CE Marking Requirements MDD & MDR - new product development covered under same scope

#1
Hi
We are developing a new product, that will be covered under our existing CE Scope (the product is the same Classification - Class IIb), will our Product Technical File require a review by the Notified Body.
Also, our current CE is only just recently under MDD (Annex II w/o 4), however, will we need to apply MDR to this new product.

I've been getting conflicting information on various websites, so not entirely sure.
Thank you in advance.
 
Elsmar Forum Sponsor

Ronen E

Problem Solver
Moderator
#2
I've been getting conflicting information on various websites, so not entirely sure.
If your strategy is asking around online, how can you ever be sure?... Why place more (or any) importance on what we'd say here?...
The only authority on this should be your NB. You are an existing customer, so they *should* respond ("should" because if you are a small customer, currently only with an MDD scope (not MDR), they probbly have higher priorities).
will our Product Technical File require a review by the Notified Body.
Definitely; the only question is whether it should happen before placing the new device on the market, or as part of the next audit. Being a class IIb device and with the next audit not anytime soon (?), I'd say you need to be proactive and approach your NB, telling them you have a new device in the pipeline, and proceed according to their instructions.
Also, our current CE is only just recently under MDD (Annex II w/o 4), however, will we need to apply MDR to this new product.
I'll take a guess that regardless of how agile you and your NB are, it's likely to be too late for MDD certification; hence it'll probably be handled under the MDR (is your NB MDR designated?).
 
Thread starter Similar threads Forum Replies Date
I MDD: CE Marking Labeling Requirements - Two questions... CE Marking (Conformité Européene) / CB Scheme 7
N CE Marking MDD & EN 980 Labeling Requirements, IFU, Packaging, etc. EU Medical Device Regulations 6
D Direct-Marking Requirements under MDR 2017/745 EU Medical Device Regulations 3
M FDA Syringe Marking requirements US Food and Drug Administration (FDA) 4
C CE marking for general IVD (self-certified) & ISO 13485 QMS requirements - auditing EU Medical Device Regulations 6
Z Direct Part Marking (DPM) requirements for Class II Medical Devices Manufacturing and Related Processes 0
S CE Marking requirements - A Medical device with different config sold under same name EU Medical Device Regulations 0
J CE Marking Requirements and IEC 60617 - Electrical Schematics Symbols CE Marking (Conformité Européene) / CB Scheme 1
L Canada Electrical Equipment Marking Language Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M Natural Rubber / Latex Product CE Marking - Requirements EU Medical Device Regulations 3
I UDI requirements vs Japanese Barcode Marking guideline Japan Medical Device Regulations 1
E ASTM D3951 Part Marking Requirements AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
F Marking Requirements on Drawings (Prints) Document Control Systems, Procedures, Forms and Templates 1
F Softening Standard Test Requirements for CE Marking Other Medical Device Related Standards 4
M Content Requirements for a Declaration of Conformity (CE Marking) CE Marking (Conformité Européene) / CB Scheme 1
A Toy Company Doesn't Conform With CE Marking Requirements CE Marking (Conformité Européene) / CB Scheme 9
N CE Marking Requirements for Class I Medical Devices EU Medical Device Regulations 2
R Power Input Marking requirements in IEC 60601 and IEC 60950 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
C FAA/PMA Part Marking Requirements Federal Aviation Administration (FAA) Standards and Requirements 2
R Class II Medical Device Marking - Different IEC 60950 & IEC 60601 Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 17
J Marking (Labeling) Requirements on Medical Equipment per IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
D CSA Marking Requirements - IEC 60601-1 and CSA 622.2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
A CE Marking and Packaging Requirements for Class I Sterilized Medical Devices EU Medical Device Regulations 8
O Understanding Requirements for CE Marking and Test Requirements CE Marking (Conformité Européene) / CB Scheme 12
A CE Marking - Annex II Requirements for Class IIa Medical Device EU Medical Device Regulations 5
D RUO (Research Use Only - Laboratory Instrument) CE Marking Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
T Requirements for marking Laboratory Equipment Electrical Products Manufacturing and Related Processes 3
S CE Marking Requirements for a Product built from Embedded Kit Parts CE Marking (Conformité Européene) / CB Scheme 11
C Medical Device Reporting Procedures compliant with FDA and CE Marking Requirements EU Medical Device Regulations 7
I CE Marking Medical Devices - Country Requirements ISO 13485:2016 - Medical Device Quality Management Systems 9
S Advice Please regarding CE Marking requirements for Printing Press CE Marking (Conformité Européene) / CB Scheme 4
ScottK Requirements for CE Marking a component to go into an assembly (PED) CE Marking (Conformité Européene) / CB Scheme 2
G UKCA and EU MDR marking UK Medical Device Regulations 5
A Distributors with stock of expired CE marking EU Medical Device Regulations 2
R CE marking of products not sold into the EU CE Marking (Conformité Européene) / CB Scheme 4
J UKCA direct marking implementation UK Medical Device Regulations 0
Y CE marking with NB identification number EU Medical Device Regulations 0
B ASTM A53 Marking of length on recut pipes by processor Manufacturing and Related Processes 11
I Is SRN required for a contract manufacturer (CE-Marking product)? EU Medical Device Regulations 2
S CE Marking components within IVD CE Marking (Conformité Européene) / CB Scheme 0
A Technical documentation ce marking pippete tips CE Marking (Conformité Européene) / CB Scheme 0
S Device with different options for a sub-component - CE marking implications EU Medical Device Regulations 2
R CE and UKCA marking Own Brand products CE Marking (Conformité Européene) / CB Scheme 0
C CE Marking - Medical Device Accessories EU Medical Device Regulations 3
J Using of CE marking - Two CE markings on the product CE Marking (Conformité Européene) / CB Scheme 5
T CE Marking - Bench mark Device from same manufacturer CE Marking (Conformité Européene) / CB Scheme 2
F Lithium polymer battery, venting and CE Marking CE Marking (Conformité Européene) / CB Scheme 4
A Migrating from whole product CE marking to modular approach CE Marking (Conformité Européene) / CB Scheme 5
H EU CE marking for Medical Device Class I EU Medical Device Regulations 2
K CE Marking Class 1 (Non sterile) medical device CE Marking (Conformité Européene) / CB Scheme 3

Similar threads

Top Bottom