CE Marking for Power Supply for Medical Device - Class IIb device per MDD, Rule 11

V

vpanchal

#1
Hello ,

I am new to the Forum,, I find it very useful. Thanks for very useful information.

I have a Question regarding Power Supply. We manufacture infusion
pumps Classifed as Class IIb device per MDD, Rule 12. We use a Switch Mode
Power Supply ( AC Input and DC Output) from a Supplier who evaluates to
IEC 60601-1 Safety Standard and Marks it with CE Mark.
Since this Power Supply is used with a Medical Device, Does this Power Supply become an accessory with the Medical Device.?
Since Power SUpply is not non invasive device, Is it Classified as Class I device or Class IIb based on its usage with pump?

What type of Declaration of Conformity does the Power Supply
Manufacturer has to provide to us? Is it DOC to MDD or Low Voltage Device?

If its MDD, then Does the Power Supply Manufacturer has to prepare a
Technical File as well as the End Product User (Pump)?

Thanks

VP
 
Elsmar Forum Sponsor

Al Rosen

Staff member
Super Moderator
#2
Your device cannot be in class IIb according to Rule 12, since Rule 12 establishes Class I.
MDD said:
3.3. Rule 12
All other active devices are in Class I.
Please clarify.
 
Last edited:
V

vpanchal

#3
I apologize.. Rule 11.. All active devices intended to admnister and/or remove medicines. body liquids or other substances to or from body are in Class IIa, unless this is done in a manner:
-that is potentially hazardous, taking account of the nature of the substances involved, of the part of the body concerned and of the mode of application in which they are in Class IIb.

once again, I apologize for stating Rule 12..

Thanks.
 

Al Rosen

Staff member
Super Moderator
#4
If the PS mfr Has declared that the PS complies with IEC-60601-1 and is in compliance with the LVD it is sufficient since your device will comply with the MDD and this PS will be a component of it, not an accessory, and classification is not an issue. You will have a description of the PS and a DoC in your device's TF. I don't think you will or could have a TF for the PS since you do not have the details required.

I'm not sure I understand:
Since Power SUpply is not non invasive device, Is it Classified as Class I device or Class IIb based on its usage with pump?
But in any case, read the attached MEDDEVs, 2.1/1, [font=Arial, Helvetica, sans-serif]Definitions of "medical devices", "accessory" and "manufacturer"[/font] and 2.4/1, [font=Arial, Helvetica, sans-serif] Classification of MD.
[/font]

Discuss it with your NB for verification since I am basing this on my experience and application.
 

Attachments

Thread starter Similar threads Forum Replies Date
P Medical Device Marking - Power Input Ports IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R Power Input Marking requirements in IEC 60601 and IEC 60950 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
J Power switch requirement and on/off touch panel marking IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
V Is CE marking required for power resistors? CE Marking (Conformité Européene) / CB Scheme 1
S CE Marking components within IVD CE Marking (Conformité Européene) / CB Scheme 0
A Technical documentation ce marking pippete tips CE Marking (Conformité Européene) / CB Scheme 0
S Device with different options for a sub-component - CE marking implications EU Medical Device Regulations 2
R CE and UKCA marking Own Brand products CE Marking (Conformité Européene) / CB Scheme 0
C CE Marking - Medical Device Accessories EU Medical Device Regulations 0
J Using of CE marking - Two CE markings on the product CE Marking (Conformité Européene) / CB Scheme 5
T CE Marking - Bench mark Device from same manufacturer CE Marking (Conformité Européene) / CB Scheme 2
F Lithium polymer battery, venting and CE Marking CE Marking (Conformité Européene) / CB Scheme 4
A Migrating from whole product CE marking to modular approach CE Marking (Conformité Européene) / CB Scheme 5
H EU CE marking for Medical Device Class I EU Medical Device Regulations 2
K CE Marking Class 1 (Non sterile) medical device CE Marking (Conformité Européene) / CB Scheme 3
J Marking "Distributed By" to product with distributors' reference-code EU Medical Device Regulations 3
S CE marking for OEM Brand CE Marking (Conformité Européene) / CB Scheme 11
E Medical Device - CE marking - Local market notifications EU Medical Device Regulations 2
U Medical Device CE Marking - Using a disposable bearing CE Marking (Conformité Européene) / CB Scheme 3
K CE Marking for Class I Medical Device? CE Marking (Conformité Européene) / CB Scheme 8
Z Is IQ necessary for laser marking validation? EU Medical Device Regulations 3
A Is it mandatory to obtain ISO13485 certificate prior to CE marking? EU Medical Device Regulations 8
G CE marking of MD and IVD set EU Medical Device Regulations 3
M CE marking for NON-EU EU Medical Device Regulations 0
NDesouza Getting Rid of Part Marking Errors Benchmarking 40
S CE marking for a product upgrade CE Marking (Conformité Européene) / CB Scheme 0
P CE Marking UK Membership CE Marking (Conformité Européene) / CB Scheme 1
J Powered wheelchairs (Class IIa) in the process of CE marking and use in hospitals CE Marking (Conformité Européene) / CB Scheme 4
D CE Marking Requirements MDD & MDR - new product development covered under same scope EU Medical Device Regulations 1
N CE marking colour for Medical Devices CE Marking (Conformité Européene) / CB Scheme 4
A EAC Marking in Medical Device Other Medical Device Regulations World-Wide 0
B Marking of Medical Electrical equipment and accessories - Cl. 7.2.2 "Identification" and Cl. 7.2.4 "Accessories" IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M Clinical investigations of medical devices - not for CE marking EU Medical Device Regulations 0
P Marking Machinery for Israel CE Marking (Conformité Européene) / CB Scheme 3
K Does company who manufacture but does not design or carry out clinical trials is responsible for CE marking for single use medical devices? ISO 13485:2016 - Medical Device Quality Management Systems 3
M Does cFDA accepts FDA approval or CE marking? China Medical Device Regulations 1
I Reducing CE marking cost using manufacturer test reports CE Marking (Conformité Européene) / CB Scheme 5
C CE marking for general IVD (self-certified) & ISO 13485 QMS requirements - auditing EU Medical Device Regulations 6
Ed Panek Is Shelf Life Change a significant change? CE Marking CE Marking (Conformité Européene) / CB Scheme 7
CPhelan Using clinical trial safety data for evidence for CE marking EU Medical Device Regulations 8
N How to Write and Compile a Technical File for CE Marking CE Marking (Conformité Européene) / CB Scheme 4
CPhelan Do you require MDSAP for CE Marking of a Medical Device or is ISO13485:2016 with clinical data sufficient? CE Marking (Conformité Européene) / CB Scheme 6
P CE Marking EN ISO 13849-1: Electrical Panel behind interlocked doors CE Marking (Conformité Européene) / CB Scheme 9
M Cleaning Process after laser marking on orthopaedic implants Other Medical Device and Orthopedic Related Topics 3
H Safety and efficacy study for CE marking CE Marking (Conformité Européene) / CB Scheme 1
S Difficulty Finding A Notified Body for CE Marking - No Capacity Registrars and Notified Bodies 5
K Post Market Clinical Follow up and CE Marking for Class IIb medical X-Ray products CE Marking (Conformité Européene) / CB Scheme 1
J CE Marking for vibrator personal pleasure toy CE Marking (Conformité Européene) / CB Scheme 16
JoshuaFroud CE Marking and Brexit - Product shipped to a distributor prior to the March deadline EU Medical Device Regulations 2
M Informational Design for new product safety marking for the no-deal Brexit scenario Medical Device and FDA Regulations and Standards News 1

Similar threads