CFR Title 21 820.72.b Calibration Procedure Accuracy and Precision Requirements
jrcook5
CFR Title 21 820.72.b states that "Calibration procedures shall include specific directions and limits for accuracy and precision." Is just stating +/- measurement tolerances adequate or is this statement to be taken literally and both must be indicated?
Al Rosen
Take it literally as with all regulatory requirements. Look at the FDA QSR Manual calibration section.
Al Rosen
jrcook5 said:
As an example, would we say, the device is 99.9988% accurate with 99.9990% precision? Or would it acceptable to say the device measurement uncertainty is +/-12 ppm?
In your previous post, you left out some of the stated requirement. For your purposes, I think the part not quoted is a key to this question. The subsection is:
Calibration. Calibration procedures shall include specific directions and limits for accuracy and precision. When accuracy and precision limits are not met, there shall be provisions for remedial action to reestablish the limits and to evaluate whether there was any adverse effect on the device's quality. These activities shall be documented.
And I have highlighted an important point of this section. You can decide if based on measurement uncertainty of the calibration, you can meet the requirement.
jrcook5
Al Rosen said:
In your previous post, you left out some of the stated requirement. For your purposes, I think the part not quoted is a key to this question. The subsection is:And I have highlighted an important point of this section. You can decide if based on measurement uncertainty of the calibration, you can meet the requirement.
I have read the regulation and the Quality Manual however, it is still unclear to me on what format the FDA is requiring. I work for a large organization and have been through a few FDA audits without any findings in the instrument control system. But some of our other facilities have been cited for not stating accuracy and precision separately.
I was interested if anyone has any experience dealing directly with this issue.
Thanks,
Jim
Chitchat
Registered Visitor
Re: CFR Title 21 820.72.b accuracy and precision
How do you demonstrate compliance to this requirement specifically in relation to precision?
"Calibration procedures shall include specific directions and limits for accuracy and precision. When accuracy and precision limits are not met, there shall be provisions for remedial action to reestablish the limits and to evaluate whether there was any adverse effect on the device's quality. These activities shall be documented."
According to the FDA QSR Manual, Equipment and Calibration “Precision has no unit of measure and only indicates a relative degree of repeatability, i.e., how closely the values within a series of replicate measurements agree with each other. Repeatability is the result of resolution and stability.”
What limits do you document for precision and how do you demonstrate a gage’s precision during routine calibrations?
normzone
I've stumbled into this issue and am looking for an example of language regarding this same issue - my example is basic mechanical measurement of a very simple part.
I don't want to reinvent the wheel if I can see a good wheel functioning and use it as a model for the wheel I need. Anybody point me to an extant document I could review?
As always, thanks for your help. This site is my second home.
:bonk:
dgriffith
According to the FDA QSR Manual, Equipment and Calibration ?Precision has no unit of measure and only indicates a relative degree of repeatability, i.e., how closely the values within a series of replicate measurements agree with each other. Repeatability is the result of resolution and stability.?
What limits do you document for precision and how do you demonstrate a gage?s precision during routine calibrations?
That document is withdrawn. Do you know what the replacement is?
normzone
I'm an ISO 9001 veteran but a novice at medical. I'll be reading the forum a lot, but any guidance would be gratefully accepted.
