#11
An easy way to remember TS16949 for the labs is that if you go outside, the lab should be ISO 17025 sometimes slang as Guide 25 or A2LA or the OEM of the device if no other services are available.



The loophole is if you do it in house. Then you don’t need 17025. What you do need though is to have the calibration of that type item listed on your own Lab Scope, defined methodology and (NIST) traceable tools for checking that type item that takes into consideration: acceptance criteria, linearity, stability & bias.



While the 1:10 ratio is a rule of thumb it basically is just saying that you want your checking tool to be 10x’s more accurate that what it is being used to check. Can you go with a lesser amount…sure. Just realize that it directly goes back to your measurement error
 
Last edited by a moderator:
Elsmar Forum Sponsor

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#12
Short answer as to whether you can use the CMM is: perhaps.

Some considerations however.

First, if the CMM is used for anything except calibration, then it is instantly suspect as a measurement standard. To assure it is acceptable, it will need to be calibrated immediately before its use as a measurement standard, and the uncertainty and traceability documented.

Second, for the items you want calibrated, the traceability and uncertainty must be documented.

If the traceability chain and the uncertainty are not documented, then there is no traceability and the exercise is pointless.

Also, as to slang, the term is accredited. Guide 25 was retired in 1999 and is not in use. A2LA is an organization, not an accreditation, and is not properly used in that manner, regardless of what one may hear on the street.

Remember that there are 6 ABs based in the U.S. that accredit general test labs, all of which are recognized internationally. 2 others are recognized, but are industry-specific.
 
Thread starter Similar threads Forum Replies Date
H Temperature Control in a Laboratory with CMM and other delicate instruments IATF 16949 - Automotive Quality Systems Standard 10
J Just accredited!!! CMM Dimensional Inspection Laboratory General Measurement Device and Calibration Topics 8
D Environmental standards for Gage and CMM laboratory General Measurement Device and Calibration Topics 1
X ISO 17025 certification for Laboratory for online gambling products ISO 17025 related Discussions 3
B Laboratory KPIs ISO 17025 related Discussions 13
M Calibration Certificate Result issued by an accredited external laboratory General Measurement Device and Calibration Topics 9
T Laboratory Verification after validation ISO 17025 related Discussions 3
Crimpshrine13 Laboratory Scope - Calibration vs. Test Methods - IATF 16949 IATF 16949 - Automotive Quality Systems Standard 3
D Using Laboratory Notebooks in R&D and Design and Development ISO 13485:2016 - Medical Device Quality Management Systems 3
G ITAR Registration for Metrology Laboratory? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 11
S Laboratory Environmental Conditions Outside of limits ISO 17025 related Discussions 4
I Laboratory Fridge / Freezer Inventory - Requirement(s) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
D Definition of "Critical Laboratory Equipment"? General Measurement Device and Calibration Topics 1
S Internal calibrations - Part of an ISO 17025 accredited testing laboratory (Automotive) ISO 17025 related Discussions 16
DMLqms Material for laboratory use or Medical Device? EU Medical Device Regulations 2
N IATF 16949:2016 7.1.5.3.2 External Laboratory - How to approve the Testing Laboratory without accreditation scope IATF 16949 - Automotive Quality Systems Standard 2
A Single use non-sterile syringe used in the oral cavity - Laboratory test advice US Food and Drug Administration (FDA) 7
B Accredited laboratory or not under MDD 93/42/EEC IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
D Does every piece of equipment used in a laboratory need to have an IQ protocol written and executed? ISO 13485:2016 - Medical Device Quality Management Systems 1
Casana IATF 16949 7.1.5.3.2 External Laboratory - On Site Calibration IATF 16949 - Automotive Quality Systems Standard 8
A Information on good laboratory practices (GLP) for the non-medical or food industry wanted Manufacturing and Related Processes 1
D Internal audit forms or checklists for a medical/veterinary laboratory General Auditing Discussions 5
N IATF 16949 7.1.5.3.2 External Laboratory - Calibration of Nikon machine in factory IATF 16949 - Automotive Quality Systems Standard 24
Q Calibration laboratory scope - Include software such as Gagetrak or Mini tab used? General Measurement Device and Calibration Topics 3
B IATF 16949 7.1.5.3.2 FAQ #7 Audit Finding - External Calibration Laboratory IATF 16949 - Automotive Quality Systems Standard 7
P Procedure for maintaining Confidentiality And Impartiality in a Laboratory Quality Management System (QMS) Manuals 2
sswaim IATF 16949 7.1.5.3.2 External Calibration Laboratory Requirelents IATF 16949 - Automotive Quality Systems Standard 15
J Good Laboratory Practice: online training course recommendations? US Food and Drug Administration (FDA) 0
E 2 participants in inter-laboratory comparison tests ISO 17025 related Discussions 3
T MDR on Good Laboratory Practice (GLP) - Pg 110, section 6.1 (b) EU Medical Device Regulations 1
D Laboratory Manual ISO/IEC 17025 Example wanted ISO 17025 related Discussions 2
D Help identify a Nationally Recognized Testing Laboratory (NRTL) (UL) certified lab General Measurement Device and Calibration Topics 3
L How to manage a Metrology Department or Calibration Laboratory General Measurement Device and Calibration Topics 3
J IATF 16949 Calibration Laboratory Scope IATF 16949 - Automotive Quality Systems Standard 22
H Internal laboratory scope requirements - IATF 16949 Cl. 7.1.5.3.1 IATF 16949 - Automotive Quality Systems Standard 1
S RSL (Remote Site Location) Laboratory - Does it need to be ISO 17025 IATF 16949 - Automotive Quality Systems Standard 5
S Jewelry vs. Normal Laboratory Balances - Accuracy and calibration General Measurement Device and Calibration Topics 2
rob73 Requirement for independent certification of device using an ISO 17025 laboratory EU Medical Device Regulations 2
G Calibration Laboratory Dimensional Temperature Question General Measurement Device and Calibration Topics 2
G ISO 17025 Calibration Laboratory Assessor's Common Questions and Procedures ISO 17025 related Discussions 11
H Need to Start an R&D Laboratory, what to do? Design and Development of Products and Processes 6
B IATF 16949 - 7.1.5.3.1 Internal calibration laboratory requirements IATF 16949 - Automotive Quality Systems Standard 9
A Pharmaceutical Manufacturer's obligations - Laboratory Test instruments General Measurement Device and Calibration Topics 0
S HVAC Thermostat for Calibration Laboratory General Measurement Device and Calibration Topics 8
V Temperature/humidity requirements in internal metrology laboratory General Measurement Device and Calibration Topics 3
C Looking for Testing Laboratory for Aluminum Products - CE Marking General Measurement Device and Calibration Topics 5
S Looking for a Threaded Ring Gage Calibration Laboratory General Measurement Device and Calibration Topics 13
J Internal Laboratory Scope and Measurement Uncertainty (MU) IATF 16949 - Automotive Quality Systems Standard 1
B IATF 16949 Cl. 7.1.5.3.2 - External Laboratory Sub-Supplier Requirements IATF 16949 - Automotive Quality Systems Standard 2
J Is a Laboratory Scope Necessary if... IATF 16949 - Automotive Quality Systems Standard 15

Similar threads

Top Bottom