Class IIB Device - IEC 62304 Software Classification

BargeSim

Starting to get Involved
#1
Hi all,

I have a class IIb medical device (a CBCT scanner for extremities) for only diagnostic purpose; is there a required software classification.

The software manages images acquisition, 3D Reconstruction and visualization on the viewer; no further decision is taken automatically.
I suppose I can assume class B, due to the risk mitigation measures (other than the software).


But my question would be:

is there any guideline defining the difference between serious injury and not serious injury? I only found some docs produced by specific Department of Health, but not an international standard definition.

Thanks
 
Elsmar Forum Sponsor

yodon

Staff member
Super Moderator
#2
In ISO 14973, they provide an example of severity levels for harm:
  • Catastrophic - death
  • Critical - permanent impairment or life-threatening injury
  • Serious - injury or impairment requiring professional medical intervention
  • Minor - temporary injury or impairment not requiring professional medical intervention
  • Negligible - inconvenience or temporary discomfort
To me, anything, well, serious or above would be a good place to draw the line. Of course, this is just an example and you could have different levels but this seems to be a pretty common way to organize.
 

Ronen E

Problem Solver
Staff member
Moderator
#3
Hi all,

I have a class IIb medical device (a CBCT scanner for extremities) for only diagnostic purpose; is there a required software classification.

The software manages images acquisition, 3D Reconstruction and visualization on the viewer; no further decision is taken automatically.
I suppose I can assume class B, due to the risk mitigation measures (other than the software).


But my question would be:

is there any guideline defining the difference between serious injury and not serious injury? I only found some docs produced by specific Department of Health, but not an international standard definition.

Thanks
Can you please quote the classification rule that resulted in the class IIb classification decision? It might help answer your question.
 

Kuldeep Singh

Involved In Discussions
#4
Hi,

I have a question for classification of software, If we control all the risks through hardware control which arises from software ,then is it possible that , we can assume our software as class A rather than Class B?
 

dgrainger

Trusted Information Resource
#6
To me CBCT = cone beam computed tomography, and will be class 2b by Rule 10. "Active devices intended to emit ionizing radiation and intended for diagnostic and therapeutic interventional radiology". Rule 9 applies to therapeutic devices.
As it is emitting ionising radiation there will always be a risk of serious injury. Class C?
 

BargeSim

Starting to get Involved
#7
Thank you all for your replies.

Further information: Classification based on rule 9: delivery of (hazardous) energy
To me CBCT = cone beam computed tomography, and will be class 2b by Rule 10. "Active devices intended to emit ionizing radiation and intended for diagnostic and therapeutic interventional radiology". Rule 9 applies to therapeutic devices.
As it is emitting ionising radiation there will always be a risk of serious injury. Class C?
Yes sorry, surely the Rule is 10.
I honestly had been thinking about class C, but my question is if it can be reduced to Class B, reducing risk, using measures other than the SW.
 

Kuldeep Singh

Involved In Discussions
#8
You can reduce the software classification by reducing the risk by hardware risk control measures. Here below the guidance given in IEC 62304 for sw classification.

1555654636400.png
 
#9
Hi All, take care for a proper architectual design including a solid planning of the hardware measure which shall mitigate your risk.
Be aware, that you have to follow the IEC/EN 60601-1 and your hardware component will be automatically a critical component. Think about, how you can detect single fault failure events and how additional fault conditions can be detected. All safety relevant and critical component shall have a proper test report which includes evidences of related production process controls of incoming inspection, in process controls and end controls.
If you want to distribute the device internationally, a certification mark (like CSA [e.g., cCSAus], UL, NEMKO, DEKRA etc.) would ease the market access and also speed up the registration in e.g. Brazil. Even US customers may benefit: If there is a certification mark, hospitals can save insurance fees as they pay less. For negotiations with the hospital purchase that can be crucial.
 

Tidge

Trusted Information Resource
#10
Hi,

I have a question for classification of software, If we control all the risks through hardware control which arises from software ,then is it possible that , we can assume our software as class A rather than Class B?
I would never recommend trying to change the classification of software (within a medical device) based on an analysis of the effectiveness of hardware risk controls. In my experience, the only clean way to have a 'lower' software safety classification (than otherwise would be naively assumed, based on the classification of the medical device) is to have a (product) system architecture that makes it clear that the software has not been allocated any functions that could contribute to the most serious harms.

In simpler terms: If you want a 'lower' classification, the design ought to be such that the software is neither a contributing factor nor a control of the risks associated with (non) serious injury.
 
