I am intrigued by this SAMM. Can you give me a little more information?
Hershal is one of the foremost experts (IMHO) on calibration. Let me encourage you to go to him name, and search through his last 100 or so posts. It will really be beneficial to you.
Most everyone else here is quite good too (except me, I'm just their agent:tg:). You have an appropriate standard, a decent procedure, the right numbers... You have a calibration. I hope you know what I mean. I just want to suggest a middle ground, reasonable approach. Calibration (and CoC) is not something everyone can do, and not something too difficult for anybody. Make sense.
As far as ISO 9001 is concern, the system is yours to decide. Basically, if it is an instrument utilized for impact, quality decision, measurement, etc., it will need a verification/ calibration, etc. to determine fitness for use. You will need to establish a program to ensure the equipment is adequately calibrated, and fit for use.
Please, I don't mean to come across too strong, so forgive me if it does. But... if you're company cannot afford a good calibration program, it needs to re-evaluate being in business (or at least going into an ISO initiative). If you are MFG. a product and you have instruments for quality assessment, IMO it's pretty important that they be accurate.
How many instruments are we talking about, and of what kind? I bet if you added it all up, you would save money sending it out for calibration. So I don't bore you do death, search on the Cove for internal calibration, and you'll find some detailed analysis by several people here.
Ok, enough of my rambling. I hope there's something here to help you.
Hello BradM,
I too, was intrigued by SAMM: Malaysian Department of Standards laboratory accreditation program, Skim Akreditasi Makmal Malaysia [SAMM].
Here are some links referring to SAMM:
http://www.standardsmalaysia.gov.my/SAMM_Leaflet 1 & 2 - Fields & Charges.pdf
http://www.agilent.com/metrology/samm_schedule.shtml
http://www.pyrometro.com.my/calibration.html
Hope this helps.
Stijloor.
Elsmar Forum Sponsor
U
Umang Vidyarthi
#12
Hi BradM & Madannc, thanks for your suggestions for carrying out in-house calibration and COC. Meanwhile, I have asked a ISO consultant for opinion on the In-house calibration and COC. She came back to me that we can have in-house calibration, but we cannot produce COC as we are not SAMM certified. In-house calibration merely means those MMDs are VERIFIED
FYI, our company is implementing ISO 9001:2000 QMS. My concern is --> is it all MMDs in the factory should be calibrated? Our company is small with around 10 ppl, we can't afford to send out all MMDs for calibration and have another spare set while MMDs sent out for calibration; we do in-house calibration. On the other hand, we sent only all MMDs used by QC Technician for external calibration.
Is this approach acceptable for ISO 9001:2000 QMS?
Your ISO consultant is absolutely right that you can not produce COC.
If you go back to the basics,the purpose of in-house calibration is to decrease the periodicity between two calibrations and increase the number of
calibrations during life-span of the tool,at lesser cost.This you have achieved.
ISO requirements are also mandatory.You will have to strike a balance between in-house calibration to be punctuated with external calibration through a certified agency.Thus you may achieve low cost while fulfilling ISO requirements as well.
In other words you may have your cake and eat it too!!;)
U
Umang Vidyarthi
#14
I just want to understand your statement. Are you saying this due to the SAMM requirement, or is there another thought regarding why you say a COC cannot be produced?
This is not the requirement of SAMM alone.There are plenty more on the globe
Point is ,any one accreditated by a recognised institution is authorised to issue a COC,no T-D-H is supposed to issue this document,'cause this will defeat the very purpose of forming the standards and/or proceedures.Now,
this is my understanding of the system,correct me if I'm wrong in my assumption. :agree: :cfingers:
Umang
BradM
Staff member
Admin
#15
Umang, I enjoy your posts and your input. I'm glad you're contributing.
I think I have mutual respect for your position, and you make a good point.
I would not say that a Certificate of Conformance or calibration is sole property of the Accredited labs. I think that internal calibration labs can (and legitimately do) issue these documents without accreditation. Internal organization A may need to use instrument 1, and requires the certificate with instrument 1 from internal organization 2 ( performing the test). Make sense? Also, I think performing internal calibrations should be done with excellence, and have good documentation for presentation.
However (big caveat, here) the documents need to be clear as to what is being reported, and what it is being reported against. It should be distinct what is done, or maybe what is not done. The information on the form may list an internal procedure, asset/ID number of instrument, uncertainty, as found/as left values, etc.
Does that seem plausible to you?
U
Umang Vidyarthi
#16
Brad,
Thanks for your appreciation.It's heartening to know that my contribution(however little) is found useful.
