CE marked medical device - do they need UKAS certified calibration provider

#1
Hi Everyone,

I'm just new to the forum - looks amazing amount of information. Don't know its take so long find it :)

Background A company's medical device is CE marked and through an audit I conducted found that the calibration provider is not UKAS certified, however they do work to 17025 guidelines and have ISO9001 and showed me a GAP analysis to get them to 17025 - however its a few years old so I don't have confidence they are actively working toward it
So, my Question is: Is having a UKAS certification a requirement for this supplier since the medical device manufacturer has ISO13485 and ships devices world wide?

Any guidance much appreciated.

Now, Im off to read the forums :)
 
Elsmar Forum Sponsor

Ed Panek

QA RA Small Med Dev Company
Leader
Super Moderator
#2
I am assuming this is a supplier of yours.

We handle suppliers as follows:

The supplier risk category
Our ability to audit suppliers technically
Any mitigation provided by certifications (ISO 17025)

For example, if it was a very low risk supplier we might just do a survey on their QMS -Done

A more complicated supplier we could do a partial survey audit with a copy of their certification

A very critical and highly technical supplier we could perform an onsite audit and focus on key areas that control the Quality of the parts they supply for our product. We might also require ISO 13485/17025/9001 certification as well and a VERY detailed SQA.
 

Enghabashy

Quite Involved in Discussions
#3
Hi Everyone,

I'm just new to the forum - looks amazing amount of information. Don't know its take so long find it :)

Background A company's medical device is CE marked and through an audit I conducted found that the calibration provider is not UKAS certified, however they do work to 17025 guidelines and have ISO9001 and showed me a GAP analysis to get them to 17025 - however its a few years old so I don't have confidence they are actively working toward it
So, my Question is: Is having a UKAS certification a requirement for this supplier since the medical device manufacturer has ISO13485 and ships devices world wide?

Any guidance much appreciated.

Now, Im off to read the forums :)
if the accreditation is requirement of your management system ' therefore you should refer to UKAS web.; search about accredit bodies regarding Lab. calibration , you can find : 1- the valid list of current Labs. , 2- the withdraw list , --& so on ; the certificate of accreditation is not enough to check only but you must refer to accreditation scope as well for each lab. scope.
* remember that : 1- many labs have limited accredit scope but providing many service out of scope of accreditation ; 2- the calibration certificate issued from any lab. must reflect the accreditation number & the logo of accreditation ; otherwise if not'---, the certificate should be without the logo & the accreditation number , if you find any deviation ' you shall stop the relation & raise notification / complaint to UKAS.
 

dwperron

Trusted Information Resource
#4
Hi Everyone,

I'm just new to the forum - looks amazing amount of information. Don't know its take so long find it :)

Background A company's medical device is CE marked and through an audit I conducted found that the calibration provider is not UKAS certified, however they do work to 17025 guidelines and have ISO9001 and showed me a GAP analysis to get them to 17025 - however its a few years old so I don't have confidence they are actively working toward it
So, my Question is: Is having a UKAS certification a requirement for this supplier since the medical device manufacturer has ISO13485 and ships devices world wide?

Any guidance much appreciated.

Now, Im off to read the forums :)
OK, let's view the vagueries here.
ISO 13485 does not reference ISO 17025, there is no requirement there.
ISO 13485, when it comes to adequacy of calibrations in section 7.6, has a guidance note:
"NOTE Further information can be found in ISO 10012. "

So lets see what ISO 10012 has to say.
In section 6.4 Outside Suppliers, there is a Guidance note that reads:
"Guidance
If an outside supplier is used for testing or calibration, the supplier should be able to demonstrate technical
competence
to a laboratory standard such as ISO/IEC 17025. Products and services provided by outside
suppliers might require verification to the specified requirements.
"
The verbiage is a bit mushy here, with the "should be able to" and "might require" wording.
There is no statement requiring accreditation to ISO 17025, just that they should be competent like a 17025 laboratory.

Basically, my guidance would be that if you are using an ISO 17025 accredited laboratory you will be safe in audits.
If you choose some other laboratory you should be prepared to have to provide "verification to the specified requirements" of ISO 10012 on your own.
 
#5
Sorry it took me a while to come back and say thank you for the guidance. Our supplier should have the certification so I'm working with them on an implementation plan for this year.

Thanks everyone for the advice - very much appreciated.
 
