Clarification regarding tests in IEC 60601-2-25

subbash

Involved In Discussions
#1
I am new to testing the ECG device as per IEC 60601-2-25 standard. I want to test a battery operated ECG device, which can work in 6 lead or 12 lead mode.

I have gone through the IEC60601-2-25 document and have few aspects that needs clarification.

1. In the indication of Inoperable ECG test. The standard says that when the saturation point is reached, we need to check that the amplitude of the signal (10 Hz, 1 mV) that is fed is reduced to less than 0.5 mV.

Clarification required: Observing the reduction in amplitude when saturation point is reached is it the only requirement or do we need to have any indicator.


2. On the ECG report we are only printing Heart rate, RR,PR interval, QRS, QT/QTc duration, QRS Axis. But there are some tests which require measurements related to amplitude and ST segment details. The standard does not mention if we have to print on the ECG record.

Clarification required: If we have to print the details on the report.

3. Recovery Time test: Is this test applicable to an ECG device which have a lead switch function only..?



Request you to provide me some suggestions and share your experience in qualifying the ECG device as per IEC60601-2-25 standard

Please also share your experience about the testing device if you have used such as Whaleteq's SECG, MECG, CMRR devices.

Thanks and Regards,
Subbash
 
Last edited:
Elsmar Forum Sponsor
#2
I think,
1. An indication is required in the ECG report to identify the device was in inoperable (saturation or overloaded)
this Indication should visible on the report (like some instruction "device inoperable", or indicated by the absence of a visible trace in the report.

2. The standard states like "if measurements are provided" so it's not mandatory to provide the measurements. (if it's related with 201.12.1.101.1 Automated measurements on ECG)

3. The Recovery Time test will be applicable for the device, but no impact due to this test as the modern ECG devices process all leads together and mostly no hardware lead switch used, (this test will have an impact if the filter required reinitialization after the lead switching)
 
#3
I think,
1. An indication is required in the ECG report to identify the device was in inoperable (saturation or overloaded)
this Indication should visible on the report (like some instruction "device inoperable", or indicated by the absence of a visible trace in the report.

2. The standard states like "if measurements are provided" so it's not mandatory to provide the measurements. (if it's related with 201.12.1.101.1 Automated measurements on ECG)

3. The Recovery Time test will be applicable for the device, but no impact due to this test as the modern ECG devices process all leads together and mostly no hardware lead switch used, (this test will have an impact if the filter required reinitialization after the lead switching)
With reference to point 3, can you please elaborate what exactly you mean by filter reinitialization ?
 
