Can enclosure of semiconductor or electrolytic capacitor provides basic insulation?
sankui
Dear all,
Our products mainly employ switching power supply (SPS) to power the equipment. In SPS, there are several semiconductors and electrolytic capacitors. some of these components are close to the transformer and may influence the safety spacings between primary and secondary. However, if the enclosure of semiconductor or electrolytic capacitor (sheet) is regarded as basic insulation, additional sleeve may not be necessary.
Can anyone tell me that whether the enclosure of semiconductor or electrolytic capacitor (sheet) provides basic insulation?
Thanks and regards.
Stijloor
Re: Can enclosure of semiconductor or electrolytic capacitor provides basic insulatio
Dear all,
Thanks and regards.
Any electronics folks that can help?
Thank you!!
Stijloor.
Al Rosen
Dear all,
Thanks and regards.
That's what basic insulation is. Supplementary insulation would be the additional sleeve and two sleeves would be reinforced insulation.
Peter Selvey
Generally this is not a good idea.
Components which provide safety insulation (with respect to electric shock) should be specifically designed for that purpose, and with the design either directly controlled by yourself (such as a custom transformer design) or a recognised component (e.g. capacitor approved by a 3rd party agency).
The main reason for this is production control: your obligation covers all products not just the sample that underwent IEC 60601-1 testing. If you are using a non-safety recognised component for a safety purpose, the only option would be test every item on the production line which is not economically feasible.
So, best to just put the insulation sheet in.
Now, there is an issue in IEC 60601-1 that the required spacings are in general higher than what is really needed. If the spacing is, say, 0.5mm less than the limit, you might reasonably argue that the insulation on the capacitor or semiconductor is enough to get you over the line. But keep in mind the real story is that the spacing alone is enough for safety (e.g. by inspecting requirements IEC 60664-1). And if it is a component, you need to be confident of the relative positions in regular production. Again, the insulation sheet may be a simpler option.
sankui
Hi Peter,
I am sorry that my question seems to confuse you. Actually, I am not talking about the components which used to bridge the primary and secondary, e.g. Y1 capacitor, transformer, optocoupler.
My question relates to the plastic enclosure of non-safety relevant semiconductors or electrolytic capacitors. Since the small size of the switching power supply, the components which do not provide any safety insulation also adjacent to the safety relevant components, e.g. the bulk capacitor closes to transformer core. This may reduce the insulation distance (assumed that primary winding to core is BI and secondary winding to core is SI, the body of bulk capacitor may bypass the distance from primary winding to core). As we know, for electrolytic capacitor, there is no certification program existing. So the electrolytic capacitor belongs to non-approved component.
What should be considered for the plastic enclosure of semiconductors or electrolytic capacitors?
Peter Selvey
I expect we are talking about the same thing. The cases that I have seen before are (values for example only):
Pri-core internal creepage = 3.2mm (creepage)
Core is conductive ferrite material
Core to secondary electrolytic capacitor 1.0mm (clearance)
Total clearance pri - sec = 3.2 + 1.0 = 4.2 mm < 5.0 limit for 2nd edition.
Then, the applicant may say that plastic covering on secondary capacitor helps to comply. Although in the test sample, 4.2mm is perfectly safe, we should reject it for DI as explained above, since there is no production control over the construction of the secondary capacitor, and there may be a question over the exact positioning of the capacitor with respect to the core (in regular production).
Safety isolation should be planned, intentional, consistent in regular production, and tending towards a bit of overkill to keep things simple.
sankui
Thanks, Peter. I just found an OSM/EE DECISION SHEET about the insulation of plastic enclosure of semiconductor component, see below.
Standard:
EN 60950:2000
EN 60950-1 :2001
Sub clause:
2.10.5
Sheet no. :
95/17
Page: 1 of 1
Subject:
Insulation of plastic enclosure of semiconductor component
Key words:
Semiconductor insulation
Meeting:
OSM/EE - 1995
Item 9.14
Question:
Type of insulation of plastic enclosure of semiconductor component
Decision:
The plastic enclosure of a semiconductor can be considered as basic insulation. These components including alternatives must be listed in the list of critical components and there must be a remark in the service documentation.
Explanatory Notes: -
From above decision sheet, it seems that the plastic enclosure of electrolytic capacitor also can be considered as basic insulation.
