Can ISO 17025 be turned on and off depending on what you pay for?

ScottK

Not out of the crisis
Staff member
Super Moderator
#1
OK that's an odd question but here's the situation:

We have a couple of pieces of equipment to get calibrated. We sent the info for a quote and the calibration house came back with a price.
Then we said "Oh - we just noticed your cert doesn't say that you are ISO 17025 accredited and we need that for this piece of equipement" and they came back and said "if you want that on the cert it's a markup".
And the markup was on the order of 100%.

I called back and said "wait wait wait... Isn't 17025 a QMS? You either have it or you don't. For all processes or none. Right?"
A manager got on the phone and proceeded to explain how expensive and how long the accreditation process was and that if we want our stuff calibrated to the "accredited process" with the "accredited equipment" we have to pay the premium.
"All calibration houses do it this way" he said.

Now - am I wrong in calling bull&*^?

I thought 17025 accredited the entire operation - not just certain processes.

Please straighten me out if I'm wrong.
 
Elsmar Forum Sponsor

Coury Ferguson

Moderator here to help
Staff member
Super Moderator
#2
OK that's an odd question but here's the situation:

We have a couple of pieces of equipment to get calibrated. We sent the info for a quote and the calibration house came back with a price.
Then we said "Oh - we just noticed your cert doesn't say that you are ISO 17025 accredited and we need that for this piece of equipement" and they came back and said "if you want that on the cert it's a markup".
And the markup was on the order of 100%.

I called back and said "wait wait wait... Isn't 17025 a QMS? You either have it or you don't. For all processes or none. Right?"
A manager got on the phone and proceeded to explain how expensive and how long the accreditation process was and that if we want our stuff calibrated to the "accredited process" with the "accredited equipment" we have to pay the premium.
"All calibration houses do it this way" he said.

Now - am I wrong in calling bull&*^?

I thought 17025 accredited the entire operation - not just certain processes.

Please straighten me out if I'm wrong.
This is strange...the calibration lab we use has 17025 accreditation and we aren't charged a premium for having the certifications reflect the 17025 accreditation.
 
S

somerqc

#3
The calibration house we use went through this process a bit over a year ago. The prices rose exactly $1 since then . It was probably more due to the overall increase in operating costs than the accreditation.

So I would agree that they are full of ........

John
 

BradM

Staff member
Admin
#4
Yes... they do those kinds of activities.

The statement goes that the paperwork may be different and stuff.

Many calibration houses may only have some/a few processes "covered", and may be doing work that is outside their scope.

I agree with you. It's like providing the data. You have it as part of the calibration process. Why am I getting charged for it as an extra?
 

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#5
Actually, that situation is not uncommon.....some (including me) consider that process unethical.....that is, an accredited lab should not charge extra for the calibrations that are under the scope.....but marketplace realities.....

The accreditation does include the QMS, but unlike 9K is not limited to the QMS.....in fact, the technical scope of accreditation requires an evaluation of the calibration process including equipment, personnel, uncertainty, certificates, and other related factors.....

Calibration that is NOT under the technical scope CANNOT receive an accredited certificate, otherwise they risk all kinds of unpleasant discussions with the AB.....

It must also be noted that many labs do in fact charge the same or a very low premium for an accredited cal. If you receive a high charge (like the 100% you mentioned), then I would look for the AB logo and the specific uncertainties for each reading to justify a high cost.....

Hope this helps.
 

ScottK

Not out of the crisis
Staff member
Super Moderator
#6
So I am wrong.
Very good.
I learned something valuable today, thank you.

Now I have to check with my PED notified body to see if their QMS being acreddited is enough or if he requires that the calibration process for our equipment used for PED parts needs be under their acreddited process.

Still... that markup seems very extreme.
 

ScottK

Not out of the crisis
Staff member
Super Moderator
#7
Continuing this line of questioning...

Is the QMS accredited PLUS certain processes?
Or are only the processes accredited?
 

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#8
The laboratory to include BOTH the QMS and technical operations (described in the scope of accreditation) are addressed by the accreditation.....the QMS without the technical operations would basically be ISO9001.
 

ScottBP

Involved In Discussions
#9
Fluke does this all the time. You get charged more for a 17025 accredited calibration with uncertainties than a standard Z540-compliant calibration, and the markup for them is on the order of 30-40%. I've been told the reasoning for the extra charge is that it takes extra time to compute and report the uncertainties versus just taking the readings.

Many cal labs charge in "tiers", where you pay one price for a certificate only, a bit more for certificate and data, and a lot more for 17025 accredited with uncertainties. There are companies who only care about getting a certificate, and unfortunately, there are cal labs out there who cater to those customers.
 

Wesley Richardson

Wes R
Trusted Information Resource
#10
Sorry to come to this discussion so late.

At a testing and calibration lab where I previously worked, that was Accredited to ISO/IEC Guide 25 and then ISO/IEC 17025, several of the suppliers that I sent our equipment to for calibration had a two-tier fee schedule. The certificate meeting the ISO/IEC 17025 requirements in one case was exactly double the non-17025 certificate.

When I audited this supplier, I then understood why. The non-17025 certificate for my device was two pages long. The ISO/IEC 17025 compliant certificate was seven pages long and had five times as many data points collected, in addition to reporting the uncertainty information. After seeing this, I then understood, and agreed with, the higher price for the calibration. By the way, even though they were Accredited to ISO/IEC 17025 for the scope of work, I still had to audit them because of our clients that did not recognize ISO/IEC 17025 in their industries.