Thread starter Similar threads Forum Replies Date
A ISO 2859 Single Sampling - Clarification regarding the sampling table Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
S Clarification regarding types of processes ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
A 60601-1 clause 8.11.3.6 Test Clarification IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Assignable causes rules clarification Statistical Analysis Tools, Techniques and SPC 4
qualprod D5 of 8D clarification, how to verify root cause Problem Solving, Root Cause Fault and Failure Analysis 26
K IPC-610 Section 10.4.2.2 clarification - Distance to be measured Various Other Specifications, Standards, and related Requirements 0
M Off-Label Use - Clarification of FDA Policy US Food and Drug Administration (FDA) 1
T Implant Card - Article 18.1(a) and MDCG 2019-8 clarification EU Medical Device Regulations 3
Q Need clarification on requirements.... Class i, gmp & 510(k) exempt Medical Device and FDA Regulations and Standards News 12
M Informational TGA Consultation: Proposed clarification of the regulatory requirements for medical device systems and procedure packs Medical Device and FDA Regulations and Standards News 2
R ASQ reference material clarification - Spiral bound materials allowed in ASQ Exam? Professional Certifications and Degrees 1
Q ISO 3310 Clarification Help - Aperture sizes for sieves used for particle sorting Other ISO and International Standards and European Regulations 2
S The Severity of a Medical Device Hazard - Risk Analysis Clarification ISO 14971 - Medical Device Risk Management 6
M 8.3.2.3 Development of products with embedded software - request for clarification IATF 16949 - Automotive Quality Systems Standard 1
M FDA News USFDA Draft Guidance – Clarification of Radiation Control Regulations for Manufacturers of Diagnostic X-Ray Equipment Medical Device and FDA Regulations and Standards News 0
T Clarification on MDR - Article 18(d) - Implant Card EU Medical Device Regulations 14
S QS, RS deflection - clarification wanted IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
S Requirements for Interval Measurement test & Frequency Response test clarification IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
J Internal Audit clarification - How to perform the audits IATF 16949 - Automotive Quality Systems Standard 6
V Clarification - Hydrogen De-embrittlement Manufacturing and Related Processes 6
K UDI Direct Marking Compliance Date Clarification and one other UDI question Other US Medical Device Regulations 0
N Applied Parts Classification Clarification - Breast Pump IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
V Clarification of Injection part shrinkage ratio Manufacturing and Related Processes 1
J ISO 9001:2015 8.2.3 - Review of Requirements (Clarification on compliance) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
B Health Canada Recall Definition - Seeking Clarification Canada Medical Device Regulations 5
Q ISO 9001:2015 - Clarification in 6.1.2 Note 1 (Options to Address Risks) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
S Clarification of threaded ring gage calibration procedure/requirements General Measurement Device and Calibration Topics 2
M Clarification on Calibration/Verification Records 7.1.5.2.1d (IATF 16949) General Measurement Device and Calibration Topics 11
B Clarification on interpretation of some EN ISO 14971:2012 & IEC 62304:2006 req's ISO 14971 - Medical Device Risk Management 46
S AS9102 - Clarification - PO asking for an Assembly at Rev B (Print at Rev C) AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
Pmarszal Clarification for 21 CFR Part 11.100 - General Requirements Other US Medical Device Regulations 14
B Clarification of ISO 9001:2015 Clause 8.5.6 Control of Changes ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
M Integrated Phased Processes - AS9100D cl. 8.1 Operational Planning - Clarification AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
A Monitoring and Measuring Resources (7.1.5) - Clarification ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
dubrizo Clarification Requested in 6.2.2 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
F TS 16949 Clause 7.2.1 - Note 2 - Recycling program - Clarification IATF 16949 - Automotive Quality Systems Standard 4
K EN ISO 15223-1:2012 Clarification or Examples on when to use Safety Symbols Other Medical Device Related Standards 3
D Clarification of Applicability of TS 16949 Requirements to a Non-Automotive Business IATF 16949 - Automotive Quality Systems Standard 13
M Request for clarification on TS 16949 Clause 5.6.1.1 IATF 16949 - Automotive Quality Systems Standard 5
Q Configuration management clarification and example AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
S Clarification in organizing required documents for ISO 27001 IEC 27001 - Information Security Management Systems (ISMS) 6
J Definition Actively Manufacturing - ISO 13485 Definition and clarification Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 8
T Early Research & Development - ISO 13485:2003 requirements Clarification ISO 13485:2016 - Medical Device Quality Management Systems 34
B Detachable Power Supply Cable Connection ESD Clarification IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
D NIST HDBK 44 Table T.3. Class III Tolerance in Divisions Clarification General Measurement Device and Calibration Topics 4
M UDI - Direct Marking and Reprocessing Clarification Other US Medical Device Regulations 12
S 21 CFR Part 820.40(b) Clarification on Required Document Approvers 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
E Clarification on Document Signatories under ISO9001 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
P ISO 27001:2013 Clause 4.1 and 4.2 Clarification and Guidance IEC 27001 - Information Security Management Systems (ISMS) 13
C Positional Tolerance - Bonus / Datum Shift / ASME Y14.5M - Clarification Various Other Specifications, Standards, and related Requirements 9

Similar threads

Top Bottom