Can I Calibrate On Demand?

imwilliam

Starting to get Involved
#1
So I'm working on getting ISO 9000 and I'm wondering about how I'm going to deal with calibration.

It's going to be prohibitively expensive and time consuming to keep every gauge in the building calibrated. I'm not even sure whether that's a requirement or not in ISO9000.

So my thought is that my policy will be to allow uncalibrated gauges to exist in an uncalibrated state, but prohibit their use in inspection. Then if work comes in that requires that gauge I can calibrate it at that time. Is this an issue with ISO9000?

I also don't want to put calibration dates on the gauges, just the gauges gauge number, I'll have to check against a master record for the gauge calibration dates and then list those dates on any inspection data/reports generated using that gauge. There's a couple of reasons for this, one of which is to drive the inspector. (usually me), back to the master gauge record to enter a little more information. The other is that again, to me a gauge with a tag that says it's out of calibration is a "violation" for lack of a better word, but I don't want the mere existence of an uncalibrated gauge to be a "violation". I also don't want to run around sticking "for reference" stickers on everything; in part because sometime a year from now a job may come in where I want to calibrate that gauge and use it.

Any thoughts or advice on this?
 
Elsmar Forum Sponsor

QChas

Involved - Posts
#2
What we have had in place for 15 years for machinists with their own tools is as follows:

1) We have sets of gauge blocks that are calibrated by an outside lab
2) Prior to using a micrometer, caliper etc the gauge is "calibrated" against the gauge blocks prior to use
3) Machinists indicates on his paperwork "Verified gauge OK" with initials and date
4) Make sure you have a work instruction / procedure documenting this process

Hope this helps!
 

mmasiddiqui

Involved In Discussions
#3
What does your company offer?
Would it not be easy to have a set of calibrated gages?
See how section 7.1.5 apply to what you make for your customers. Not sure if 7.1.5.2 applies to you or not.
 

imwilliam

Starting to get Involved
#5
Hey QChas, that's exactly the sort of thing I'm hoping will work.

Mmasiddiqui, Some gauges will have to be calibrated on a regular schedule and by an outside lab, surface plates, cylinder squares and gage blocks as QChas mentioned among others. I should have made that more clear. However, something like gauge pins can be very expensive/or time consuming to maintain in calibration at all times. We also have specialized tooling, some micrometers that aren't used very often that I wouldn't want to keep "in calibration"

And as Sidney pointed out, according to 7.1.5.2 a) I don't have to. Thanks for that Sidney. I am warming up to this stuff, but the standard still devolves into a wall of text rather quickly for me. I'm sure it will get better over time.

Also looks like 7.1.5.2 b) requires only that I identify the gauge so that the status can be determined, not that I need to put "status", the calibration dates themselves, on the gauge.

Thanks everyone.
 

QChas

Involved - Posts
#6
Hey QChas, that's exactly the sort of thing I'm hoping will work.

Mmasiddiqui, Some gauges will have to be calibrated on a regular schedule and by an outside lab, surface plates, cylinder squares and gage blocks as QChas mentioned among others. I should have made that more clear. However, something like gauge pins can be very expensive/or time consuming to maintain in calibration at all times. We also have specialized tooling, some micrometers that aren't used very often that I wouldn't want to keep "in calibration"

And as Sidney pointed out, according to 7.1.5.2 a) I don't have to. Thanks for that Sidney. I am warming up to this stuff, but the standard still devolves into a wall of text rather quickly for me. I'm sure it will get better over time.

Also looks like 7.1.5.2 b) requires only that I identify the gauge so that the status can be determined, not that I need to put "status", the calibration dates themselves, on the gauge.

Thanks everyone.
Just remember to document what you do so you can reference the process
 

mmasiddiqui

Involved In Discussions
#7
Hey QChas, that's exactly the sort of thing I'm hoping will work.

Mmasiddiqui, Some gauges will have to be calibrated on a regular schedule and by an outside lab, surface plates, cylinder squares and gage blocks as QChas mentioned among others. I should have made that more clear. However, something like gauge pins can be very expensive/or time consuming to maintain in calibration at all times. We also have specialized tooling, some micrometers that aren't used very often that I wouldn't want to keep "in calibration"

And as Sidney pointed out, according to 7.1.5.2 a) I don't have to. Thanks for that Sidney. I am warming up to this stuff, but the standard still devolves into a wall of text rather quickly for me. I'm sure it will get better over time.

Also looks like 7.1.5.2 b) requires only that I identify the gauge so that the status can be determined, not that I need to put "status", the calibration dates themselves, on the gauge.

Thanks everyone.
You can argue both the sides. That was the reason, I asked you about the product you manufacture for your customer. What is your capability? How frequently do you have to measure? You know your process better than me, so you can chose the option of calibration before measurement. As Qchas pointed, you will surely need to have that in your procedure.
I would be cautious during your audit when the external auditor is watching your folks on the line should not forget to calibrate before use. Else, you may have some issues with Audit findings.
Good Luck! Wish you success.
Ninja is good with this stuff, I am surprised he has not responded to this post.
 

imwilliam

Starting to get Involved
#8
I would be cautious during your audit when the external auditor is watching your folks on the line should not forget to calibrate before use. Else, you may have some issues with Audit findings.
.
It's a good point. I'm thinking that some sort of notation like QChas mentioned should help keep everyone on track.
 

imwilliam

Starting to get Involved
#10
So a common scenario I'm trying to cover here is a new job coming in where the customer needs a quick turn around, think in on a Friday and out on a Monday, that is best checked with a gauge pin. Usually I can get a pin next day, but getting that pin calibrated by a lab in time in a situation like this is next to impossible.

