Class II Implant Steam Sterilizer Validation

E

ehoqa

#1
Hi,

We make Class II implants that we will be sterilized by a new steam sterilizer. The new sterilizer was calibrated and validated by an external lab. Upon reviewing the validation, I'm wondering how others have handled such results:

I think it's standard to use 121?C for X minutes and during the validation to run 3 half-cycles and 3 full cycles with some 10^6 BIs. In the end all of the BIs were dead, so it was overall successful.

What happens when some of the readings from the sterilizer (printed out on the tape) show 120?C during exposure? Would this cause any concern (or auditors to question) that at some point the sterilization was operating at less than 121?C? How would one justify that the 120?C readings are acceptable because it was validated that way? Is it enough to state in our procedure or work instruction that 120? is part of the acceptable temperature range?

Similarly, do we need to explain why the external lab's temperature probes at various locations of the sterilizer read less than 121?C at some point during exposure, particularly at the "cold spots"?

Thank you in advance... this is my first time really dealing with a full validation of any sort.
 
Elsmar Forum Sponsor

BradM

Leader
Admin
#3
Were the Bi's placed at the hot and cold spots during the chamber? Did you set a uniformity requirement for the mapping study; such as ?1?C?

Do you have a validated load setup, where the load probes will be placed consistently in the load?
 

pkost

Trusted Information Resource
#4
I've held off responding to this as it's not my area, but I can give some general pointers:

Question: are you a manufacturer or a healthcare provider (or their service provider)?

If you are a manufacturer you are in a position to specify your sterilisation parameters, you just have to validate the process is capable of assuring sterility, you can therefore account for temperature deviations and temperature uniformity in this.

If you are a healthcare provider you should follow the instructions provided by the manufaturer otherwise you cannot rely on their assurance of sterility. You should therefore validate your equipment for their reqiurements. I fully appreciate that there is a termperature rise period and non uniformity in the chamber, it may be possible to counter this by running at a higher temperature, but alternatively you could contact the manufacturer and ask for their advise
 
Thread starter Similar threads Forum Replies Date
M Real-time-aging Validation - ISO / ASTM Standard - Class 111 Orthopedic Implant Other Medical Device and Orthopedic Related Topics 1
D Declaration of Conformity for a Class III Medical Device Implant EU Medical Device Regulations 5
R Class I MDSW CE Mark CE Marking (Conformité Européene) / CB Scheme 6
N Class III PMA shelf life extension US Medical Device Regulations 2
A Class 2b medical device - best pathway to obtain CE mark? (2022, MDR 2017/745) CE Marking (Conformité Européene) / CB Scheme 2
A Class 2a CE Mark Regulatory Pathway CE Marking (Conformité Européene) / CB Scheme 1
normhowe Surprising shoe policies for Class 7 Cleanrooms US Medical Device Regulations 5
L Class I under MDD EU Medical Device Regulations 1
C Legacy device under Class A sterile devices EU Medical Device Regulations 0
N FDA class 1 US Food and Drug Administration (FDA) 6
P Self Certified Class 1 MDR 2017/745 Design Change CE Marking (Conformité Européene) / CB Scheme 2
M UDI IVD Medical Device Class III US Food and Drug Administration (FDA) 1
M Class Exempt Medical Device Recall Question US Food and Drug Administration (FDA) 3
J Clinical Evaluation For Instruments - Class I, Class Ir, Class Im, and Class IIa - EU MDR EU Medical Device Regulations 4
XRAY_3121 Class II heating pad - does the power supply have to have IEC 60601-1 testing?? US Medical Device Regulations 5
T Testing for proving food safety on USP class vi material RoHS, REACH, ELV, IMDS and Restricted Substances 0
N FDA class 2 Device QS Requirements US Food and Drug Administration (FDA) 2
D Class I Shelf-life requirements EU Medical Device Regulations 5
K Class I Medical Device Registration EU Medical Device Regulations 11
dgrainger Informational Common Specifications for certain class D IVDs EU Medical Device Regulations 0
S BASIC UDI-DI for a class IIa device marketed in a Procedure kit EU Medical Device Regulations 1
M From Class 1 to Acessory EU Medical Device Regulations 4
Q Class I Medical Device with measuring function (UK MDR 2002 / MDD) UK Medical Device Regulations 3
R Class II Active Medical Device SYSTEMS and COMPONENTS Canada Medical Device Regulations 2
S Drawing Requirements for Catalog Items (Class II Medical Device) US Food and Drug Administration (FDA) 3
M Labeling IVD Class III Medical Device and FDA Regulations and Standards News 0
L Class II Accessory Labeling Conundrum US Medical Device Regulations 19
H Mapping QMS+business processes in a medical device with class I and IIa+b Process Maps, Process Mapping and Turtle Diagrams 3
S Labeling requirements for class 1, fashion sunglasses US Food and Drug Administration (FDA) 0
B Can a software that receive data from a MD be classified as Class I?or is not a MD? EU Medical Device Regulations 5
N Shelf life guidance EU MDR, class 1 and class 2 Medical Devices EU Medical Device Regulations 2
S Can a product option change the product class? CE Marking (Conformité Européene) / CB Scheme 2
D Class III license: Add new device to the license system Canada Medical Device Regulations 1
P Servicing of Class IIb devices under MDD EU Medical Device Regulations 2
marmotte Help with Surgical Class I + 510(k) exemption US Food and Drug Administration (FDA) 0
R Is "SAFE" permitted in class 1 medical device marketing claims? US Medical Device Regulations 1
D Shelf life of non sterile class IIb product. ISO 13485:2016 - Medical Device Quality Management Systems 1
ChrisM Class 1 non-sterile device, need certified QMS or not? EU Medical Device Regulations 2
S Reintroduce packaging component/med device from white room back to class 7 clean room Manufacturing and Related Processes 1
Steve Prevette I need a template to solve world hunger, a world class Quality Program, and make lots of money Coffee Break and Water Cooler Discussions 5
Edward Reesor Class I non-sterile "Shelf Life" labeling (EU/UK) CE Marking (Conformité Européene) / CB Scheme 4
T Class III device and shelf life data requirements US Medical Device Regulations 7
D Is sampling allowed for MDR Class III products? EU Medical Device Regulations 0
M Custom made class 1 Medical devices EU Medical Device Regulations 3
A Class IIb implantable-TD sampling plan EU Medical Device Regulations 0
D Proof QMS compliance for class I ISO 13485:2016 - Medical Device Quality Management Systems 1
M Metal parts bonding/grounding on Class II, internally powered with functional earth? IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
T FDA UDI Question - Class II Medical Device Other US Medical Device Regulations 2
B UKRP to what level should you audit Class I Technical Documentation? UK Medical Device Regulations 0
CharlieUK EU representative details on Class I device CE Marking (Conformité Européene) / CB Scheme 3

Similar threads

Top Bottom