Thread starter Similar threads Forum Replies Date
B Registration of a CE-marked medical device in Japan -Will they accept conformity with GSPRs? Japan Medical Device Regulations 0
S Medical Device Software Distribution for a CE-marked medical device via a USB memory stick EU Medical Device Regulations 8
L Can a hair transplant apparatus be CE-marked as medical device? EU Medical Device Regulations 4
F MDD Article 12 Labeling Requirements for CE Marked Medical Device - Help CE Marking (Conformité Européene) / CB Scheme 4
S Australian Registration of Non-CE Marked Medical Device CE Marking (Conformité Européene) / CB Scheme 2
S OBL CE Marked Medical Device Labels EU Medical Device Regulations 5
JoCam Validation of a change to a CE Marked Medical Device EU Medical Device Regulations 4
D Australian Registration of a CE Marked Medical Device Other Medical Device Regulations World-Wide 2
D Canada Medical Device - Licencing Process of a CE Marked Device Canada Medical Device Regulations 2
E Can I use CE marked Li-Ion external power bank for medical device? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
G Can a Distributer in USA sell a medical device (CE marked) to Europe? CE Marking (Conformité Européene) / CB Scheme 2
E Changes to IFU (Instructions for Use) on a CE Marked Medical Device ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
R Dialysis Machine is CE Marked as a Medical Device - Accessory or not? EU Medical Device Regulations 4
K Change Circuit Board to CE marked product on CE Marked Medical Device EU Medical Device Regulations 2
J CE Marking a CE marked Class I Medical Device - To aid approval/registration in China EU Medical Device Regulations 7
S Export of CE marked medical devices outside of Europe EU Medical Device Regulations 1
M Informational DSVG 04 – Breast Implants – Guidance on the vigilance system for CE-marked medical devices Medical Device and FDA Regulations and Standards News 0
M Informational EU – DSVG 03 – Cardiac Implantable Electronic Devices (CIED) – Guidance on the vigilance system for CE-marked medical devices Medical Device and FDA Regulations and Standards News 0
R Clinical validation/investigation in Egypt - Medical Devices sevices that are not marked yet Other Medical Device Regulations World-Wide 4
H IS there any database for list of CE marked Medical devices and there current status EU Medical Device Regulations 8
P Requirements for CE Marked Class I Medical Devices for EU and Canada ISO 13485:2016 - Medical Device Quality Management Systems 6
M Registration of Medical Devices (CE Marked) in Barbados Other Medical Device Regulations World-Wide 2
N Fast Track for Medical Devices Registration in Mexico that are already CE Marked Other Medical Device Regulations World-Wide 11
E Requirements for distribution of Class I and Class IIa CE Marked Medical Devices EU Medical Device Regulations 3
B Medical Devices that are CE marked and/or 510(k) - Registration in India? Other Medical Device Regulations World-Wide 21
R Essential info on the name place of a CE marked medical product EU Medical Device Regulations 1
N Non CE marked medical devices /clinical investigation EU Medical Device Regulations 2
A Sales in Europe of CE marked Medical devices with sister company EU Medical Device Regulations 11
R CE MARKED DEVICE REGISTER SEPARATELY (LOCALLY) IN EU COUNTRIES? CE Marking (Conformité Européene) / CB Scheme 2
D CE marked MD Notification to Spain & Italy EU Medical Device Regulations 4
JoCam Non CE marked device for customer review Other Medical Device Regulations World-Wide 0
R CE marked product sold in UK no EU rep CE Marking (Conformité Européene) / CB Scheme 2
M Selling non-CE marked devices for evaluation EU Medical Device Regulations 4
Q Understanding "free movement" for CE-marked products EU Medical Device Regulations 2
L Class IIb devices marked EC according to the 93/42/CE - Portugal CE Marking (Conformité Européene) / CB Scheme 3
A How to make sure whether the predicated device is marked with EC certificate? EU Medical Device Regulations 8
D CE Marked product considered a Drug in India Other Medical Device Regulations World-Wide 0
A Selling CE marked self-test in Malta - any additional requirements? EU Medical Device Regulations 0
K Addition of a test reader to existing CE marked product. Other ISO and International Standards and European Regulations 2
ernieto CE marked IVD - Does the "IVD" symbols needs to be in the IFUs? CE Marking (Conformité Européene) / CB Scheme 2
H IVD in US Phase III Clinical Trial - Non-CE Marked and Non-FDA Approved Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 3
T ISO13485 cert expired - Product still CE Marked CE Marking (Conformité Européene) / CB Scheme 1
S A CE Marked Product that does not comply with the Standard EU Medical Device Regulations 7
M EN ISO 15197:2015 - Glucose meters that were CE marked to EN ISO 2013 Other Medical Device Related Standards 2
Y How to distribute CE marked product for Japanese market? EU Medical Device Regulations 3
V Need of EU representative if CE marked Med Device not marketed in Europe EU Medical Device Regulations 8
P CE Marked IVD Labelling Canada - Conformity symbol? Canada Medical Device Regulations 6
V Dimension marked with "frame" on 2D drawing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
M Can a product be reviewed and CE Marked in accordance with ISO 15197:2013 EU Medical Device Regulations 2
Q Wiring Specs to Supplier - How wires are spooled, labeled, marked, etc. Manufacturing and Related Processes 3

Similar threads

Top Bottom