CBE (Change Being Effected) 510(k) Submission

monoj mon

Quite Involved in Discussions
#1
Recently we have added one "Contraindication" to our IFU of a previously 510(k) cleared device. The FDA guidance "Deciding When to Submit a 510(k) for a Change to an Existing Device" suggest (in the Figure 2, flowchart A, Point A2) to submit a "change being effected (CBE) 510(k)". However, after searching for a while, I have found a very little or no information on the content/structure of a CBE 510(k) notification. It is something new type of submission for us and any information regarding what should be the content of CBE 510(k) would be highly appreciated. Besides "Updated Labeling" what else should be submitted? Does this submission require the submitter to pay the MDUFA Fee?
 
Elsmar Forum Sponsor

med_cert

Involved In Discussions
#2
For each significant change you need to submit 510(k) Application before you implement this change to your device/system/labeling. Once FDA approves this change, you can modify your device/labeling. It is different for contraindication.
CBE refers to the basic 510(k) application only that prior to 510(k) submission you can state this new indication to your official labeling for US. The regular fee for 510(k) have to be paied. The content of CBE 510(k) should be the same as regular 510(k). You are allowed to mark chapters that are the same as in previous 510(k) submission.
 

Ronen E

Problem Solver
Staff member
Moderator
#3
Just a thought - if a new contraindication is now deemed necessary without any other change to the device, that would represent narrowing of the allowed/recommended use of existing devices that might be on the market / in the supply chain. I would think the Manufacturer would need to consider in such a case any safety aspects and take appropriate action to modify or collect devices with the old labeling (without the added contraindication) or to at least inform customers and users.
 
Thread starter Similar threads Forum Replies Date
G Belgium - How do I get an enterprise CBE Number Other Medical Device Regulations World-Wide 0
I What revalidation is necessary if we change parts in a WFI loop ? Qualification and Validation (including 21 CFR Part 11) 1
Y Exporting data to the cloud is a "Significant Change"? EU Medical Device Regulations 5
A Design Change/ECO Related Question 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
S Reseller Request to Change UPC on Medical Device via Re-labeling Medical Device and FDA Regulations and Standards News 1
S Change management with FAA Federal Aviation Administration (FAA) Standards and Requirements 0
R Revalidation under VDmax25 following material change ISO 13485:2016 - Medical Device Quality Management Systems 2
D Change to labelling - does it require a new 510(k)? US Food and Drug Administration (FDA) 5
A How can we change a device trade name in 510(k) summary? US Food and Drug Administration (FDA) 0
L Change Notification Requirements for class 1 devices Medical Device and FDA Regulations and Standards News 1
R CB/Auditor Requiring a change in scope ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
Ed Panek Notified Body Substantial Change MDD EU Medical Device Regulations 1
A Purchasing Controls on Change Management of Literature/Labeling ISO 13485:2016 - Medical Device Quality Management Systems 3
A For software change - New Channel of interoperability CE Marking (Conformité Européene) / CB Scheme 5
JoCam IEC 60601-1 and 60601-1-2 retest after PCBA change IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
N ISO 13485 7.3.9 Change control in medical device software ISO 13485:2016 - Medical Device Quality Management Systems 6
A ISO 13485 procedure change and reflect to legacy manufacture items ISO 13485:2016 - Medical Device Quality Management Systems 2
R AS9102 FAI Change in Material / Process Supplier AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
D Reports under change management | ISO 13485:2016 & ISO 9001:2015 ISO 13485:2016 - Medical Device Quality Management Systems 3
Q Time change via change of Time Zone Qualification and Validation (including 21 CFR Part 11) 3
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Change in Constitution / Ownership of firm -------ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
U Power supply - Change From 160 to 300 Watts - Significant change? EU Medical Device Regulations 2
N Sterilization Protocol Change in Validation Process and further impacts ISO 13485:2016 - Medical Device Quality Management Systems 1
M Label change - 510(k) Medical Device and FDA Regulations and Standards News 5
M EMS change management ISO 14001:2015 Specific Discussions 2
E Document Change Request forms (or no forms)? ISO 13485:2016 - Medical Device Quality Management Systems 9
J Significant change related to design and intended use EU Medical Device Regulations 3
J Need Change Control Yes/No Decision Tree Template ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
S Cleaning process center change ISO 13485:2016 - Medical Device Quality Management Systems 4
U NOC - What is considered a "design change" EU Medical Device Regulations 5
qualprod What do CB´s change when doing a partial moving ? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
Y Algo change significancy Medical Information Technology, Medical Software and Health Informatics 3
Revision36 Engineering Change Order Industry Standards Other Medical Device and Orthopedic Related Topics 4
Z Iterative development and FDA change requests IEC 62304 - Medical Device Software Life Cycle Processes 13
D ECO (Engineering Change Order) process questions ISO 13485:2016 - Medical Device Quality Management Systems 7
R Medical Device - Change manufacturing plant Design and Development of Products and Processes 6
Tagin Template or Checklist for Process Change Document Control Systems, Procedures, Forms and Templates 5
R Do we need issue ECN (Engineering Change Notice) towards updated Material Specification? Design and Development of Products and Processes 2
A What does this line from MDCG 2020-3 (MDR art. 120 substantial change) mean to you? EU Medical Device Regulations 4
M Change management procedure when 7.3 is not applicable ISO 13485:2016 - Medical Device Quality Management Systems 5
D Change Approval Requirements - Does every change need formal customer approval? Design and Development of Products and Processes 17
Q Change the shape of the raw material FAI AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
D Engineering Change Order (ECO) metrics to track ISO 13485:2016 - Medical Device Quality Management Systems 6
P Notified Body Approval of Change EU Medical Device Regulations 2
F Change to Formulation due to Reach (China Medical Device Regulations) China Medical Device Regulations 0
E Change in control plan - Do I have to do sampling? IATF 16949 - Automotive Quality Systems Standard 1
S Forced ServiceNow validation - No change in our current user and functional requirements IT (Information Technology) Service Management 4
M Address change for a company with CE/ISO13485 EU Medical Device Regulations 2
G Need to change KPI we called NC parts (maximum 3%.) to FTQ (first time quality) IATF 16949 - Automotive Quality Systems Standard 4

Similar threads

Top Bottom