Manufacturers of devices classified as class I, other than custom-made or investigational devices, shall declare the conformity of their products by issuing the EU declaration of conformity referred to in Article 17 after drawing up the technical documentation set out in Annex II. If the devices are placed on the market in sterile condition, are reusable surgical instruments or have a measuring function, the manufacturer shall apply the procedures set out in Annex VIII, Chapter I (Quality Management System) and Chapter III (Administrative provisions), or in Part A of Annex X. However, the involvement of the notified body shall be limited:
(a) in the case of devices placed on the market in sterile condition, to the aspects concerned with establishing, securing and maintaining sterile conditions,
(b) in the case of devices with a measuring function, to the aspects concerned with the conformity of the devices with the metrological requirements;
(c) in the case of reusable surgical instruments, to the aspects related to the reuse of the device, in particular cleaning, disinfection, sterilization, maintenance and functional testing and the related instructions for use.