Class I devices in the extended MDR transition period

regularengineer

Starting to get Involved
Are all Class I devices under MDD expected to be included in the MDR transition period through 2028?

From what I have seen, only certain Class I devices included in the extension (sterile/measuring function/reusable). My team is working on a product that is Class I medical device software under MDD, and we are wondering if we can continue to leverage the MDD certificate/CE mark through this transition period.

I am aware that Class I devices were to be transitioned in 2021, but wondering if anyone is aware of some "wiggle room" so to speak.

Thanks in advance
 

KShaw

Involved In Discussions
Hey there,

The extension will only apply to legacy devices, i.e. those scenarios mentioned under Article 120:
  • Devices that did not require NB intervention under the MDD but require under the MDR.
  • Devices covered by a valid Directive CE certificate.
So "simple" Class I devices will not be covered by the extension. In your case, if this is a Class I SaMD, you will need to immediately comply with the MDR.

If you do have a valid MDD certificate for your products, then you could still reap the benefits of the extension. But beware of the T&Cs..

Best of luck,
Ken
 

iam1235

Starting to get Involved
We are in the same case as you, we will try to contact our NB in addition in our case, if you are in France, there is the new repository of interoperability that comes into force so if we are in the same case we must see if with the integration of these new features we are supposed to go to MDR. Any one can tell us what should we do ? thank you
 

Raisin picker

Quite Involved in Discussions
For basic class I devices (not Ir, Is, Im), you do not need a NB. So no need there to extend the transition period. Just finalize your documents (they should already be done), sign your MDR DoC, and you're good.

Sorry for being direct, but the transition period is not for lazy manufacturers, but to cover for overworked NBs.
 

iam1235

Starting to get Involved
Software under MDD will change class to IIa according to the RDM, so to be certified it needs a NB. In this case it is necessary to see the conditions defined (for example not to have significant changes) to see if the manufacturer can benefit from the transition period.
 

regularengineer

Starting to get Involved
Hey there,

The extension will only apply to legacy devices, i.e. those scenarios mentioned under Article 120:
  • Devices that did not require NB intervention under the MDD but require under the MDR.
  • Devices covered by a valid Directive CE certificate.
So "simple" Class I devices will not be covered by the extension. In your case, if this is a Class I SaMD, you will need to immediately comply with the MDR.

If you do have a valid MDD certificate for your products, then you could still reap the benefits of the extension. But beware of the T&Cs..

Best of luck,
Ken
Thanks Ken! I'm a little unclear as I understand we had a CE certificate under MDD - wouldn't this mean we are a legacy device under Article 120. That does seem to be the case based on your last line as well, but I'm also unsure about these T&Cs! It would seem that the fact this extension only applies to a certain subset of Class I's (Is/Ir/Im) may be one of those T&Cs, but just trying to be sure.

We really do not feel that the software falls under Class IIa under MDR, particularly when reviewing MDCG 2021-24 and its description of Rule 11.
 

regularengineer

Starting to get Involved
For basic class I devices (not Ir, Is, Im), you do not need a NB. So no need there to extend the transition period. Just finalize your documents (they should already be done), sign your MDR DoC, and you're good.

Sorry for being direct, but the transition period is not for lazy manufacturers, but to cover for overworked NBs.
Thanks! And agree on the last point - we are mindful that the NBs are overworked and certainly do not intend to "exploit" this situation, but nevertheless there is an opportunity for us to continue to get our product to patients and we want to take advantage.

I want to be crystal clear though: as we have a class I MDD certificate, do we fall into the transition period?
 

Newbie22

Starting to get Involved
Refer to Article
Hey there,

The extension will only apply to legacy devices, i.e. those scenarios mentioned under Article 120:
  • Devices that did not require NB intervention under the MDD but require under the MDR.
  • Devices covered by a valid Directive CE certificate.
So "simple" Class I devices will not be covered by the extension. In your case, if this is a Class I SaMD, you will need to immediately comply with the MDR.

If you do have a valid MDD certificate for your products, then you could still reap the benefits of the extension. But beware of the T&Cs..

Best of luck,
Ken

___________

I agree with Ken.

Your Class I device is not considered a legacy device, as per by Article 120 thus not covered by the extended transitional agreement.