I do agree that Issuance of certificate should not be the sole property of Accreditated labs,but 1.what about the laid down proceedure by ISO? 2.And more importantly you need a governer or observer to certify that your internal
calibration is in line with the laid down proceedures.This is where the accreditated labs come into play.If everyone is efficient enough to control the standards,then there is no need for any standard et all.Your own internal calibration may be the 'Best' in the world,yet it needs to be certified by a neuteral body,and that's where & why accreditation is needed.
The 'big Caveat' you have stated is literally a big caveat.Who will supervise that the documents reporting for /against are correct?The information given therein is right or wrong?and so on.This is the sole reason for a neuteral body's involvement.
As for the plausibilty,yesss.A group of companies dealing within their domain,
can resort to a volley ball game,and yet maintain the quality better than those working overtime on International Standards.
But in such a case,how will you determine as to who is better or worse?
So I opine,that set standards(even mediocre) are better than 'Superb' non standards.
Looking forwards for your respectable views.
Umang:bigwave:
E
Esthyl - 2010
#17
BradM, sorry for confusing you. I didn't realize that SAMM is only used in M'sia, I thought it's as universal as the word 'ISO'. Stijloor, thanks for the search.
All these posts are helpful to me, that's true that many experts here...
Stijloor
Staff member
Super Moderator
#18
Hello Esthyl,
I want to thank you for asking and BradM for following up. Glad to do the "research", really no big deal. I learned a few things while doing this. Isn't this great? A win for all.
Stijloor.
BradM
Staff member
Admin
#19
Again, you make good points. I see you hold high passion for good metrology practices. My pat answer has become one highly similar to Hershal's, and many others. In general, find a qualified lab and have them perform the calibration.
However, the real world is just not simple. As far as ISO, they are international standards that require you do develop and maintain a program. Every company is required to develop an effective calibration program. That may involve internal calibration/external calibration, one or the other. It is perfectly acceptable to establish an internal calibration program in-house. ISO does not have a police squad. The organization must have a competent internal auditing program to determine that the established procedures are being followed and serving management's desires. Competent external auditors are usually pretty good at observing processes, and determining if the calibration program is effective and in compliance with organizational procedures.
Saying all this, I do fully believe in competent/ qualified labs and their service. I just want to be careful that we all don't fall into this "ISO mandates this or that" when in reality, it doesn't.
More times than not, unless you have greater than a sizeable number of instruments, it's worth it to send them to a competent lab. After adding everything up, it's not costing you any more than maintaining a competent internal lab.
Esthyl, I learn something everyday too. I did not know what SAMM is, and I had to learn.
To you and Umang, while basic components of good quality and metrology are universal, there are differences in industries, compliance, requirements, culture, etc. that may require subtle adjustments to practice. I really don't think Umang or I are right or wrong, we just have a different realm of experience and practice.
harry
Super Moderator
#20
Just to add on to Brad's post.
1. SAMM is the accreditation body in Malaysia. ISO 9001 only requires that calibrations be traceable to a National or international standard - something which any non accreditated labs or well equipped in house labs should be able to carry out (refer to post #8 by Hershal)
2. Non accreditated labs also issue calibration certificates. However, they cannot use the accreditation body's logo (SAMM) in their certificate. Therefore the issue that only accreditated labs can issue certificates is not not right. At least in Malaysia, SAMM has power to police their accreditated members but not the non-accreditated ones. Use of accreditated labs is of-course preferred but not a necessity.
3. In-house labs normally record their calibration results in some sort of forms but the contents are similar to certificates issued by external labs but as pointed out by Hershal, you need to be able to calibrated (manipulate the instrument and adjustment) plus determine uncertainties and have tracebilities to qualify as 'calibration' instead of verification. If you read through the calibration thread, you will find that even ISO 17025 certified labs sometimes fumble with the subject of determining uncertainties. In addition, you may need to prove the competency of your calibration technicians.
4. In the light of the above, I think it's cheaper for the original poster (Esthyl) to send their MDDs' for outside calibration. If you insist on in-house, you need to purchase the calibration blocks (with certificate traceable to relevant standards), have a lab with controls on temperature and humidity, etc and employ qualified technicians.
5. Esthyl had not come back to let us know what industry she's in. Normally, not all MDDs' need to be send out for calibration. Relevant advice will be given when she responds.
