CE Mark Approval Design Verification and Validation Questions

S

Sandeshgk7

Hello all,

I need a help regarding Design verification protocol

Right now we are working on a product for CE approval.

Our product have in-house manufactured and bought out CE approved parts, My questionis
1) While writing verification steps for the complex requirement of bought out parts is it acceptable to refer the user manual or product spec document and the qualify the requirement or do we really need to inspect the product to qualify the requirment requirement ?

2) Please share an example of validation and verification protocol?

Thanks in advance
 
S

Sandeshgk7

Hello Ronen E

Yes it is medical device, Radiography device in particular.
 

Ronen E

Problem Solver
Moderator
Hello all,

I need a help regarding Design verification protocol

Right now we are working on a product for CE approval.

Our product have in-house manufactured and bought out CE approved parts, My questionis
1) While writing verification steps for the complex requirement of bought out parts is it acceptable to refer the user manual or product spec document and the qualify the requirement or do we really need to inspect the product to qualify the requirment requirement ?

2) Please share an example of validation and verification protocol?

Thanks in advance

Hi,

If a requirement you are verifying is met by an off-the-shelf, CE marked product alone, without interaction with other parts, and that characteristic is also in that product's specification, I would say it would suffice to refer to the original product (component) specification in your verification. In all other cases you should probably run a test.

Cheers,
Ronen.
 
S

Sandeshgk7

Thank you for your valuable answer.

I have query regarding validation protocol

i am using below format to write the test cases for Design validation

Column No 1 User need ID
Column No 2 User need Statement
Column No 3 Verification method ( Inspection,test,demonstration etc)
Column No 4 Acceptance criteria
Column No 5 Verification Method
Column No 6 observed results


My question is

Do i need to write the detailed step by step Work Instructions for validating the user need in the column no 5 or can i take out that column ?

Really help full if you could share any example verification and validation protocol with similar information


Thanks in adavance
 

Ronen E

Problem Solver
Moderator
I have 3 comments:

1. Form doesn't matter as long as the essential information is communicated clearly.
2. I think you are confusing verification and validation.
3. Yes, you need to include detailed test instructions in the validation protocol (and also in verification protocols).
 

Mark Meer

Trusted Information Resource
Hi Sandeshgk7,

Verification and validation have been discussed a lot on this forum, but I'll try to summarize here as best I can with some simple examples.

Design Verification
Design verification activities should output documented evidence that design requirements are met. For example:
  • Design input requirement: "Device shall measure 4cm +/- 0.5cm in length"
  • Design verification plan: "Use calipers to measure the length of the device, Acceptance criteria: measurement is 4cm +/- 0.5cm"
  • Design verification output: "EUT Pass. Measured length: 4.1cm"

Design Validation
Design validation activities should output evidence that the product does what it's supposed to, typically involving representative subjects using the device in actual or simulated use conditions. For example (say the product is instant rice):
  • Product requirement: "Users should be able to follow instructions and cook the rice"
  • Validation plan: "Select X subjects, have them read and follow the instructions. Inspect the resulting rice to see how well it's cooked on a scale of 1 (uncooked) - 3 (perfectly cooked)".
  • Acceptance criteria: Average result is 2.75 or higher.

Hope this helps somewhat.
If you want more details, I'd suggest searching this forum for discussion threads on verification and validation.

Good luck!
MM
 

Ronen E

Problem Solver
Moderator
Mark's post above is a good intro. I'd like to add some comments (in no way intended to fully cover the subject):

1. Design verification is about comparing the design output (product specifications and actual products) with the design input (design engineering requirements), while design validation is about checking the final product against the user needs/requirements in a real-life or simulated environment.

2. Design verification is typically about demonstrating compliance while design validation is about proving compliance.

3. Both need to address statistical considerations. Protocols must specify sample size and should address sampling strategy and analysis methods. Item #2 above implies that validations require more rigorous statistical methods than verifications (this is true also because many times validations try to address more "fluid" or subjective requirements)

4. Validations should show that requirements will be consistently met - that means accounting for variability and noises in the environment (including users) and in the product itself.
 
Last edited:
S

Sandeshgk7

Hello Mark Meer

I have one more question for you.

As per out previous discussion we don't need to verify the requirements of the CE approved parts during verification, just refereeing to product specification is sufficient.

My question is

1) Do we really need validate the user needs of CE approved parts?

2) If we don't need to validate them ,then what needs to written in the validation method column ( Detailed work instruction) column? i,e How to write work instruction steps to say that product spec meets the specified user need?

Please use below examples.

1) Radiography generator shall be able to produce high frequency voltage.
2) Maximise the image resolution of the X-Ray.
 
