Changing calibration frequency

qcman

Registered Visitor
I am not sure if this belongs here or in the ISO forum. Anyway after reviewing the system I am going to change our annual calibration of some 1100 gages we have in the system to every 2 years. I could do this as each one comes due and change the cal sticker then however, this would take a year to complete. I am wondering if I can just send out a directive to the departments stating that we will be adding 1 year to the due date currently on the calibration stickers of each gage. We would also note this in the software program ( Gagetrak) we use. Would the latter result in a non conformance in our next ISO audit?
 

Jim Wynne

Leader
Admin
I think that somewhere in your documentation you would need to specify that the controlling date is the one found in the calibration record, but I can't help but feel that you're looking for trouble in one way or another. You will also need to be able to justify, for each device, the extension.
 

Ninja

Looking for Reality
Trusted Information Resource
If you document the decision and the basis for it...and show it to the auditor if a question comes up...you're clean.

What a wonderful reason to not use calibration stickers...take the next year (or two) to remove them and not replace them as the gages come around for calibration.
 

Ninja

Looking for Reality
Trusted Information Resource
...for each device...spot on.

When we extended cal for FOG gages, we did it as a family...but wouldn't extend the "we just don't feel like spending the money, and these haven't failed or drifted in 15 years" across to micrometers without supporting justification.
 

qcman

Registered Visitor
Jim

After reviewing the system as it is now I am comfortable with the justification for the extension. I am curious as to what trouble you think may be ahead with this?

Ninja

I feel the same way about the stickers and they may go as well.
 

Ninja

Looking for Reality
Trusted Information Resource
Not answering for Jim...he doesn't need the help...
But your OP can easily be read to think that you are being a bit blase and just extending interval because you feel like it.
Might not be what you meant...but it is short posts on a forum. That's why I also broached the "we just don't feel like spending the money ".
 

qcman

Registered Visitor
Good point Ninja. Looking at previous recalibrations I see virtually no change in the gages. This along with the fact that gages are protected from damage (it does happen) it seems silly to do it annually.
 

Mike S.

Happy to be Alive
Trusted Information Resource
1100 gages is a lot.

Make sure your controlling internal procedure supports what you're going to do, and as they said above I'd justify the interval increase decision in writing. Make sure you use "risk-based thinking".

Look at the entire process from an auditor's standpoint and the users' standpoint and see if you can find any holes or weaknesses, then address them.
 

Tagin

Trusted Information Resource
After reviewing the system as it is now I am comfortable with the justification for the extension. I am curious as to what trouble you think may be ahead with this?

Speaking for myself, I see a lot of risk in potentially producing up to 2 years of nonconforming product, and with 1100 gages, well, that sounds like a big volume of potentially nonconforming product. The costs, the impact to customers, damage to reputation could be enormous.

But I don't know the nature of your gages and how impervious they are to change, precisions involved, etc.

Or perhaps, the gages could be categorized into ABC categories, based on usage & product volume; where A gets calibrated every 6mo, B gets calibrated every year, C every 2 years, for example.

An auditor may well query you on your process for the last part of 7.1.5.2:
The organization shall determine if the validity of previous measurement results has been adversely affected when measuring equipment is found to be unfit for its intended purpose, and shall take appropriate action as necessary.
 
I had to do this also, I wrote up documentation for each gage showing how it held tight for 5 years, then extended the calibration to 2 years for about 25% of the gages. I also tagged them on the label with a number -00- for master gages, -01- for QC gages, -02 for shop gages, etc. the -02 gages stayed on annual calibration, I extended some of the rarely used QC gages and master gages (gage block sets). The more they are used the closer the calibrations have to be.
 
