Hi all..

Quick question; I'm getting myself in a tizz...

Is it possible to have an IVD/Medical Device with both the CE mark AND the FDA mark on the same label?

Assuming you have met the necessary MEDDEV/IVDD/MDR/IVDR requirements for CE and relevant FDA requirements.

(it's been a long day and my head is about to implode; think I'm reading too much into this)

Huge thank you in advance :agree1:
 
Elsmar Forum Sponsor

yodon

Staff member
Super Moderator
#2
Um, there is no FDA mark.

You can have the CE mark on devices distributed in the US (presuming you have appropriate FDA clearance to market).
 

Ronen E

Problem Solver
Staff member
Moderator
#3
It is forbidden to make representations or imply that the device is approved by FDA or the likes, unless it was the subject of a successful PMA.

I usually advise to avoid any use of the FDA logo in device labelling.
 
Last edited:

fondantcookie

Starting to get Involved
#4
So IF a product has been through a successful PMA and has met the requirements of the MEDDEV, you can then technically put the CE mark and FDA logo on the same label? And market it within the EU and the US?

This isn't something we are looking to do I'm just enquiring from a knowledge perspective about how it all works.

Cheers
 

pkost

Trusted Information Resource
#5
you can have a single label that is suitable for both US and EU markets

The FDA will not stop you having a CE mark on your label; there is no equivalent US mark for the EU to object to.

I would avoid putting the FDA logo on your product (even for sale in the US); it is meaningless at best and a regulatory risk at worst
 

fondantcookie

Starting to get Involved
#6
Thanks for that :)

The reason I was getting in a tizz was because at my last company (based in the UK) we went through a successful FDA PMA and updated the labels to state "FDA" approved... I think more as a marketing decision rather than a requirement.
 

Ronen E

Problem Solver
Staff member
Moderator
#7
Thanks for that :)

The reason I was getting in a tizz was because at my last company (based in the UK) we went through a successful FDA PMA and updated the labels to state "FDA" approved... I think more as a marketing decision rather than a requirement.
I think that in that case it was OK to state "FDA approved" however I would still refrain from using the FDA logo in that statement. They just don't like manufacturers doing it (to say the least).
 

Ronen E

Problem Solver
Staff member
Moderator
#10
It is not OK to indicate "FDA approved" - this is considered misbranding according to FDA guidance. See HHS Publication FDA 89-4203 Labeling Regulatory Requirements for Medical Devices (go to the section starting on page 4 entitled False or Misleading Labeling, particularly the second paragraph).
That's a pretty old guidance. I think they've changed the rule for PMAs in particular, not so long ago actually, but I'll have to dig up a little to get the actual source. When you think of it, a PMA is actually an approval, while 510(k)s, IDEs etc. are not.
 
Last edited:
Thread starter Similar threads Forum Replies Date
T Signal Words for CE Mark and FDA Labels US Food and Drug Administration (FDA) 7
E How to get FDA Approval for Class II Medical Device - We have CE Mark, 13485, 9001 US Food and Drug Administration (FDA) 12
W Is it Legal for Printing CE Mark in Label? FDA Aspect 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 22
L CE Mark - FDA - Class I traction systems - What is involved and how long does it take EU Medical Device Regulations 5
K Two FDA questions regarding UDI (and potentially 803.52 MDR) Other US Medical Device Regulations 0
R Importers - For the FDA Registration/ Listing, is the CMO the importer? Medical Device and FDA Regulations and Standards News 6
Z Iterative development and FDA change requests IEC 62304 - Medical Device Software Life Cycle Processes 2
P FDA Approved Product Contact Parts ISO 13485:2016 - Medical Device Quality Management Systems 6
O Clarifying FDA definition of "finished device" and “capable of functioning” 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
A FDA guidance on non-sterile Medical Device Packaging Medical Device and FDA Regulations and Standards News 6
JoCam FDA Registration for Sub-contract manufacturers Medical Device and FDA Regulations and Standards News 2
M FDA News FDA Releases Draft Guidance Clarifying Application of ISO 10993-1 Biocompatibility Standard Medical Device and FDA Regulations and Standards News 0
M Is IEC 60601-1-2 required by FDA for all electronic medical devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
E Contract manufacturer FDA requirements foreign company US Food and Drug Administration (FDA) 6
P Choice of FDA reviewer US Food and Drug Administration (FDA) 6
J Is Device Tracking Card an actual requirement under FDA regulation? Other US Medical Device Regulations 0
T Definition Human Use (FDA) Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 7
K FDA 510k electrosurgery 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
J Leveraging FDA 510k Clearance for International Registrations Other Medical Device Regulations World-Wide 2
shimonv FDA News FDA guidance on Multiple Function Device Products (8/2020) Other US Medical Device Regulations 1
N FDA UDI - Label vs. Labeling - Does the insert need to include UDI? Other US Medical Device Regulations 1
J FDA notification of address change US Food and Drug Administration (FDA) 2
N Usability testing required for FDA IDE (investigational device exemption)? Human Factors and Ergonomics in Engineering 3
A FDA Class Classification for a cabinet 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
L FDA Biocompatibility Requirements - Transitory Contact Other US Medical Device Regulations 1
Edward Reesor FDA DM and /or Class I Life Saving US Food and Drug Administration (FDA) 0
Watchcat FDA will not tolerate fraud…meaning what, exactly? Other US Medical Device Regulations 6
K FDA Premarket Cybersecurity Guidance - 4 questions Other US Medical Device Regulations 5
gunnyshore Form FDA 3500A MedWatch eSubmitter 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
Watchcat FDA Webinar Series - N95 Respirators Other US Medical Device Regulations 0
M Off-Label Use - Clarification of FDA Policy US Food and Drug Administration (FDA) 1
J Sub-supplier change from manual to automated process - same specs - Report to FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
C FDA Establishment registration - Buying some medical devices from another manufacturer Medical Device and FDA Regulations and Standards News 5
I How to classify a medical device based on FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
dinaroxentool Question about FDA Classification of a Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
S The US FDA requirements on Disposal of a medical device US Food and Drug Administration (FDA) 1
S FDA Requirements for Medical Device Label Reconciliation 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
S What is considered a "core algorithm"? (From an FDA guidance document) Medical Information Technology, Medical Software and Health Informatics 4
B FDA-Medical Device Reporting (MDR )procedure compliant with 21CFR section 803 US Food and Drug Administration (FDA) 0
Stoic Are any medical device companies using the 2011 FDA process validation guidance instead of GHTF/SG3/N99-10:2004? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
D ISO 13485, FDA 21 CFR 820 and Auditing the Accounting Department ISO 13485:2016 - Medical Device Quality Management Systems 5
S FDA's E submitter - For clinical electronic thermometer Medical Device and FDA Regulations and Standards News 6
J FDA wants electrical safety testing on battery powered medical device US Food and Drug Administration (FDA) 11
A FDA and NB audit of Engineering Drawings in DHF and DMR. Medical Device and FDA Regulations and Standards News 1
D FDA De Novo Cover Letter - What is expected US Food and Drug Administration (FDA) 3
Ed Panek 21 CFR Part 820 - FDA Label Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
J Mislabeling - Consider this an FDA notified recall? CFR 806.10 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
O Hospital-Developed, In-House Device - FDA Clearance? Other US Medical Device Regulations 19
D FDA Approved Class II Medical Device? Other US Medical Device Regulations 11
R When does the FDA consider a component a medical device? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 17

Similar threads

Top Bottom