CE Marking on Multiple-Components Medical Products
H
hugbe
Hi Experts,
I would like to get a better idea on how to CE mark a medical product consisting of multiple components. As an example: a laser focusing hand piece consisting of a hand piece body, different focusing lenses, different sleeves matching in length with the focal length of the lenses.
Conformity assessment is performed with all the components. A final product is configured by the user. When he is ordering a focusing hand piece with a focal length of - let me say 20 mm and 40 mm, he will receive a ready assembled focusing hand piece and in addition an extra lense and a compatible sleeve. In the definition of the MDD the lense and sleeves are accessories and should be classified independent of the final product. However - what is the classification of a lense or a sleeve? And - in addition to it - what is about the marking? Shall all components carry a CE mark or just the Main Product?
Next Example:
A laparoscopic instrument which is dismountable for cleaning reasons. You may find an outer shaft, an inner shaft, a luer-lock connector and a closing cap for a luer connector. In addition you may have different insert tubings and a distal actuator. The user may use it with or without an inner shaft, may use different actuators and so on. How to mark the instrument and what is about the classification of the components?
Simple Question:
What is about a closing cap for a luer-lock connector? It is an accessory in my understanding and so would have to be classified separately? This seems very strange to me - and what is about the CE mark? Does it have to have a CE mark?
Thank you all for your qualified comments and proposals.
Bernhard
K
Krishna Priya
Hello experts,
I'm also having the same question with our product & accessories.
Experts help us to have the better understanding on this topic.
We have received the CE marking with Accessories which is tested with the main system. If we want to sell that accessories separately, whether we have to get different EC certificate to place CE mark for that accessory? in that case how we will classify the device?
Thanking you for the clarification.
With Regards,
Krishna Priya T
shimonv
There are far too many questions here, so lets get down to the basics:
In the MDD, Article 1.1 says:
"This Directive shall apply to medical devices and their accessories. For the purposes of this Directive, accessories shall be treated as medical devices in their own right. Both medical devices and accessories shall hereinafter be termed devices."
Article 1.2 defines the meaning of accessory:
"‘accessory’ means an article which whilst not being a device is intended specifically by its manufacturer to be used together with a device to enable it to be used in accordance with the use of the device intended by the manufacturer of the device;"
Let me point out two things:
1. Accessories are treated like medical devices. That means that they must bear the CE mark on them. If it's a class I accessory just the CE mark; if its a class II or above then the CE mark with the identification number of the notified body.
2. Your declaration of conformity is an important document for this purpose. It identifies the classification of your device and accessories. In most cases that I have seen the manufacturers have two declarations of conformity - one for the device with higher classification, and one for the accessories with lower classification (usually class I).
Chew on this and see if it makes sense to you.
Cheers,
Shimon
Ronen E
Problem Solver
Moderator
Before plunging into questions about how to handle Accessories, one needs to make sure that they're actually dealing with Accessories (in the official sense).
Q1: Who makes the so-called "accessory"? Their intentions are paramount in the determination of whether or not that item is officially an Accessory to a Medical Device.
Q2: Is the so-called accessory intended by its Manufacturer for a specific device (or a finite list of specific devices - not a type of devices)? If not, it's not an Accessory.
Q3: Is the so-called accessory intended by its Manufacturer to ENABLE the Device/Devices from Q2 to achive its / their intended use (as its / their own Manufacturer/s has/have called out)? If not, it's not an Accessory. Please note that "enable" means that it/they wouldn't be able to achive the intended use(s) without that item!
Is it actually an Accessory, or a Comoponent (or a Part)?
The official Accessory definition can be read neatly on some components / parts, too. Unfortunately, existing regulations provide little clarification or guidance to distinguish the two categories. Based on my experience and general knowledge of regulations and guidance, I've developed my own rule to tell Accessories from Components (or Parts):
A Component is an object that was already offered for sale as an integral part* of another medical device (excluding a medical device which is "a System or Procedure Pack" in the regulatory sense), and is also being offered for sale on its own, for use with the same device model; whilst an item that may be offered for sale for use with/in a medical device model, without previously being integrally included in any sale configuration of the same device model, may be** an Accessory.
*) "Integral part" doesn't necessarily mean that the object is assembled in/to the medical device; it only means that they are offered for sale as one unit, under the same PN, labelling, packaging etc.
**) To actually be an Accessory in the regulatory sense it'd have to satisfy additional conditions, as stated in the regulatory definition of an Accessory to a medical device - I didn't include these conditions in the above definition for readability reasons.
Challenges are welcome.
K
Krishna Priya
Hello Sirs,
Thank you.
Kindly help me on the below Queries.
1. In the Accessories, CE mark can be placed in accordance with the certificate which is provided for the main system along with the accessories?
2. Can we sell the Accessories separately with the CE mark - with the reference to the main system certifications?
shimonv
Hi Krishna,
I will try to answer your questions even though I have no knowledge about your product configuration. Take it with limited warranty :)
Generally speaking - yes. Accessories are really not much different than devices, expect for their supporting role. The MDD treats accessories like devices.
Yes, providing that the accessories are included in the system Technical File and covered in the Declaration of Conformity.
BR,
Shimon
Ronen E
Problem Solver
Moderator
In general, whatever item that is listed in an EC certificate (be it a medical device or an accessory to a medical device) can, and probably should, be CE marked and can be sold on its own.
Accessories to Medical Devices (in the official sense) that are not explicitly included in any EC certificate's scope - be it their "parent" device's certificate or their own - can't and shouldn't be CE marked or sold on their own.
The above applies to Accessories that are classified higher than class I non-sterile & non-measuring. Plain class I (NSNM) accessories require either being explicitly included in a parent-device EC certificate's scope, or being notified to (registered with) a member state's Competent Authority in their own right (like class I devices).
