Class II Medical Device Marking - Different IEC 60950 & IEC 60601 Requirements

Roland chung

Trusted Information Resource
#1
Hello forums,

In IEC 60601-1, Class II equipment with functional earth (e.g. internal screen for EMC purpose) should also mark a Class II symbol on the label.
However, the same construction is forbidden to mark with Class II symbol in IEC 60950 (see clause 2.6.2 for details).

Would it be a problem when such a 60950 compliant equipment became an item of ME system?

Regards,
Roland
 
Elsmar Forum Sponsor

ca_moni

Starting to get Involved
#3
Our notified body , many years ago, refused to accept a label with the class II mark and we had to remove it.
The device was class II, but it was powered by a commercial (medical grade) power adapter on which the connection to earth was used for the EMC screen. The output of the power adapter was a two-pole concentric connector, so no way that there were a protective earthing of the device: nevertheless, they stated (in brief): "If there are three pins in the mains plug, then it is a class I".

:2cents:
 

Roland chung

Trusted Information Resource
#4
Thank you for sharing experience.

It seems IEC 60601-1 is clear to the classification of equipment with functional earth. It is reasonable to classify such equipment to Class II, because of the equipment does not rely on basic insulation plus protective earthing for protection against electric shock.
 

Peter Selvey

Staff member
Super Moderator
#5
ca_moni mentioned:

The device was class II, but it was powered by a commercial (medical grade) power adapter on which the connection to earth was used for the EMC screen.
I would be suspicious in this case. In 25 years I have seen a couple of cases of Class II equipment with a functional earth, which is then confirmed to have 4kV/8mm spacing from mains parts as required in IEC 60601-1. But these had custom designed power supplies, not off the shelf. I never saw an off the shelf Class II "Medical grade" power supply with a functional earth. I expect it not to be commercially viable because the application is rare, and the potential for confusion is also high.

For the original question, I think it is not an issue in terms of compatibility with IEC 60601-1.

A device labelled as Class I (i.e. no double box symbol) but has all internal earth connections double insulated from mains parts (8mm/4kV) and only used for functional purpose still complies with IEC 60601-1. While unusual, there is no point of non-compliance in the standard, and no safety issues.

Also, this is the best option. The potential confusion from labelling a 3 pin device as Class II is not worth it. If the earth is used only for EMC, then it still needs to be connected to earth anyway to comply with regulations. The user does not know the internal construction so it makes no difference to them if the internal earthing is functional or protective.
 

Roland chung

Trusted Information Resource
#6
Thank you for your opinion.

We actually have an equipment with functional earth only. The equipment is enclosed by plastic. Yes, it is indeed a 3 pins device. But the earth conductor in the power supply cord is directly connected to the interior coated surface which made for EMC purposes.

If such equipment is classified as Class I, it really conflicts with the definition and clause 8.6.9.
 

Peter Selvey

Staff member
Super Moderator
#7
Clause 8.6.9 would be not applicable if the equipment is classified as Class I. So, it's not a problem.

The only I see would be the definition:
3.95 PROTECTIVE EARTH TERMINAL
terminal connected to conductive parts of CLASS I equipment for safety purposes. This terminal is intended to be connected to an external protective earthing system by a PROTECTIVE EARTH CONDUCTOR
Some people might say if there is no safety related connections then the terminal cannot be called a protective earth terminal. But even the metal parts of the terminal itself (e.g. the end of an appliance inlet that extends inside the equipment, or the terminal block for connecting the cord) should be considered an internal connection.

Also, looking at the definitions and requirements in the standard, in particular figure 4, it may be that a functional earth terminal is different to the pin in the appliance inlet or terminal for L/N/E cord connection. So the NB's original comment may be correct (3 pin device must be Class I).
 

Roland chung

Trusted Information Resource
#8
So, you meant the protective earth connection in my case is only from the pin of plug to pin of appliance inlet, the functional earth connection is from pin of appliance inlet to interior coated surface.

So, the 25A/0.1 ohm test can be performed, but only between the pin of plug and pin of appliance inlet.

If it is ture, the concept of Class II equipment with functional earth (clause 8.6.9) becomes superfluous.
 

Peter Selvey

Staff member
Super Moderator
#9
This is actually new for me, but reading the standard again, it appears that Class II equipment can have a functional earth terminal, but that terminal is not part of the mains supply connection. It appears to be a separate terminal somewhere on the outside of the equipment. Figure 4 shows the terminal well apart from the mains supply connection. Also, it is nowhere written that the "functional earth terminal" can use the supply mains terminals, it's just something we assumed.

In this case, 8.6.9 is still useful (not superfluous).

Note there is a difference here between "functional earth terminal" and "functional earth connection". It is OK to connect to the protective earth terminal for functional purpose (this is very common). But in that case, the equipment must be Class I. If the functional part is not tested for 0.1ohm / 25A, it needs to be double insulated. 8.6.9 does not apply, because it is not Class II.
 

Roland chung

Trusted Information Resource
#10
Reading the clause 8.6.9 again, it is clear the third conductor of supply cord is used as functional earth connection. The situation described in the 8.6.9 is indeed a 3 pins device.

