We are planning to change the layout of your labels. Is it required to report this change to the NB?

What we are planning to do is also to box every part that is being supplied to the customer, this means that we have one big shipping box where the product is being fitted. In this big box we have smaller boxes containing different parts of the device. What is the minimal labeling requirement on these? Is it sufficient to have the name of the company, name of manufacturer, name of the part (and a picture) together with the adress and name of legal manufacturer? Is there a requirement to have the CE mark on any of the boxes?

I had found a document stating that we need to write out the weight of the package being shipped. However, I can't find that document right now and I need advice on if someone knows where to find this information (preferably) or if someone knows the requirements on when to write out the weight of any package or box? (For safety reasons of course)

Another question concerns spare parts, we have the main part which is the medical device itself and we are planning on changing our spare parts handling. So we are going to ship this part as a spare part and we haven't done this before. What is the minimal required information to supply together with this spare part? Do we need to send the IFU or can we just put in a quick guide showing how to mount the part and refer to the IFU (that they should already be having)?

This is a class IIa medical device in EU.

Thanks in advance!
 
Elsmar Forum Sponsor

Ronen E

Problem Solver
Staff member
Moderator
#2
Hi,

You have listed many questions in a single post. Those are not very difficult questions, however some of them are also not trivial to answer. It's not impossible that a single Cove member would know all, or most of the answers. It's also not impossible that several members together could have answered all your questions. Nevertheless, the fact is that no one replied. In my opinion the reason is that the scope of your post exceeds the "normal" extent of a public discussion forum post. It looks more like an email that a retainer consultant would receive from a client.

If I had to responsibly answer your post, it would take me more than "5 minutes", and I suspect the same applies to most members here. Hopefully I'm wrong - my reply has "bumped" this thread for another chance, and maybe someone would answer at least some of your questions, off the top of their head. If not, and you're still interested in some focused help, please feel free to PM me.

Cheers,
Ronen.
 

somashekar

Staff member
Super Moderator
#3
We are planning to change the layout of your labels. Is it required to report this change to the NB?

What we are planning to do is also to box every part that is being supplied to the customer, this means that we have one big shipping box where the product is being fitted. In this big box we have smaller boxes containing different parts of the device. What is the minimal labeling requirement on these? Is it sufficient to have the name of the company, name of manufacturer, name of the part (and a picture) together with the adress and name of legal manufacturer? Is there a requirement to have the CE mark on any of the boxes?

I had found a document stating that we need to write out the weight of the package being shipped. However, I can't find that document right now and I need advice on if someone knows where to find this information (preferably) or if someone knows the requirements on when to write out the weight of any package or box? (For safety reasons of course)

Another question concerns spare parts, we have the main part which is the medical device itself and we are planning on changing our spare parts handling. So we are going to ship this part as a spare part and we haven't done this before. What is the minimal required information to supply together with this spare part? Do we need to send the IFU or can we just put in a quick guide showing how to mount the part and refer to the IFU (that they should already be having)?

This is a class IIa medical device in EU.

Thanks in advance!
As it is your NB who is providing you the gateway to EU, it is good to discuss any and all types of label changes with your NB. Your additional questions will also be answered in that process.
 

pkost

Trusted Information Resource
#4
We are planning to change the layout of your labels. Is it required to report this change to the NB?
What class of product? IF class III then yes, anything lower then if it is layout only, I would consider it a small change

What we are planning to do is also to box every part that is being supplied to the customer, this means that we have one big shipping box where the product is being fitted. In this big box we have smaller boxes containing different parts of the device. What is the minimal labeling requirement on these? Is it sufficient to have the name of the company, name of manufacturer, name of the part (and a picture) together with the adress and name of legal manufacturer? Is there a requirement to have the CE mark on any of the boxes?
What do you do currently? This sounds like more than just a change of label layout?....
Is this kit being put together under article 12? If so then each item would usually be considered a device in it's own right and therefore labelled appropriately.
If you are not going down the article 12 route and it is essentially an IKEA kit for your device, then the required information must be on the outer box and I would risk assess the level of information on the inner boxes; as a starter for 10 I would say the internal boxes need:
  • Item name/identifier
  • Statement that it is intended for the main product
  • manufacturer details
  • lot details
  • warning symbols e.g. keep dry

I had found a document stating that we need to write out the weight of the package being shipped. However, I can't find that document right now and I need advice on if someone knows where to find this information (preferably) or if someone knows the requirements on when to write out the weight of any package or box? (For safety reasons of course)
weight is not required by the device directive unless your risk assessment determines it should be. the weight requirement is more likely to be customer driven (I used to demand weight on outer shippers from our suppliers), the NHS may similarly require a weight indication for their health and safety requirements

You may want to take a look at directive 76/211/EEC and see if that applies (it refers to selling measured quantities e.g. buying 50ml perfume)



Another question concerns spare parts, we have the main part which is the medical device itself and we are planning on changing our spare parts handling. So we are going to ship this part as a spare part and we haven't done this before. What is the minimal required information to supply together with this spare part? Do we need to send the IFU or can we just put in a quick guide showing how to mount the part and refer to the IFU (that they should already be having)?
I'm confused by "the main part which is the medical device itself", are the other parts accessories?? in which case the directive treats them as devices in their own right.

