I am a little confused about the requirements of the MOP.
Should the insulation between the primary and secondary circuit be 2 MOPP despite the fact that the applied part is the enclosure part made of insulation?
The secondary circuit is connected to the PE as is the enclosure. Thus applied part's only contact with secondary is through earthed enclosure.
It is unclear do 2 MOPP are really needed between primary and secondary circuits (or should be 2 MOOP?). In this case the non-medical PSU is used. Or just use medical grade PSU?
And how to determine MOP between applied part (Type B) and secondary? The applied part is integrated in the metal enclosure which is earthed so as the secondary circuit.
Benjamin Weber
In my oppinion you have to implement the following MOP:
  • 1 MOOP between L and N (see cl. 8.9.1)
  • 2 MOOP between accessible enclosure and mains voltage (1 MOOP is aachieved by PE)
  • 2 MOOP between accessible enclosure and secondary (1 MOOP is achieved by PE)
    • I hear you say: "Why? It's below 60 Vdc so it is SELV!" (btw: There is no such term as SELV defined in 60601-1)
    • According to cl. 8.4.2 c) this voltage is only safe, if the corresponding power is less than 240 VA for 1 minute (in NC and SFC). So it depneds on the maximum power output of the PSU and if this is limited to the power value even in SFC.
  • 2 MOPP between the patient connection of the applied and mains voltage
    • If the applied part is only the insulated part, PE cannot be considered as MOPP (only metal parts can be protectively earthed)
    • If the whole enclosure part at the right side can be seen as the applied part and the insulating part is the patient connection only, PE could be considered as 1 MOPP
  • 2 MOPP between the patient connection of the applied part and the secondary voltage
    • see both additional bullet points above
Regarding your question on the 2 MOPP to mains in particular:
  • You can implement these at the AC-DC-converter. But you will still need 2 MOPP regarding the secondary voltage.
  • If the insulating material is strong enough for 2 MOPP regarding mains voltage (i.e 4 kV dieelctric strength), then you could fulfill the 2 MOPP for mains and secondary at one point.
EMengineer
Thank you Benjamin for clarification.
Here is the updated insulation diagram:
According your suggestion:
  • 1 MOOP between L and N (see cl. 8.9.1) - Point A
  • 2 MOOP between accessible enclosure and mains voltage (1 MOOP is achieved by PE) - Point B, in this case enclosure is of insulating material thus 2 MOOP.
  • 2 MOOP between accessible enclosure and secondary (1 MOOP is achieved by PE) -Point C
  • 2 MOPP between the patient connection of the applied and mains voltage - Point D
All other parts are clear, here is the part where I got confused originally. I tried to apply 2 MOPP requirements for insulation in PSU between primary and secondary, but a.c. and c.d. come a little short (around 7 mm). If I consider earth as 1 MOP does it mean that earthed secondary circuit could be separated from mains by c.d and a.c for 1 MOPP?
As you mentioned:
At first I considered this case, the only possible contact to mains is through earthed enclosure. In this case I could not decide whether I should apply 2 MOPP (a.c and c.d) between PSU primary and secondary (secondary is connected to enclosure through earth), or should I consider earth as 1 MOP.
Applied part is glass where patient puts his body part (through that glass camera takes pictures). But patient connection is in the middle of the glass, so does the edges of the glass plane can be treated as an additional insulating measure?:
In that case this additional distance would sum up with distance between mains part and secondary thus forming adequate distance for 2 MOPP between mains and patient?