To cover the "I can get a pin but can't get it calibrated" I have a submicron micrometer so we can calibrate to Z class gauge pins in house.

However, sometimes I can't get the pin, particularly if it's a half size. So, in the range where the vast majority of our works falls, up to .5. I have a set of gauge pins including 1/2 sizes in both plus and minus. But, I don't want to calibrate all these pins, just the ones I need as I need them.

However, once they're calibrated they're calibrated for a given time frame, a year or whatever.

If the job looks like it's going to repeat, I buy job specific pins, put them in a handle and give the assembly it's own gauge number.

Hope this explains what I'm really trying to cover here when I write about "calibrating on demand".
 
Thread starter Similar threads Forum Replies Date
S Calibrating our own equipment, can we? Micrometers to calibrate vernier calipers General Measurement Device and Calibration Topics 13
F Crimper Calibration - How do you Control & Calibrate your crimpers? Calibration Frequency (Interval) 13
P How to Calibrate a HP 3458A Digital Multimeter with Fluke 5080A calibrator General Measurement Device and Calibration Topics 1
B Choosing not to calibrate (IATF 16949) IATF 16949 - Automotive Quality Systems Standard 6
K New Kilogram - The process we use to calibrate the standards General Measurement Device and Calibration Topics 6
D How to Calibrate any Digital Sphygmomanometer under use for 2 years in a Hospital Forum News and General Information 1
L Re-calibrate after software update? Calibration Frequency (Interval) 3
J To Calibrate or not to calibrate electronics rack components Manufacturing and Related Processes 5
T How to Calibrate Taper Gages General Measurement Device and Calibration Topics 1
N Calibrate temperature on refrigerated Centrifuges? General Measurement Device and Calibration Topics 2
B Can we calibrate a standard pin with a gauge? General Measurement Device and Calibration Topics 2
D Lesser expensive to calibrate various test weights General Measurement Device and Calibration Topics 5
A How to Calibrate Laser Beam Tester & Radiation Monitor General Measurement Device and Calibration Topics 3
D How to Calibrate and Verify a Heating Oven General Measurement Device and Calibration Topics 3
P How do you calibrate a VNA (Vector Network Analyzer) cal kit? General Measurement Device and Calibration Topics 4
V How to Calibrate Ring Force Gauges Service Industry Specific Topics 1
bobdoering Who says you can't calibrate a steel rule! Funny Stuff - Jokes and Humour 14
S How can I Calibrate a Biosafety Cabinet? General Measurement Device and Calibration Topics 1
C How to Calibrate a Viscosity Tester General Measurement Device and Calibration Topics 3
0 Is it possible to use a threaded ring gage to calibrate the M8 gage? Calibration and Metrology Software and Hardware 4
C How to Calibrate a Medical Infrared Thermometer General Measurement Device and Calibration Topics 3
T How do I calibrate NPT Thread Plug Gages? General Measurement Device and Calibration Topics 9
E Do you need a Certification to Calibrate Gauges? General Measurement Device and Calibration Topics 2
B Calibrate Hole Presence Detection Device? General Measurement Device and Calibration Topics 5
J What do you use to Calibrate 40" Calipers General Measurement Device and Calibration Topics 10
A Guideline to Calibrate Chamber Temp. & Pressure of a Medical Steam Sterilizer General Measurement Device and Calibration Topics 8
S How to calibrate AOI (Automated Optical Inspection) for IOPQ Qualification and Validation (including 21 CFR Part 11) 3
S Laboratory to Calibrate Objective Micrometer needed General Measurement Device and Calibration Topics 4
C Tolerance Estimation of a Load Cell I want to Calibrate General Measurement Device and Calibration Topics 9
D Procedure on How to Calibrate a External Micrometer General Measurement Device and Calibration Topics 1
B ISO/TS 16949 Section 7.6.3 - Use of Manufacturer to Calibrate IATF 16949 - Automotive Quality Systems Standard 21
W How to verify/calibrate a Neptune Water Meter General Measurement Device and Calibration Topics 4
N Laboratory that is able to Calibrate Dew Point Meter up to 60?Cdp - Needed General Measurement Device and Calibration Topics 1
C How to Calibrate and Verify Torque Multiplier Gages General Measurement Device and Calibration Topics 6
C How to calibrate or control patterns for fiberglass General Measurement Device and Calibration Topics 4
N Scale with Internal Calibration - Do I need to calibrate? How to document this? Calibration Frequency (Interval) 8
I Need help finding an Accredited Lab that can Calibrate Brunson 76-RH Survey Transits General Measurement Device and Calibration Topics 5
M Controlling Temperature and Time - Do I need to calibrate the thermostat? IATF 16949 - Automotive Quality Systems Standard 4
D How much Tolerance do I use to Calibrate a 150kgs capacity Scale? General Measurement Device and Calibration Topics 5
M To Calibrate or Not to Calibrate various Weight Scales ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 24
N Thermocouple Inhomogeneity - Need to calibrate a roll of 50m thermocouple type K General Measurement Device and Calibration Topics 3
J Calibrate before use required? Inspection Gages and Measurement Equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
R Traceability for Mercury Argon Calibration Source to Calibrate a Spectrometer? General Measurement Device and Calibration Topics 5
P How to Calibrate Laser Micrometer when no Calibration Company in the Country? General Measurement Device and Calibration Topics 3
S 3 Point Calibration or Calibrate to Extreme High and Low? General Measurement Device and Calibration Topics 9
C Seeking a holder/jig to calibrate large OD micrometer calipers (up to 23") General Measurement Device and Calibration Topics 8
S Is it OK to calibrate a measurement device tighter than manufacturer specifications? General Measurement Device and Calibration Topics 18