Cannot create Rejects for TMV on a medical instrument with an LCD

S

SteHa

Hello All,
Hope I can get some suggestions.

In the process of developing a protocol for an attribute TMV on a medical instrument with an LCD. Part of the Final functional test includes a LCD test.

I've been informed that we are not able to create rejects and rarely come across them in testing.

What do others do in TMV situations where rejects cannot be created?

Thanks,
Steve
 

shimonv

Trusted Information Resource
Hi Steve,
Why do you need to create a reject? why not simply conform that you get the correct / expected output?

Shimon
 
S

SteHa

Hi Shimon,
We are required to perform a Test Method Validation so that we have objective evidence that the test method is suitable for its intended purpose and capable of producing valid results.

Rejectable parts are a requirement to demonstrate that the test method can actually detect rejects. If the test method cannot consistently detect rejects we will have type 2 errors in the field which is not good for the patients or healthcare provider.

Thanks
 

shimonv

Trusted Information Resource
I see. I'm afraid it's is not feasible to find a solution without detailed information on the requirement, the risks and what exactly are you trying to test and when exactly do you do the test.

Shimon
 

yodon

Leader
Super Moderator
You mention that you did have some rejects in the past - don't presume those are available for use in testing?

What is the nature of the test? Visual? Electronic? Are there ways you might be able to "fake" a reject without actually create an actual reject?
 
S

SteHa

Hi Again,
Although finding a solution for my specific situation would be great, it would also be valuable to hear if anyone else has run into this type of problem with TMV.

I would be surprised if this has never been encountered before.

Thanks,
Steve
 
S

SteHa

You mention that you did have some rejects in the past - don't presume those are available for use in testing?

What is the nature of the test? Visual? Electronic? Are there ways you might be able to "fake" a reject without actually create an actual reject?

Hi Yodon,
The test is visual. Looking for brightness changes and missing segments.

I believe they encountered a couple of failures many years ago but those units are no longer available.

The FW drives the segments lighting up and we do not have access to the FW to change it.

Tomorrow I will be getting a walk through on how the test is conducted and hopefully I can work with the tech to see if there are any other ideas on creating some type of failure.

Thanks,
Steve
 

Ronen E

Problem Solver
Moderator
If such failures are so rare in the field, based on a risk-management approach such testing is most likely redundant and represents inefficiency. I would document such risk analysis; remove the test; and suffice with monitoring and trending (the latter only if necessary) until otherwise indicated.

QC means finding and eliminating bad parts after they’ve been made —> waste.
QA means making sure bad parts won’t be made in the first place.
It seems that you have effective QA in that respect, and thus QC is much less necessary (and somewhat of a waste in itself). Congratulations!

Cheers,
Ronen.
 

normzone

Trusted Information Resource
I have a bad idea - if the failure mode is one that can be manually induced, you could damage a piece of product accordingly for a test piece.

I could be way off base, but the flaw mode reminds me of damages done to LCDs by pressure, causing bright spots, and by pixel failures, causing .... well, little dots of unintended color or no color.
 
S

SteHa

Hi Ronen,
I appreciate your insight. Having very good quality is a great problem to have!

I've learned more today and there is the possibility to create some of the failure modes to run the TMV. We are also going to work with the contract manufacturer of the instrument to see if they have any fall out for LCDs in areas that are harder to create. They might be seeing more than we do.

