Hello Everyone,

I need some guidance regarding regulations for a class II electronic and software driven monitoring device that we intend to market in the USA. The device integrates with other monitoring devices to give the user lots of monitoring options. The device has a GUI where the user selects different functions but for commercial reasons some of these functions cannot be implemented in the USA. Our software engineer can easily `grey out` these non-available options on the GUI menu but could the visibility of these options (even if they are greyed out) cause us a problem with our 510(k) ? Our engineer says its too much of a problem to remove the options completely from the menu. Can someone guide me to any FDA publication that speaks to this type of situation ?

Thanks in advance,

Frodo.
 
Elsmar Forum Sponsor

kreid

Involved In Discussions
#2
I think you could make the case that the greying out of non-available options does not impact the use of the device and so it could be acceptable to the FDA.

You would need to back this stance up with:
  • evidence that the greyed options cannot be exercised through any means (the greyed buttons or other ways) and
  • evidence that they do not cause confusion (you would need to carry out usability studies on the US variant).
 

mihzago

Trusted Information Resource
#3
I agree with kreid.
I would also add to make sure that the "greyed-out" functionality is not required for the proper operation of the cleared device in the US. In other words, is the functionality or the data provided by the greyed-out options required to meet the intended use of the device cleared in the US.
 

yodon

Staff member
Super Moderator
#4
I'm not saying this *would* be a problem but certainly something to be aware of and prepared for.

The FDA frowns on "dead" code and by graying out options, you're pretty much advertising that there's code in the system that is unreachable through normal operations. Of course, the FDA is concerned about the anomalous cases, for example, what happens if the code somehow vectors into this dead code space.

If you actually compile out all that unavailable code (i.e., it's not represented in the binaries) you'd be in pretty good shape, I think.

Either way, if you address this in your risk assessment, you might be better off.
 
Thread starter Similar threads Forum Replies Date
T First 510(k) submission - Class II software as medical device US Food and Drug Administration (FDA) 4
Q Need clarification on requirements.... Class i, gmp & 510(k) exempt Medical Device and FDA Regulations and Standards News 12
L A Taiwan company want to sell Class I medical device (510(k) exempt) on Amazon, should we register with FDA? US Food and Drug Administration (FDA) 4
lilybef Number of lots required for stability testing of a Class II device in the 510(k) Qualification and Validation (including 21 CFR Part 11) 3
S Who needs a 510(k) in this situation? Class II medical device for a US Orthodontist Other US Medical Device Regulations 4
N Class I 510(k) exempt - Applicable Standards 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
M Importing a Registered Class II Device, the Applying Our Label - Need New 510(k)? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
I FDA Exempts Product Codes from 510(k) Process (Class II Medical Devices) US Food and Drug Administration (FDA) 0
R OTC with Rx Predicate - 510(k) for a Class II medical device Other US Medical Device Regulations 6
T Level of Concern for Class II Medical Device Software 510(k) Submission 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
A Class II Medical Device 510(k) Purchase - Documentation Transfer Requirements Other US Medical Device Regulations 4
M Software for a Class II Medical Device - Required Advice for 510(k) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 10
R When is a 510(k) needed - Class I Medical Device Other Medical Device and Orthopedic Related Topics 5
P PMA called for a class III-510(k)..what next? US Food and Drug Administration (FDA) 4
T Advertising a Class II Medical Device pending 510(k) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
M 510(K) Premarket Notification Approval for Cardiovascular Class II Medical Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
B 510(k) Third Party Review of Class II Medical Device US Food and Drug Administration (FDA) 13
A Research in Mexico, anyone? US class I (non exempt) 510(K) Dental ISO 13485:2016 - Medical Device Quality Management Systems 1
A What is required to export Class I (non exempt) 510(k) approved product to Mexico ISO 13485:2016 - Medical Device Quality Management Systems 3
A Ownership of 510(k) to transfer or submit new one - Class 1 medical device ISO 13485:2016 - Medical Device Quality Management Systems 4
S Interesting Discussion FDA 510(k) required? US company Private Labeling an FDA Class II Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 34
I Class VI certification on seals, o-ring and plastic for a WFI system Qualification and Validation (including 21 CFR Part 11) 0
N Class I accessories covered by EC certificate EU Medical Device Regulations 0
L Change Notification Requirements for class 1 devices Medical Device and FDA Regulations and Standards News 1
M Issues with leakage current testing for a CLASS II device since no PE and FE is present IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
J Med Device Class for a Service Organization ISO 13485:2016 - Medical Device Quality Management Systems 2
shimonv Clinical evaluation report for class I device EU Medical Device Regulations 3
D Control Number for Class III and IV devices Canada Medical Device Regulations 0
M Class II type machine , and its compliance with 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 14
A Clinical assessment sample size - Medical device Class IIb implantable (93/42 directive) EU Medical Device Regulations 2
J Should a Class 1 medical device with an option to measure body weight be considered Class 1m? EU Medical Device Regulations 0
P ISO 14644 Class 8 Cleanroom Air Filter Requirements Other Medical Device Related Standards 4
J Instructions for use for Class I devices under MDR EU Medical Device Regulations 1
D Electrical Medical Devices class I EU Medical Device Regulations 0
K MDR Class 1 Self Certification - Competent Authority or Notified Body? EU Medical Device Regulations 6
K "World Class Product" based QM. I need advice. Quality Management System (QMS) Manuals 14
L Class IIb devices marked EC according to the 93/42/CE - Portugal CE Marking (Conformité Européene) / CB Scheme 3
J Biocompatbility testing on Class 1 device requirements Other US Medical Device Regulations 12
J Reusable surgical Instrument is under class IIa? EU Medical Device Regulations 4
H EU CE marking for Medical Device Class I EU Medical Device Regulations 2
E Any sample of a full software life cycle IEC 62304 report ( any class )? IEC 62304 - Medical Device Software Life Cycle Processes 1
JoCam Certified QMS for MDR - Class I medical device manufacturers EU Medical Device Regulations 4
N Is there a need for clinical test of Class IIa products (for MDR)? EU Medical Device Regulations 2
K CE Marking Class 1 (Non sterile) medical device CE Marking (Conformité Européene) / CB Scheme 3
G Voluntary Class I Device Filing Other US Medical Device Regulations 4
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A ISO 13485 for Class 1 Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 7
N Validating Software before getting approved as Class 2 device US Food and Drug Administration (FDA) 5
D Can a diagnostic device be class I? EU Medical Device Regulations 1
A AMSQQP416E Class 2 type 2 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0

Similar threads

Top Bottom