Our company wants to apply to US FDA for an OTC classification of our Class II medical device. I have the following questions:

1. Should we submit a 510(k) application? - our device already has a 510(k) certification, but as a Class II device.
2. What if there was no other device of the same type in the OTC classification? Can we still apply and request to be considered for an OTC classification, even if there is no predicate device with OTC classification ?
3. Is there any guidance documents available from FDA on this subject?

Thanks,
Sarang
 
Elsmar Forum Sponsor

mihzago

Trusted Information Resource
#2
Yes, you will have to submit a 510(k). This is not a change in classification (though FDA might create a new product code); rather, it's an extension of your indications for use.

There are a number of guidance documents for a submission of a 510(k), but I don't think there is one specific to going from Rx to OTC.
You should be able to find some info in one of these guidance docs.

https://www.fda.gov/downloads/Medic...onandGuidance/GuidanceDocuments/UCM284443.pdf
https://www.fda.gov/downloads/Medic...onandGuidance/GuidanceDocuments/ucm514771.pdf

I'm also providing a link to a usability guidance, because one of the main things you will have to provide is usability data that the lay users can properly use the device.
https://www.fda.gov/downloads/MedicalDevices/.../UCM259760.pdf
 
Thread starter Similar threads Forum Replies Date
R OTC with Rx Predicate - 510(k) for a Class II medical device Other US Medical Device Regulations 6
E Guidance/Tips to go Rx to OTC for Class II Medical Device Other US Medical Device Regulations 6
A What is Classification of Class A Certificate? OTC health care and consumer products Various Other Specifications, Standards, and related Requirements 1
D Is sampling allowed for MDR Class III products? EU Medical Device Regulations 0
M Custom made class 1 Medical devices EU Medical Device Regulations 3
A Class IIb implantable-TD sampling plan EU Medical Device Regulations 0
D Proof QMS compliance for class I ISO 13485:2016 - Medical Device Quality Management Systems 1
M Metal parts bonding/grounding on Class II, internally powered with functional earth? IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
T FDA UDI Question - Class II Medical Device Other US Medical Device Regulations 2
B UKRP to what level should you audit Class I Technical Documentation? UK Medical Device Regulations 0
CharlieUK EU representative details on Class I device CE Marking (Conformité Européene) / CB Scheme 3
A Import a MD Class IIa in the EU from China EU Medical Device Regulations 11
M MD Class I transition period MDD to MDR - changes? EU Medical Device Regulations 1
A Class 1 medical device - Thailand Other Medical Device Regulations World-Wide 0
J Shoe Covers - medical device class I EU Medical Device Regulations 3
L "Shelf-Life" Class I Non-Sterile Products Expiration Date? CE Marking (Conformité Européene) / CB Scheme 3
S Stability sampling for Class 1 medical devices ISO 13485:2016 - Medical Device Quality Management Systems 1
T Including a class 1 camera in the 510k application of class 2 therapeutic technology? US Food and Drug Administration (FDA) 18
S Class I Equipment with mobile power inverter IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
shimonv Can you submit a traditional application for a class I device? Other US Medical Device Regulations 7
D Faxback form versus amendment License for Class II Canada Medical Device Regulations 1
K Ground Bond Test for Class I Medical Electrical Equipment - calibration problems IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J The necessity and benefit of ISO13485:2016 for a manufacturer of class I devices ISO 13485:2016 - Medical Device Quality Management Systems 3
L Class II (Double Insulated) Power Supply with Output Tied to Earth: FE or PE? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S IPC-A-610 Class 3 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
S Class 1 and 2 changes in CM standards (aerospace industry) Federal Aviation Administration (FAA) Standards and Requirements 0
S Class 1 vs Class 2 changes Federal Aviation Administration (FAA) Standards and Requirements 0
Fjalar Spare parts for discontinued MDD compliant class I medical device EU Medical Device Regulations 4
Y MDR requirements for Class I accessories EU Medical Device Regulations 2
J EU Declaration of Conformity for Class I EU Medical Device Regulations 3
H Class II a vs "software safety class A" IEC 62304 - Medical Device Software Life Cycle Processes 4
K Classification Rationale for class 1 EU Medical Device Regulations 3
I Class VI certification on seals, o-ring and plastic for a WFI system Qualification and Validation (including 21 CFR Part 11) 0
N Class I accessories covered by EC certificate EU Medical Device Regulations 0
L Change Notification Requirements for class 1 devices Medical Device and FDA Regulations and Standards News 1
M Issues with leakage current testing for a CLASS II device since no PE and FE is present IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
J Med Device Class for a Service Organization ISO 13485:2016 - Medical Device Quality Management Systems 2
shimonv Clinical evaluation report for class I device EU Medical Device Regulations 18
D Control Number for Class III and IV devices Canada Medical Device Regulations 1
M Class II type machine , and its compliance with 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 14
K Technical file for MHRA - looking for assistance with self declaration Class I device CE Marking (Conformité Européene) / CB Scheme 8
A Clinical assessment sample size - Medical device Class IIb implantable (93/42 directive) EU Medical Device Regulations 2
J Should a Class 1 medical device with an option to measure body weight be considered Class 1m? EU Medical Device Regulations 0
P ISO 14644 Class 8 Cleanroom Air Filter Requirements Other Medical Device Related Standards 4
J Instructions for use for Class I devices under MDR EU Medical Device Regulations 1
D Electrical Medical Devices class I EU Medical Device Regulations 0
K MDR Class 1 Self Certification - Competent Authority or Notified Body? EU Medical Device Regulations 6
K Interesting Discussion "World Class Product" based QM. I need advice. Quality Management System (QMS) Manuals 14
L Class IIb devices marked EC according to the 93/42/CE - Portugal CE Marking (Conformité Européene) / CB Scheme 3
J Biocompatbility testing on Class 1 device requirements Other US Medical Device Regulations 12

Similar threads

Top Bottom