Hi everyone,
For some reasons I won't developed here our NB has informed us that we have "surclassified" one of our devices we market. In their opinion, this device (an accessory actually) is a Class I device per the annex IX of the MDD.
This accessory is intended to be used with our main parent device, a class IIb therapeutic ultrasound device. The purpose of this accessory is similar to the coupling gel of an ultrasound scanner: enabling the transmission of the ultrasound energy from the machine head to the tissues. However, in our case, this accessory is not gel but a plastic pouch filled with a specific liquid inserted between the transducer and the skin.
Based on the annex IX, I actually agree that this accessory is a Class I device per rule 1. Especially because it cannot be classified as an active medical device, according to following part of the definition of an active device:
"Medical devices intended to transmit energy, substances or other elements between an active medical device and the patient, without any significant change, are not considered to be active medical devices"
Here is my problem. The annex IX should be a risk-based classification scheme. Therefore if our accessory becomes a Class I (non sterile/no measuring function) device, it means that there is no/little risk associated with its use and then it can be self-certified.
But there are risks associated with this accessory. Of course the risks originate from the ultrasound parent device, but if this accessory is not properly designed, verified, validated and manufactured, the parent device may harm patients.
Some parts of the validation of the parent device are done with this accessory, but still, putting this accessory out of the CE mark/ISO 13485 audit scope seems hazardous to me.
Your opinion will be much appreciated
Thank you !