:bigwave:
Q Books that have a combination of ISO 13485 and FDA Title 21 CFR Part 820 ISO 13485:2016 - Medical Device Quality Management Systems 9
D CFR Title 14: Aeronautics and Space PART 120—DRUG AND ALCOHOL TESTING PROGRAM Federal Aviation Administration (FAA) Standards and Requirements 3
J AS9100 & recent amendments to FAA Title 14 (14 CFR) Federal Aviation Administration (FAA) Standards and Requirements 2
A TUV Audits - 21 CFR 820 General Auditing Discussions 6
S Records - Do's and don't' of record entries (FDA - 21 CFR 820) Records and Data - Quality, Legal and Other Evidence 13
T 21 CFR 820.20 - Quality Planning Requirements? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
S ISO 9001:2015 vs 21 CFR Part 211 matrix Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 0
M 21 CFR 820 vs 21CFR820 vs 21 CFR Part 820 Document Control Systems, Procedures, Forms and Templates 3
A 21 CFR 820 - Risk Management - Looking for some guidance US Food and Drug Administration (FDA) 3
C 21 CFR 803 malfunction detected during packaging - report? Other US Medical Device Regulations 1
J 21 CFR 821 Medical Device Tracking Requirement 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
P Equipment 21 CFR 820.70(g) - User Requirements Document for Off the shelf equipment 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
Anonymous16-2 21 CFR Part 11 - Steps to take if we want to validate an electronic system Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 2
A 21 CFR part 11 - section 11.100 - Electronic Signature Certification Other US Medical Device Regulations 6
E 21 CFR 211.103 Calculation of Yield Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 5
D ISO 13485, FDA 21 CFR 820 and Auditing the Accounting Department ISO 13485:2016 - Medical Device Quality Management Systems 5
Q 21 CFR 821 Medical Device Tracking 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
L Wearables 21 CFR Part 11 compliance Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
Ed Panek 21 CFR Part 820 - FDA Label Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
J Mislabeling - Consider this an FDA notified recall? CFR 806.10 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
L UDI Requirments Gap Analysis - EU MDR vs FDA CFR EU Medical Device Regulations 8
P 21 CFR 807.81 When a premarket notification submission is required Other US Medical Device Regulations 0
F 21 CFR Part 11 - Implicit requirements - Validation plan for a Software as a Service Other US Medical Device Regulations 1
C 21 CFR Part 11- What about handwritten signatures? Other US Medical Device Regulations 1
Z 21 CFR and Annex 11 mapping document Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
Ed Panek Do Cloud services require 21 CFR Part 11 compliance? Qualification and Validation (including 21 CFR Part 11) 7
QIE FDA 21 CFR Part 11 "Meaning of Signature" Other US Medical Device Regulations 6
J Business Intelligence and 21 CFR Part 11 Compliance Qualification and Validation (including 21 CFR Part 11) 1
R Addressing training requirements - 21 CFR Part 820.25 (1) & (2) Other US Medical Device Regulations 4
R How to improve a Validation program and procedures to FDA (21 CFR part 820) & ISO13485 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
Marc Problem with 21 CFR Part 820 - US FDA Quality System Regulations (QSR) sub-forum link - 2 May 2019 Elsmar Cove Forum Suggestions, Complaints, Problems and Bug Reports 1
R 21 CFR Part 820 Contract Manufacturer of Medical Device Component 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
S 21 CFR 810 - Medical Device Recall Authority - Contract Manufacturing Other US Medical Device Regulations 1
S Is Adobe Sign - E-signature for QMS documents - 21 cfr part 11 compliant? ISO 13485:2016 - Medical Device Quality Management Systems 2
D Design developer - 21 CFR 807 - do we need to register? Other US Medical Device Regulations 0
C Signature manifestations - 21 CFR Part 11 Qualification and Validation (including 21 CFR Part 11) 4
C Informational CFR 21 Definition for Data or Raw Data Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 5
I GMP 21 CFR Part 11 Electronic Records Compliance Project Help Qualification and Validation (including 21 CFR Part 11) 9
M Suggestions for Electronic Signature Software (FDA 21 CFR Part 11 Compliant) Qualification and Validation (including 21 CFR Part 11) 12
D Use of password managers on validated computer systems (21 CFR Part 11) Medical Information Technology, Medical Software and Health Informatics 2
Q 21 CFR Part 111 - Requirements for Dietary Supplement Manufacturing Water Quality Document Control Systems, Procedures, Forms and Templates 1
N Change Control - Compliance with FDA 21 CFR Part 820 Document Control Systems, Procedures, Forms and Templates 3
D 21 CFR Part 820 (Subpart A) - Question about "Authority" 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
K 510K statement Questions 21 CFR 807.93 - Discontinued or never released for sale US Food and Drug Administration (FDA) 0
A Alternative to keeping hard copy records that is 21 CFR part 11 compliant Document Control Systems, Procedures, Forms and Templates 0
R RUO audits - CFR 820.22 and ISO 13485 8.2.4 Internal Auditing 3
B 21 CFR 803 Definition for 'Substantial Harm" Other US Medical Device Regulations 1
A Training as well as certifications for ISO 13485:2016, 21 CFR 820... Training - Internal, External, Online and Distance Learning 5
K Production Unit Labeling - 21 CFR 820.184.e 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
D 21 CFR Part 11 - Electronic Signature Management SOP Other US Medical Device Regulations 0