Changing the internal audit schedule

M

Mark Smith

#1
I have had conflicting responses to this question. Is there any requirment for either ISO 9001 or the QSR that the internal audit schedule be a controlled document? I find for various reasons that I have to change the dates (never moved up, always moved back) of several internal audits per year. Currently I initial and date on the schedule if I need to reschedule but if the schedule is controlled, I would need to initiate a change order each time and that could be a headache. comments?
 
Elsmar Forum Sponsor
T

Tom Goetzinger

#2
We have a controlled form to be used for our Internal Audit schedule, with a second page that is used to explain why changes in the schedule take place. The intention is to reduce or eliminate changes in the schedule for insignificant reasons. Our procedures grants the Management Representative the authority to change the schedule as required, and only the Management Rep can change it. Perhaps you just need to change the way your procedures are written to allow you to do what is required.
 
L

Laura M

#3
We did something similar...allow mgmt. rep. the authority. If the dept scheduled for audit requested the change, they needed to put it in writing w/ mgmt rep approval. Chnages initiated by the internal auditors were just initialed by mgmt. rep.
 
R

Russ Jackson

#4
Whether or not you maintain the schedule as a controlled document will probably depend on how you have it set up originally and how the schedule is being used. You may need to communicate changes to the schedule to any persons who received copies of the original schedule; you can do this in controlled document form or by communication via memo or e-mail, keeping copies as evidence that you did so. If you are the principal user of the schedule and notify participants (assigned auditors and the area to be audited) of an impending audit, then you should not formalize the control of the schedule. However, as Tom pointed out you should revise your audit procedure to ensure that it defines what you need to do to meet your needs. Perhaps note that you are authorized to make changes to the schedule and that those affected will be notified of changes.
 
J

Jackie Jolly

#5
We had an outside source come in and perform an audit to help us get ready for our QS Registration. There was a finding on Internal Audits for not having the Audit Schedule under document control. My concern with this was everytime I had to revise it, the rev. level would have to change. My Management Rep said as long as all I was doing was revising the information & not the form it's self and had approval on it, we were covered. We received our registration with out a hitch. At least for now. It seems our registar is pretty inonsistant from one auditor to the next. But that's another story!
 
J

James Gutherson

#6
Let's think about the concept of document control. The idea is that all the people using the document have the upto date information. If they are filling out a form, then the layout and form fields are the important information, but in this case the important information is the audit schedule it's self. You do want every body to be operating to the same schedule don't you?
This is not a shot at you Jackie but at your Management Rep who should know better.
By the way "revision level" means just that, when the document is revised, you change the revision level, by definition.

The elements in ISO/QS 9K are all there for a reason. They are actually there to help your business, not to create hurdles for hurdles sake, nobody wins then. Think about the intent of the element, how you address this in your business and where it might be usefull in other areas. (Shoot me down if you think I'm wrong, I can handle it.)
 
M

Mark Smith

#7
I think in this case James I must disagree. The audit schedule that I refer to is not a circulated document. It resides in my audit files and no one else uses it. I inform department Managers by memo of upcoming audits as prompted by the schedule. It was written originally at the end of last year and in it I identified which departments would be audited during which months of the coming year. However, as with most plans, there were unforseen circumstances that required me to make changes to the schedule. My SOP for internal audits states that "Scheduled audits may be rescheduled if approved (signified by initialing of the schedule) by the Director of Q.A." In such instances, it should not be necessary for me to initiate a document change order and revise the document. The format of the schedule is a controlled document template in the way it is structured for two years out by month and department name. Are we in agreement now?
 

barb butrym

Quite Involved in Discussions
#8
think about it..yes you need to have it under control, yes its a working document with updates, yes it needs to be public in most cases...so where does that leave you?

In your audit program procedures/documents say how you will control the schedule and who has access/authority to change. Typically the format is under doc control, the working info is what changes...how do you know when it was changed for instance, and by who...come up with something real. What they want to know is that the audits are planned....scheduled in advance..not "GEE think I'll audit this stuff this week...." AND that updates reflect reviews/CA etc. based on need. The revision control feature in word tables will do that nicely......

Typically I have a planned schedule by quarter, then update for assigned auditors, they schedule the dates and there's room for followup additions etc...and the closure details....

[This message has been edited by barb butrym (edited 03 December 1999).]
 

Marc

Fully vaccinated are you?
Leader
#9
Not to get into a pissing match, but I have *several* clients whose internal audit schedule is nothing more than penciled in dates on a standard, every day calendar on a wall in an office (granted these are relatively small [<200 souls] companies). It is controlled - it is in the office of the person controlling it. No other employee is expected to come into the office and change dates (why, prey tell, would they short of the intent of sabatoge which we all know one can only go so far to prevent). On the other hand, they outsource internal audits so there's not typically any changes in the schedule.

I think this is a case where folks are over reaching in defining document control. Control alone is not so complex as it seems some folks think. If I have a document on my computer and I back it up and if I change it I let everyone know about changes who should know about changes, it is controlled. I know many companies where some process documents are controlled locally by the engineer responsible for the line.

All you have to do is be smart and make sure your level 2 acknowledges local control.