  • 2 MOPP between the patient connection of the applied part and the secondary voltage - Point E
L Choosing the correct Distribution for Acceptance Sampling Inspection, Prints (Drawings), Testing, Sampling and Related Topics 19
S Choosing the correct Elisa Test Food Safety - ISO 22000, HACCP (21 CFR 120) 1
C Difference between softwares in choosing correct Taguchi designs Using Minitab Software 16
S Help needed in choosing the method of calculating the minimum sample size Internal Auditing 12
T Choosing a Notified Body for MDR - SGS/BSI/DNV/DQS-MED Registrars and Notified Bodies 0
Q Choosing Nonconformities to Report ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 26
C Quantifying risk in choosing the number of parts, operators and replicates in a GR&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
Q Old products new class - Dental Devices - Choosing tests EU Medical Device Regulations 2
M Choosing Auditors - ISO 9001 / ISO 27001 (UK) IEC 27001 - Information Security Management Systems (ISMS) 2
T AS9100D - Scope of QMS for New Company - Only Choosing a Function Subset Due to Management AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 14
E Choosing an ISO 9001 registrar with auditors familiar with our industry Registrars and Notified Bodies 10
M Dilemma about choosing the most applicable clause related to Risk ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 19
J Choosing QMS Software for Aerospace Company Quality Assurance and Compliance Software Tools and Solutions 5
B Choosing not to calibrate (IATF 16949) IATF 16949 - Automotive Quality Systems Standard 6
Q Choosing between ISO 9001 (2015) & TL 9000 certifications ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
H On choosing touchscreen displays and ensuring IEC 60601 compliance IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
Q Choosing In-Process Inspection Characteristics Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
S Choosing a suitable type of Elisa to Test my Sample Misc. Quality Assurance and Business Systems Related Topics 3
K Choosing a Six Sigma training organization Six Sigma 4
H Choosing between RABQSA Lead Auditor or ASQ CQA Certification Professional Certifications and Degrees 2
S Choosing ISO 9001 Training and if I need the training for work ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
S CE Marking choosing between module A, B CE Marking (Conformité Européene) / CB Scheme 2
A Getting Stuck Choosing a Sampling Plan Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
M Choosing the right Industry Financial Report Manufacturing and Related Processes 1
M Advice for choosing Rechargeable Lithium Batteries to be used in a Medical Device Other Medical Device Related Standards 16
L Choosing a Statistical Test for dissertation results! Statistical Analysis Tools, Techniques and SPC 4
I Choosing a X-Ray Fluorescence Measuring Device for Chrome Coating Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
J Choosing the Best Standard - ISO 9001 or ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 14
B Choosing Inspection Level and AQL AQL - Acceptable Quality Level 1
I Choosing the alpha level in an ANOVA Study Statistical Analysis Tools, Techniques and SPC 3
M Tolerance vs. Study Variation - Choosing the right one Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
S Choosing a Higher EER Air Conditioner After Work and Weekend Discussion Topics 7
T Choosing a Mentor Career and Occupation Discussions 3
K Selection Criteria for choosing a Management Representative for QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
A Choosing Document Control Software Document Control Systems, Procedures, Forms and Templates 13
S Choosing Parts for Attribute MSA (Measurement System Analysis) in 4th Edition, pg 132 Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
bio_subbu Choosing a laboratory for biocompatibility test ISO 13485:2016 - Medical Device Quality Management Systems 8
M Choosing a Laser Micrometer - Your recommendations? General Measurement Device and Calibration Topics 8
G Short TQM training - Choosing of most important points Training - Internal, External, Online and Distance Learning 3
B Choosing number of subgroups and subgroup size Statistical Analysis Tools, Techniques and SPC 4
D Choosing a Vision System - Recommendations for a User Friendlier System Quality Tools, Improvement and Analysis 3
D Need Help on choosing dimensional standards / equipment General Measurement Device and Calibration Topics 2
J Choosing Samples for Gage R&R - Randomly picked samples show very little variation Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 13
Tim Folkerts DOE: Choosing a Design - Factors that affect the choice of design The Reading Room 0
B ISO10012:2003 Question - Choosing or assessing the capability of a piece of equipment Other ISO and International Standards and European Regulations 1
G Choosing a Sampling plan for Start Up Fabless IC Company - Guarantee a 200 DPM limit Inspection, Prints (Drawings), Testing, Sampling and Related Topics 9
gard2372 Choosing between a Small Company or a Large Company Career and Occupation Discussions 17
D Choosing a notified body for a new start up company in the UK - Medical Device area Registrars and Notified Bodies 6
G Choosing Supplier Evaluation Methods - Determining what a Critical Supplier is Supplier Quality Assurance and other Supplier Issues 31
Steve Prevette Choosing Performance Indicators Registered Visitor Articles Archive 0