I think the Figure 4 does not necessarily correspond to clause 8.6.9. It is actually just an example.
 
Thread starter Similar threads Forum Replies Date
A Clinical assessment sample size - Medical device Class IIb implantable (93/42 directive) EU Medical Device Regulations 2
J Should a Class 1 medical device with an option to measure body weight be considered Class 1m? EU Medical Device Regulations 0
H EU CE marking for Medical Device Class I EU Medical Device Regulations 2
JoCam Certified QMS for MDR - Class I medical device manufacturers EU Medical Device Regulations 4
K CE Marking Class 1 (Non sterile) medical device CE Marking (Conformité Européene) / CB Scheme 3
A ISO 13485 for Class 1 Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 7
P Anyone have an Idea on UAE Medical device registeration- Class B with FDA only Other Medical Device Regulations World-Wide 0
F Mobile app regulations - Class II medical device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
K CE Marking for Class I Medical Device? CE Marking (Conformité Européene) / CB Scheme 8
D Class IIB Medical Device Accessories - OTS Tablets Other Medical Device Regulations World-Wide 1
A Interpretation of GMP Requirements for class 1 medical device manufacturer (device GMP exempt, only General controls applicable) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
H MDD Article 12 Labeling for Class IIa Medical Device - Please Advise EU Medical Device Regulations 3
A Medical Device Vigilance decision tree for Japan for class 2 devices. Japan Medical Device Regulations 1
Edward Reesor EU Authorized Representative for a Class I Medical Device CE Marking (Conformité Européene) / CB Scheme 11
W Direct to customer export of medical device (class I: prescription lenses + frame) US Food and Drug Administration (FDA) 2
C New Class III medical device application for Health Canada Canada Medical Device Regulations 6
T First 510(k) submission - Class II software as medical device US Food and Drug Administration (FDA) 4
JoCam Failure to test Class I medical device to IEC 60601-1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S How to register class IIA medical device accessories EU Medical Device Regulations 1
U Medical device trial budget - 2 hospitals for Class IIa medical device in Sweden Other Medical Device and Orthopedic Related Topics 2
A Class medical device (MDD) - Classification help EU Medical Device Regulations 1
M Notified Body audits -Class Ia medical device product EU Medical Device Regulations 4
N Can a (class I) medical device be manufactured in a GMP certified site with no ISO13485? EU Medical Device Regulations 18
U Do we need clinical trial data for Class IIa medical device under MDR EU Medical Device Regulations 7
R Countries that accept 510k approvals - Class II nonsterile medical device Other Medical Device Regulations World-Wide 2
B Class III Medical Device Accelerated conditions Other Medical Device Related Standards 1
D FDA Approved Class II Medical Device? Other US Medical Device Regulations 11
N CE Mark for Class 1 medical device CE Marking (Conformité Européene) / CB Scheme 11
D Class II medical device - When should a complaint be closed? Customer Complaints 6
D Software as an accessory to a Class I medical device EU Medical Device Regulations 4
C ETO Sterilised Class II Medical Device - Required Temperature Storage ISO 13485:2016 - Medical Device Quality Management Systems 1
A Does Class 1 Medical Device need to be certified to MDSAP? Canada Medical Device Regulations 5
N Can we take a part from 510k cleared medical device and use it in class I device? Other US Medical Device Regulations 3
J Class 1 Medical Device - Using a UPC over the UDI? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
N Selling a class II Medical Device in Canada in Pharmacies Elsmar Xenforo Forum Software Instructions and Help 10
S Acrylic in Class II b Active Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
L A Taiwan company want to sell Class I medical device (510(k) exempt) on Amazon, should we register with FDA? US Food and Drug Administration (FDA) 4
S Conformity Assessment Route question for Class IIa medical device under MDR EU Medical Device Regulations 3
K Inventory Control - Class II (IIb for CE) medical device ISO 13485:2016 - Medical Device Quality Management Systems 6
S Software Release Note - Class A stand alone software medical device IEC 62304 - Medical Device Software Life Cycle Processes 2
J FDA Class 1 Medical Device UDI requirements Other US Medical Device Regulations 19
L Class IIa Medical Device, CE Mark & 510k - Labeling Content CE Marking (Conformité Européene) / CB Scheme 4
M How is Class I Medical Device Compliance Enforced? EU Medical Device Regulations 7
K Calibration of Dose Area Product (DAP) Meter in our Class IIb X-Ray Medical Device General Measurement Device and Calibration Topics 1
Z Class 2 medical device - Process Validation - Test sterile or non-sterile units? Qualification and Validation (including 21 CFR Part 11) 4
C Medical device manufacturing (class 2 ISO 13485:2016) - Is a Deviation allowed? Other Medical Device Related Standards 5
S Understanding UDI requirements - Class 2 medical device (hearing aids) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
N Per the FDA regulations can a medical device be class I despite that it is sterile? Other Medical Device Related Standards 4
M Include IFU in Class II Medical Device Marketing Brochure? US Food and Drug Administration (FDA) 2
J Steps when changing material for class 1 medical device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1

Similar threads

Top Bottom