If the main part is a medical device in its own right it should be labelled as a
 
Thread starter Similar threads Forum Replies Date
R Impact of changing manufacturer post clinical investigation CE Marking (Conformité Européene) / CB Scheme 3
D Clothes changing in pharmaceutical factory Manufacturing and Related Processes 8
M Changing supplier of critical raw material (III class device) Other Medical Device and Orthopedic Related Topics 1
Kuldeep Singh Changing the Address in 510(K) Approved Product 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
T AS9100 - Changing PEARs target AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 41
C FDA on changing acceptance criteria re: analytical method validation US Food and Drug Administration (FDA) 1
V Statistical basis and justification while comparing / changing sampling plans Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 11
K Changing from Procedure based to Process based Competency Evaluations General Auditing Discussions 5
J Steps when changing material for class 1 medical device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
Marc Email Addresses - Changing your email address Elsmar Xenforo Forum Software Instructions and Help 0
supadrai OEM Changing the name of the manufacturer - length of process? Is 1-3 months correct? Other Medical Device Regulations World-Wide 0
R Changing Document Control software. Must I transfer EVERYTHING? Document Control Systems, Procedures, Forms and Templates 3
A IMDS - Changing the Locked Weight Value RoHS, REACH, ELV, IMDS and Restricted Substances 1
D Changing Templates Approval Field Document Control Systems, Procedures, Forms and Templates 5
P Romer Arm - Profile of a Surface Changing General Measurement Device and Calibration Topics 4
D Changing software classification via software - IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 3
L MINITAB 17 - Changing Data Using Minitab Software 3
S Process validation when changing location ISO 13485:2016 - Medical Device Quality Management Systems 10
D CAPA vs. Risk Assessment - Changing a product material for better performance ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
C Changing Registrars Registrars and Notified Bodies 15
S Keeping track of the changing Medical Device Standards Other Medical Device Related Standards 3
J Changing of our New Hire Orientation Training Records Training - Internal, External, Online and Distance Learning 12
A Changing a Form Template & Updating Records ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
D Is anyone changing their ISO 14001:2004 EMS Policy? ISO 14001:2015 Specific Discussions 8
Q Changing Terminology in ISO 9001:2015 - Documentation Content ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
V Changing Indication of Use for a Class IIa Medical Device EU Medical Device Regulations 2
B Changing out Thermocouple - Is new calibration required? General Measurement Device and Calibration Topics 1
F Rules to respect when changing from Certification Body ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
MDD_QNA Clinical data requirements - Changing a Class I Medical Device to Class II EU Medical Device Regulations 5
shimonv Changing the Label of a Medical Device product in the field Other US Medical Device Regulations 5
E Information required about changing an off shelf part 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
M Changing FMEA template to include Detection Ranking for the PFMEA FMEA and Control Plans 2
WCHorn Changing Sampling Accept Number to Zero, Arbitrarily AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 15
M Amending CFDA Registration for changing site of shipment China Medical Device Regulations 2
S Gage R&R on Potentially Changing Parts - Guidance Needed Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
R Re-registration of medical device in Russia - "changing holder" Other Medical Device Regulations World-Wide 1
R Changing CE Mark on Documentation ISO 13485:2016 - Medical Device Quality Management Systems 2
K Changing the Auditor within a Notified Body ISO 13485:2016 - Medical Device Quality Management Systems 4
B Process Capability - Changing Limits to Improve Cpk or find Root Causes? Capability, Accuracy and Stability - Processes, Machines, etc. 6
M Changing Notified Body - Anyone have stories to share CE Marking (Conformité Européene) / CB Scheme 3
R Changing ISO 9001 Certification Scope Quality Manager and Management Related Issues 4
W Depth Gage Spindle moved when changing Contact Point Calibration and Metrology Software and Hardware 6
M Changing a Chinese Legal Agent China Medical Device Regulations 5
B Changing dimensional tolerances to approve rejected product ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
Q How to handle customer request for changing specifications 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
T Change Control Requirements when changing Suppliers IATF 16949 - Automotive Quality Systems Standard 4
P Changing from contract manufactured to self manufactured 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
N Document Control - Changing Company Logo Document Control Systems, Procedures, Forms and Templates 10
L Changing Quality Objectives ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
P Changing the Trade Name of a cleared Medical Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2

Similar threads

Top Bottom