Class II Medical Device Conformity Assessment Route according to Annex V only

S

silentassassin

#1
Hi folks,

We recently got our CE marking for several of our IIa Medical Devices.
The certificate provided by the NB states that the conformity assessment route was according to Annex V ONLY even the scope of the audit included examination of the technical file.
I think that no class IIa medical can be CE marked according to Annex V only; it should be either according to Annex II (section 4 excluded) Full Quality assurance or according to Annex VII Technical File Examination AND Annex V Production Quality Assurance.
Your opinion is appreciated

Thanks in advance
 

Attachments

Elsmar Forum Sponsor

Marcelo

Inactive Registered Visitor
#2
Re: Conformity Assessment Route according to Annex V only

There's no route of Annex V only. However, the drawing you attached is wrong.

The Annex VII route (declaration of conformity) is performed by the manufacturer, not the NB. The NB would only perform only Annex IV, V or VI, so you would only have a certificate for either of those.

Regarding the the analysis of the device design (Annex II - 4. Examination of the design of the product) it is only required for Class II devices, and you have to lodge a separate application (and you will have a separate certificate).

For other classes, the technical file is reviewed only to the extent of meeting the QMS requirements (for example, the requirements in item 3 of the Annexes-routes).
 
S

silentassassin

#3
Hi Marcelo,

Thanks for your response.
So as I understand from your explanation that a certificate issued according to Annex V is correct ? (see attachment)

Can provide me with some references further explaining this assessment routes
 

Attachments

Marcelo

Inactive Registered Visitor
#4
Hi Marcelo,

Thanks for your response.
So as I understand from your explanation that a certificate issued according to Annex V is correct ? (see attachment)

Can provide me with some references further explaining this assessment routes
See below:

Article 11
Conformity assessment procedures

2. In the case of devices falling within Class IIa, other than devices which are custom-made or intended for clinical investigations, the manufacturer shall, in order to affix the CE marking, follow the procedure relating to the EC declaration of conformity set out in Annex VII, coupled with either:
(a) the procedure relating to the EC verification set out in Annex IV; or
(b) the procedure relating to the EC declaration of conformity set out in Annex V (production quality assurance);
or
(c) the procedure relating to the EC declaration of conformity set out in Annex VI (product quality assurance).
Please note that for Annex V, the manufacturer must "lodge an application for assessment of his quality system with a notified body", which in the end will give you the certificate for this Annex.

Annex VII does not have this requirement because the manufacturer himself performs the activity (including drafting the declaration of conformity that covers both annexes, as detailed in Annex VII 6.1.).
 

Marcelo

Inactive Registered Visitor
#5
Another thing - you as the manufacturer should have chosen the route you want between the 4 possible routes, so this should have been discussed in the beginning of the process. Did you not chose the route?
 
S

silentassassin

#6
Thank you for your feedback, much appreciated !
Well during the audit they insisted to certify us according to Annex V.
we still don't get why ?!?!
 

Marcelo

Inactive Registered Visitor
#8
One of the reasons may be that they sold you something and then noticed that another solution would be better (usually for them, because it may require less work and keep gaining the same amount).
 
