CE Classification of Phacoemulsification device
M
Medevice
#1
I cannot decide whether a phacoemulsification device is class IIa or IIb (MDD Rule 9). I think it is probably IIa but it could be IIb.
I have read the guidance documents and they are not clear whether occular contact is potentially hazardous given any hazard would result from misuse rather than a device characteristic.
Can anyone help me with a reference, or direct experience where a Notified Body has accepted one or other classisfication?
Thanks. :confused:
Elsmar Forum Sponsor
G
Gert Sorensen
#2
From your post it is not clear to me what your device does. But the rules that apply, as far as I can see, are:
2. Invasive devices
2.1. Rule 5
All invasive devices with respect to body orifices, other than surgically invasive devices and which are not intended for connection to an active medical device:
- are in Class I if they are intended for transient use,
- are in Cla!s lla if they are intended for short-term use, except if they are used in the oral cavity as far as the pharvnx, in an car canal up to the car drum or in a nasal cavity, in which case they are in Class I,
- are in Class IIb if they are intended for long-term use, except if they are used in the oral cavity as far as the phar).nx, in an ear canal up to the car drum or in a nasal caviry and are not liable to be absorbed by the mucous membrane, in which case they are in Class IIa.
All invasive devices with respect to body orifices, other than surgically invasive devices, intended for connection to an active medical device in Class lla or a higher class, are in Class IIa.
2.2. Rule 6
All surgically invasive devices intended for transient use are in Class lIa unless they arc:
- intended specifically to diagnose, monitor or correct a defect of the heart or of the central circulatory system through direct contact with these parts of the body, in which case they are in Class III,
- reusable surgical instruments, in which case they are in Class I,
- intended to supply energy in the form of ionizing radiation in which case they are in Class llb,
- intended to have a biological effect or to he wholly or mainly absorbed in which case they are in Class IIb,
- intended to administer medicines by means of a delivery system, if this is done in a manner that is potentially hazardous taking account of the mode of application, in which they are in Class lIb.
So, I'm guessing, yours would be a Class IIa
M
Medevice
#3
Phacoemulsification devices remove cataracts by the point administration of ultrasonic energy and generally by the concurrent low pressure suction removal of the emulsified material.
G
Gert Sorensen
#4
2.3. Rule 7
All surgically invasive devices intended for short-term use are in Class IIa unless they are intended:
- either specifically to diagnose, monitor or correct a defect of the heart or of the central circulatory system through direct contact with these parts of the body, in which case they are in Class III,
- or specifically for use in direct contact with the central nervous system, in which case they are in Class III,
- or to supply energy in the form of ionizing radiation in which case they are in Class IIb,
- or to have a biological effect or to be wholly or mainly absorbed in which cast they are in Class III,
- or to undergo chemical change in the body, except if the devices are placed in the teeth, or to administer medicines, in which case they are in Class lIb.
I'm guessing again: Ultrasonic is not the same as ionizing radiation? If so, then I do not believe rule 7 applies.
M
Medevice
#6
Yes Al, that was my initial ipinion, but I think Rule 9 may apply. The guidance document MEDDEV 2.4-1 specifically mentions "surgical ultrasound devices" as an example of Class IIb.
M
Medevice
#8
You are probably right Al. I have confirmed that it is IIb in Australia so it will probably be IIb in Europe as well.
