Class II Implant Steam Sterilizer Validation

E

ehoqa

#1
Hi,

We make Class II implants that we will be sterilized by a new steam sterilizer. The new sterilizer was calibrated and validated by an external lab. Upon reviewing the validation, I'm wondering how others have handled such results:

I think it's standard to use 121?C for X minutes and during the validation to run 3 half-cycles and 3 full cycles with some 10^6 BIs. In the end all of the BIs were dead, so it was overall successful.

What happens when some of the readings from the sterilizer (printed out on the tape) show 120?C during exposure? Would this cause any concern (or auditors to question) that at some point the sterilization was operating at less than 121?C? How would one justify that the 120?C readings are acceptable because it was validated that way? Is it enough to state in our procedure or work instruction that 120? is part of the acceptable temperature range?

Similarly, do we need to explain why the external lab's temperature probes at various locations of the sterilizer read less than 121?C at some point during exposure, particularly at the "cold spots"?

Thank you in advance... this is my first time really dealing with a full validation of any sort.
 
Elsmar Forum Sponsor

BradM

Staff member
Admin
#3
Were the Bi's placed at the hot and cold spots during the chamber? Did you set a uniformity requirement for the mapping study; such as ?1?C?

Do you have a validated load setup, where the load probes will be placed consistently in the load?
 

pkost

Trusted Information Resource
#4
I've held off responding to this as it's not my area, but I can give some general pointers:

Question: are you a manufacturer or a healthcare provider (or their service provider)?

If you are a manufacturer you are in a position to specify your sterilisation parameters, you just have to validate the process is capable of assuring sterility, you can therefore account for temperature deviations and temperature uniformity in this.

If you are a healthcare provider you should follow the instructions provided by the manufaturer otherwise you cannot rely on their assurance of sterility. You should therefore validate your equipment for their reqiurements. I fully appreciate that there is a termperature rise period and non uniformity in the chamber, it may be possible to counter this by running at a higher temperature, but alternatively you could contact the manufacturer and ask for their advise
 
Thread starter Similar threads Forum Replies Date
M Real-time-aging Validation - ISO / ASTM Standard - Class 111 Orthopedic Implant Other Medical Device and Orthopedic Related Topics 1
D Declaration of Conformity for a Class III Medical Device Implant EU Medical Device Regulations 5
D Proof QMS compliance for class I ISO 13485:2016 - Medical Device Quality Management Systems 1
M Metal parts bonding/grounding on Class II, internally powered with functional earth? IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
T FDA UDI Question - Class II Medical Device Other US Medical Device Regulations 1
B UKRP to what level should you audit Class I Technical Documentation? UK Medical Device Regulations 0
CharlieUK EU representative details on Class I device CE Marking (Conformité Européene) / CB Scheme 3
A Import a MD Class IIa in the EU from China EU Medical Device Regulations 11
M MD Class I transition period MDD to MDR - changes? EU Medical Device Regulations 1
A Class 1 medical device - Thailand Other Medical Device Regulations World-Wide 0
J Shoe Covers - medical device class I EU Medical Device Regulations 3
L "Shelf-Life" Class I Non-Sterile Products Expiration Date? CE Marking (Conformité Européene) / CB Scheme 3
S Stability sampling for Class 1 medical devices ISO 13485:2016 - Medical Device Quality Management Systems 1
T Including a class 1 camera in the 510k application of class 2 therapeutic technology? US Food and Drug Administration (FDA) 18
S Class I Equipment with mobile power inverter IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
shimonv Can you submit a traditional application for a class I device? Other US Medical Device Regulations 7
D Faxback form versus amendment License for Class II Canada Medical Device Regulations 1
K Ground Bond Test for Class I Medical Electrical Equipment - calibration problems IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J The necessity and benefit of ISO13485:2016 for a manufacturer of class I devices ISO 13485:2016 - Medical Device Quality Management Systems 3
L Class II (Double Insulated) Power Supply with Output Tied to Earth: FE or PE? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S IPC-A-610 Class 3 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
S Class 1 and 2 changes in CM standards (aerospace industry) Federal Aviation Administration (FAA) Standards and Requirements 0
S Class 1 vs Class 2 changes Federal Aviation Administration (FAA) Standards and Requirements 0
Fjalar Spare parts for discontinued MDD compliant class I medical device EU Medical Device Regulations 4
Y MDR requirements for Class I accessories EU Medical Device Regulations 2
J EU Declaration of Conformity for Class I EU Medical Device Regulations 3
H Class II a vs "software safety class A" IEC 62304 - Medical Device Software Life Cycle Processes 4
K Classification Rationale for class 1 EU Medical Device Regulations 3
I Class VI certification on seals, o-ring and plastic for a WFI system Qualification and Validation (including 21 CFR Part 11) 0
N Class I accessories covered by EC certificate EU Medical Device Regulations 0
L Change Notification Requirements for class 1 devices Medical Device and FDA Regulations and Standards News 1
M Issues with leakage current testing for a CLASS II device since no PE and FE is present IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
J Med Device Class for a Service Organization ISO 13485:2016 - Medical Device Quality Management Systems 2
shimonv Clinical evaluation report for class I device EU Medical Device Regulations 18
D Control Number for Class III and IV devices Canada Medical Device Regulations 1
M Class II type machine , and its compliance with 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 14
K Technical file for MHRA - looking for assistance with self declaration Class I device CE Marking (Conformité Européene) / CB Scheme 8
A Clinical assessment sample size - Medical device Class IIb implantable (93/42 directive) EU Medical Device Regulations 2
J Should a Class 1 medical device with an option to measure body weight be considered Class 1m? EU Medical Device Regulations 0
P ISO 14644 Class 8 Cleanroom Air Filter Requirements Other Medical Device Related Standards 4
J Instructions for use for Class I devices under MDR EU Medical Device Regulations 1
D Electrical Medical Devices class I EU Medical Device Regulations 0
K MDR Class 1 Self Certification - Competent Authority or Notified Body? EU Medical Device Regulations 6
K Interesting Discussion "World Class Product" based QM. I need advice. Quality Management System (QMS) Manuals 14
L Class IIb devices marked EC according to the 93/42/CE - Portugal CE Marking (Conformité Européene) / CB Scheme 3
J Biocompatbility testing on Class 1 device requirements Other US Medical Device Regulations 12
J Reusable surgical Instrument is under class IIa? EU Medical Device Regulations 4
H EU CE marking for Medical Device Class I EU Medical Device Regulations 2
E Any sample of a full software life cycle IEC 62304 report ( any class )? IEC 62304 - Medical Device Software Life Cycle Processes 4
JoCam Certified QMS for MDR - Class I medical device manufacturers EU Medical Device Regulations 8

Similar threads

Top Bottom