Wes R.
 
Thread starter Similar threads Forum Replies Date
M Transferring ISO 17025 from one company to another ISO 17025 related Discussions 1
B ISO 17025:2017 risk management Risk Management Principles and Generic Guidelines 0
K Integrating ISO 9001:2015 with ISO 17025:2017 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
X ISO 17025 certification for Laboratory for online gambling products ISO 17025 related Discussions 3
R ISO 17025 - ISO Guides 33 & 80 ISO 17025 related Discussions 1
T ISO 17025: Lockout/Tagout Requirement ISO 17025 related Discussions 1
R ISO 17025 vertical audit checklist wanted Document Control Systems, Procedures, Forms and Templates 2
A ISO 17025 - Methods validation and clients ISO 17025 related Discussions 3
S Thoughts on managing ISO 9001, 13485, IATF 16949 and 17025 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 33
T ISO 17025:2017 requirement 5.7.b. about maintenance the integrity of the management system ISO 17025 related Discussions 1
R Who is the customer in the ISO/IEC 17025:2017? ISO 17025 related Discussions 1
C Compliance with ISO 17025 requirement 8.4.2 - Controls - Records recovery ISO 17025 related Discussions 4
G ISO 17025-2017 Management Review reporting items - Inputs ISO 17025 related Discussions 1
S High voltage testing - ISO 17025 - 7.2.2 Validation of methods and 7.3 Sampling ISO 17025 related Discussions 3
S Internal calibrations - Part of an ISO 17025 accredited testing laboratory (Automotive) ISO 17025 related Discussions 16
I Approved Suppliers ISO/IEC 17025:2017 and used test equipment ISO 17025 related Discussions 6
W Issuing reports under ISO 17025:2017 ISO 17025 related Discussions 2
S The (E) in ISO/IEC 17025:2017(E) ISO 17025 related Discussions 3
T ISO 17025:2017 Clause 4.2.2 - The difference between "be notified" and "be informed" ISO 17025 related Discussions 4
G ISO 17025:2017 Uninterrupted chain of traceability for test laboratories ISO 17025 related Discussions 6
DuncanGibbons Who are ISO/IEC 17065 and 17025 applicable to? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
V IS/ISO/IEC 17025:2017 Clause 7, sub clause 7.11 Control of data and information management ISO 17025 related Discussions 1
V IS/ISO/IEC 17025:2017 Clause 4.1 Impartiality ISO 17025 related Discussions 3
F ISO 17025:2017 Clause 7.7 Ensuring the validity of results - Threshold ISO 17025 related Discussions 9
A ISO 17025 Requirement 6.2 - Department Manager Testing Qualifications ISO 17025 related Discussions 1
F ISO 17025 8.6.2 Customer Feedback Analysis ISO 17025 related Discussions 5
F ISO 17025 8.8 Internal Audits in a segmented company ISO 17025 related Discussions 5
F Faulty Equipment and ISO 17025 General Measurement Device and Calibration Topics 1
F Online courses - Metrology, standardization, calibration and ISO 17025 Training - Internal, External, Online and Distance Learning 3
F Proficiency Testing for ISO 17025: AC Current Ranges ISO 17025 related Discussions 1
F ISO 17025:2017 Format for Procedure and Records Document Control Systems, Procedures, Forms and Templates 3
F ISO 17025:2017 Clause 7.4.1 - Requirements for Procedures ISO 17025 related Discussions 4
F Complying with ISO 17025 by UNIDO ISO 17025 related Discussions 0
F Measurement Audit and ILC for ISO 17025 Clause 7.7.2 - Comparison with results of other laboratories ISO 17025 related Discussions 0
F ISO 17025:2017 mandatory documentation requirements ISO 17025 related Discussions 1
F Linking an ISO 31000 Risk management SOP to ISO 17025 ISO 17025 related Discussions 2
K ISO 17025:2017 clause 7.6.2 - Performing calibration of its own equipment shall evaluate the measurement uncertainty ISO 17025 related Discussions 6
C ISO 17025 2017, Requirement 6.6.3 - Communicate requirements to external providers ISO 17025 related Discussions 4
D Risk Assessment Procedure in accordance with ISO 17025:2017 ISO 17025 related Discussions 5
A Validation of Calibration Methods and Traceability - ISO 17025 ISO 17025 related Discussions 3
J ISO 17025 - Please share a sample procedure for monitoring the validity of results ISO 17025 related Discussions 1
J ISO 17025 Documented Procedure for 6.2.5 - Determining competency ISO 17025 related Discussions 4
G ISO 17025:2017 7.1.2 - Should I produce a document for the customer? ISO 17025 related Discussions 8
G ISO 9001 and ISO 17025 - Measuring tools not in calibration system General Measurement Device and Calibration Topics 3
S Can anybody share a sample risk assessment prepared based on ISO 17025:2017? ISO 17025 related Discussions 15
U ISO 17025:2017 Clause 7.7 Ensuring the Validity of Calibration Results ISO 17025 related Discussions 6
G From an ISO 17025 auditor perspective must micrometer calibration check anvil flatness? General Measurement Device and Calibration Topics 4
S ISO 17025 Requirements for Mobile On-site Calibration ISO 17025 related Discussions 2
C Must your reference standard provider be ISO17034 certified to meet your testing lab's ISO 17025 certification requirements? Other ISO and International Standards and European Regulations 2
S The procedure for replicate check in calibration lab as per ISO 17025 General Measurement Device and Calibration Topics 2