I'm dealing with a private label heating pad that the manufacturer tested the power supply to IEC 62368-1 instead of 60601-1. I was under the impression that the power supply was a component of the medical device and therefore should be tested to a medical device standard? These power supplies run hot. Does anyone know if they are required to be tested to 60601-1? And I have another heating pad that is ETL certified based on UL130 - but is registered as a Class II medical device. It too has not have IEC 60601-1 testing. Any comments would be appreciated!
Thanks,
Matthew
Elsmar Forum Sponsor
yodon
Leader
Super Moderator
#2
Not an expert here but maybe this will get the ball rolling.
When you submit your *system* for 60601-1 testing, the power supply will go with it. If the lab reviews the power supply test data (ostensibly from an accredited lab) and sees that it meets the 60601-1 requirements, then I believe they could accept it without further testing.
XRAY_3121
#3
Hi Yodon,
I appreciate the response! I guess I'm trying to understand if it's acceptable to test the power supply to a different standard. My opinion (if I were the manufacturer) is that it is inherently part of the device. I think the manufacturers are viewing it as a separate component. The power supply is certified by UL, but I question the standard it was tested to.
Can they test the PS to an IT standard, test the heating pad to proper ISO standard, and sell together as a single product?
yodon
Leader
Super Moderator
#4
May well depend on the intended use environment. If hospital, probably not (as the snippet mentions "business equipment"). You can probably ask your test lab.
ca_moni
Involved In Discussions
#5
This question would have been better addressed in IEC 60601-1 subforum, anyway ... as far as I know, the use of power supply certified according to an appropriate standard (not IEC 60601-1) is allowed : the medical device (here the heating pad) shall be compliant with IEC 60601-1 and the combination of the power supply and the medical device shall be regarded as a ME system and shall comply with the related prescriptions of the IEC 60601-1 standard.
XRAY_3121
#6
Thanks cs_moni, that is very helpful. I was not aware there was a 60601-1 sub forum, I appreciate the tip.
Thread starter Similar threads Forum Replies Date
B Painted surface of a protective Class I equipment. CE Marking (Conformité Européene) / CB Scheme 1
B Post-Market Surveillance Requirements for Canada - class IIa/IIb device Canada Medical Device Regulations 0
N Direct Part Marking of the "CE" symbol- Class I Device EU Medical Device Regulations 1
A PMSR or PSUR for legacy MDD-class I becoming MDR class-IIa EU Medical Device Regulations 7
T PMCF Surveys for Class II a devices: basic requisites EU Medical Device Regulations 0
T PMCF Surveys for Class II a devices : basic requisites CE Marking (Conformité Européene) / CB Scheme 0
D Conformity Assessment for a System Class IIb EU Medical Device Regulations 0
F Philippines: Old Certificate of exemption for Class B device and new LTO. How to proceed? Other Medical Device Regulations World-Wide 0
R Class I MDSW CE Mark CE Marking (Conformité Européene) / CB Scheme 6
N Class III PMA shelf life extension US Medical Device Regulations 2
A Class 2b medical device - best pathway to obtain CE mark? (2022, MDR 2017/745) CE Marking (Conformité Européene) / CB Scheme 2
A Class 2a CE Mark Regulatory Pathway CE Marking (Conformité Européene) / CB Scheme 1
normhowe Surprising shoe policies for Class 7 Cleanrooms US Medical Device Regulations 5
L Class I under MDD EU Medical Device Regulations 1
C Legacy device under Class A sterile devices EU Medical Device Regulations 0
N FDA class 1 US Food and Drug Administration (FDA) 6
P Self Certified Class 1 MDR 2017/745 Design Change CE Marking (Conformité Européene) / CB Scheme 2
M UDI IVD Medical Device Class III US Food and Drug Administration (FDA) 1
M Class Exempt Medical Device Recall Question US Food and Drug Administration (FDA) 3
J Clinical Evaluation For Instruments - Class I, Class Ir, Class Im, and Class IIa - EU MDR EU Medical Device Regulations 4
T Testing for proving food safety on USP class vi material RoHS, REACH, ELV, IMDS and Restricted Substances 0
N FDA class 2 Device QS Requirements US Food and Drug Administration (FDA) 2
D Class I Shelf-life requirements EU Medical Device Regulations 5
K Class I Medical Device Registration EU Medical Device Regulations 11
dgrainger Informational Common Specifications for certain class D IVDs EU Medical Device Regulations 0
S BASIC UDI-DI for a class IIa device marketed in a Procedure kit EU Medical Device Regulations 1
M From Class 1 to Acessory EU Medical Device Regulations 4
Q Class I Medical Device with measuring function (UK MDR 2002 / MDD) UK Medical Device Regulations 3
R Class II Active Medical Device SYSTEMS and COMPONENTS Canada Medical Device Regulations 2
S Drawing Requirements for Catalog Items (Class II Medical Device) US Food and Drug Administration (FDA) 3
M Labeling IVD Class III Medical Device and FDA Regulations and Standards News 0
L Class II Accessory Labeling Conundrum US Medical Device Regulations 19
H Mapping QMS+business processes in a medical device with class I and IIa+b Process Maps, Process Mapping and Turtle Diagrams 3
S Labeling requirements for class 1, fashion sunglasses US Food and Drug Administration (FDA) 0
B Can a software that receive data from a MD be classified as Class I?or is not a MD? EU Medical Device Regulations 5
N Shelf life guidance EU MDR, class 1 and class 2 Medical Devices EU Medical Device Regulations 2
S Can a product option change the product class? CE Marking (Conformité Européene) / CB Scheme 2
D Class III license: Add new device to the license system Canada Medical Device Regulations 1
P Servicing of Class IIb devices under MDD EU Medical Device Regulations 2
marmotte Help with Surgical Class I + 510(k) exemption US Food and Drug Administration (FDA) 0
R Is "SAFE" permitted in class 1 medical device marketing claims? US Medical Device Regulations 1
D Shelf life of non sterile class IIb product. ISO 13485:2016 - Medical Device Quality Management Systems 1
ChrisM Class 1 non-sterile device, need certified QMS or not? EU Medical Device Regulations 2
S Reintroduce packaging component/med device from white room back to class 7 clean room Manufacturing and Related Processes 1
Steve Prevette I need a template to solve world hunger, a world class Quality Program, and make lots of money Coffee Break and Water Cooler Discussions 5
Edward Reesor Class I non-sterile "Shelf Life" labeling (EU/UK) CE Marking (Conformité Européene) / CB Scheme 4
T Class III device and shelf life data requirements US Medical Device Regulations 7
D Is sampling allowed for MDR Class III products? EU Medical Device Regulations 0
M Custom made class 1 Medical devices EU Medical Device Regulations 3
A Class IIb implantable-TD sampling plan EU Medical Device Regulations 0
Similar threads
Top Bottom