Has the device been a Class Is or Im then Article 120 3b would have been applicable

3(b) 31 December 2028, for class IIb devices other than those covered by point (a) of this paragraph, for class IIa devices, and for class I devices placed on the market in sterile condition or having a measuring function.

Had the Class I device been up-classified under the MDR, and the DoC been written prior to 26th May 2021 then 120 (3) b would been applicable.

3b. Devices for which the conformity assessment procedure pursuant to Directive 93/42/EEC did not require the involvement of a notified body, for which the declaration of conformity was drawn up prior to 26 May 2021 and for which the conformity assessment procedure pursuant to this Regulation requires the involvement of a notified body, may be placed on the market or put into service until 31 December 2028.

Refer to:

PE-CONS 1/23 – 2023/0005(COD)
REGULATION (EU) 2023/... OF THE EUROPEAN PARLIAMENT AND OF THE COUNCIL of ... amending Regulations (EU) 2017/745 and (EU) 2017/746 as regards the transitional provisions for certain medical devices and in vitro diagnostic medical devices

Also, when applying MDR Article 120, the risk class applicable is the risk class of the device as per Directive 93/42/EEC not MDR 2017/745 :)

I hope the above helps
 

KShaw

Involved In Discussions
Thanks! And agree on the last point - we are mindful that the NBs are overworked and certainly do not intend to "exploit" this situation, but nevertheless there is an opportunity for us to continue to get our product to patients and we want to take advantage.

I want to be crystal clear though: as we have a class I MDD certificate, do we fall into the transition period?

Taking into consideration that your software is not sterile, is your MDD certificate due to some measuring function your software has? Is the Notified-Body issued certificate still valid?
 

iam1235

Starting to get Involved
Refer to Article


___________

I agree with Ken.

Your Class I device is not considered a legacy device, as per by Article 120 thus not covered by the extended transitional agreement.

Has the device been a Class Is or Im then Article 120 3b would have been applicable

3(b) 31 December 2028, for class IIb devices other than those covered by point (a) of this paragraph, for class IIa devices, and for class I devices placed on the market in sterile condition or having a measuring function.

Had the Class I device been up-classified under the MDR, and the DoC been written prior to 26th May 2021 then 120 (3) b would been applicable.

3b. Devices for which the conformity assessment procedure pursuant to Directive 93/42/EEC did not require the involvement of a notified body, for which the declaration of conformity was drawn up prior to 26 May 2021 and for which the conformity assessment procedure pursuant to this Regulation requires the involvement of a notified body, may be placed on the market or put into service until 31 December 2028.

Refer to:

PE-CONS 1/23 – 2023/0005(COD)
REGULATION (EU) 2023/... OF THE EUROPEAN PARLIAMENT AND OF THE COUNCIL of ... amending Regulations (EU) 2017/745 and (EU) 2017/746 as regards the transitional provisions for certain medical devices and in vitro diagnostic medical devices

Also, when applying MDR Article 120, the risk class applicable is the risk class of the device as per Directive 93/42/EEC not MDR 2017/745 :)

I hope the above helps

Hello, Thank you for these details.
If I understand correctly, self-certified medical device software according to MDD does not benefit from the transition period of 2028?
 
