SBS - The Best Value in QMS software

Class II Sterile Product Shelf Life Labeling - Should we do a recall?

D

deuce64

#1
Hoping for opinions on this scenario:
this is for US jurisdiction, FDA. No sales Outside USA.
A class II, sterile product that we acquired was labeled for 2 years shelf life. the previous company (510k) owner did a shelf life study to extend the product to 5 years shelf life. Upon acquisition, we began relabeling product that was already built with 5 years shelf life - the correct expiration date is this coming july.
During a recent random review of finished goods, we found that 1 batch was mis labeled as if it had a 6 year shelf life (i.e. instead of label reading it would expire this coming july, it is now labeled to expire the following july.)
Normally this would be a case-closed recall, but here is our thought. Firstly, there is an obvious internal CAPA for the break down in procedure/process that allowed this to happen. and then there is a Health Hazard Analysis. We've written the analysis and it could say that, 'of course, in July of this year, there is a risk to the population (potentially breached barrier or non-usable material) however, before July, the product is not a risk. We've since begun accelerated age testing and in May of this year we will have an answer for 6 year shelf life. If the testing passes, the mislabeled product is good, if it fails, we must recall.' and of course, this would happen prior to the july 2015 date. Curious to know people's reaction to this strategy. Would FDA take a hard stance that the product should have been recalled right away....or because we took action to investigate, prior to any risks to humans, our actions were acceptable.
thanks for thoughts.
 
Elsmar Forum Sponsor
M

MIREGMGR

#2
Re: Class II sterile product Shelf Life - Should we do a recall?

I understand your argument for waiting...but I don't think FDA would buy it. You (inadvertently) distributed product that was mislabeled at the time of distribution according to the information then available to you. Ergo, a field action is required for that batch.

That normally would be a recall for various reasons, but in theory it could be a correction i.e. relabeling in the field, if the batch is small enough and you know where it all is, and the economics justify.
 

TWA - not the airline

Trusted Information Resource
#3
Another thought: How fast does your product typically turn over in the field? Maybe this lot has long since been used up. If you could confirm this beforehand with the customers that did receive product from this specific lot, then IMHO there is no recall/field action. If you send the "Dear doctor" letter first and then find out that there is nothing left to recall, then the damage is done...
As an afterthought: How sure are you that this is a singular event? Did you check the copies of the labeling in your DHRs?
 
D

deuce64

#4
Thanks miregmar and TWA.
I appreciate your opinions. There is certainly work to be done on the CAPA side to verify that this is indeed a one-off situation and further, to improve our systems and processes so this does not happen again.

Just to pressure test your opinions and see if they change - The issue has been detected after selling product to about 10 customers over the course of about 5 months. (a slow moving product). The issue was detected as we were selling the last units from this batch, and therefore, this issue was not instantly detected....it was detected quite some time after release of the product....so the misdeed of distributing mislabeled product has occurred no matter how you slice it. So, the product was violative yesterday, and the day before, and the day before etc. We've put our arms around the issue with the preliminary action of quarantining current product, have aged product coming out in May, and have high confidence that the testing will pass with flying colors (very simple product configuration). Where the scope of customers/product in field is so small, a FSCA would take just a short time to execute (1 month or less) and therefore risk is even further mitigated in the slim chance we would have to perform the FSCA. i haven't found any regulation that specifies when the timing of a FSCA must be executed...only the circumstances which trigger an FSCA. In the circumstance where a product is mislabeled with the wrong diameter, it wold be text book, but here, this label has the potential to be corrected without field action.
Still believe a recall right now is required? thanks
 
