Class IIB Device - IEC 62304 Software Classification
BargeSim
Starting to get Involved
#1
Hi all,
I have a class IIb medical device (a CBCT scanner for extremities) for only diagnostic purpose; is there a required software classification.
The software manages images acquisition, 3D Reconstruction and visualization on the viewer; no further decision is taken automatically.
I suppose I can assume class B, due to the risk mitigation measures (other than the software).
But my question would be:
is there any guideline defining the difference between serious injury and not serious injury? I only found some docs produced by specific Department of Health, but not an international standard definition.
Thanks
Elsmar Forum Sponsor
yodon
Staff member
Super Moderator
#2
In ISO 14973, they provide an example of severity levels for harm:
  • Catastrophic - death
  • Critical - permanent impairment or life-threatening injury
  • Serious - injury or impairment requiring professional medical intervention
  • Minor - temporary injury or impairment not requiring professional medical intervention
  • Negligible - inconvenience or temporary discomfort
To me, anything, well, serious or above would be a good place to draw the line. Of course, this is just an example and you could have different levels but this seems to be a pretty common way to organize.
Ronen E
Problem Solver
Staff member
Moderator
#3
Hi all,
Thanks
Can you please quote the classification rule that resulted in the class IIb classification decision? It might help answer your question.
Kuldeep Singh
Involved In Discussions
#4
Hi,
I have a question for classification of software, If we control all the risks through hardware control which arises from software ,then is it possible that , we can assume our software as class A rather than Class B?
dgrainger
Trusted Information Resource
#6
To me CBCT = cone beam computed tomography, and will be class 2b by Rule 10. "Active devices intended to emit ionizing radiation and intended for diagnostic and therapeutic interventional radiology". Rule 9 applies to therapeutic devices.
As it is emitting ionising radiation there will always be a risk of serious injury. Class C?
BargeSim
#7
Thank you all for your replies.
Further information: Classification based on rule 9: delivery of (hazardous) energy
Yes sorry, surely the Rule is 10.
I honestly had been thinking about class C, but my question is if it can be reduced to Class B, reducing risk, using measures other than the SW.
Kuldeep Singh
#8
You can reduce the software classification by reducing the risk by hardware risk control measures. Here below the guidance given in IEC 62304 for sw classification.
#9
Hi All, take care for a proper architectual design including a solid planning of the hardware measure which shall mitigate your risk.
Be aware, that you have to follow the IEC/EN 60601-1 and your hardware component will be automatically a critical component. Think about, how you can detect single fault failure events and how additional fault conditions can be detected. All safety relevant and critical component shall have a proper test report which includes evidences of related production process controls of incoming inspection, in process controls and end controls.
If you want to distribute the device internationally, a certification mark (like CSA [e.g., cCSAus], UL, NEMKO, DEKRA etc.) would ease the market access and also speed up the registration in e.g. Brazil. Even US customers may benefit: If there is a certification mark, hospitals can save insurance fees as they pay less. For negotiations with the hospital purchase that can be crucial.
Tidge
#10
Hi,
I would never recommend trying to change the classification of software (within a medical device) based on an analysis of the effectiveness of hardware risk controls. In my experience, the only clean way to have a 'lower' software safety classification (than otherwise would be naively assumed, based on the classification of the medical device) is to have a (product) system architecture that makes it clear that the software has not been allocated any functions that could contribute to the most serious harms.
In simpler terms: If you want a 'lower' classification, the design ought to be such that the software is neither a contributing factor nor a control of the risks associated with (non) serious injury.
