Certification services for ISO 13485, CMDCAS 13485, and CE Markings

S

sabrina.qa

#1
I would like to apologize off the bat if this is terribly uninformed.

I've inherited a QMS with certifications for ISO 13485, CMDCAS 13485, and CE Markings.

If we are certified for CMDCAS 13485 anyway, which is ISO 13485 + CMDR requirements, is it necessary for us to also have the cert for the original flavour ISO 13485?
 
Elsmar Forum Sponsor
G

Gert Sorensen

#2
Re: Certification services

If you are located outside of the EU and provide products for the EU I believe that you will need a "clean" ISO 13485 certification. The Canada version CAN/CSA ISO 13485 is in terms of compliance not the same as the EN ISO 13485.
 
S

sabrina.qa

#3
Re: Certification services

From what I understand it is BSI EN ISO 13485 that we're certified to, not CAN/CSA ISO 13485, and our CMDCAS cert is because the QMS addresses the CMDR requirements.
 
Last edited by a moderator:
S

sabrina.qa

#5
Thanks WAF64

I hope this isn't me being dense

but I understood the EC Certificate (Directive 93/42/EEC, Annex V) was what I needed to sell in the EU. (1 certificate)

And then an ISO 13485 certificate issued by a CMDCAS recognised registrar to sell in Canada. (2nd certificate) (Does not specify CAN/CSA) (says SCC)

We have a 3rd certificate and I am (clearly) struggling with the need to have it. It is EN ISO 13485 (says UKAS) - who will refuse the SCC cert above if that is offered in lieu of this one?
 

Ronen E

Problem Solver
Moderator
#6
You should ask the NB who issued your EC certificate if they will accept the SCC one. Please also consider that you might need ISO 13485 certification for other markets.

Normally when being EC certified you get 2 certificates - one for EN ISO 13485 and the EC certificate. Usually they both originate from the same NB audit (but not necessarily).

Cheers,
Ronen.
 

WAF64

Starting to get Involved
#7
Sabrina,
if you want to sell medical devices in EU you are required to meet the quality system requirements outlined in the directive (Annex II and Annex V specifically), but how you are going to do so is left to you.
ISO 13485 is a voluntary tool available to you to meet the quality system requirement. However, although ISO 13485 is voluntary, it is actually considered “the standard” for companies that sell medical devices in Europe. If you are in compliance with ISO 13485 you are presumed to meet the quality system requirements of the MDD 93/42 EEC directive.

Finally, the conformity assessment route may vary based on your device class, so I strongly suggest you to contact your notified body to check out possible alternative conformity assessment routes.
 
S

sabrina.qa

#8
You should ask the NB who issued your EC certificate if they will accept the SCC one. Please also consider that you might need ISO 13485 certification for other markets.

Normally when being EC certified you get 2 certificates - one for EN ISO 13485 and the EC certificate. Usually they both originate from the same NB audit (but not necessarily).

Cheers,
Ronen.
Hi Ronen,

The same NB issued all 3 certs from the same audit.

My two ISO 13485 certificates look like this:

First one -

"the management system of [company/address] has been assessed and certified as meeting the requirements of

ISO 13485:2003
EN ISO 13485:2012

for the following activities

(activities)
(expiration)
(NB address) (UKAS stamp)"

Second one -

"the management system of [company/address] has been assessed and certified as meeting the requirements of

ISO 13485:2003

for the following activities

(identical activities)
(expiration)
(registrar address) (SCC stamp)
CMDCAS recognised registrar"


In what market would the second one be rejected?

It still seems to me that I should be able to satisfy all regulatory authorities with the EC certificate and the second ISO 13485 certificate. Two ISO 13485 certificates seem like overkill!
 

Ronen E

Problem Solver
Moderator
#10
Hi Ronen,

The same NB issued all 3 certs from the same audit.

My two ISO 13485 certificates look like this:

First one -

"the management system of [company/address] has been assessed and certified as meeting the requirements of

ISO 13485:2003
EN ISO 13485:2012

for the following activities

(activities)
(expiration)
(NB address) (UKAS stamp)"

Second one -

"the management system of [company/address] has been assessed and certified as meeting the requirements of

ISO 13485:2003

for the following activities

(identical activities)
(expiration)
(registrar address) (SCC stamp)
CMDCAS recognised registrar"


In what market would the second one be rejected?

It still seems to me that I should be able to satisfy all regulatory authorities with the EC certificate and the second ISO 13485 certificate. Two ISO 13485 certificates seem like overkill!
If they all came from the same audit and you didn't pay extra for the 3rd certificate, why would you care?
 