Spare parts (=Components) of a CE-marked device don't need to (and probably shouldn't) be CE marked, but they can still be sold on their own, as such - spare parts. Their packaging and accompanying paperwork should, however, clarify that these are components of/for duly CE-marked Medical Devices; in the presence of such clarification I would consider CE-marking their shipping/sales packages, as necessary. I wouldn't CE-mark the component itself, though.
T Multiple Applied Parts - Cardiac Floating (CF), Body Floating (BF) Marking EU Medical Device Regulations 3
H EU CE marking for Medical Device Class I EU Medical Device Regulations 2
K CE Marking Class 1 (Non sterile) medical device CE Marking (Conformité Européene) / CB Scheme 3
J Marking "Distributed By" to product with distributors' reference-code EU Medical Device Regulations 3
S CE marking for OEM Brand CE Marking (Conformité Européene) / CB Scheme 10
E Medical Device - CE marking - Local market notifications EU Medical Device Regulations 2
U Medical Device CE Marking - Using a disposable bearing CE Marking (Conformité Européene) / CB Scheme 3
K CE Marking for Class I Medical Device? CE Marking (Conformité Européene) / CB Scheme 8
Z Is IQ necessary for laser marking validation? EU Medical Device Regulations 3
A Is it mandatory to obtain ISO13485 certificate prior to CE marking? EU Medical Device Regulations 8
G CE marking of MD and IVD set EU Medical Device Regulations 3
M CE marking for NON-EU EU Medical Device Regulations 0
NDesouza Getting Rid of Part Marking Errors Benchmarking 40
S CE marking for a product upgrade CE Marking (Conformité Européene) / CB Scheme 0
P CE Marking UK Membership CE Marking (Conformité Européene) / CB Scheme 1
J Powered wheelchairs (Class IIa) in the process of CE marking and use in hospitals CE Marking (Conformité Européene) / CB Scheme 4
D CE Marking Requirements MDD & MDR - new product development covered under same scope EU Medical Device Regulations 1
N CE marking colour for Medical Devices CE Marking (Conformité Européene) / CB Scheme 4
A EAC Marking in Medical Device Other Medical Device Regulations World-Wide 0
B Marking of Medical Electrical equipment and accessories - Cl. 7.2.2 "Identification" and Cl. 7.2.4 "Accessories" IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M Clinical investigations of medical devices - not for CE marking EU Medical Device Regulations 0
P Marking Machinery for Israel CE Marking (Conformité Européene) / CB Scheme 3
K Does company who manufacture but does not design or carry out clinical trials is responsible for CE marking for single use medical devices? ISO 13485:2016 - Medical Device Quality Management Systems 3
M Does cFDA accepts FDA approval or CE marking? China Medical Device Regulations 1
I Reducing CE marking cost using manufacturer test reports CE Marking (Conformité Européene) / CB Scheme 5
C CE marking for general IVD (self-certified) & ISO 13485 QMS requirements - auditing EU Medical Device Regulations 6
Ed Panek Is Shelf Life Change a significant change? CE Marking CE Marking (Conformité Européene) / CB Scheme 7
CPhelan Using clinical trial safety data for evidence for CE marking EU Medical Device Regulations 8
N How to Write and Compile a Technical File for CE Marking CE Marking (Conformité Européene) / CB Scheme 4
CPhelan Do you require MDSAP for CE Marking of a Medical Device or is ISO13485:2016 with clinical data sufficient? CE Marking (Conformité Européene) / CB Scheme 6
P CE Marking EN ISO 13849-1: Electrical Panel behind interlocked doors CE Marking (Conformité Européene) / CB Scheme 9
M Cleaning Process after laser marking on orthopaedic implants Other Medical Device and Orthopedic Related Topics 3
H Safety and efficacy study for CE marking CE Marking (Conformité Européene) / CB Scheme 1
S Difficulty Finding A Notified Body for CE Marking - No Capacity Registrars and Notified Bodies 5
K Post Market Clinical Follow up and CE Marking for Class IIb medical X-Ray products CE Marking (Conformité Européene) / CB Scheme 1
J CE Marking for vibrator personal pleasure toy CE Marking (Conformité Européene) / CB Scheme 16
JoshuaFroud CE Marking and Brexit - Product shipped to a distributor prior to the March deadline EU Medical Device Regulations 2
M Informational Design for new product safety marking for the no-deal Brexit scenario Medical Device and FDA Regulations and Standards News 1
E Safety work instruction if the equipment have the CE Marking and a Manual with safety instruction Various Other Specifications, Standards, and related Requirements 16
Z Direct Part Marking (DPM) requirements for Class II Medical Devices Manufacturing and Related Processes 0
M Medical Device News FDA Guidance - UDI: Policy Regarding Compliance Dates for Class I and Unclassified Devices and Certain Devices Requiring Direct Marking Other US Medical Device Regulations 0
J Virtual Manufacturer and CE Marking EU Medical Device Regulations 8
N CE marking for distributed product CE Marking (Conformité Européene) / CB Scheme 1
S CE Marking requirements - A Medical device with different config sold under same name EU Medical Device Regulations 0
J Class 1 CE Marking - Can this be removed from the label on the product? CE Marking (Conformité Européene) / CB Scheme 4
I Tool Control Marking Voiding Warranty AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
C Data Matrix and DPM (direct part marking) UDI Standards - ISO/IEC TR 29158 Other US Medical Device Regulations 2
K UDI Direct Marking Compliance Date Clarification and one other UDI question Other US Medical Device Regulations 0
T CE Marking & Compliance to MDD, EMC, RoHS, and GPSD REACH and RoHS Conversations 4
J Installation Qualification (IQ) and Operational Qualification (OQ) for Laser Marking 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2