Thread starter Similar threads Forum Replies Date
B Changing out Thermocouple - Is new calibration required? General Measurement Device and Calibration Topics 1
B Effect of Calibration Laboratory Changing Ownership & Moving on ISO17025 registration ISO 17025 related Discussions 4
R Changing our DMS/ECM to a Wiki. Advice on best practices and changes of paradigm? Document Control Systems, Procedures, Forms and Templates 0
T New process capability analysis required when changing a product on the same line? IATF 16949 - Automotive Quality Systems Standard 6
N Changing PRRC in Eudamed EU Medical Device Regulations 1
S Changing UK Responsible Person UK Medical Device Regulations 0
S Changing PCB Assembler and impact on QMS and MDD CE Certificate EU Medical Device Regulations 2
B How to change hour format to text but not changing the number ? Excel .xls Spreadsheet Templates and Tools 1
A Changing Notified Body EU Medical Device Regulations 6
H Traceability after changing the machine IATF 16949 - Automotive Quality Systems Standard 13
M Changing Investigational Device from Non-Sterile to Sterile - what are the implications? Other Medical Device and Orthopedic Related Topics 3
S Changing revision status of a reviewed document Manufacturing and Related Processes 12
S Sterilization validation after changing sterilization process provider Qualification and Validation (including 21 CFR Part 11) 3
R Impact of changing manufacturer post clinical investigation CE Marking (Conformité Européene) / CB Scheme 3
D Clothes changing in pharmaceutical factory Manufacturing and Related Processes 8
M Changing supplier of critical raw material (III class device) Other Medical Device and Orthopedic Related Topics 1
K Changing the Address in 510(K) Approved Product 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
T AS9100 - Changing PEARs target AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 41
C FDA on changing acceptance criteria re: analytical method validation US Food and Drug Administration (FDA) 1
V Statistical basis and justification while comparing / changing sampling plans Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 11
K Changing from Procedure based to Process based Competency Evaluations General Auditing Discussions 5
J Steps when changing material for class 1 medical device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
Marc Email Addresses - Changing your email address Elsmar Xenforo Forum Software Instructions and Help 0
supadrai OEM Changing the name of the manufacturer - length of process? Is 1-3 months correct? Other Medical Device Regulations World-Wide 0
R Changing Document Control software. Must I transfer EVERYTHING? Document Control Systems, Procedures, Forms and Templates 3
A IMDS - Changing the Locked Weight Value RoHS, REACH, ELV, IMDS and Restricted Substances 1
D Changing Templates Approval Field Document Control Systems, Procedures, Forms and Templates 5
P Romer Arm - Profile of a Surface Changing General Measurement Device and Calibration Topics 4
D Changing software classification via software - IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 3
L MINITAB 17 - Changing Data Using Minitab Software 3
S Process validation when changing location ISO 13485:2016 - Medical Device Quality Management Systems 10
Don Fardie CAPA vs. Risk Assessment - Changing a product material for better performance ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
C Changing Registrars Registrars and Notified Bodies 15
MDD_QNA Changing the Layout of Medical Device Label(s) CE Marking (Conformité Européene) / CB Scheme 3
S Keeping track of the changing Medical Device Standards Other Medical Device Related Standards 3
J Changing of our New Hire Orientation Training Records Training - Internal, External, Online and Distance Learning 12
A Changing a Form Template & Updating Records ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
D Is anyone changing their ISO 14001:2004 EMS Policy? ISO 14001:2015 Specific Discussions 8
Q Changing Terminology in ISO 9001:2015 - Documentation Content ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
V Changing Indication of Use for a Class IIa Medical Device EU Medical Device Regulations 2
F Rules to respect when changing from Certification Body ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
MDD_QNA Clinical data requirements - Changing a Class I Medical Device to Class II EU Medical Device Regulations 5
shimonv Changing the Label of a Medical Device product in the field Other US Medical Device Regulations 5
E Information required about changing an off shelf part 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
M Changing FMEA template to include Detection Ranking for the PFMEA FMEA and Control Plans 2
WCHorn Changing Sampling Accept Number to Zero, Arbitrarily AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 15
M Amending CFDA Registration for changing site of shipment China Medical Device Regulations 2
S Gage R&R on Potentially Changing Parts - Guidance Needed Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
R Re-registration of medical device in Russia - "changing holder" Other Medical Device Regulations World-Wide 1
R Changing CE Mark on Documentation ISO 13485:2016 - Medical Device Quality Management Systems 2

Similar threads

Top Bottom