Can I calibrate Optical Comparators internally?
  • Thread starter Bob Gallinger - 2007
  • Start date
Do I have to get my opticals calibrated externally or can I not write good calibration proceedures and using my tracable gauge bloack etc. calibrate internally. I have 4-5 of them and I could save some money by calibrating them myself. Anyone have good writen procedures that could be shared ???
Thanks.
Hershal
Metrologist-Auditor
Re: Optical Comparators - Can I calibrate internally?
Of course you can calibrate internally, but I strongly suggest having an accredited cal lab calibrate them for you.
The idea of having metrology professionals do your calibration is the same as having an appropriately trained and qualified professional to inspect or program a CNC or occupy any other highly technical position.
Since you are in Canada then check this link to find a lab closest to you:
http://palcan.scc.ca/SpecsSearch/CalibrationSearchForm.do
Hope this helps.
Hershal
I understand that having external personal would be better but I have the standards and know how to measure accurately on an optical. If the optical fails my internal calibration then I'll call an external Lab to adjust, but 9 out-of 10 times the optical is fine but we pay just to have a cert. Somebody out here must calibrate/verify their own opticals internally.
Duke Okes
It's entirely your organization's decision whether to do internally or externally. As long as whoever does it has appropriate equipment, knowledge & skills, and calibration is traceable to an external standard that is acceptable.
jrubio
It depends.
If you work in the Quality Anena, The calibration is "voluntary", only mandatory because your customer want it or because your proceddures call for them.
But if you are within a Security Arena Structure under industrial Laws you are not authorized to calibrate. Only notified Labs
This rules applies in Spain and Europe, not sure in USA.
:bonk: :bonk:
Ron Rompen
Bob:
The short answer is yes.....there is no requirement (that I know of) that you must have an external 'expert' perform calibration of any of your equipment.
Depending on what brand of OC and software you have, the equipment used for calibration will vary significantly. If you have something like QC4000 or 5000, you want to be able to check a LOT of things (length, height, diameter, position, etc etc etc) for accuracy and (over time) stability. Mituotoyo now sells (not sure of the price) a calibration master that is VERY appropriate for calibration of OC's, or you can design one and have it manufactured for you by any decent tool shop/wire room.
Sorry I no longer have my procedure available to me, but I will see if I can dig up my draft copy, and will forward it on to you.
G
Grizz1345
Bob
I do the calibrations on our optical comparators here. I use a glass scale and glass reticle for checking magnification. If I were to run into a problem I would not hesitate to call for outside help. My standards are NIST traceable and I have had no issues in any audit.
I have attempted to attach my procedure.
Hope all of this helps.
jane_ackerman - 2010
My question would be... what about angles?
With gage blocks you can build various lengths to check throughout the resolution of the gage... but you could only verify a 90 degree angle... not any minutes or seconds in between, nor any other various angles throughout the resolution.
Re: Optical Comparators - Can I calibrate Optical Comparators internally?
I'm well aware of angles and plan on purchasing/certifying angle blocks.
I still need to research on classes/standards of angle blocks.
Thanks for all your guys help.
Sorry if I am beating a dead horse here... but what would you use to certify your angle (and radius) gages besides an optical comparator?
