can not be used?
Marcelo
The idea is that you must not use another equipment with a "general" plug in this socket (a household electrical appliance, for example), you do have to use some kind of proprietary of different plug than the accepted mains plug. This will guarantee that only equipment designated to be used with the device will be used (please note that are other clause which require tests for the current or define that you do need to test if youo clearly define the connecting device).
For example, in Brazil right now we have a new plug according to a national standard. In the case, any other plug which is not a plug according to this standard can be used.
E
eephotoki
Hi Marcelo,
So if we have a powerstrip that other manufactures can plug there equipment into for example an IV pump which uses a standard three prong (Nema 5-15) this isn't allowed right?
Because our UL inspector passed our equipment which has a standard (Nema 5-15) powerstrip and the rationale was it's only accesible with a tool even if it's only accesible with a tool that doesn't prevent them from plugging in say an IV pump or any other device.
E
eephotoki
I meant to say that the rationle our UL inspector wrote on our report was that the Powerstrip is only accessible with a tool, that's why he passed it.
Marcelo
Tha main problem here is not only the fact that other equipment "can" be connected, but also what is meant to be connected (as can be seen from the leakage current rationale).
What equipment you intend this to be connected with? Did you test your equipment with them (using IEC 60601-1-1)? Are you sure that these intended equipment won´t make your equipment exceed leakage current limits?
From what you said, the rationale for passing the tests does not look like sound engineering (If it was just the "connection problem" i would agree that using a tool would prevent the problem, but being a leakage current problem, if using a tool would allow leakage current limits to be passed, this would not be ok)
(also, as this situation is not detailed in clause 57.2 e), the only accepted option would be using clause 3.4, which permits other types of construction if justified. In that case, i would take a look into the equipment risk management file to verify that indeed this alternative construction was analysed and that the risk remains aceptable - and i think this was not done, nor you could say that the risk would be acceptable :))
M Use of an Internal Powerstrip in a Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P Ethernet switch with powerstrip for conference table After Work and Weekend Discussion Topics 0
A Applicability of Photobiological Safety Evaluation for LED used in medical devices Reliability Analysis - Predictions, Testing and Standards 1
R Tools used for measuring (New to metrology) General Measurement Device and Calibration Topics 9
L Requirements for Cables used for Low-Voltage DC Mains (12-36VDC) IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
N Voltage Doubler fail - can gas tubes be used? IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
N Maximum mains voltage used when applying 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
W How long do you keep information about equipment no longer used? Document Control Systems, Procedures, Forms and Templates 2
Q PPT used as Design Review ISO 13485:2016 - Medical Device Quality Management Systems 3
G Calibration of "Master Parts" Used as Gauges Calibration Frequency (Interval) 5
L Does a backdate form format can be changed if wrong revision is used? Document Control Systems, Procedures, Forms and Templates 8
R Nanomaterials used as coating on class I MD - rule 19 EU Medical Device Regulations 4
E SIP/SOP requirements for USB port used for charging IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A What does this sentence "this symbol shall be used in the orientation shown" mean in ISO 780:2015? Other Medical Device Related Standards 4
D Can PFMEA be used in disposition of NC material? ISO 14971 - Medical Device Risk Management 4
K Software Validation for Measurement Tools used in Process Validation ISO 13485:2016 - Medical Device Quality Management Systems 2
John Broomfield Informational Terms used in quality management [Deleted] Misc. Quality Assurance and Business Systems Related Topics 6
G Devices from IQ, OQ or PQ process to be used for verification, validation and summative? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
Q LOT or Serial Number Symbol not used when the information is contained in the UDI? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
I Approved Suppliers ISO/IEC 17025:2017 and used test equipment ISO 17025 related Discussions 6
R Foam mattresses used in hospitals - compliance with MDR requirements? EU Medical Device Regulations 6
DuncanGibbons Looking for example aerospace part CAD files to be used for a case study Career and Occupation Discussions 2
I Custom software services to be used by medical software ISO 13485:2016 - Medical Device Quality Management Systems 1
A Single use non-sterile syringe used in the oral cavity - Laboratory test advice US Food and Drug Administration (FDA) 7
R PMCF - centers/countries used CE Marking (Conformité Européene) / CB Scheme 0
K EU MDR Rule 11 - Does the 'Risk logic' used in Rule 11 conflict with that used in the other rules? EU Medical Device Regulations 2
G Can calibrated device values be used instead of manufacturers spec? General Measurement Device and Calibration Topics 2
M Informational US FDA – URGENT/11 Cybersecurity Vulnerabilities in a Widely-Used Third-Party Software Component May Introduce Risks During Use of Certain Medical Dev Medical Device and FDA Regulations and Standards News 0
D What is the best software used for the pharma compliance management? Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 0
D Does every piece of equipment used in a laboratory need to have an IQ protocol written and executed? ISO 13485:2016 - Medical Device Quality Management Systems 1
I Recommendation about h and k values when CUSUM charts are used Statistical Analysis Tools, Techniques and SPC 4
R Please share your method used for PEMS Validation IEC 62304 - Medical Device Software Life Cycle Processes 5
M Can Brazil approval be used for other countries in Latin America? Other Medical Device Regulations World-Wide 2
S Do we need to validate Software used in Drug discovery and development process? Qualification and Validation (including 21 CFR Part 11) 2
O Validation approach for a Photostability Chamber (used for fluid therapy and injectable drug products) Qualification and Validation (including 21 CFR Part 11) 1
K Verifying gas tight glass syringe used for making standards General Measurement Device and Calibration Topics 4
J Regs for non-medical product to be used in hospitals - 61326 CE Marking (Conformité Européene) / CB Scheme 1
J Is Design Validation required for fixtures or test systems used to manufacture a product? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
M Informational US FDA final guidance Characterization of Ultrahigh Molecular Weight Polyethylene (UHMWPE) Used in Orthopedic Devices Medical Device and FDA Regulations and Standards News 0
K First Article Inspection (FAI) - Equipment used - Dimensions vs. Go/No-Go ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
Q ISO 3310 Clarification Help - Aperture sizes for sieves used for particle sorting Other ISO and International Standards and European Regulations 2
N What subcommands can be used with "Stop" in Minitab 18? Using Minitab Software 8