Thread starter Similar threads Forum Replies Date
A Clinical assessment sample size - Medical device Class IIb implantable (93/42 directive) EU Medical Device Regulations 2
D Class IIB Medical Device Accessories - OTS Tablets Other Medical Device Regulations World-Wide 1
K Inventory Control - Class II (IIb for CE) medical device ISO 13485:2016 - Medical Device Quality Management Systems 6
K Calibration of Dose Area Product (DAP) Meter in our Class IIb X-Ray Medical Device General Measurement Device and Calibration Topics 1
H Classification of a Class IIb Medical Device with 3 Components CE Marking (Conformité Européene) / CB Scheme 1
A Technical file requirement for Class IIb medical device CE Marking (Conformité Européene) / CB Scheme 3
MDD_QNA Class I Medical Device into a higher class (IIa or IIb) CE Marking (Conformité Européene) / CB Scheme 7
L What are Accredited laboratories? (Medical Device Class IIb Related) EU Medical Device Regulations 1
E When to submit a Technical File for a Class IIb Medical Device EU Medical Device Regulations 3
B Sri Lanka Class IIb Medical Device Registration Other Medical Device Regulations World-Wide 4
L Labeling Requirements in Part A Class IIb Medical Device Technical File CE Marking (Conformité Européene) / CB Scheme 2
M Labeling a Small Class IIb Medical Device EU Medical Device Regulations 8
G Cost for Generating one Class IIb Medical Device GMDN Code Other Medical Device and Orthopedic Related Topics 2
B Class IIb Medical Device using Annex II route - Need to Update ISO 13485 Scope? ISO 13485:2016 - Medical Device Quality Management Systems 2
S Medical Device Technical Files (Class IIa & IIb) ? Design Development Documentation EU Medical Device Regulations 7
S Is using Annex V of 93/42/EEC for CE Certification incorrect for a Class IIb Device? EU Medical Device Regulations 7
I Medical Device Directive, Conversion from class IIb to class III EU Medical Device Regulations 4
V CE Marking for Power Supply for Medical Device - Class IIb device per MDD, Rule 11 EU Medical Device Regulations 3
L Class IIb devices marked EC according to the 93/42/CE - Portugal CE Marking (Conformité Européene) / CB Scheme 3
S We are looking for EU authorised Representative for both Class I and Class IIb devices EU Medical Device Regulations 7
J Custom-made implants (Class IIb) and pre-commercialization clinical trials in Spain EU Medical Device Regulations 4
K Post Market Clinical Follow up and CE Marking for Class IIb medical X-Ray products CE Marking (Conformité Européene) / CB Scheme 1
A Level of details required for Class IIb Product Verification and Validation CE Marking (Conformité Européene) / CB Scheme 1
mikinnear Evidence - Class IIb Disinfectant Sanitiser EU Medical Device Regulations 2
C Sampling of Class IIb Implantables under EU MDR - Assessments to the EU MDR EU Medical Device Regulations 4
R Risk Analysis of Class IIb Disinfectant ISO 14971 - Medical Device Risk Management 6
J Canada Bulk Packaging and IFU requirements for Class IIa and IIb Medical Devices Canada Medical Device Regulations 2
C Bulgarian regulatory requirements for export of class IIb medical devices EU Medical Device Regulations 4
A Endoscopy Software - Class I or Class IIa or Class IIb CE Marking (Conformité Européene) / CB Scheme 4
M MDD 93/42/EEC, Annex IX - Class IIa vs IIb: "Potentially Hazardous" definition EU Medical Device Regulations 2
P Valid NB Technical Review Nonconformance? New Class IIb Product EU Medical Device Regulations 4
S GMDN Code (Class IIb) for Male Condom Other Medical Device and Orthopedic Related Topics 4
somashekar Multisite CE certificate - Class IIB medical devices per MDD Annex 2 EU Medical Device Regulations 4
G New intended use - Validation of a class IIb implantable (non-active) ISO 13485:2016 - Medical Device Quality Management Systems 5
shimonv Clinical evaluation report for class I device EU Medical Device Regulations 3
D Control Number for Class III and IV devices Canada Medical Device Regulations 0
M Class II type machine , and its compliance with 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 14
J Should a Class 1 medical device with an option to measure body weight be considered Class 1m? EU Medical Device Regulations 0
P ISO 14644 Class 8 Cleanroom Air Filter Requirements Other Medical Device Related Standards 4
J Instructions for use for Class I devices under MDR EU Medical Device Regulations 1
D Electrical Medical Devices class I EU Medical Device Regulations 0
K MDR Class 1 Self Certification - Competent Authority or Notified Body? EU Medical Device Regulations 6
K "World Class Product" based QM. I need advice. Quality Management System (QMS) Manuals 14
J Biocompatbility testing on Class 1 device requirements Other US Medical Device Regulations 12
J Reusable surgical Instrument is under class IIa? EU Medical Device Regulations 4
H EU CE marking for Medical Device Class I EU Medical Device Regulations 2
E Any sample of a full software life cycle IEC 62304 report ( any class )? IEC 62304 - Medical Device Software Life Cycle Processes 1
JoCam Certified QMS for MDR - Class I medical device manufacturers EU Medical Device Regulations 4
N Is there a need for clinical test of Class IIa products (for MDR)? EU Medical Device Regulations 2
K CE Marking Class 1 (Non sterile) medical device CE Marking (Conformité Européene) / CB Scheme 3
Similar threads
Top Bottom