Can I use a Laboratory CMM for Gauge Calibration
Bmwtone
I have a need to have large Attribute Fixture Gauges calibrated (up to 2000mm x 1000mm x 500mm). Does anyone know if I can use a fully certified Testing Laboratory using a large CMM to calibrate these types of gauges? Our TS auditor suggests he/she prefers we follow the conventional measurement calibration techniques using hard gauging such as mics, verniers etc!
I very much look forward to any valued response
Stijloor
Re: Can I use a Lab CMM for Gauge Calibration
Welcome to The Cove Forums! :bigwave: :bigwave:
#1. Your "TS Auditor" is not allowed to provide advice on how to do certain things.
#2. A CMM can be used to verify/calibrate your attribute gage.
Stijloor.
Ron Rompen
Just to add on to what Stij has already (correctly) stated.
Using your CMM for calibration of other measuring equipment is perfectly acceptable. However, keep in mind that the required accuracy of your CMM should be following the 1:10 rule as well; i.e. if you are calibrating your gage to ± 0.0005", then your CMM should be accurate to ± 0.00005".
Big Jim
Two thoughts come to mind. I may be wrong, as I'm not involved with TS16949.
First, isn't it a 1:4 rule, not 1:10 rule?
Second, I think that you must use a calibration lab accredited to ISO 17025 if you are registered to TS 16949.
F
falconer65
BMW,
We use our CMM for everything we can.
Jim,
Yeah, I had to look it up myself. Here it is paraphrased
7.6.3.1 Internal laboratory
Accreditation to 17025 may be used but is not mandatory.
Seems like I remember 4:1 being what I was taught years ago. Nowadays it seems everyone wants 10:1 to show capability.
ram4302
Hi
You can calibrate the attribute gauge by using CMM after ensuring its accuracy as explained by our members.
You have create procedure for internal calibration.
:thanx:
JAltmann
The 4:1 and 10:1 uncertianty ratios are not explicit rules, but good guidelines to help ensure your not commiting type 1 or type 2 errors. That being said the 4:1 rule tend to be the standard under the ISO 17025 standard for laboratory calibrations.
The 10:1 rule comes from the MSA side of things, currently under AIAG's MSA manual but i know dates back to the 50's (i think) in regards to reducing potential errors.
Not all cases will allow for 10:1 realistically with todays tighter tolerances. Basically if you can not preform to these levels you need to mae some provisions for this, either multiple inspections or some form of statistical studies. You must then be able to defend this stance if challenged.
Mr.Happy
We use a 5:1 (or 1:5) for calibrations.
Also we conduct Gage rings calibration on our CMM.
I have an instruction for it and a Excel sheet for writing down the (history)data.
However it is not in English, if you are interested I can make a translation.
For double check you can send one Gage out for calibration and then compare whit your own results. And later on you can do a MSA study.
Cheers
Mr.Happy
Bmwtone
Many thanks for all of your responses and inputs. What you have highlighted is exactly how I was thinking. I will have to get the specs from the Testing Lab for their CMM capabilities but I am sure that I can progress along these lines.
Curious - When you say "Testing Lab" are you referring to and outside, independent lab that does you calibrations or an internal quality lab?
If it's an outside lab, there shouldn't be a problem as they are expected to know what they are doing and supply you with proper certifications.
If it is an internal lab - Some of the factors mentioned by others come into play.
Assuming an Internal lab:
To me the bottom line is - can you check the gages with sufficient accuracy and precision, covering all the possible "faults" (roundness, flatness etc) so as to ensure the gage functions properly for it's intended purpose?
This requires that your equipment be sufficiently precise, but also that the personnel doing the check be sufficiently capable - That they understand how the gage is used, the potential failure modes etc.
If the answers to the above are yes, then by all means go ahead and do the check in house...