F MOPP in battery powered device with metal enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
H Battery Enclosure Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
C Lithium battery powered Medical Device - 15.3.3 Impact test, broken enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
T Plastic of an infusion pump enclosure and ISO 10993 ISO 13485:2016 - Medical Device Quality Management Systems 2
L How to Measure Paint Thickness for an Enclosure Manufacturing and Related Processes 5
D Definition Enclosure - IEC 60601-1 definition Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 2
R IEC 60601-2-49 - Touch Current (Enclosure Leakage Current) requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC/EN 60335-2-30 cl.30 Plastic Enclosure on Fan Heater question Various Other Specifications, Standards, and related Requirements 2
E Application of Foil to Enclosure (Recessed Connector) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R Stand-Alone Power Supply with Metal Enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
A IEC 60601 Ground Bond Test Requirements for a Plastic Enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
W Insulation requirements between F-type applied part and enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R Is any part of earthed (grounded) enclosure required to meet, say, 25A/0.1ohm? IEC 60601 - Medical Electrical Equipment Safety Standards Series 18
S Sharp Edge testing method for Plastic or Sheet Metal Enclosure Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
J Impact using two different materials for product enclosure in radiation and immunity IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
T Enclosure Flammability Rating of Molded-In Custom Color Liquid Concentrates IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
I Connection of Secondary Circuits to Medical Weight Scale Enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
D Flame Rating for Internal Plastic Support inside of a Fire Enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
somashekar Lead acid Batteries inside DG set acoustic enclosure - What Hazard and Risk and how ? Occupational Health & Safety Management Standards 5
J Protective Earthing for Medical Device with two wire power cord in metal enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
T Flame Rating Class for Plastic Materials in Enclosure of Medical Equipment EU Medical Device Regulations 9
T Are Coating, Painting and Labels part of the Enclosure of ME Equipment Other ISO and International Standards and European Regulations 3
K What are the current ISO/ TS 16949 Semiconductor Manufacturer Requirements IATF 16949 - Automotive Quality Systems Standard 12
L FMEA Rankings for Semiconductor Products FMEA and Control Plans 9
H How is a Semiconductor / Electronic Part Manufacturing APQP and PPAP 1
I Should chemical use within a semiconductor facility be an aspect? ISO 14001:2015 Specific Discussions 2
Q Semiconductor Inductor Failure - Will HAAS/HALT or any other test detect it? Manufacturing and Related Processes 4
R Semiconductor Reliability Testing for System - Operating Life Test Reliability Analysis - Predictions, Testing and Standards 11
A Cleanroom class in semiconductor Other ISO and International Standards and European Regulations 8
T Mil Std Certification for a Semiconductor Burn-In Process provider Various Other Specifications, Standards, and related Requirements 2
N ISO 9001 Implementation - Fabless Semiconductor (ASIC) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
Q Definition Excursion - Semiconductor, Definition for Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 4
Jen Kirley Cleanroom Standard for Semiconductor Fabrication Wanted Various Other Specifications, Standards, and related Requirements 3
R TS Certification Scope - Semiconductor subcontractor providing assembly and testing IATF 16949 - Automotive Quality Systems Standard 14
A Supplier Quality Management Strategies - Semiconductor Assembly and Test Services Supplier Quality Assurance and other Supplier Issues 6
antoine.dias What are the current automotive semiconductor requirements? IATF 16949 - Automotive Quality Systems Standard 2
K Predictive Quality Indicators - Manufacture of semiconductor optoelectronic products Quality Tools, Improvement and Analysis 2
D Should the clean room be called a laboratory? Semiconductor industry IATF 16949 - Automotive Quality Systems Standard 7
T Can a Fabless Semiconductor Company achieve TS16949 Certification (Registration)? IATF 16949 - Automotive Quality Systems Standard 48
Q Defunct SAC (Semiconductor Assembly Council) Standards Various Other Specifications, Standards, and related Requirements 11
Q FMEA for Inspection and Test required by TS 16949? Semiconductor Sector FMEA and Control Plans 6
D Is Wafer Probe (On Wafer Test) a Laboratory? Semiconductor Industry IATF 16949 - Automotive Quality Systems Standard 8
Q Concerning Special Characteristics Interpretation - Semiconductor commodity standard IATF 16949 - Automotive Quality Systems Standard 3
W ISO/TS16949 Certificate Withdrawn by CB - Burn-in services to semiconductor companies IATF 16949 - Automotive Quality Systems Standard 7
C Semiconductor Subcontractor Quality System Supplier Quality Assurance and other Supplier Issues 6
D Gage RR to train operators? QS-9000 Semiconductor Supplement 4.11.4.S Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
X Modeling and Analysis of Semiconductor Manufacturing (MASM) Conference Statistical Analysis Tools, Techniques and SPC 3
R Is Gage R&R with measurement data possible in a Semiconductor Fab? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
N TS16949 Quality manual for a fabless semiconductor company Quality Management System (QMS) Manuals 0
V TS 16949 Semiconductor Supplement Status - Obsolete? IATF 16949 - Automotive Quality Systems Standard 1