Don't over complicate this issue, folks. Let's look at the intent. The intent is to ensure audits are scheduled. If some audits are rescheduled, that is not an issue. Big deal, that's business and that's life. The issue is whether you get them done or not, and if not that you do a corrective action.
 

barb butrym

Quite Involved in Discussions
#10
YUP local control ..define it and then do it.

Some use a simple scheduling program like "outlook"

As an auditor, I would look for a planned schedule.....updated with closure status of some kind (not always on the schedule...but somewhere close by or it would prompt me to ask many more questions) and follow up schedules if needed. Then pull the reports and match to see if they were carried out on time and as planned, and then followed up/closed. Facilitates the audit to have it all handy, although not a requirement...BUT bear in mind 99% of auditors will build their confidence in a company through the internal audit program....if its top shelf, and efficient its in the company's best interest...SO whats top shelf and efficient?
Planned audits, clearly scheduled, executed and closed. Trained auditors, detailed audit trails that can be used to recreate the scenario (n/c).
 
Thread starter Similar threads Forum Replies Date
F Changing the Frequency of Internal Audits - ISO/TS 16949 Internal Audit Requirements Internal Auditing 5
C Changing Internal Audit Program - Regional Manager Online Audits Internal Auditing 2
N Internal auditing - Changes after recently changing to 9001:2000 Internal Auditing 4
B Internal Audits - Changing from Check Lists to Process Approach Internal Auditing 13
M Changing Positions: System Quality Assurance vs. Internal Quality Auditor Career and Occupation Discussions 20
N Changing PRRC in Eudamed EU Medical Device Regulations 1
S Changing UK Responsible Person UK Medical Device Regulations 0
S Changing PCB Assembler and impact on QMS and MDD CE Certificate EU Medical Device Regulations 2
B How to change hour format to text but not changing the number ? Excel .xls Spreadsheet Templates and Tools 1
A Changing Notified Body EU Medical Device Regulations 6
H Traceability after changing the machine IATF 16949 - Automotive Quality Systems Standard 13
M Changing Investigational Device from Non-Sterile to Sterile - what are the implications? Other Medical Device and Orthopedic Related Topics 3
Q Changing calibration frequency Calibration and Metrology Software and Hardware 18
S Changing revision status of a reviewed document Manufacturing and Related Processes 12
S Sterilization validation after changing sterilization process provider Qualification and Validation (including 21 CFR Part 11) 3
R Impact of changing manufacturer post clinical investigation CE Marking (Conformité Européene) / CB Scheme 3
D Clothes changing in pharmaceutical factory Manufacturing and Related Processes 8
M Changing supplier of critical raw material (III class device) Other Medical Device and Orthopedic Related Topics 1
K Changing the Address in 510(K) Approved Product 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
T AS9100 - Changing PEARs target AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 41
C FDA on changing acceptance criteria re: analytical method validation US Food and Drug Administration (FDA) 1
V Statistical basis and justification while comparing / changing sampling plans Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 11
K Changing from Procedure based to Process based Competency Evaluations General Auditing Discussions 5
J Steps when changing material for class 1 medical device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
Marc Email Addresses - Changing your email address Elsmar Xenforo Forum Software Instructions and Help 0
supadrai OEM Changing the name of the manufacturer - length of process? Is 1-3 months correct? Other Medical Device Regulations World-Wide 0
R Changing Document Control software. Must I transfer EVERYTHING? Document Control Systems, Procedures, Forms and Templates 3
A IMDS - Changing the Locked Weight Value RoHS, REACH, ELV, IMDS and Restricted Substances 1
D Changing Templates Approval Field Document Control Systems, Procedures, Forms and Templates 5
P Romer Arm - Profile of a Surface Changing General Measurement Device and Calibration Topics 4
D Changing software classification via software - IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 3
L MINITAB 17 - Changing Data Using Minitab Software 3
S Process validation when changing location ISO 13485:2016 - Medical Device Quality Management Systems 10
Don Fardie CAPA vs. Risk Assessment - Changing a product material for better performance ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
C Changing Registrars Registrars and Notified Bodies 15
MDD_QNA Changing the Layout of Medical Device Label(s) CE Marking (Conformité Européene) / CB Scheme 3
S Keeping track of the changing Medical Device Standards Other Medical Device Related Standards 3
J Changing of our New Hire Orientation Training Records Training - Internal, External, Online and Distance Learning 12
A Changing a Form Template & Updating Records ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
D Is anyone changing their ISO 14001:2004 EMS Policy? ISO 14001:2015 Specific Discussions 8
Q Changing Terminology in ISO 9001:2015 - Documentation Content ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
V Changing Indication of Use for a Class IIa Medical Device EU Medical Device Regulations 2
B Changing out Thermocouple - Is new calibration required? General Measurement Device and Calibration Topics 1
F Rules to respect when changing from Certification Body ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
MDD_QNA Clinical data requirements - Changing a Class I Medical Device to Class II EU Medical Device Regulations 5
shimonv Changing the Label of a Medical Device product in the field Other US Medical Device Regulations 5
E Information required about changing an off shelf part 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
M Changing FMEA template to include Detection Ranking for the PFMEA FMEA and Control Plans 2
WCHorn Changing Sampling Accept Number to Zero, Arbitrarily AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 15
M Amending CFDA Registration for changing site of shipment China Medical Device Regulations 2

Similar threads

Top Bottom