Class II PSU (200W) BF Power Supply Configuration Isolation Requirements
S
Sampie
#1
Hi all
I am struggling with the isolation requirements for a product that we are developing. It is a CLASS II BF system (from spec). My first problem comes in where I have to find a CLASS II PSU (200W). Most seem to be CLASS I at that power rating (they have earth connections). Can I use a CLASS I power supply in a CLASS II product. The input to output isolation is 4000V 2xMOPP but the input and output to ground (I assume earth) is only 1xMOPP 1500V. Is the input and output isolation enough or do I need more isolation. Medical DC/DC converters above 20W+ is also rare if I wanted to use one for an additional isolation layer? Can someone point me in the right direction? :cfingers:
Kind regards
Sampie
Elsmar Forum Sponsor
Peter Selvey
Staff member
Super Moderator
#2
If the power supply is sold as Class I, then there is nothing much you can do, you need to connect the earth for compliance.
Under the 3rd edition (2005), it is possible to use a 950 power supply, and then provide additional isolation to the patient elsewhere in the device. This additional isolation can be in a special isolation PCB (e.g. with small DC/DC converter, optocouplers, isolating out just function which needs to contact the patient), or could be in the probe or just the nature of contact.
But if you want to be compatible with the 2nd edition (1988), you need to use a 601 power supply.
S
Sampie
#3
Thanks for your feedback. I think I will try to get the specification changed to CLASS I BF. Is this possible for a system where sub dural electrical stimulation is applied? We will still have 2 MOPPs of patient protection for the mains and earth. I have seen references in the 60601-2-10:2001 (Part 2.10: Particular requirements for the safety of nerve and muscle stimulators)
that they mention CLASS I systems but I am not sure if that is acceptable.
regards
Sampie
Peter Selvey
Staff member
Super Moderator
#4
Absolutely. Either Class I or II is OK for the standard, if intended for professional use (home use is limited to Class II under IEC 60601-1-10).
200W is not a small power supply. Class II power supplies tend to be smaller, 100W or less. I guess EMC gets more difficult as the power goes up, and earthing helps to keep the noise down. That might be why it's difficult to find a Class II power supply.
Thread starter Similar threads Forum Replies Date
shimonv Clinical evaluation report for class I device EU Medical Device Regulations 3
D Control Number for Class III and IV devices Canada Medical Device Regulations 0
M Class II type machine , and its compliance with 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 14
A Clinical assessment sample size - Medical device Class IIb implantable (93/42 directive) EU Medical Device Regulations 2
J Should a Class 1 medical device with an option to measure body weight be considered Class 1m? EU Medical Device Regulations 0
P ISO 14644 Class 8 Cleanroom Air Filter Requirements Other Medical Device Related Standards 4
J Instructions for use for Class I devices under MDR EU Medical Device Regulations 1
D Electrical Medical Devices class I EU Medical Device Regulations 0
K MDR Class 1 Self Certification - Competent Authority or Notified Body? EU Medical Device Regulations 6
K "World Class Product" based QM. I need advice. Quality Management System (QMS) Manuals 14
L Class IIb devices marked EC according to the 93/42/CE - Portugal CE Marking (Conformité Européene) / CB Scheme 3
J Biocompatbility testing on Class 1 device requirements Other US Medical Device Regulations 12
J Reusable surgical Instrument is under class IIa? EU Medical Device Regulations 4
H EU CE marking for Medical Device Class I EU Medical Device Regulations 2
E Any sample of a full software life cycle IEC 62304 report ( any class )? IEC 62304 - Medical Device Software Life Cycle Processes 1
JoCam Certified QMS for MDR - Class I medical device manufacturers EU Medical Device Regulations 4
N Is there a need for clinical test of Class IIa products (for MDR)? EU Medical Device Regulations 2
K CE Marking Class 1 (Non sterile) medical device CE Marking (Conformité Européene) / CB Scheme 3
G Voluntary Class I Device Filing Other US Medical Device Regulations 4
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A ISO 13485 for Class 1 Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 7
N Validating Software before getting approved as Class 2 device US Food and Drug Administration (FDA) 5
D Can a diagnostic device be class I? EU Medical Device Regulations 1
A AMSQQP416E Class 2 type 2 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
S IVD risk class II devices for Brazil and MDSAP Other Medical Device Regulations World-Wide 0
P Anyone have an Idea on UAE Medical device registeration- Class B with FDA only Other Medical Device Regulations World-Wide 0
F Mobile app regulations - Class II medical device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
S Medical watch Class II (AP Type CF) with USB connection IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
L Radiology software Class I exemption Medical Device and FDA Regulations and Standards News 3
K CE Marking for Class I Medical Device? CE Marking (Conformité Européene) / CB Scheme 8
A Class I (exempt) testing requirements Other Medical Device Related Standards 0
Z Class I marketing regulation and supply chain Other Medical Device Regulations World-Wide 0
S We are looking for EU authorised Representative for both Class I and Class IIb devices EU Medical Device Regulations 7
S How is the service life for class I reusable medical devices described in MDR? EU Medical Device Regulations 7
S How is the service life described for class I reusable medical devices in MDR? Elsmar Xenforo Forum Software Instructions and Help 1
R Class 1 Convenience kits for home teeth impression, NOT smile direct or other for alignment purposes Canada Medical Device Regulations 0
J Confusing on Harmonic Emissions/Voltage Fluctuations/Flicker test for CISPR 11 Class A Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J Intervertebral body fusion cage, MDR Class II or Class III? EU Medical Device Regulations 4
T Do I need a qualified compiler for class B software? IEC 62304 - Medical Device Software Life Cycle Processes 3
R Nanomaterials used as coating on class I MD - rule 19 EU Medical Device Regulations 4
D Amendment Class II - Please confirm if the submission is by email (Canada) - Help Canada Medical Device Regulations 0
D Class IIB Medical Device Accessories - OTS Tablets Other Medical Device Regulations World-Wide 1
K Doubt on insulation class of a device IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
D Where to find Guidance Class II Amendment Canada Medical Device Regulations 1
A First CE-mark class III implantable device under MDD EU Medical Device Regulations 8
P ISO Class 8 particle count (annual certification vs monitoring) ISO 13485:2016 - Medical Device Quality Management Systems 4
I IVD - 98/79EC - class IIa - Annex to the EC from the NB states the CM as one of the facilities? CE Marking (Conformité Européene) / CB Scheme 2
Z MDR Class Ir deadline (Reusable surgical instruments) EU Medical Device Regulations 27
D Adding a Class I to a sample box Canada Medical Device Regulations 5
T Class A PRTD beyond 300°C General Measurement Device and Calibration Topics 0
Similar threads
Top Bottom