Clarifying FDA definition of "finished device" and “capable of functioning”

outdoorsNW

Quite Involved in Discussions
#1
We are an electronics contract mfg., high mix, low volume. We are looking at targeting medical electronics companies. We are at an early stage and am trying to determine where we may need to hire outside experts and what we should avoid until we have a better understanding of the requirements.

For the moment I think we should avoid producing “finished medical devices” because of the added regulatory requirements. However the definition of what is a finished medical device leaves room for interpretation.

From the FDA regulations: “820.3 Definitions. (l) Finished device means any device or accessory to any device that is suitable for use or capable of functioning, whether or not it is packaged, labeled, or sterilized.” [Emphasis added]

How is “capable of functioning” defined?

When I searched both the web and elsmar, I did not find anything clarifying this.

For instance, with an electronic item, the device may have been fully assembled but not tested. If you have a good customer design and a good manufacturing process, an untested device usually will be fully functional. If the design responsible customer is responsible for final testing, is this considered “capable of functioning” when we send it to our customer?

What if the scenario changes and the device needs calibration or some other setup before use, but may function with reduced performance, such as reduced accuracy?
 
Elsmar Forum Sponsor

monoj mon

Involved In Discussions
#3
How is “capable of functioning” defined?
Let me give you a simple example. Suppose you manufacture a suction liner which is suitable as an accessory to any manual or electronic suction device. If you manufacture every single component of the suction liner and assemble it at your facility, it will be considered as the finished device. Because, in that form anyone can use your suction liner to fit it with any type of suction device. However, if you just manufacture the fluid containing part while rest of the parts (such as the lid, tubing etc.) are manufactured by some one else and all the components are sent to your client's company for assembly then none of the third party manufacturers are considered to be manufacturing a "Finished Device" as the individual components can not be used by anyone as a "Device".
For instance, with an electronic item, the device may have been fully assembled but not tested.
In your case, as you have said, you fully assemble the device. That means, anyone can pick your product and attempt to use it as a "Device" (refer to the FDA's definition of device) which would mean that your device is "Capable of functioning".
If the design responsible customer is responsible for final testing, is this considered “capable of functioning” when we send it to our customer?
No, that is not what it means by "capable of functioning". It means, the product can be used as intended even though it fail and harm the patient.
What if the scenario changes and the device needs calibration or some other setup before use, but may function with reduced performance, such as reduced accuracy?
Exactly, if anyone can pick your device and perform the intended use even though with reduced accuracy or failure, still your device will be considered "capable of function".

This is why, it is important for the third party manufacturer of "finished device" to register and list the device with the FDA. Registration and Listing will make FDA aware of your existence and ultimately your facility will be inspected in the future.
 
Thread starter Similar threads Forum Replies Date
M FDA News FDA Releases Draft Guidance Clarifying Application of ISO 10993-1 Biocompatibility Standard Medical Device and FDA Regulations and Standards News 0
M Clarifying the many different Medical Device national and international standards Other Medical Device Related Standards 4
K Two FDA questions regarding UDI (and potentially 803.52 MDR) Other US Medical Device Regulations 0
R Importers - For the FDA Registration/ Listing, is the CMO the importer? Medical Device and FDA Regulations and Standards News 6
Z Iterative development and FDA change requests IEC 62304 - Medical Device Software Life Cycle Processes 2
P FDA Approved Product Contact Parts ISO 13485:2016 - Medical Device Quality Management Systems 6
A FDA guidance on non-sterile Medical Device Packaging Medical Device and FDA Regulations and Standards News 6
JoCam FDA Registration for Sub-contract manufacturers Medical Device and FDA Regulations and Standards News 2
M Is IEC 60601-1-2 required by FDA for all electronic medical devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
E Contract manufacturer FDA requirements foreign company US Food and Drug Administration (FDA) 6
P Choice of FDA reviewer US Food and Drug Administration (FDA) 6
J Is Device Tracking Card an actual requirement under FDA regulation? Other US Medical Device Regulations 0
T Definition Human Use (FDA) Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 7
K FDA 510k electrosurgery 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
J Leveraging FDA 510k Clearance for International Registrations Other Medical Device Regulations World-Wide 2
shimonv FDA News FDA guidance on Multiple Function Device Products (8/2020) Other US Medical Device Regulations 1
N FDA UDI - Label vs. Labeling - Does the insert need to include UDI? Other US Medical Device Regulations 1
J FDA notification of address change US Food and Drug Administration (FDA) 2
N Usability testing required for FDA IDE (investigational device exemption)? Human Factors and Ergonomics in Engineering 3
A FDA Class Classification for a cabinet 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
L FDA Biocompatibility Requirements - Transitory Contact Other US Medical Device Regulations 1
Edward Reesor FDA DM and /or Class I Life Saving US Food and Drug Administration (FDA) 0
Watchcat FDA will not tolerate fraud…meaning what, exactly? Other US Medical Device Regulations 6
K FDA Premarket Cybersecurity Guidance - 4 questions Other US Medical Device Regulations 5
gunnyshore Form FDA 3500A MedWatch eSubmitter 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
Watchcat FDA Webinar Series - N95 Respirators Other US Medical Device Regulations 0
M Off-Label Use - Clarification of FDA Policy US Food and Drug Administration (FDA) 1
J Sub-supplier change from manual to automated process - same specs - Report to FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
C FDA Establishment registration - Buying some medical devices from another manufacturer Medical Device and FDA Regulations and Standards News 5
I How to classify a medical device based on FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
dinaroxentool Question about FDA Classification of a Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
S The US FDA requirements on Disposal of a medical device US Food and Drug Administration (FDA) 1
S FDA Requirements for Medical Device Label Reconciliation 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
S What is considered a "core algorithm"? (From an FDA guidance document) Medical Information Technology, Medical Software and Health Informatics 4
B FDA-Medical Device Reporting (MDR )procedure compliant with 21CFR section 803 US Food and Drug Administration (FDA) 0
Stoic Are any medical device companies using the 2011 FDA process validation guidance instead of GHTF/SG3/N99-10:2004? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
D ISO 13485, FDA 21 CFR 820 and Auditing the Accounting Department ISO 13485:2016 - Medical Device Quality Management Systems 5
S FDA's E submitter - For clinical electronic thermometer Medical Device and FDA Regulations and Standards News 6
J FDA wants electrical safety testing on battery powered medical device US Food and Drug Administration (FDA) 11
A FDA and NB audit of Engineering Drawings in DHF and DMR. Medical Device and FDA Regulations and Standards News 1
D FDA De Novo Cover Letter - What is expected US Food and Drug Administration (FDA) 3
Ed Panek 21 CFR Part 820 - FDA Label Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
J Mislabeling - Consider this an FDA notified recall? CFR 806.10 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
O Hospital-Developed, In-House Device - FDA Clearance? Other US Medical Device Regulations 19
D FDA Approved Class II Medical Device? Other US Medical Device Regulations 11
R When does the FDA consider a component a medical device? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 17
N Risk Management besides mandated FDA requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
M Informational US FDA – Ear, Nose, and Throat Devices; Classification of the Self-Fitting Air-Conduction Hearing Aid Medical Device and FDA Regulations and Standards News 0
L UDI Requirments Gap Analysis - EU MDR vs FDA CFR EU Medical Device Regulations 8
M Informational US FDA – Modifications to the List of Recognized Standards, Recognition List Number: 052 Medical Device and FDA Regulations and Standards News 0

Similar threads

Top Bottom