Thread starter Similar threads Forum Replies Date
K Japan Classification of Phacoemulsification Device ISO 13485:2016 - Medical Device Quality Management Systems 0
K Question on MDR classification EU Medical Device Regulations 4
L Programmer accessories classification EU Medical Device Regulations 0
R ELISA reader - IVDR classification EU Medical Device Regulations 8
F FDA classification for a mobile app Medical Information Technology, Medical Software and Health Informatics 3
D Classification of product for clinical trials EU Medical Device Regulations 14
A IEC 62304 safety classification, External Controls and off-label use related risks IEC 62304 - Medical Device Software Life Cycle Processes 5
R Classification of USB DC operated medical equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
D Surgical Instrument Tray Classification EU Medical Device Regulations 7
S Classification of a product according to MDR EU Medical Device Regulations 3
L Classification under MDR EU Medical Device Regulations 1
shimonv Classification of a cloud- base viewer for the output from a medical device US Food and Drug Administration (FDA) 7
K Biological indicator system classification under MDR EU Medical Device Regulations 2
A FDA Class Classification for a cabinet 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
dinaroxentool Question about FDA Classification of a Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
N What is our product classification? (Does Unclassified classification still exists) Other US Medical Device Regulations 14
nadhar2 Classification of Action Items Misc. Quality Assurance and Business Systems Related Topics 3
A Class medical device (MDD) - Classification help EU Medical Device Regulations 1
D IEC 62304 Risk Classification - With and without hardware control IEC 62304 - Medical Device Software Life Cycle Processes 2
M Classification of Clean Rooms Other Medical Device Related Standards 9
A ISO 10993-7:2008/ Amd.1:2019 - Classification of special populations Other Medical Device Related Standards 3
DuncanGibbons Classification of aerospace parts depending on their risk and criticality etc. Federal Aviation Administration (FAA) Standards and Requirements 3
M Barrier cream classification EU Medical Device Regulations 2
N Medical Device Classification under MDR - Rule 21 EU Medical Device Regulations 11
M Risk Classification For Supplier - Clinical Research Organisation (CRO) Supply Chain Security Management Systems 3
D Classification of Syringe (nozzle) of needle free injection system. EU Medical Device Regulations 8
S CE Mark - Classification Confusion EU Medical Device Regulations 12
S Medical Device Product Classification in South Korea Other Medical Device Regulations World-Wide 1
M Informational US FDA – Ear, Nose, and Throat Devices; Classification of the Self-Fitting Air-Conduction Hearing Aid Medical Device and FDA Regulations and Standards News 0
D MDR 2017/745 Classification Guide EU Medical Device Regulations 8
M Informational EU – Guidance on Qualification and Classification of Software in Regulation (EU) 2017/745 – MDR and Regulation (EU) 2017/746 – IVDR Medical Device and FDA Regulations and Standards News 2
dgrainger Informational MDCG 2019-11: Guidance on Qualification and Classification of Software in Regulation (EU) 2017/745 – MDR and Regulation (EU) 2017/746 – IVDR Medical Device and FDA Regulations and Standards News 0
M Informational EU MDR Classification Rule 11 – what??? Medical Device and FDA Regulations and Standards News 9
M Informational US FDA Final Guidance – Acceptance Review for De Novo Classification Requests Medical Device and FDA Regulations and Standards News 1
A Fire hazard classification - Clause in IEC 60601 about gauze and electricity ISO 14971 - Medical Device Risk Management 0
J MDR Annex VIII, Rule 6 Classification - Implication for lower risk CV products? CE Marking (Conformité Européene) / CB Scheme 3
A Device Classification - India CDCSO - How to classify the device? Other Medical Device Related Standards 1
B ASTM F2924-14 /ASTM F3001 - Room temperature classification criteria Other US Medical Device Regulations 0
M Informational EU – MANUAL ON BORDERLINE AND CLASSIFICATION IN THE COMMUNITY REGULATORY FRAMEWORK FOR MEDICAL DEVICES Version 1.22 (05-2019) Medical Device and FDA Regulations and Standards News 2
E Wristwatch Skin-proximal glucose sensor - Classification under MDR EU Medical Device Regulations 4
K MEDDEV 2.4/1 rev. 9 Classification EU Medical Device Regulations 1
M Informational USFDA – Medical Devices; Anesthesiology Devices; Classification of the Ventilatory Electrical Impedance Tomograph Medical Device and FDA Regulations and Standards News 0
N Medical Device Re-Classification EU Medical Device Regulations 2
B Class IIB Device - IEC 62304 Software Classification IEC 62304 - Medical Device Software Life Cycle Processes 13
C Suppliers re-classification from Critical to Significant or from Significant to Non-critical Supply Chain Security Management Systems 0
J Medical Device re-classification and marketing ISO 13485:2016 - Medical Device Quality Management Systems 13
M APQC PCF (Process Classification Framework) and ISO 9001 - Processes Based Approach ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
M Informational TGA – Several proposed changes to classification to better align with the EU MDR Medical Device and FDA Regulations and Standards News 0
N Medical Device Accessory Classification - Software as a Medical Device Other Medical Device Regulations World-Wide 5
G EU MDR 2017/745 Rule 11 interpretation - Re-classification of a Software as Medical Device Other Medical Device Related Standards 0
Similar threads