Peace
James
H Temperature Control in a Laboratory with CMM and other delicate instruments IATF 16949 - Automotive Quality Systems Standard 10
J Just accredited!!! CMM Dimensional Inspection Laboratory General Measurement Device and Calibration Topics 8
D Environmental standards for Gage and CMM laboratory General Measurement Device and Calibration Topics 1
X ISO 17025 certification for Laboratory for online gambling products ISO 17025 related Discussions 3
B Laboratory KPIs ISO 17025 related Discussions 13
M Calibration Certificate Result issued by an accredited external laboratory General Measurement Device and Calibration Topics 9
T Laboratory Verification after validation ISO 17025 related Discussions 3
Crimpshrine13 Laboratory Scope - Calibration vs. Test Methods - IATF 16949 IATF 16949 - Automotive Quality Systems Standard 3
D Using Laboratory Notebooks in R&D and Design and Development ISO 13485:2016 - Medical Device Quality Management Systems 3
G ITAR Registration for Metrology Laboratory? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 11
S Laboratory Environmental Conditions Outside of limits ISO 17025 related Discussions 4
I Laboratory Fridge / Freezer Inventory - Requirement(s) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
D Definition of "Critical Laboratory Equipment"? General Measurement Device and Calibration Topics 1
S Internal calibrations - Part of an ISO 17025 accredited testing laboratory (Automotive) ISO 17025 related Discussions 16
DMLqms Material for laboratory use or Medical Device? EU Medical Device Regulations 2
N IATF 16949:2016 7.1.5.3.2 External Laboratory - How to approve the Testing Laboratory without accreditation scope IATF 16949 - Automotive Quality Systems Standard 2
A Single use non-sterile syringe used in the oral cavity - Laboratory test advice US Food and Drug Administration (FDA) 7
B Accredited laboratory or not under MDD 93/42/EEC IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
D Does every piece of equipment used in a laboratory need to have an IQ protocol written and executed? ISO 13485:2016 - Medical Device Quality Management Systems 1
Casana IATF 16949 7.1.5.3.2 External Laboratory - On Site Calibration IATF 16949 - Automotive Quality Systems Standard 8
A Information on good laboratory practices (GLP) for the non-medical or food industry wanted Manufacturing and Related Processes 1
D Internal audit forms or checklists for a medical/veterinary laboratory General Auditing Discussions 5
N IATF 16949 7.1.5.3.2 External Laboratory - Calibration of Nikon machine in factory IATF 16949 - Automotive Quality Systems Standard 24
Q Calibration laboratory scope - Include software such as Gagetrak or Mini tab used? General Measurement Device and Calibration Topics 3
B IATF 16949 7.1.5.3.2 FAQ #7 Audit Finding - External Calibration Laboratory IATF 16949 - Automotive Quality Systems Standard 7
P Procedure for maintaining Confidentiality And Impartiality in a Laboratory Quality Management System (QMS) Manuals 2
sswaim IATF 16949 7.1.5.3.2 External Calibration Laboratory Requirelents IATF 16949 - Automotive Quality Systems Standard 15
J Good Laboratory Practice: online training course recommendations? US Food and Drug Administration (FDA) 0
E 2 participants in inter-laboratory comparison tests ISO 17025 related Discussions 3
T MDR on Good Laboratory Practice (GLP) - Pg 110, section 6.1 (b) EU Medical Device Regulations 1
D Laboratory Manual ISO/IEC 17025 Example wanted ISO 17025 related Discussions 2
D Help identify a Nationally Recognized Testing Laboratory (NRTL) (UL) certified lab General Measurement Device and Calibration Topics 3
L How to manage a Metrology Department or Calibration Laboratory General Measurement Device and Calibration Topics 3
J IATF 16949 Calibration Laboratory Scope IATF 16949 - Automotive Quality Systems Standard 22
H Internal laboratory scope requirements - IATF 16949 Cl. 7.1.5.3.1 IATF 16949 - Automotive Quality Systems Standard 1
S RSL (Remote Site Location) Laboratory - Does it need to be ISO 17025 IATF 16949 - Automotive Quality Systems Standard 5
S Jewelry vs. Normal Laboratory Balances - Accuracy and calibration General Measurement Device and Calibration Topics 2
rob73 Requirement for independent certification of device using an ISO 17025 laboratory EU Medical Device Regulations 2
G Calibration Laboratory Dimensional Temperature Question General Measurement Device and Calibration Topics 2
G ISO 17025 Calibration Laboratory Assessor's Common Questions and Procedures ISO 17025 related Discussions 11
H Need to Start an R&D Laboratory, what to do? Design and Development of Products and Processes 6
B IATF 16949 - 7.1.5.3.1 Internal calibration laboratory requirements IATF 16949 - Automotive Quality Systems Standard 9
A Pharmaceutical Manufacturer's obligations - Laboratory Test instruments General Measurement Device and Calibration Topics 0
S HVAC Thermostat for Calibration Laboratory General Measurement Device and Calibration Topics 8
V Temperature/humidity requirements in internal metrology laboratory General Measurement Device and Calibration Topics 3
C Looking for Testing Laboratory for Aluminum Products - CE Marking General Measurement Device and Calibration Topics 5
S Looking for a Threaded Ring Gage Calibration Laboratory General Measurement Device and Calibration Topics 13
J Internal Laboratory Scope and Measurement Uncertainty (MU) IATF 16949 - Automotive Quality Systems Standard 1
B IATF 16949 Cl. 7.1.5.3.2 - External Laboratory Sub-Supplier Requirements IATF 16949 - Automotive Quality Systems Standard 2
J Is a Laboratory Scope Necessary if... IATF 16949 - Automotive Quality Systems Standard 15