Checklist for ISO 11607-2:2006 - Medical Device Packaging - Information needed
meimei
I'm working in suppling packaing material for medical device company, doing Blown film and web printing.
Recently, one of our new customer requested us to fill up a checklist for ISO11607-2:2006.
Will need some advise from anyone who know this standard, how will this standard applicable to my industry.
:thanx:
Best Regards
Meimei
Jimmy the Brit
Hello meimei,
If you are in the business of sterile medical device packaging then this document is absolutely vital for you.
ISO11607 (2006) "Packaging for terminally sterilized medical devices" part 1 defines the requirements for all medical packaging materials, sterile barrier systems and packaging systems. This covers preformed barriers, microbial retention properties, sterilization process compatibility, labelling compatibility, storage, transport, performance testing and design of packaging systems.
Part 2 states the validation requirements for forming, sealing and assembly of packaging operations. This defines IQ,OQ and PQ expectations as well as process control considerations.
Both of these documents are supported by AAMI TIR22:2007, which offers guidance on the interpretation of 11607. This is an awesome document with very clear instructions about how to demonstrate compliance with both parts of the standard. If you are not already familiar with them, it would also be valuable for you to look at some of the ASTM methods for packaging validation, shelf life validation, defect detection etc.. as these are called out by 11607.
As a purchaser of medical device packaging I would not even look at a supplier that did not claim compliance with ISO11607.
I hope this helps. Regards,
Jimmy
meimei
Hi Jimmy,
:thanks: for the reply.
Think I have to read the standard in detail to understand it.
Rgrds
MeiMei
meimei
Dear Jimmy,
I had read through the ISO 11607 standard. To my understanding, part 1 is more appliable to my industry, which we supply the top and bottom web to medical device company for their packaging.
AS for part 2, I'm not sure about it. Will you able to elevate more?
Thanks in advance
Rgrds
Meimei
Jimmy the Brit
Hi Meimei,
I agree with your primary compliance responsibility being with Part 1 of ISO11607, however I assume you sell these packaging materials on the basis that, under the correct time, temperature and pressure, they will form an impermeable (or at least bacterial retentive) seal. In that case you still have to validate the integrity of the potential seal, and for that you will need part 2, as well as numerous ASTM methods, and probably ISO5636 part 5, the Gurley test of air permeance of sterile barrier system paper packaging.
I recommend that you read AAMI TIR22:2007 and choose the methods that would make sense for you to comply with, assuming you make some form of label claim about the type of seal your materials can produce.
Hope this helps,
Regards,
Jimmy
adlcisa
our new customer wants to pack its IV Administration Sets only in microforate PE pouches; I'm loocking for a Standards for PE Packaging, but I found only for nonwoven materials or PA/PE film AND paper medical grade...
where I can find an equivalent Standard, but for only PE made pouches?
many thanks in advance,
kind regards,
Adlcisa
P
powens
Hello, I hope that this thread is close enough to my question. I have 2 parts to this question. Please assume that the client or company in question makes a medical device and packages it in double Tyvek pouch:
1) I wanted to see if anyone has a shared or public domain line by line, or at least key section guide to 11607 Part 1 and Part 2 (what are specific responses/practices to a given section from 11607?
I guess that this is going to have only the guide/primer/suggested practices for a given line or section. We cant have people actually distributing a "marked up" copy of 11607, that wouldn't fly.
2) The second part, is what are the key auditor question or drill in spots on Part 1 and Part 2.
I am sure that I can add some clarification to this thread.
Pat
X
X.Liu
I have a question associated with this standard but I could not find very clear answer.
It is about the label on top of microbial barrier system. Our product has a product label on the lid of pouch. Paper base, backside adhesive and ink/paint could be toxic and product could be badly conteminated if no proper design and validation is carried out. What tests we should consider? Biocompatibility?
P
powens
Hello all,
Please feel free to calibrate my response.
The start up way:
1) Build your intended 10993 series product samples.
2) Package your product samples in your sterile barrier.
3) Gas or Irradiate product (typ 2X).
4) Send packaged product to test lab for 10993 series testing. Lab personnel removes product from sterile barrier and test product for selected 10993 series test.
To me this seems defendable. Everything that is going to touch the patient and the sterile staff has been exposed to the inside of the sterile barrier and it's subsequent processing.
You could really argue that your test samples should have been exposed to accelerated aging conditions to maximize any outgassing etc. effects (real time duplication testing of course).
Big company way:
1) I don't know. I haven't worked on big company packaging bio compatibility project. Do big companies send out just packaging materials and labeling for seperate biocompat testing? Why? Is this a specific requirment of 11607?
If the outside of the sterile barrier does not touch the product or the sterile field staff, then it should be very low risk to the patient. Is the concern more for the circulating staff? Or very worst case, imagined toxic airbourne toxic molecules floating across the room from icorrectly chosen toxic label glue?
Thoughts, comments, calibrations?
I hope that this helps,
Pat
X
X.Liu
As I am looking at the similar issues, I dug a bit about it. AAMI TIR22:2007 discussed safety consideration of packaging material (E.3). My understanding is that this is not only the sterile barrier, but also the carton, label, IFU, etc. Bare in mind that the sterile barrier is not a insolation of chemical, toxic out-leach over the years of shelf life from carton, paper, ink, glue can penatrate and contaminate the product.
