Dear all,
Can anyone help me in understanding a situation cited below:
During our 60601-1-2 testing, Surge test failed at 2KV and the System shut down. However by risk assessment it was concluded that as the system shut off (failed safe), Basic safety was not compromised and did not hamper the essential performance, it was decided to consider surge failure as not a failure and still compliant to Surge test requirement.
my query is:
1. Is it acceptable
2. Do we need to inform this to the end user of the system. and if yes how?

Thanks with regards
Dileep Joshi
 
Elsmar Forum Sponsor

Marc

Hunkered Down for the Duration
Staff member
Admin
#2
I personally can't help, but here is a quick "Bump" - My Thanks in advance to anyone who can help with this one.
 
#3
1) This is acceptable if you're not saying that system shall withstand 2kV surge (line to ground?), which is what it sounds like you've concluded based on your risk management.

If the system has been designed to pass 2kV surge and you have that as a traceable requirement, even though Basic Safety wasn't affected Essential Performance would be and I don't think this would be compliant in that case.

2) This should be addressed in your Intended Use documentation.
 

Pads38

Trusted Information Resource
#4
This is rather dependant upon how you have classified your Essential Performance.

The Compliance Criteria, at 6.2.1.10, state that 'resets' and 'cessation of intended operation' shall not be allowed.

But that applies only if they are associated with Essential Performance - you would have to argue that your device has no Essential Performance (not catching fire is considered Basic Safety!).

Regulators and test houses are usually very reluctant to allow the "no Essential Performance" argument - they would suggest that a medical device with no Essential Performance has no medical purpose, so is not a medical device!

If you have some inherent sensitivity to surge you might be able to use the justification set out in 6.2.1.1
Lower IMMUNITY COMPLIANCE LEVELS are allowed provided they are justified based on significant physical, technological or physiological limitations.
Look to the part 2 Particular Standards for guidance on this allowance.

In such a case I would expect to see prominent warnings within the Operators Manual and Risk Management would have to carefully take this into account, along with probabilities that the device would only be connected to 'low-surge' mains supplies (perhaps you, as manufacturer, supervise the installation process and supplies are only ever dedicated supplies fed from local hospital sub-stations).

Or you could add some extra surge protection!

(Clause numbering is from the 2007 version).
 
Last edited:
Thread starter Similar threads Forum Replies Date
Sidney Vianna What do these things have in common? Surge in new ISO Committees & Standards Other ISO and International Standards and European Regulations 7
B Surge Marks in Coextrusion Process Manufacturing and Related Processes 2
quality1 Standard Home Surge Protector Testing Requirements Reliability Analysis - Predictions, Testing and Standards 3
Crimpshrine13 Laboratory Scope - Calibration vs. Test Methods - IATF 16949 IATF 16949 - Automotive Quality Systems Standard 2
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R Accelerated Aging - Creating test samples - Implantable medical device Question Other Medical Device Related Standards 4
E Mentor for Test Method Validation (TMV) Design and Development of Products and Processes 2
L ASQ's Biomedical Auditor Course Test ASQ - American Society for Quality 1
W Regarding Requirement for Hi pot & Leakage Current test Station Set-up IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
R Regulatory Affairs Certification (RAC) prep - Practice test versus actual exam Professional Certifications and Degrees 0
R SAT (System Accuracy Test) temperature according to AMS2750E AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
R MSA for ATE (Automatic Test Equipment Embedded Software) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
D Test summary report example for design validation wanted - ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
S Documenting Design Verification Test Results (ISO 9001) Design and Development of Products and Processes 1
A Earth leakage current test for Permanently installed Large scale MEE IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
A Bench test report translation to English for 510(k) submission Other US Medical Device Regulations 2
J Production Line Test Brasil - High Voltage Dielectric Strenght Test Other Medical Device Regulations World-Wide 5
W Seeking Guidance Verification Test Strategy for Class B Medical Devices IEC 62304 - Medical Device Software Life Cycle Processes 1
J 510(k) for a control kit for an external IVD test kit 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
K Cord anchorage pull test IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Minitab assistant: 1-Sample % Defective Test Using Minitab Software 1
N Is capability applicable for a destructive test? Capability, Accuracy and Stability - Processes, Machines, etc. 9
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K Dielectric strength test as per IEC 60601-1 -Infant incubator IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M Test method validation - Is MSA (MSA1, MSA2, MSA3 and linearity) a good solution? Medical Device and FDA Regulations and Standards News 1
B Operator protection - When to apply table 7, Dielectric strength test voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Medical Device Airworthiness, DO-160 Test Planner IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
J Audit Finding For Not Retaining Test Results ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
W ASTM F1929 dye penetration test - Validation for in-house testing ISO 13485:2016 - Medical Device Quality Management Systems 13
S IVF medical devices standards and test methods Medical Device and FDA Regulations and Standards News 4
I Approved Suppliers ISO/IEC 17025:2017 and used test equipment ISO 17025 related Discussions 6
N Heavy tare to simulate high load in a routine sensitivity test of a floor scale? General Measurement Device and Calibration Topics 3
bryan willemot Looking for a certified test report for aerospace parts AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
A Information about Medical Device Test Laboratories Other Medical Device and Orthopedic Related Topics 4
B Standard for Dielectric Strength Test IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Definition of customer in ITP (Inspection & Test Plan) Contract Review Process 3
J EN61010 Electrical Withstand Test IEC 62304 - Medical Device Software Life Cycle Processes 0
Z MDD ER 7.5, "phthalate test"? Aspiration tips EU Medical Device Regulations 3
G ISO 17025:2017 Uninterrupted chain of traceability for test laboratories ISO 17025 related Discussions 6
A Single use non-sterile syringe used in the oral cavity - Laboratory test advice US Food and Drug Administration (FDA) 7
M Proper method to test flow of a regulator Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
C CQA Test Questions Bank General Auditing Discussions 1
G Is impact test according to IEC 60601-1 applicable for HAND-HELD equipment if also classified as BODY-WORN? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Watchcat Identification of Test Sample in Test Reports? Design and Development of Products and Processes 22
S Safety Class A with no test? IEC 62304 - Medical Device Software Life Cycle Processes 1
A IEC 60601-2-2 - Defib-proof test amendment to base standard IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K The correct way of doing a 2 Sample T test Reliability Analysis - Predictions, Testing and Standards 7
J Good Documentation Practices Self Test? ISO 13485:2016 - Medical Device Quality Management Systems 6
B EU: Alcohol test kits for breast milk IVDD EU Medical Device Regulations 2

Similar threads

Top Bottom