S How to calibrate AOI (Automated Optical Inspection) for IOPQ Qualification and Validation (including 21 CFR Part 11) 3
A Looking for a third party to calibrate our measure equipment - South Carolina General Measurement Device and Calibration Topics 6
I Can I Calibrate On Demand? General Measurement Device and Calibration Topics 9
S Calibrating our own equipment, can we? Micrometers to calibrate vernier calipers General Measurement Device and Calibration Topics 13
F Crimper Calibration - How do you Control & Calibrate your crimpers? Calibration Frequency (Interval) 13
P How to Calibrate a HP 3458A Digital Multimeter with Fluke 5080A calibrator General Measurement Device and Calibration Topics 1
B Choosing not to calibrate (IATF 16949) IATF 16949 - Automotive Quality Systems Standard 6
K New Kilogram - The process we use to calibrate the standards General Measurement Device and Calibration Topics 6
D How to Calibrate any Digital Sphygmomanometer under use for 2 years in a Hospital Forum News and General Information 1
L Re-calibrate after software update? Calibration Frequency (Interval) 3
J To Calibrate or not to calibrate electronics rack components Manufacturing and Related Processes 5
T How to Calibrate Taper Gages General Measurement Device and Calibration Topics 1
N Calibrate temperature on refrigerated Centrifuges? General Measurement Device and Calibration Topics 2
B Can we calibrate a standard pin with a gauge? General Measurement Device and Calibration Topics 2
D Lesser expensive to calibrate various test weights General Measurement Device and Calibration Topics 5
A How to Calibrate Laser Beam Tester & Radiation Monitor General Measurement Device and Calibration Topics 3
D How to Calibrate and Verify a Heating Oven General Measurement Device and Calibration Topics 3
P How do you calibrate a VNA (Vector Network Analyzer) cal kit? General Measurement Device and Calibration Topics 4
V How to Calibrate Ring Force Gauges Service Industry Specific Topics 1
bobdoering Who says you can't calibrate a steel rule! Funny Stuff - Jokes and Humour 14
S How can I Calibrate a Biosafety Cabinet? General Measurement Device and Calibration Topics 1
C How to Calibrate a Viscosity Tester General Measurement Device and Calibration Topics 3
0 Is it possible to use a threaded ring gage to calibrate the M8 gage? Calibration and Metrology Software and Hardware 4
C How to Calibrate a Medical Infrared Thermometer General Measurement Device and Calibration Topics 3
T How do I calibrate NPT Thread Plug Gages? General Measurement Device and Calibration Topics 9
E Do you need a Certification to Calibrate Gauges? General Measurement Device and Calibration Topics 2
B Calibrate Hole Presence Detection Device? General Measurement Device and Calibration Topics 5
J What do you use to Calibrate 40" Calipers General Measurement Device and Calibration Topics 10
A Guideline to Calibrate Chamber Temp. & Pressure of a Medical Steam Sterilizer General Measurement Device and Calibration Topics 8
S Laboratory to Calibrate Objective Micrometer needed General Measurement Device and Calibration Topics 4
C Tolerance Estimation of a Load Cell I want to Calibrate General Measurement Device and Calibration Topics 9
D Procedure on How to Calibrate a External Micrometer General Measurement Device and Calibration Topics 1
B ISO/TS 16949 Section 7.6.3 - Use of Manufacturer to Calibrate IATF 16949 - Automotive Quality Systems Standard 21
W How to verify/calibrate a Neptune Water Meter General Measurement Device and Calibration Topics 4
N Laboratory that is able to Calibrate Dew Point Meter up to 60?Cdp - Needed General Measurement Device and Calibration Topics 1
C How to Calibrate and Verify Torque Multiplier Gages General Measurement Device and Calibration Topics 6
C How to calibrate or control patterns for fiberglass General Measurement Device and Calibration Topics 4
N Scale with Internal Calibration - Do I need to calibrate? How to document this? Calibration Frequency (Interval) 8
I Need help finding an Accredited Lab that can Calibrate Brunson 76-RH Survey Transits General Measurement Device and Calibration Topics 5
M Controlling Temperature and Time - Do I need to calibrate the thermostat? IATF 16949 - Automotive Quality Systems Standard 4
D How much Tolerance do I use to Calibrate a 150kgs capacity Scale? General Measurement Device and Calibration Topics 5
M To Calibrate or Not to Calibrate various Weight Scales ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 24
N Thermocouple Inhomogeneity - Need to calibrate a roll of 50m thermocouple type K General Measurement Device and Calibration Topics 3
J Calibrate before use required? Inspection Gages and Measurement Equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
R Traceability for Mercury Argon Calibration Source to Calibrate a Spectrometer? General Measurement Device and Calibration Topics 5
P How to Calibrate Laser Micrometer when no Calibration Company in the Country? General Measurement Device and Calibration Topics 3
S 3 Point Calibration or Calibrate to Extreme High and Low? General Measurement Device and Calibration Topics 9
C Seeking a holder/jig to calibrate large OD micrometer calipers (up to 23") General Measurement Device and Calibration Topics 8
S Is it OK to calibrate a measurement device tighter than manufacturer specifications? General Measurement Device and Calibration Topics 18
B Difference between Calibration & Verification and How to calibrate Go/No Go gauge? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4