Choosing a laboratory for biocompatibility test
bio_subbu
Super Moderator
#1
Dear Forum users
This is regarding getting biocompatibility testing of devices done from an external laboratory as per ISO 10993. What all the pre-requisites of choosing/ selecting/ qualifying an external laboratory for getting biocompatibility test done, including accreditation to ISO/IEC 17025 or equivalent but not just limited to. Is that a requirement mandatory by regulation (if I would market them to such specific countries, and if so, grateful if you explain such country specific regulation) or it is an added value to our product testing by such reputation?
Thanks and Regards
S. Subramaniam, INDIA
Elsmar Forum Sponsor
GStough
Staff member
Super Moderator
#3
Dear Forum users
Thanks and Regards
I'm not sure if there are specific requirements for choosing an external lab for biocompatibility testing, but we use SATRA (located in the UK) for ours. We chose one that is a "one-stop shop" where we can get the most value from the service for the fees we pay.
There are others, as well. I think NAMSA is one. Perhaps someone else has other suggestions...:bigwave:
M
MIREGMGR
#4
EU:
EN ISO 17025 has mandatory applicability to laboratory organizations that provide EN ISO 10993 biocompatibility testing to medical device makers that are MDD compliant. You should consider only labs that can provide a 17025 certificate with appropriate scope for the period of time during which they performed testing for you.
FDA:
Laboratory organizations that provide ISO 10993 biocompatibility testing to medical device makers are at least strongly encouraged to be compliant with Good Laboratory Practices (GLP), 21 CFR 58. You should consider only labs that can certify in their results-report for you that they have conformed to GLP during the testing they performed for that report.
I don't know what requirements apply in China or Russia.
Japan utilizes 10993-based biocompatibility standards, but I don't know what requirements they have for labs.
I believe most if not all other nations will accept EU or FDA biocompatibility testing results, or some combination of the two.
bio_subbu
Super Moderator
#5
Thank you so much for your informative post. Recently, I have received a quotation from one of leading laboratory which is being providing the testing services for biocompatibility. The laboratory holding NABL accreditation (NABL provides laboratory accreditation services to laboratories that are performing tests / calibrations in accordance with ISO/IEC 17025 and ISO 15189:2003 for medical laboratories). But not being followed the FDA regulation 21 CFR Part 58. They quoted for both GLP and Non-GLP studies. For GLP studies, quoted little higher then Non- GLP, I inquired with the respective person about the reason for the difference, and they says that for GLP, In-house quality assurance people will audit for your entire biocompatibility testing process.
I am not interested to go with GLP since they are biased people auditing for the same. So what is your opinion on this?
Regards
S. Subramaniam
M
MIREGMGR
#7
I understand how you could have a theoretical ethical concern based on that explanation. In my experience such ethical issues are not relevant to the work of a well-selected, reputable lab. Also, I don't think they provided you with a very good explanation of GLP (21 CFR 58).
Generally the point of GLP is to assure the application of a standardized set of conditions and practices so that all testing to a given standard can be regarded as "apples vs. apples", i.e. technically comparable. This greatly simplifies the FDA's scientific workload by allowing them to judge the testing results without first having to conduct a technical review of the testing conditions and practices used in order to determine how those condtions and practices may have affected the results.
I suppose a lab of marginal reputability might do things differently for GLP and non-GLP testing. In my experience, reputable labs generally don't have any significant differences in their actual practices between the two project approaches, but they nonetheless charge more for GLP testing because they must prepare and archivally retain additional records and because some operations and recordkeeping processes must be witnessed and signed-off.
During a 510(k) process I usually ascertain what it will take to smoothly interact with the FDA, as disclosed in various guidance documents, and whenever possible I follow those guidelines. They want GLP, so I deliver GLP.
Z
zhang126
#9
WuXi AppTec, Inc.
St. Paul Facility
WuXi AppTec, Inc.
2540 Executive Drive
St. Paul, MN 55120
Toll-free: 888- 794-0077
Tel: 651- 675-2000
Fax: 651-675-2005
Driving Directions
Nearby Lodging
Atlanta Facility
WuXi AppTec, Inc.
1265 Kennestone Circle
Marietta, GA 30066
Toll-free: 888-847-6633
Tel: 770-514-0262
Fax: 770-514-0294
Driving Directions
Nearby Lodging
Philadelphia Facility
WuXi AppTec, Inc.