Thread starter Similar threads Forum Replies Date
A Clinical assessment sample size - Medical device Class IIb implantable (93/42 directive) EU Medical Device Regulations 2
D Class IIB Medical Device Accessories - OTS Tablets Other Medical Device Regulations World-Wide 1
K Inventory Control - Class II (IIb for CE) medical device ISO 13485:2016 - Medical Device Quality Management Systems 6
K Calibration of Dose Area Product (DAP) Meter in our Class IIb X-Ray Medical Device General Measurement Device and Calibration Topics 1
H Classification of a Class IIb Medical Device with 3 Components CE Marking (Conformité Européene) / CB Scheme 1
A Technical file requirement for Class IIb medical device CE Marking (Conformité Européene) / CB Scheme 3
MDD_QNA Class I Medical Device into a higher class (IIa or IIb) CE Marking (Conformité Européene) / CB Scheme 7
L What are Accredited laboratories? (Medical Device Class IIb Related) EU Medical Device Regulations 1
E When to submit a Technical File for a Class IIb Medical Device EU Medical Device Regulations 3
B Sri Lanka Class IIb Medical Device Registration Other Medical Device Regulations World-Wide 4
L Labeling Requirements in Part A Class IIb Medical Device Technical File CE Marking (Conformité Européene) / CB Scheme 2
M Labeling a Small Class IIb Medical Device EU Medical Device Regulations 8
G Cost for Generating one Class IIb Medical Device GMDN Code Other Medical Device and Orthopedic Related Topics 2
B Class IIb Medical Device using Annex II route - Need to Update ISO 13485 Scope? ISO 13485:2016 - Medical Device Quality Management Systems 2
S Medical Device Technical Files (Class IIa & IIb) ? Design Development Documentation EU Medical Device Regulations 7
S Is using Annex V of 93/42/EEC for CE Certification incorrect for a Class IIb Device? EU Medical Device Regulations 7
I Medical Device Directive, Conversion from class IIb to class III EU Medical Device Regulations 4
V CE Marking for Power Supply for Medical Device - Class IIb device per MDD, Rule 11 EU Medical Device Regulations 3
L Class IIb devices marked EC according to the 93/42/CE - Portugal CE Marking (Conformité Européene) / CB Scheme 3
S We are looking for EU authorised Representative for both Class I and Class IIb devices EU Medical Device Regulations 7
J Custom-made implants (Class IIb) and pre-commercialization clinical trials in Spain EU Medical Device Regulations 4
K Post Market Clinical Follow up and CE Marking for Class IIb medical X-Ray products CE Marking (Conformité Européene) / CB Scheme 1
A Level of details required for Class IIb Product Verification and Validation CE Marking (Conformité Européene) / CB Scheme 1
mikinnear Evidence - Class IIb Disinfectant Sanitiser EU Medical Device Regulations 2
C Sampling of Class IIb Implantables under EU MDR - Assessments to the EU MDR EU Medical Device Regulations 4
R Risk Analysis of Class IIb Disinfectant ISO 14971 - Medical Device Risk Management 6
J Canada Bulk Packaging and IFU requirements for Class IIa and IIb Medical Devices Canada Medical Device Regulations 2
C Bulgarian regulatory requirements for export of class IIb medical devices EU Medical Device Regulations 4
A Endoscopy Software - Class I or Class IIa or Class IIb CE Marking (Conformité Européene) / CB Scheme 4
M MDD 93/42/EEC, Annex IX - Class IIa vs IIb: "Potentially Hazardous" definition EU Medical Device Regulations 2
P Valid NB Technical Review Nonconformance? New Class IIb Product EU Medical Device Regulations 4
S GMDN Code (Class IIb) for Male Condom Other Medical Device and Orthopedic Related Topics 4
somashekar Multisite CE certificate - Class IIB medical devices per MDD Annex 2 EU Medical Device Regulations 4
G New intended use - Validation of a class IIb implantable (non-active) ISO 13485:2016 - Medical Device Quality Management Systems 5
M Class II type machine , and its compliance with 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
J Should a Class 1 medical device with an option to measure body weight be considered Class 1m? EU Medical Device Regulations 0
P ISO 14644 Class 8 Cleanroom Air Filter Requirements Other Medical Device Related Standards 4
J Instructions for use for Class I devices under MDR EU Medical Device Regulations 1
D Electrical Medical Devices class I EU Medical Device Regulations 0
K MDR Class 1 Self Certification - Competent Authority or Notified Body? EU Medical Device Regulations 6
K "World Class Product" based QM. I need advice. Quality Management System (QMS) Manuals 14
J Biocompatbility testing on Class 1 device requirements Other US Medical Device Regulations 12
J Reusable surgical Instrument is under class IIa? EU Medical Device Regulations 4
H EU CE marking for Medical Device Class I EU Medical Device Regulations 2
E Any sample of a full software life cycle IEC 62304 report ( any class )? IEC 62304 - Medical Device Software Life Cycle Processes 1
JoCam Certified QMS for MDR - Class I medical device manufacturers EU Medical Device Regulations 4
N Is there a need for clinical test of Class IIa products (for MDR)? EU Medical Device Regulations 2
K CE Marking Class 1 (Non sterile) medical device CE Marking (Conformité Européene) / CB Scheme 3
G Voluntary Class I Device Filing Other US Medical Device Regulations 4
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3

Similar threads

Top Bottom