Thread starter Similar threads Forum Replies Date
N Registrars / Certification Bodies offering consultancy and training services? Registrars and Notified Bodies 17
G TS 16949 Certification for Burn-in services company for semiconductors? IATF 16949 - Automotive Quality Systems Standard 1
M Gravimetric blender for resin in automotive industry - periodical calibration via external lab with certification Reliability Analysis - Predictions, Testing and Standards 0
S ISO 13485 certification training Training - Internal, External, Online and Distance Learning 1
K Sales of MDD devices with expired CE that are pending MDR certification EU Medical Device Regulations 4
W Incoming Certification Requirements AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
B Multisite Certification Requirements for ISO 27001 IEC 27001 - Information Security Management Systems (ISMS) 8
PowerJunkie Cancelling AS9100 Certification - How and Who to notify? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
H Regarding UL Certification CE Marking (Conformité Européene) / CB Scheme 6
Q ISO 13485 certificate needed if I go for MDSAP certification? ISO 13485:2016 - Medical Device Quality Management Systems 3
S Eligibility for certification to IATF16949 IATF 16949 - Automotive Quality Systems Standard 8
M ASQ certification study material Professional Certifications and Degrees 1
A New to QMS and regulatory... advice on training/certification? ISO 13485:2016 - Medical Device Quality Management Systems 1
D Is IATF certification required when customer doesn't require it? IATF 16949 - Automotive Quality Systems Standard 19
A Loss of certification because of outdated information in the Clinical Evaluation Report CE Marking (Conformité Européene) / CB Scheme 8
Sidney Vianna IAQG News Down with OPMT - Long live CERTIFICATION OVERSIGHT AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
H Relation of VDE mark with CE certification CE Marking (Conformité Européene) / CB Scheme 11
T Adding to certification AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
T Is NRTL certification mandatory in the US ? Registrars and Notified Bodies 2
H EAC-MED Certification CE Marking (Conformité Européene) / CB Scheme 0
Aliken Recommendation for the ISO 13485 certification company ISO 13485:2016 - Medical Device Quality Management Systems 7
A List of documents to prepare for Re-certification ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
F Attendance possibility for a certification audit ISO 13485:2016 - Medical Device Quality Management Systems 7
V ADDING NEW MEDICAL DEVICE / Product, WHEATHER THIS AFFECTS EXISTING ISO 13485 CERTIFICATION? ISO 13485:2016 - Medical Device Quality Management Systems 4
K SGS Certification - Advice on their Performance Service Industry Specific Topics 9
V ISO 9001 or ISO 17065 required for an organization issuing 5 star rating certification of digital experience? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
P How to keep a ISO 9001:2015 certification when a company is sold ? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
A Scope of ISO 13485 certification as legal manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 4
J CMO in Canada do they need MDSAP/13485 certification? ISO 13485:2016 - Medical Device Quality Management Systems 1
R ISO 13485 certification scope ISO 13485:2016 - Medical Device Quality Management Systems 5
A AS9100D new certification cost?? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
N ISCC Plus - The international Sustainability Carbon Certification Sustainability, Green Initiatives and Ecology 2
S Which certification to get? Professional Certifications and Degrees 3
J Implementation, establishment and certification of IATF 16949. IATF 16949 - Automotive Quality Systems Standard 6
D Regulatory Affairs Certification (RAC) Devices Exam - Autumn 2022 study Professional Certifications and Degrees 3
A MDR Article 16 Certification of Distributors by a Notified Body? EU Medical Device Regulations 8
T NDT/NDE Level 2 Technicians Certification Oil and Gas Industry Standards and Regulations 20
M PSE Certification Japan Medical Device Regulations 5
C IATF16949 certification eligibility IATF 16949 - Automotive Quality Systems Standard 6
D IATF 16949 Certification and Letter of Conformance expiry (Help!!!) IATF 16949 - Automotive Quality Systems Standard 8
H API monogram 4F - Certification process Oil and Gas Industry Standards and Regulations 2
J Recommendations for certification bodies ISO 13485:2016 - Medical Device Quality Management Systems 4
S Certification of equipment after move General Measurement Device and Calibration Topics 14
J Help to understand and response to API AAR during the re-certification audit Oil and Gas Industry Standards and Regulations 17
G Medical Device Auditor (CMDA) certification exam by ASQ - looking for input Career and Occupation Discussions 3
D Automotive Customer asking for ISO 14001 Certification from suppliers ISO 14001:2015 Specific Discussions 3
I Foreign manufacturer registered place of business in EU for CE certification EU Medical Device Regulations 1
eldercare Multi-Site ISO/AS Certification Requirement for some sites ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 22
Casana ISO 17025 - Contractor certification? IATF 16949 - Automotive Quality Systems Standard 5
blackholequasar ISO 13485 certification prior to Medical Device Manufacturing... worth it? ISO 13485:2016 - Medical Device Quality Management Systems 4

Similar threads