The FDA identifies "Product Performance Qualification" which includes functional testing. Functional requirements are considered design inputs. Sometimes it is not realistic to identify all design outputs needed to relate back to your design inputs. For example: if you have bond strength of 3 lbs. as your design input, we don't measure the dimensions, adhesive properties and surface finish needed to guarantee a good part, we just do the pull-test.

Therefore when you perform product performance qualification as part of process validation, are you really doing design verification since you are verifying the design inputs?:confused:
 
Elsmar Forum Sponsor

yodon

Leader
Super Moderator
#2
We have, for one client in the past, used process validation results to verify design requirements. No question was raised by the FDA. We showed the linkage from design to test, showed the test had pre-approved acceptance criteria and recorded the results.
 

Randy

Super Moderator
#3
Wow, you have asked a good question, and seeing that I just finished creating an 8-hour training course on Design Verification & Validation (per ISO 9001, 7.3) that I have to deliver next week I'll wade in.

Based upon my extensive research (lots of hours googling and from responses supplied by many of my friends here) I fell comfortable saying "YES"!:yes::agree1:

Verification and validation are at times used synonymously and often the two "processes" may overlap so much that you can't hardly tell one from another. What you have described is one of those cases or areas of overlap...we can't say the design really meets the specifications until we actually test it in operation.
 
R

Roland Cooke

#4
Never allow arbitrary/semi-arbitrary definitions to get in the way of logic and common-sense.

BUT

as with everything, it is key to not only be compliant, but to be be able to sell the fact that you are compliant (not always the same thing!)
 
Thread starter Similar threads Forum Replies Date
G Mfr. Process Validation BEFORE Design Transfer? Other Medical Device and Orthopedic Related Topics 1
I Are suppliers required to hand over process validation reports? ISO 13485:2016 - Medical Device Quality Management Systems 20
S Performance Qualification and Process Validation ISO 13485:2016 - Medical Device Quality Management Systems 5
D Machine rebuilds versus process re-validation IATF 16949 - Automotive Quality Systems Standard 1
G Process Validation Before/After Sterilization? Design and Development of Products and Processes 3
B Vision system process validation Manufacturing and Related Processes 1
B SPECIAL PROCESS VALIDATION & REVALIDATION Qualification and Validation (including 21 CFR Part 11) 5
B How to satisfy clause 5.7.1.5 process validation for valve production API 6D Oil and Gas Industry Standards and Regulations 13
Melissa Process Validation of Rotary Heat Sealer Speeds Design and Development of Products and Processes 4
C Validation of process for releasing the UDI EU Medical Device Regulations 4
C Validation of process for production and servicing 5.7.1.5 API Spec Q1, 9th Edition Oil and Gas Industry Standards and Regulations 9
mustomutlu Process Validation Final Report Other Medical Device and Orthopedic Related Topics 2
N Sterilization Protocol Change in Validation Process and further impacts ISO 13485:2016 - Medical Device Quality Management Systems 1
R PCBA process validation Qualification and Validation (including 21 CFR Part 11) 2
S Sterilization validation after changing sterilization process provider Qualification and Validation (including 21 CFR Part 11) 3
D Questions regarding process validation ISO 13485:2016 - Medical Device Quality Management Systems 16
A Validation of Forced Aeration Process ISO 13485:2016 - Medical Device Quality Management Systems 3
T ISO 13485 - Process validation at critical suppliers ISO 13485:2016 - Medical Device Quality Management Systems 7
K Software Validation for Measurement Tools used in Process Validation ISO 13485:2016 - Medical Device Quality Management Systems 2
P Design verification driven by new equipment. How is this different than process validation? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
Stoic Are any medical device companies using the 2011 FDA process validation guidance instead of GHTF/SG3/N99-10:2004? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
G Devices from IQ, OQ or PQ process to be used for verification, validation and summative? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
N Design Verification & Process Validation - Statistical sample sizes Design and Development of Products and Processes 2
E Equipment Qualification - IQ/OQ per ISO 13485:2016 section 7.5.6 Process validation ISO 13485:2016 - Medical Device Quality Management Systems 7
B API Q1 5.7.1.5 Process Validation - Machining, heat treating, Manganese Phosphate and Zinc Plating Manufacturing and Related Processes 2
N Validation procedure, Major NC CAP - Equipment Validation process is not effective ISO 13485:2016 - Medical Device Quality Management Systems 5
D ISO13485:2016 7.5.6 Process Validation Responsibilities ISO 13485:2016 - Medical Device Quality Management Systems 1
Z Who is responsible for writing the validation of a new process? Manufacturing and Related Processes 7
M Requirement for manufacturing process validation in IATF 16949 IATF 16949 - Automotive Quality Systems Standard 8
E Sampling plan for orthopedic implant - Process Validation Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
Z Class 2 medical device - Process Validation - Test sterile or non-sterile units? Qualification and Validation (including 21 CFR Part 11) 9
T ISO 13485:2016 - Processes exempt from process validation ISO 13485:2016 - Medical Device Quality Management Systems 12
L Injection Moulding Process (Special Process) Validation Food Safety - ISO 22000, HACCP (21 CFR 120) 2
K Is Mold qualification / process validation always required? Manufacturing and Related Processes 7
R Process Validation on CNC Lathe and Milling machining process Design and Development of Products and Processes 1
Q Release of the first batch of a cleared product before process validation 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
R Process Validation sample size selection Statistical Analysis Tools, Techniques and SPC 0
P Auditing "process validation" process 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
E Cleaning Validation Process of Non-Sterile Implants and Instruments Other Medical Device Related Standards 2
W Design and Process Validation - Different? ISO 13485:2016 - Medical Device Quality Management Systems 6
A Supplier - Process Validation Manufacturing and Related Processes 3
M Medical Device Process validation, Validation of excel spreadsheets used for process Other Medical Device and Orthopedic Related Topics 4
V Tyvek packaging for Medical device process validation Qualification and Validation (including 21 CFR Part 11) 2
L Process Definition and Process Validation 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
S Can we sell units that were chosen for process validation? ISO 13485:2016 - Medical Device Quality Management Systems 6
S How do I do a Process Validation for soldering process Manufacturing and Related Processes 5
S Process validation when changing location ISO 13485:2016 - Medical Device Quality Management Systems 10
A Process Validation of QMS Software ISO 13485: 2016 Cl. 4.1.6 ISO 13485:2016 - Medical Device Quality Management Systems 26
P Root Cause for a Process Validation Batch Failure 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
GoSpeedRacer Validation of Processes 7.5.6 - Do cosmetic welds need process validation? ISO 13485:2016 - Medical Device Quality Management Systems 3

Similar threads

Top Bottom