Thanks
Steve
 
Thread starter Similar threads Forum Replies Date
C Document Control using Wiki - Cannot Create all Documents as Wiki Pages Document Control Systems, Procedures, Forms and Templates 10
D How to address the content deviation of 'cannot apply criteria of risk acceptability prior to...' ISO 14971 - Medical Device Risk Management 1
K END CALIBRATION CANNOT BE PERFORMED Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 2
supadrai Possible to Submit a 510(k) if you have never, do not and cannot produce the device for which you're submitting the application? US Food and Drug Administration (FDA) 4
I Cannot Export multiple sections from Excel to PDF without losing pagination Excel .xls Spreadsheet Templates and Tools 23
L How to evaluate the process capability of a data set that is non-normal (cannot be transformed and does not fit any known distribution)? Capability, Accuracy and Stability - Processes, Machines, etc. 14
BradM Cannot stay logged in Elsmar Cove Forum Suggestions, Complaints, Problems and Bug Reports 16
K What is supposed to happen when a CA cannot be closed out or the root cause... General Auditing Discussions 1
insect warfare Cannot determine what type of medical device this product is.... 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
K Anyone who audits 8.2.2 cannot audit other clauses of the standard? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 31
rob73 If NB's (notified bodies) cannot be trusted how can we? EU Medical Device Regulations 39
K Corrective Action cannot be completed within the agreed time frame Nonconformance and Corrective Action 7
X Cannot Verify until Customer uses Product - Clause 7.5.2 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
L Preventive Action should not / cannot be taken when rectifying a problem ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 94
W What can/cannot be placed on packaging/advertising relating to ISO/TS? Registrars and Notified Bodies 1
M 'Searched' virus? Cannot delete folder - HELP please After Work and Weekend Discussion Topics 33
A Changes to QRI (Quick Reference Instruction)? What I can or cannot change ISO 13485:2016 - Medical Device Quality Management Systems 2
T Control Chart on Paper vs. Software where trend cannot be seen (Alarm only) Statistical Analysis Tools, Techniques and SPC 8
J Is it true that the Quality Manual cannot be a Procedure? Quality Management System (QMS) Manuals 13
C What can I do if I cannot meet PPAP deadline? APQP and PPAP 4
D New thread ring cannot be threaded onto the truncated part of the master General Measurement Device and Calibration Topics 4
E MSA Manual says the sum of EV,AV,GRR, and PV cannot equal 100%? Why? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
D Disaster Recovery Plan - I cannot find references to same in ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 3
C What is the best practice to label small tools (small pins) which cannot be scribed? General Measurement Device and Calibration Topics 5
A Goals cannot be reached, don't adjust the goals, adjust the action steps Coffee Break and Water Cooler Discussions 8
Casana Auditor changing ISO 9001 AUDIT scope midstream - Cannot exclude all of 7.3 anymore General Auditing Discussions 26
V MINITAB and non-normal unilateral tolerance data - Cannot confirm Ppk values Using Minitab Software 15
M ISO 9001 Clause 7.6 Measuring Devices cannot be excluded!? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 48
R 17025 Calibration Requirements - External calibration service cannot calibrate ISO 17025 related Discussions 4
T Some Suppliers to TS 16949 companies told they cannot register to TS 16949 IATF 16949 - Automotive Quality Systems Standard 4
Casana Distributors cannot convert to TS? IATF 16949 - Automotive Quality Systems Standard 2
L I looking for a book, but cannot remember the name? . . . . HELP! Book, Video, Blog and Web Site Reviews and Recommendations 3
T Scope and Purpose - For some reason cannot seem to write anything - ISO 9001 - 7.2.2 Document Control Systems, Procedures, Forms and Templates 8
M Gage R&R and Capability - Between 10% and 30% and cannot be improved Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
Q Nonconformance vs. Nonconformity - I cannot find major and minor, either! Nonconformance and Corrective Action 36
A Matrix to determine what's needed to create a QMS v ISO 9001 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
Q Looking to create a new job title Misc. Quality Assurance and Business Systems Related Topics 15
S Do we have to create a new PPAP to our customer if our Sub Contractor had to change? APQP and PPAP 10
N 8.3.3.3 Special characteristics - Auditor said we are required to create our own, if no customer char. exist. IATF 16949 - Automotive Quality Systems Standard 10
V How many hrs does it take to create a Technical File (TD)? EU Medical Device Regulations 4
D ISO 13485 - 7.3.6 Design and development verification - Do most folks create a separate SOP? ISO 13485:2016 - Medical Device Quality Management Systems 6
J Question: How to create an IMDS RoHS, REACH, ELV, IMDS and Restricted Substances 3
M Do we need to create a new CER or can we just update the existing CER EU Medical Device Regulations 3
J Create your own symbol? Other Medical Device Related Standards 7
D How to create IQ, OQ, PQ protocol for existing equipment Other Medical Device Related Standards 2
O How to create robustness checklist from p diagram FMEA and Control Plans 11
T Change control and configuration management - When to create a new model/part number? Other Medical Device and Orthopedic Related Topics 0
R How to create a Work Breakdown Structure? Registrars and Notified Bodies 2
H IMDS for material - Can my Customer create one for my material? RoHS, REACH, ELV, IMDS and Restricted Substances 4
A How do I create a 3^k factorial design with factors be treated as continuous where I Using Minitab Software 0

Similar threads

Top Bottom