I have a question regarding a change in labelling for an FDA cleared device.
If during Post market feedback a new risk has been identified, therefore resulting in an added precaution on the IFU, that however does not change the intended use or instructions for use.
Would this require a new 510(k)?

Thanks!
 
Elsmar Forum Sponsor

Watchcat

Trusted Information Resource
#2
A precaution advises provider and/or patient of specific circumstances of use under which use of the product might pose additional risk relative to other circumstances of use within the scope of its intended use. How provider and patient adjust their use under the precaution is usually left to them to decide, but a precaution can be expected to change use, or there would be no point in including it.

I'm not sure about your statement that adding a precaution to the IFU does not change the instructions for use. IFU = instructions for use. If you add something to the IFU, you have added something to the instructions for use. An addition is a change. Perhaps you meant indication for use. I don't want to get into the difference between intended use and indication for use, because I think this is a swamp. Maybe someone else has their hip boots on and will be happy to wade in.

I also don't want to advise you regarding the need for a 510(k), nor what flavor of 510(k) you might submit. Just want to disagree with the notion that you can add a precaution on use of the product without changing its use, whatever you want to call it...intended, indication, instructions for.
 
Last edited:

Ed Panek

QA RA Small Med Dev Company
Staff member
Super Moderator
#3
We utilize a form based on FDA guidance

A1
Does the change affect the indications for use statement?
Yes – See A1.1 / No – See Question A2
A2
Does the change add or delete a contraindication?
Yes -New 510K / No – See Question A3
A3
Is this a change in the warnings or precautions?
Yes – See A1.1 / No – See Question A4
A4
Could the change affect the directions for use?
Yes – See A1.1 / No – Documentation
A1.1
Is it a change from a device labeled for single use only to a device labeled as reusable?
Yes – New 510K / No – See Question A1.2
A1.2
Is it a change from Rx to over the counter OTC use?
Yes – New 510K / No – See Question A1.3
A1.3
Is it a change to the device name or to solely improve readability or clarity?
Yes – Documentation / No – See Question A1.4
A1.4
Does the change describe a new disease, condition, or pt pop that the device is intended in diagnosing, treating, preventing, curing or mitigating?
Yes – Documentation / No – See Question A1.5
A1.5
Does a risk-based assessment identify any new risks or significantly modified existing risks?
Yes – 510K see chart / No – Documentation​
 
#5
Hi and thanks for the answers. I went also through the guidance and I end up with documentation. However I am not really sure if this at all applies to my situation. The change made does not change the indication (thanks Watchcat) and doesn't introduce any new risks, all other parameters are in the cleared range. However, the change made results in a product variant to the main product e.g. one part is 2mm longer than in the main product. The product variant has also a slight change in the name as well... does this also fall under the the same guidance? Because yes it is a change to the preduct but results in a new product variant (under the same product family)
 

Ed Panek

QA RA Small Med Dev Company
Staff member
Super Moderator
#6
Is this 2mm longer product specific for a user population that should not purchase the 2mm shorter version?
 
Thread starter Similar threads Forum Replies Date
TheMightyWife Address Change - Medical Device Packaging Labelling Requirements EU Medical Device Regulations 7
N API Q1, Management of Change & Document Change Request Oil and Gas Industry Standards and Regulations 2
adir88 Substantial Change to QMS (Annex IX, Chapter I) EU Medical Device Regulations 0
A Eudamed actor registrations if you change AR for MDR product EU Medical Device Regulations 0
C Change Management Process ISO 14971 - Medical Device Risk Management 8
S ISO13485:2016 7.3.9 design and development change ISO 13485:2016 - Medical Device Quality Management Systems 3
N Change Control for processes & products Manufacturing and Related Processes 7
Q Change Control Numbering system Document Control Systems, Procedures, Forms and Templates 6
D Laboratories listed in the Declaration of Conformity change Canada Medical Device Regulations 1
M Change control on Tracking Sheet ISO 13485:2016 - Medical Device Quality Management Systems 11
D EU MDR - Change of device name (legacy device) EU Medical Device Regulations 4
L Change Log in (controlled) Forms? (ISO 9001:2015) Document Control Systems, Procedures, Forms and Templates 6
H Control of Change i.e. process and WI IATF 16949 - Automotive Quality Systems Standard 9
dgrainger Informational MHRA's Software and AI as a Medical Device Change Programme UK Medical Device Regulations 0
K Is implementing Electronic Signatures a Significant Change to the QMS? ISO 13485:2016 - Medical Device Quality Management Systems 10
A Management of change procedure for ISO 45001 Occupational Health & Safety Management Standards 5
N Cleanroom Classification Change Other Medical Device Related Standards 7
D Engineering Change order vs. Change Control ISO 13485:2016 - Medical Device Quality Management Systems 3
Watchcat Change Request Documents - What's Typical These Days? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 14
NDesouza Management of Change (MOC) Manufacturing and Related Processes 11
H Customer Specific Change in a SaaS SAMD IEC 62304 - Medical Device Software Life Cycle Processes 1
D GAMP CAT 5 Change Control Qualification and Validation (including 21 CFR Part 11) 9
lanley liao Questions regarding the change of critical suppliers and key personnel Oil and Gas Industry Standards and Regulations 5
I What revalidation is necessary if we change parts in a WFI loop ? Qualification and Validation (including 21 CFR Part 11) 1
Y Exporting data to the cloud is a "Significant Change"? EU Medical Device Regulations 5
A Design Change/ECO Related Question 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
S Reseller Request to Change UPC on Medical Device via Re-labeling Medical Device and FDA Regulations and Standards News 5
S Change management with FAA Federal Aviation Administration (FAA) Standards and Requirements 0
R Revalidation under VDmax25 following material change ISO 13485:2016 - Medical Device Quality Management Systems 2
A How can we change a device trade name in 510(k) summary? US Food and Drug Administration (FDA) 0
L Change Notification Requirements for class 1 devices Medical Device and FDA Regulations and Standards News 1
R CB/Auditor Requiring a change in scope ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
Ed Panek Notified Body Substantial Change MDD EU Medical Device Regulations 1
A Purchasing Controls on Change Management of Literature/Labeling ISO 13485:2016 - Medical Device Quality Management Systems 3
A For software change - New Channel of interoperability CE Marking (Conformité Européene) / CB Scheme 5
JoCam IEC 60601-1 and 60601-1-2 retest after PCBA change IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
N ISO 13485 7.3.9 Change control in medical device software ISO 13485:2016 - Medical Device Quality Management Systems 6
A ISO 13485 procedure change and reflect to legacy manufacture items ISO 13485:2016 - Medical Device Quality Management Systems 2
R AS9102 FAI Change in Material / Process Supplier AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
D Reports under change management | ISO 13485:2016 & ISO 9001:2015 ISO 13485:2016 - Medical Device Quality Management Systems 3
Q Time change via change of Time Zone Qualification and Validation (including 21 CFR Part 11) 3
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
M Change in Constitution / Ownership of firm -------ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
U Power supply - Change From 160 to 300 Watts - Significant change? EU Medical Device Regulations 2
N Sterilization Protocol Change in Validation Process and further impacts ISO 13485:2016 - Medical Device Quality Management Systems 1
M Label change - 510(k) Medical Device and FDA Regulations and Standards News 5
M EMS change management ISO 14001:2015 Specific Discussions 2
E Document Change Request forms (or no forms)? ISO 13485:2016 - Medical Device Quality Management Systems 9
J Significant change related to design and intended use EU Medical Device Regulations 3
J Need Change Control Yes/No Decision Tree Template ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1

Similar threads

Top Bottom