Thread starter Similar threads Forum Replies Date
P Approval process for modifications to a medical device with a CE Mark CE Marking (Conformité Européene) / CB Scheme 5
JoCam Cost of maintaining CE Mark Regulatory Approval Other Medical Device Regulations World-Wide 3
E How to get FDA Approval for Class II Medical Device - We have CE Mark, 13485, 9001 US Food and Drug Administration (FDA) 12
J Class III Medical Device with CE Mark Approval - India Registration Process ISO 13485:2016 - Medical Device Quality Management Systems 10
C CE mark - Trying to determine if our product needs CE approval CE Marking (Conformité Européene) / CB Scheme 8
A CE Mark - How do you keep data and test reports? CE Marking (Conformité Européene) / CB Scheme 2
A CE IVD Mark - Emission certifications for unexpensive product CE Marking (Conformité Européene) / CB Scheme 1
A Positive Experiences with CE Mark Consultants? CE Marking (Conformité Européene) / CB Scheme 1
A CE mark symbol should be on the package? EU Medical Device Regulations 1
A Medical Device including a smartphone - CE mark ? CE Marking (Conformité Européene) / CB Scheme 4
V SaMD/MDSW - Class I - CE Mark CE Marking (Conformité Européene) / CB Scheme 1
H Is it a requirement for run charts to have inspection data or can it have just a pass/fail check mark Records and Data - Quality, Legal and Other Evidence 4
R Class I MDSW CE Mark CE Marking (Conformité Européene) / CB Scheme 6
A Class 2b medical device - best pathway to obtain CE mark? (2022, MDR 2017/745) CE Marking (Conformité Européene) / CB Scheme 2
A Class 2a CE Mark Regulatory Pathway CE Marking (Conformité Européene) / CB Scheme 1
A CE-Mark regulatory advice - US based Organization CE Marking (Conformité Européene) / CB Scheme 6
H Regarding EAC-MED mark Other Medical Device Regulations World-Wide 0
H Relation of VDE mark with CE certification CE Marking (Conformité Européene) / CB Scheme 13
B Spare parts of medical device need CE mark? EU Medical Device Regulations 0
K Not keeping CE mark for some devices CE Marking (Conformité Européene) / CB Scheme 1
H Regarding the validity of EMC & Safety test reports without NABL & ILAC-MRA mark. CE Marking (Conformité Européene) / CB Scheme 2
G Remove & Reapply Laser Mark on Steel Manufacturing and Related Processes 1
T Does marketing company require CE mark if manufacturer has CE mark on product? EU Medical Device Regulations 5
S Clinical trial requirements for UKCA mark in UK UK Medical Device Regulations 4
H Notified Body Number with CE Mark CE Marking (Conformité Européene) / CB Scheme 4
H Production Validation- CE Mark ISO 13485:2016 - Medical Device Quality Management Systems 3
Leigh76 CE Mark on Kit label AND inner label? EU Medical Device Regulations 1
T CE-IVD mark on packaging EU Medical Device Regulations 1
J Validity of CE mark on distributed product when company ceases trading EU Medical Device Regulations 3
T CE Marking - Bench mark Device from same manufacturer CE Marking (Conformité Européene) / CB Scheme 2
V New Product Model CE Mark Certification under MDR EU Medical Device Regulations 0
P Violation of CE mark - Re-use of single-use Products CE Marking (Conformité Européene) / CB Scheme 2
K One CE Mark holder - multiple manufacturing sites EU Medical Device Regulations 3
A First CE-mark class III implantable device under MDD EU Medical Device Regulations 8
F List of countries that 'prefer' a CE Mark CE Marking (Conformité Européene) / CB Scheme 1
C Surgical mask stability testing (CE mark) EU Medical Device Regulations 2
M MDR EU Distributor for our CE Mark product in the for Netherlands EU Medical Device Regulations 6
D Does Manufacture can submit CE mark application under MDD with NB for his New product after May 2020? EU Medical Device Regulations 4
Ed Panek CE Mark Requirements on Kits (off the shelf product requirements met) CE Marking (Conformité Européene) / CB Scheme 2
R IVDR CE Mark for Class A products - Declaration of Conformity CE Marking (Conformité Européene) / CB Scheme 2
G UVC sterilizer CE Mark CE Marking (Conformité Européene) / CB Scheme 5
C Does a CE mark infer meeting all applicable standards? CE Marking (Conformité Européene) / CB Scheme 4
M CE Mark and 510(k) -Global registration requirements for every country Other Medical Device Regulations World-Wide 3
A CE Mark and PMCF study EU Medical Device Regulations 3
T CE mark start date for a device EU Medical Device Regulations 7
P API monogram mark requirements Oil and Gas Industry Standards and Regulations 1
S CE Mark - Classification Confusion EU Medical Device Regulations 12
S Until when can we CE-mark and sell MDD class 1 devices? CE Marking (Conformité Européene) / CB Scheme 23
N CE Mark for Class 1 medical device CE Marking (Conformité Européene) / CB Scheme 11
Y CE Mark - Placing the CE mark on medical device labels EU Medical Device Regulations 2

Similar threads

Top Bottom