Thread starter Similar threads Forum Replies Date
J In-House Calibration Certificate wording - Confidence Level or Level of Confidence? Document Control Systems, Procedures, Forms and Templates 3
V Who should sign the Certificate of Attendance for In-House Training Training - Internal, External, Online and Distance Learning 7
M Medical device certificate in Australia - ARTG certificate Other Medical Device Regulations World-Wide 0
S Certification Body Changeover - Transfer of Certificate Registrars and Notified Bodies 9
J How to keep MDD certificate valid when legal manufacturer has liquidity problem EU Medical Device Regulations 0
D ISO 17021 Certificate Registrars and Notified Bodies 1
B Free Sales Certificate for Non Medical Devices Other Medical Device Related Standards 2
Louddogsbark When your 13485:2016 certificate has been pulled ISO 13485:2016 - Medical Device Quality Management Systems 2
lanley liao The validity of API Spec Q1 certificate Oil and Gas Industry Standards and Regulations 2
B Countries with no need for FSC (Free sales certificate) Other Medical Device Regulations World-Wide 0
M Calibration Certificate Result issued by an accredited external laboratory General Measurement Device and Calibration Topics 9
U Certificate of Free Sale Other US Medical Device Regulations 7
A Scope of ISO 13485 certificate ISO 13485:2016 - Medical Device Quality Management Systems 1
D ISO 9001 certificate issued by QMS International for 10 years - legit? Registrars and Notified Bodies 17
E Adding models to an approved Russian Registration Certificate Other Medical Device Regulations World-Wide 1
S Philippines CMDL (Certificate of Medical Device Listing) Elsmar Cove Forum ToS and Forum Policies 0
A Is it mandatory to obtain ISO13485 certificate prior to CE marking? EU Medical Device Regulations 8
R External Audit and Certificate prorogation due to the pandemic General Auditing Discussions 10
A How to make sure whether the predicated device is marked with EC certificate? EU Medical Device Regulations 8
I Is the contract mfg. mentioned in legal manufacturer EC certificate? CE Marking (Conformité Européene) / CB Scheme 4
SocalSurfer AS9100 new certificate, but need QMS software, help Quality Assurance and Compliance Software Tools and Solutions 2
C Certificate of Conformance Form - COC for each customer a controlled document? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
B Certificate Suspension? Shut down due to Covid-19 IATF 16949 - Automotive Quality Systems Standard 14
L Appears to be a certificate mill certificate ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
D ISO 9001 & AS9100 Certificate issued by a consultant for a Distributor Registrars and Notified Bodies 11
J CE Certificate and additional processing EU Medical Device Regulations 4
bryan willemot Chemical / physical certificate of conformance AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
B IATF certificate suspended - Who do we notify? IATF 16949 - Automotive Quality Systems Standard 6
F Calibration certificate content - Average Value and True Value ISO 17025 related Discussions 6
H What is different between PED certificate and CPR certificate? Manufacturing and Related Processes 2
C Migrate ISO 9001:2015 Certificate Registrars and Notified Bodies 4
M How to read a calibration certificate for thermo hygrometer General Measurement Device and Calibration Topics 2
V Calibration certificate verification Qualification and Validation (including 21 CFR Part 11) 3
A Calibration Certificate Results General Measurement Device and Calibration Topics 3
A What if Contract Manufacturers does not have an ISO 13485 certificate? Where will the NB audit take place, at legal mfg. site or contract mfg. site? Other Medical Device Regulations World-Wide 3
E Calibration Certificate Content Requirements General Measurement Device and Calibration Topics 3
M IATF 16949:2016 clause 8.4.2.3 - We don't have ISO 9001:2015 certificate IATF 16949 - Automotive Quality Systems Standard 26
S Boeing D6-51991 Product Acceptance Software certificate and Artefact? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
J COC (Certificate of Conformity) when NOT purchasing medical devices direct from the Manufacturer Other Medical Device Related Standards 2
P List of ISO certificate registrars around the world - ISO certification databases ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
F EC in EC certificate - Is it an EC Certificate or a CE certificate? CE Marking (Conformité Européene) / CB Scheme 3
M Accredited Calibration Sevice Provider using computerized system to issue calibration certificate Qualification and Validation (including 21 CFR Part 11) 3
K Inspection Certificate EN 10204 2.1 - Distribution company Other ISO and International Standards and European Regulations 4
M Accredited Calibration Sevice Provider using computerized system to issue calibration certificate General Measurement Device and Calibration Topics 2
R Certificate of compliance to RoHS/REACH/WEEE - any such animal? REACH and RoHS Conversations 18
D CB Certificate - Distributor for a product with electrical components CE Marking (Conformité Européene) / CB Scheme 4
P Non-Active Suppliers on CE Certificate EU Medical Device Regulations 2
A Interpretation of Article 16 (2b) - Packaging, samples and Certificate EU Medical Device Regulations 10
R List of countries that require only CE Mark certificate for product registration Other ISO and International Standards and European Regulations 2
K CFG (certificate of foreign government) Requestor Other US Medical Device Regulations 0
Similar threads
Top Bottom