Thread starter Similar threads Forum Replies Date
L "Shelf-Life" Class I Non-Sterile Products Expiration Date? CE Marking (Conformité Européene) / CB Scheme 3
K CE Marking Class 1 (Non sterile) medical device CE Marking (Conformité Européene) / CB Scheme 3
K Sterile barrier system symbols - Label redlines for a Class III device (Kit Box) EU Medical Device Regulations 0
Z Class 2 medical device - Process Validation - Test sterile or non-sterile units? Qualification and Validation (including 21 CFR Part 11) 4
N Per the FDA regulations can a medical device be class I despite that it is sterile? Other Medical Device Related Standards 4
A Medical Device Class 2A, non sterile bodyworn packaging requirements CE Marking (Conformité Européene) / CB Scheme 7
N Registering a Class 1 Sterile Medical Device - Currently selling a Non-Sterile 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 10
J Non-Sterile Class 2 Medical Device Shelf Life Other Medical Device Related Standards 3
J Change to Outer Packaging Material/Supplier - Sterile, Class IV Device Other Medical Device Related Standards 4
R Shipping Stability for Class III IVD (Non Sterile) Medical Device Other Medical Device Related Standards 7
J Biocompatibility for Class I Non-Sterile Medical Device EU Medical Device Regulations 2
W Chinese Distributor Making Counterfeit Products - Class I sterile products China Medical Device Regulations 2
R How to go about package validation on a class II, non-sterile device US Food and Drug Administration (FDA) 4
T Transportation of Class III Sterile Implants - Sterile Filler (Hyaluronic Acid) EU Medical Device Regulations 1
Q Expiry Dates on a Non-Sterile Class 1 Medical Device (5 Year Shelf Life). Other Medical Device and Orthopedic Related Topics 11
A Need to register with CA for Class I (Non-Sterile and Non-Measuring) Medical Device EU Medical Device Regulations 4
M Packaging Requirements for Non-Sterile Class II Medical Devices EU Medical Device Regulations 11
M Class I non-sterile, non-measuring: Requirements fulfiled? EU Medical Device Regulations 4
J Requirements for Own Branding Class I Non-Sterile Device? ISO 13485:2016 - Medical Device Quality Management Systems 1
M Process Validation/Equipment Qualification - Class 1A (Sterile) Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 12
M Class I Sterile Application-TGA - What application form do we need to fill in? Other Medical Device and Orthopedic Related Topics 2
M ISO 14971:2007 Risk Management - Class I A Sterile Medical Device ISO 14971 - Medical Device Risk Management 4
K Requirements for Exporting from US to EU - Class I MD non-sterile, non-measuring EU Medical Device Regulations 7
J Addition of a Class 1 (Sterile) Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 3
D Selling Class IIa medical device (sterile syringes) in USA ISO 13485:2016 - Medical Device Quality Management Systems 2
A Non-sterile liquid bandage as Class 1 - Anyone concur or disagree? EU Medical Device Regulations 17
A Documentation required for launch of a non-sterile exempt class 1 medical device ISO 13485:2016 - Medical Device Quality Management Systems 7
S Stability sampling for Class 1 medical devices ISO 13485:2016 - Medical Device Quality Management Systems 1
T Including a class 1 camera in the 510k application of class 2 therapeutic technology? US Food and Drug Administration (FDA) 18
S Class I Equipment with mobile power inverter IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
shimonv Can you submit a traditional application for a class I device? Other US Medical Device Regulations 7
D Faxback form versus amendment License for Class II Canada Medical Device Regulations 1
K Ground Bond Test for Class I Medical Electrical Equipment - calibration problems IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J The necessity and benefit of ISO13485:2016 for a manufacturer of class I devices ISO 13485:2016 - Medical Device Quality Management Systems 3
L Class II (Double Insulated) Power Supply with Output Tied to Earth: FE or PE? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S IPC-A-610 Class 3 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
S Class 1 and 2 changes in CM standards (aerospace industry) Federal Aviation Administration (FAA) Standards and Requirements 0
S Class 1 vs Class 2 changes Federal Aviation Administration (FAA) Standards and Requirements 0
Fjalar Spare parts for discontinued MDD compliant class I medical device EU Medical Device Regulations 4
Y MDR requirements for Class I accessories EU Medical Device Regulations 2
J EU Declaration of Conformity for Class I EU Medical Device Regulations 3
H Class II a vs "software safety class A" IEC 62304 - Medical Device Software Life Cycle Processes 3
K Classification Rationale for class 1 EU Medical Device Regulations 3
I Class VI certification on seals, o-ring and plastic for a WFI system Qualification and Validation (including 21 CFR Part 11) 0
N Class I accessories covered by EC certificate EU Medical Device Regulations 0
L Change Notification Requirements for class 1 devices Medical Device and FDA Regulations and Standards News 1
M Issues with leakage current testing for a CLASS II device since no PE and FE is present IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
J Med Device Class for a Service Organization ISO 13485:2016 - Medical Device Quality Management Systems 2
shimonv Clinical evaluation report for class I device EU Medical Device Regulations 3
D Control Number for Class III and IV devices Canada Medical Device Regulations 1

Similar threads

Top Bottom