Thread starter Similar threads Forum Replies Date
H EU CE marking for Medical Device Class I EU Medical Device Regulations 0
JoCam Certified QMS for MDR - Class I medical device manufacturers EU Medical Device Regulations 4
K CE Marking Class 1 (Non sterile) medical device CE Marking (Conformité Européene) / CB Scheme 3
A ISO 13485 for Class 1 Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 7
P Anyone have an Idea on UAE Medical device registeration- Class B with FDA only Other Medical Device Regulations World-Wide 0
F Mobile app regulations - Class II medical device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
K CE Marking for Class I Medical Device? CE Marking (Conformité Européene) / CB Scheme 7
D Class IIB Medical Device Accessories - OTS Tablets Other Medical Device Regulations World-Wide 1
A Interpretation of GMP Requirements for class 1 medical device manufacturer (device GMP exempt, only General controls applicable) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
H MDD Article 12 Labeling for Class IIa Medical Device - Please Advise EU Medical Device Regulations 3
A Medical Device Vigilance decision tree for Japan for class 2 devices. Japan Medical Device Regulations 1
Edward Reesor EU Authorized Representative for a Class I Medical Device CE Marking (Conformité Européene) / CB Scheme 11
W Direct to customer export of medical device (class I: prescription lenses + frame) US Food and Drug Administration (FDA) 2
C New Class III medical device application for Health Canada Canada Medical Device Regulations 6
T First 510(k) submission - Class II software as medical device US Food and Drug Administration (FDA) 4
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S How to register class IIA medical device accessories EU Medical Device Regulations 1
U Medical device trial budget - 2 hospitals for Class IIa medical device in Sweden Other Medical Device and Orthopedic Related Topics 2
A Class medical device (MDD) - Classification help EU Medical Device Regulations 1
M Notified Body audits -Class Ia medical device product EU Medical Device Regulations 4
N Can a (class I) medical device be manufactured in a GMP certified site with no ISO13485? EU Medical Device Regulations 18
U Do we need clinical trial data for Class IIa medical device under MDR EU Medical Device Regulations 7
R Countries that accept 510k approvals - Class II nonsterile medical device Other Medical Device Regulations World-Wide 2
B Class III Medical Device Accelerated conditions Other Medical Device Related Standards 1
D FDA Approved Class II Medical Device? Other US Medical Device Regulations 11
N CE Mark for Class 1 medical device CE Marking (Conformité Européene) / CB Scheme 11
D Class II medical device - When should a complaint be closed? Customer Complaints 6
D Software as an accessory to a Class I medical device EU Medical Device Regulations 4
C ETO Sterilised Class II Medical Device - Required Temperature Storage ISO 13485:2016 - Medical Device Quality Management Systems 1
A Does Class 1 Medical Device need to be certified to MDSAP? Canada Medical Device Regulations 5
N Can we take a part from 510k cleared medical device and use it in class I device? Other US Medical Device Regulations 3
J Class 1 Medical Device - Using a UPC over the UDI? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
N Selling a class II Medical Device in Canada in Pharmacies Elsmar Xenforo Forum Software Instructions and Help 10
S Acrylic in Class II b Active Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
L A Taiwan company want to sell Class I medical device (510(k) exempt) on Amazon, should we register with FDA? US Food and Drug Administration (FDA) 4
S Conformity Assessment Route question for Class IIa medical device under MDR EU Medical Device Regulations 3
K Inventory Control - Class II (IIb for CE) medical device ISO 13485:2016 - Medical Device Quality Management Systems 6
S Software Release Note - Class A stand alone software medical device IEC 62304 - Medical Device Software Life Cycle Processes 2
J FDA Class 1 Medical Device UDI requirements Other US Medical Device Regulations 19
L Class IIa Medical Device, CE Mark & 510k - Labeling Content CE Marking (Conformité Européene) / CB Scheme 4
M How is Class I Medical Device Compliance Enforced? EU Medical Device Regulations 7
K Calibration of Dose Area Product (DAP) Meter in our Class IIb X-Ray Medical Device General Measurement Device and Calibration Topics 1
Z Class 2 medical device - Process Validation - Test sterile or non-sterile units? Qualification and Validation (including 21 CFR Part 11) 4
C Medical device manufacturing (class 2 ISO 13485:2016) - Is a Deviation allowed? Other Medical Device Related Standards 5
S Understanding UDI requirements - Class 2 medical device (hearing aids) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
N Per the FDA regulations can a medical device be class I despite that it is sterile? Other Medical Device Related Standards 4
M Include IFU in Class II Medical Device Marketing Brochure? US Food and Drug Administration (FDA) 2
J Steps when changing material for class 1 medical device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
S Clinical evaluation for a class IIa medical device - EU MDR Requirements EU Medical Device Regulations 3
A Class I medical device and Category III PPE - Long length gloves EU Medical Device Regulations 5

Similar threads

Top Bottom