4751 League Island Blvd
Philadelphia, PA 19112
Toll-free: 800-622-8820
Tel: 215-218-5500
Fax: 215-218-5990
Thread starter Similar threads Forum Replies Date
Q Choosing Nonconformities to Report ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 26
C Quantifying risk in choosing the number of parts, operators and replicates in a GR&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
Q Old products new class - Dental Devices - Choosing tests EU Medical Device Regulations 2
M Choosing Auditors - ISO 9001 / ISO 27001 (UK) IEC 27001 - Information Security Management Systems (ISMS) 2
T AS9100D - Scope of QMS for New Company - Only Choosing a Function Subset Due to Management AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 14
E Choosing an ISO 9001 registrar with auditors familiar with our industry Registrars and Notified Bodies 10
M Dilemma about choosing the most applicable clause related to Risk ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 19
J Choosing QMS Software for Aerospace Company Quality Assurance and Compliance Software Tools and Solutions 5
B Choosing not to calibrate (IATF 16949) IATF 16949 - Automotive Quality Systems Standard 6
Q Choosing between ISO 9001 (2015) & TL 9000 certifications ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
H On choosing touchscreen displays and ensuring IEC 60601 compliance IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
Q Choosing In-Process Inspection Characteristics Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
L Choosing the correct Distribution for Acceptance Sampling Inspection, Prints (Drawings), Testing, Sampling and Related Topics 19
S Choosing a suitable type of Elisa to Test my Sample Misc. Quality Assurance and Business Systems Related Topics 3
S Choosing the correct Elisa Test Food Safety - ISO 22000, HACCP (21 CFR 120) 1
K Choosing a Six Sigma training organization Six Sigma 4
H Choosing between RABQSA Lead Auditor or ASQ CQA Certification Professional Certifications and Degrees 2
S Choosing ISO 9001 Training and if I need the training for work ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
S CE Marking choosing between module A, B CE Marking (Conformité Européene) / CB Scheme 2
A Getting Stuck Choosing a Sampling Plan Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
M Choosing the right Industry Financial Report Manufacturing and Related Processes 1
M Advice for choosing Rechargeable Lithium Batteries to be used in a Medical Device Other Medical Device Related Standards 16
L Choosing a Statistical Test for dissertation results! Statistical Analysis Tools, Techniques and SPC 4
I Choosing a X-Ray Fluorescence Measuring Device for Chrome Coating Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
J Choosing the Best Standard - ISO 9001 or ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 14
B Choosing Inspection Level and AQL AQL - Acceptable Quality Level 1
I Choosing the alpha level in an ANOVA Study Statistical Analysis Tools, Techniques and SPC 3
M Tolerance vs. Study Variation - Choosing the right one Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
S Choosing a Higher EER Air Conditioner After Work and Weekend Discussion Topics 7
T Choosing a Mentor Career and Occupation Discussions 3
K Selection Criteria for choosing a Management Representative for QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
A Choosing Document Control Software Document Control Systems, Procedures, Forms and Templates 13
S Choosing Parts for Attribute MSA (Measurement System Analysis) in 4th Edition, pg 132 Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
C Difference between softwares in choosing correct Taguchi designs Using Minitab Software 16
M Choosing a Laser Micrometer - Your recommendations? General Measurement Device and Calibration Topics 8
Marc Forum User Name - Choosing a forum User Name (aka 'handle' or 'screen name') Elsmar Cove Forum ToS and Forum Policies 0
G Short TQM training - Choosing of most important points Training - Internal, External, Online and Distance Learning 3
B Choosing number of subgroups and subgroup size Statistical Analysis Tools, Techniques and SPC 4
D Choosing a Vision System - Recommendations for a User Friendlier System Quality Tools, Improvement and Analysis 3
D Need Help on choosing dimensional standards / equipment General Measurement Device and Calibration Topics 2
J Choosing Samples for Gage R&R - Randomly picked samples show very little variation Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 13
Tim Folkerts DOE: Choosing a Design - Factors that affect the choice of design The Reading Room 0
B ISO10012:2003 Question - Choosing or assessing the capability of a piece of equipment Other ISO and International Standards and European Regulations 1
G Choosing a Sampling plan for Start Up Fabless IC Company - Guarantee a 200 DPM limit Inspection, Prints (Drawings), Testing, Sampling and Related Topics 9
gard2372 Choosing between a Small Company or a Large Company Career and Occupation Discussions 17
D Choosing a notified body for a new start up company in the UK - Medical Device area Registrars and Notified Bodies 6
G Choosing Supplier Evaluation Methods - Determining what a Critical Supplier is Supplier Quality Assurance and other Supplier Issues 31
Steve Prevette Choosing Performance Indicators Registered Visitor Articles Archive 0
C Questions to Ask Prospective Registrars Before Choosing Registrars and Notified Bodies 3
L Choosing a Registrar - Need registrar selection advice Registrars and Notified Bodies 6
Similar threads
Top Bottom