Thread starter Similar threads Forum Replies Date
J Is ISO 13485 certification necessary for Class I (non-sterile) medical devices? ISO 13485:2016 - Medical Device Quality Management Systems 7
D Correct form to amend medical devices class III (IMDRF vs Form 2016) Canada Medical Device Regulations 0
O Is a Certified 13485 for class I devices needed ISO 13485:2016 - Medical Device Quality Management Systems 3
T PMCF Surveys for Class II a devices: basic requisites EU Medical Device Regulations 0
T PMCF Surveys for Class II a devices : basic requisites CE Marking (Conformité Européene) / CB Scheme 0
C Legacy device under Class A sterile devices EU Medical Device Regulations 0
N Shelf life guidance EU MDR, class 1 and class 2 Medical Devices EU Medical Device Regulations 2
P Servicing of Class IIb devices under MDD EU Medical Device Regulations 2
M Custom made class 1 Medical devices EU Medical Device Regulations 3
adztesla Stability sampling for Class 1 medical devices ISO 13485:2016 - Medical Device Quality Management Systems 1
J The necessity and benefit of ISO13485:2016 for a manufacturer of class I devices ISO 13485:2016 - Medical Device Quality Management Systems 3
L Change Notification Requirements for class 1 devices Medical Device and FDA Regulations and Standards News 1
D Control Number for Class III and IV devices Canada Medical Device Regulations 1
J Instructions for use for Class I devices under MDR EU Medical Device Regulations 1
drwallice Electrical Medical Devices class I EU Medical Device Regulations 0
L Class IIb devices marked EC according to the 93/42/CE - Portugal CE Marking (Conformité Européene) / CB Scheme 3
S IVD risk class II devices for Brazil and MDSAP Other Medical Device Regulations World-Wide 0
S We are looking for EU authorised Representative for both Class I and Class IIb devices EU Medical Device Regulations 7
S How is the service life for class I reusable medical devices described in MDR? EU Medical Device Regulations 7
S How is the service life described for class I reusable medical devices in MDR? Elsmar Xenforo Forum Software Instructions and Help 1
F UDI - Unit of Use and Class I, single-use devices EU Medical Device Regulations 5
A Medical Device Vigilance decision tree for Japan for class 2 devices. Japan Medical Device Regulations 1
K Contamination Control - Class Is medical devices (Clause 6.4.2 ISO 13485:2016 (E)) ISO 13485:2016 - Medical Device Quality Management Systems 12
W Seeking Guidance Verification Test Strategy for Class B Medical Devices IEC 62304 - Medical Device Software Life Cycle Processes 1
Q Old products new class - Dental Devices - Choosing tests EU Medical Device Regulations 2
M Post Market Surveillance for Class II Medical Devices Medical Device and FDA Regulations and Standards News 3
M Special Controls and Class II Medical Devices Medical Device and FDA Regulations and Standards News 18
J ISO 14971 applied to ISO 13485? Low risk class 1 devices ISO 13485:2016 - Medical Device Quality Management Systems 5
rob73 MDCG guidance on class 1 devices and MDR EU Medical Device Regulations 0
S Until when can we CE-mark and sell MDD class 1 devices? CE Marking (Conformité Européene) / CB Scheme 23
T Rumours that class 1 NS, NM, NR devices will have a new deadline for MDR compliance EU Medical Device Regulations 3
H Clinical evidence for Class II medical devices in EU and US EU Medical Device Regulations 6
N How to look up all Class 1 medical devices on sale in Canada Canada Medical Device Regulations 3
E Different requirements for Class II and Class III Medical Devices Other ISO and International Standards and European Regulations 5
Dobby1979 Applicable MDR Annex's for Class IIa Medical Devices EU Medical Device Regulations 0
Q MDR Effectivity date for Legacy Class I Medical Devices EU Medical Device Regulations 5
M Informational ANVISA – Notification for Class 1 devices – Resolução da Diretoria Colegiada – RDC nº 270 de 28/02/2019 Medical Device and FDA Regulations and Standards News 0
A Distributing Replacement Parts for Class II Medical Devices Other Medical Device Related Standards 13
Y Informational Change control process - Major vs Minor change - Active class III medical devices ISO 13485:2016 - Medical Device Quality Management Systems 13
Z Direct Part Marking (DPM) requirements for Class II Medical Devices Manufacturing and Related Processes 0
S Design Control for Class I Medical Devices 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
M Medical Device News FDA Guidance - UDI: Policy Regarding Compliance Dates for Class I and Unclassified Devices and Certain Devices Requiring Direct Marking Other US Medical Device Regulations 0
S What is the significance of UDI date requirements for Class 1 devices? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
B Can Class IIa Dental Medical Devices be sold into the EU without CE EU Medical Device Regulations 2
J MDR 2020 for Class 1 Medical Devices EU Medical Device Regulations 8
S Importing Medical devices class I in EU Other Medical Device and Orthopedic Related Topics 5
S Scope change in Japan - Class I disposable medical devices Japan Medical Device Regulations 0
D Technical file structure/content requirements for Class IIa devices & new MDR EU Medical Device Regulations 14
N Sampling for stability studies - Class II surgically invasive devices Other Medical Device and Orthopedic Related Topics 1
M Who polices/surveils Class I Medical Devices in the EU? EU Medical Device Regulations 6

Similar threads

Top Bottom