M ISO 13485:2016 internal audit checklist Medical Device and FDA Regulations and Standards News 5
R ISO 17025 vertical audit checklist wanted Document Control Systems, Procedures, Forms and Templates 2
J Audit Checklist for Integrated Management System for ISO 9001:2015, ISO 14001 & OHSAS18001 (IMS) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
A ISO 10002:2018 Checklist Needed (Complaints Handling) Customer Complaints 5
V Generic ISO 50001 checklist Various Other Specifications, Standards, and related Requirements 6
G ISO 22442 Supplier Audit Checklist - Medical devices utilizing animal tissues General Auditing Discussions 0
R Anyone who can share a checklist for ISO 29134? IEC 27001 - Information Security Management Systems (ISMS) 7
R ISO 9001:2015 and EMS 14001:2015 combined audit checklist General Auditing Discussions 5
A Audit Checklist for ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 19
B Checklist for ISO/TS16949 for review and comment IATF 16949 - Automotive Quality Systems Standard 5
S Looking for a checklist comparing ISO 13485:2016 and 21 CFR Part 820 ISO 13485:2016 - Medical Device Quality Management Systems 14
S Audit checklist for "Design Controls" per ISO 13485 wanted Internal Auditing 3
Richard Regalado ISO/IEC 27001 Mandatory Documentation Checklist IEC 27001 - Information Security Management Systems (ISMS) 1
C Checklist comparison between ISO9001:2008 and ISO 3834 Quality Management System (QMS) Manuals 1
S ISO 9001 Audit Checklist - Red, Yellow and Green status Quality Manager and Management Related Issues 10
D Internal Audit Checklist for the latest ISO/TS 16949 standard IATF 16949 - Automotive Quality Systems Standard 11
N ISO 20000 Implementation Guide and Compliance Checklist wanted IT (Information Technology) Service Management 2
C ISO 27001:2013 ISMS Internal Audit Checklist/Questionnaire IEC 27001 - Information Security Management Systems (ISMS) 24
L Needed ISO 31000 And ISO 22301 Checklist Risk Management Principles and Generic Guidelines 2
J Essential Requirements Checklist w/o ISO 13485 CE Marking (Conformité Européene) / CB Scheme 7
S Information Technology Department (IT) Checklist for ISO 9001:2008 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
M Risk Management (ISO 14971:2007) Internal Audit Checklist ISO 14971 - Medical Device Risk Management 7
Z ISO/IEC 20000 Checklist and/or Implementation Road Map IT (Information Technology) Service Management 11
G ISO 9001 Internal Audit Checklist ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 21
V ISO 9001 Checklist for Electromagnet Manufacturing ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
D Gap Analysis Checklist for ISO 13485:2003 2012 ISO 13485:2016 - Medical Device Quality Management Systems 13
S Essential Gap Analysis Checklist Components for ISO Certification ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
I IVD (in vitro diagnostic) Checklist for ISO 14971:2009 Annex H ISO 14971 - Medical Device Risk Management 3
T One Supplier Audit Checklist to cover ISO 9001:2008 and 13485:2003 Supplier Quality Assurance and other Supplier Issues 1
T I'm seeking ISO 13485:2003 Supplier Audit Checklist Supplier Quality Assurance and other Supplier Issues 1
J ISO 9001 Checklist Type Internal Audit Forms? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
S Checklist for ISO 27001 ISMS Internal Audit IEC 27001 - Information Security Management Systems (ISMS) 2
B Checklist Contents - Do ISO clause references need to be included? Internal Auditing 4
D ISO TS 29001 (Petroleum Petrochemical Natural Gas) - Internal Audit Checklist example Other ISO and International Standards and European Regulations 3
C ISO 17024 - Personal Certifications Programs Checklist Other ISO and International Standards and European Regulations 27
T ISO 14155 Good Clinical Investigation Practices Compliance Checklist? EU Medical Device Regulations 1
K ISO/IEC 23026 Audit checklist wanted Other ISO and International Standards and European Regulations 2
K ISO 22000 Audit Checklist wanted Food Safety - ISO 22000, HACCP (21 CFR 120) 6
H ISO 13485 & 21 CFR 820. Healthcare Plant Readiness Checklist ISO 13485:2016 - Medical Device Quality Management Systems 5
K ISO 28000 Internal Audit Checklist wanted Supply Chain Security Management Systems 10
R Product Audit Checklist according to ISO/TS 16949 IATF 16949 - Automotive Quality Systems Standard 24
A ISO 27001:2005 ISMS Internal Audit Checklist/Questionnaire IEC 27001 - Information Security Management Systems (ISMS) 14
P Internal ISO 9001 Audit - Process Audit Checklist Internal Auditing 31
B Internal Audit Checklist fo ISO 9001 Internal Auditing 1
AnaMariaVR2 ISO 17025:2005 Checklist ISO 17025 related Discussions 8
I ISO 20000 Checklist for Gap Analysis and Internal Audit Other ISO and International Standards and European Regulations 2
S GMP Checklist for Distribution Activities? ISO 13485 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
C Auditing Checklist for ISO 11135 and ISO 11137 (regulations for sterilization) Other Medical Device Related Standards 9
N All-or-Nothing Rating System - Internal Audit Checklist - ISO 9001 Internal Auditing 5