SBS - The best value in QMS software

Class II to OTC conversion of a medical device

S

ssarangapani

#1
Our company wants to apply to US FDA for an OTC classification of our Class II medical device. I have the following questions:

1. Should we submit a 510(k) application? - our device already has a 510(k) certification, but as a Class II device.
2. What if there was no other device of the same type in the OTC classification? Can we still apply and request to be considered for an OTC classification, even if there is no predicate device with OTC classification ?
3. Is there any guidance documents available from FDA on this subject?

Thanks,
Sarang
 
Elsmar Forum Sponsor

mihzago

Trusted Information Resource
#2
Yes, you will have to submit a 510(k). This is not a change in classification (though FDA might create a new product code); rather, it's an extension of your indications for use.

There are a number of guidance documents for a submission of a 510(k), but I don't think there is one specific to going from Rx to OTC.
You should be able to find some info in one of these guidance docs.

https://www.fda.gov/downloads/Medic...onandGuidance/GuidanceDocuments/UCM284443.pdf
https://www.fda.gov/downloads/Medic...onandGuidance/GuidanceDocuments/ucm514771.pdf

I'm also providing a link to a usability guidance, because one of the main things you will have to provide is usability data that the lay users can properly use the device.
https://www.fda.gov/downloads/MedicalDevices/.../UCM259760.pdf
 
Thread starter Similar threads Forum Replies Date
R OTC with Rx Predicate - 510(k) for a Class II medical device Other US Medical Device Regulations 6
E Guidance/Tips to go Rx to OTC for Class II Medical Device Other US Medical Device Regulations 6
A What is Classification of Class A Certificate? OTC health care and consumer products Various Other Specifications, Standards, and related Requirements 1
J Biocompatbility testing on Class 1 device requirements Other US Medical Device Regulations 1
J Reusable surgical Instrument is under class IIa? EU Medical Device Regulations 3
H EU CE marking for Medical Device Class I EU Medical Device Regulations 0
E Any sample of a full software life cycle IEC 62304 report ( any class )? IEC 62304 - Medical Device Software Life Cycle Processes 1
JoCam Certified QMS for MDR - Class I medical device manufacturers EU Medical Device Regulations 4
N Is there a need for clinical test of Class IIa products (for MDR)? EU Medical Device Regulations 2
K CE Marking Class 1 (Non sterile) medical device CE Marking (Conformité Européene) / CB Scheme 2
G Voluntary Class I Device Filing Other US Medical Device Regulations 4
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A ISO 13485 for Class 1 Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 7
N Validating Software before getting approved as Class 2 device US Food and Drug Administration (FDA) 5
D Can a diagnostic device be class I? EU Medical Device Regulations 1
A AMSQQP416E Class 2 type 2 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
S IVD risk class II devices for Brazil and MDSAP Other Medical Device Regulations World-Wide 0
P Anyone have an Idea on UAE Medical device registeration- Class B with FDA only Other Medical Device Regulations World-Wide 0
F Mobile app regulations - Class II medical device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
S Medical watch Class II (AP Type CF) with USB connection IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
L Radiology software Class I exemption Medical Device and FDA Regulations and Standards News 3
K CE Marking for Class I Medical Device? CE Marking (Conformité Européene) / CB Scheme 7
A Class I (exempt) testing requirements Other Medical Device Related Standards 0
Z Class I marketing regulation and supply chain Other Medical Device Regulations World-Wide 0
S We are looking for EU authorised Representative for both Class I and Class IIb devices EU Medical Device Regulations 7
S How is the service life for class I reusable medical devices described in MDR? EU Medical Device Regulations 7
S How is the service life described for class I reusable medical devices in MDR? Elsmar Xenforo Forum Software Instructions and Help 1
R Class 1 Convenience kits for home teeth impression, NOT smile direct or other for alignment purposes Canada Medical Device Regulations 0
J Confusing on Harmonic Emissions/Voltage Fluctuations/Flicker test for CISPR 11 Class A Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J Intervertebral body fusion cage, MDR Class II or Class III? EU Medical Device Regulations 4
T Do I need a qualified compiler for class B software? IEC 62304 - Medical Device Software Life Cycle Processes 3
R Nanomaterials used as coating on class I MD - rule 19 EU Medical Device Regulations 4
D Amendment Class II - Please confirm if the submission is by email (Canada) - Help Canada Medical Device Regulations 0
D Class IIB Medical Device Accessories - OTS Tablets Other Medical Device Regulations World-Wide 1
K Doubt on insulation class of a device IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
D Where to find Guidance Class II Amendment Canada Medical Device Regulations 1
A First CE-mark class III implantable device under MDD EU Medical Device Regulations 8
P ISO Class 8 particle count (annual certification vs monitoring) ISO 13485:2016 - Medical Device Quality Management Systems 4
I IVD - 98/79EC - class IIa - Annex to the EC from the NB states the CM as one of the facilities? CE Marking (Conformité Européene) / CB Scheme 2
Z MDR Class Ir deadline (Reusable surgical instruments) EU Medical Device Regulations 27
D Adding a Class I to a sample box Canada Medical Device Regulations 5
T Class A PRTD beyond 300°C General Measurement Device and Calibration Topics 0
F How to register class I device to all European Competent Autorities(CA) when we are European manufacturer EU Medical Device Regulations 4
K Sterile barrier system symbols - Label redlines for a Class III device (Kit Box) EU Medical Device Regulations 0
D Class 3 Device - Change of supplier of material Canada Medical Device Regulations 6
A FDA Class Classification for a cabinet 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
W Protective earth for Class I ME equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K Class III Kit Box - Are translations mandatory under EU MDR? EU Medical Device Regulations 5
Edward Reesor FDA DM and /or Class I Life Saving US Food and Drug Administration (FDA) 0
A Interpretation of GMP Requirements for class 1 medical device manufacturer (device GMP exempt, only General controls applicable) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4

Similar threads

Top Bottom