Class IIb Medical Device using Annex II route - Need to Update ISO 13485 Scope?

B

brandtpeters

#1
Hello, we wish to register a new Class IIb medical device. We wish to use the Annex II route. We have created a technical file in order to self declare.
This new medical device isn't covered by the scope of our existing iso13485 certificate, therefore do we have to update the 13485 scope before we can attach a CE mark?

Thanks
 
Elsmar Forum Sponsor

chris1price

Trusted Information Resource
#2
I dont think you have to wait until you have updated your ISO13485 certificate, you just have to operate the ISO13485 system for the product.

However, you do have to wait until your Notified Body has updated the scope of your CE Certificate. Only then can you sign your Declaration of Conformity and sell the product.

Chris
 

Peter Selvey

Leader
Super Moderator
#3
As Chris mentioned, the ISO 13485 certificate is not the key point here.

Your Notified Body should have issued you with an "Annex II certificate" or some kind of certificate attesting to compliance with relevant requirements of Annex II. For a Class IIb device, this Annex II certificate needs to be updated before you can legally market the device, as described in Article 11 of the directive.

They may also issue an ISO 13485 certificate but this is voluntary at least with respect to the European MDD. But, the processing of certificates is usually combined so you may as well update both certificates at the same time.
 
Thread starter Similar threads Forum Replies Date
A Clinical assessment sample size - Medical device Class IIb implantable (93/42 directive) EU Medical Device Regulations 2
D Class IIB Medical Device Accessories - OTS Tablets Other Medical Device Regulations World-Wide 1
K Inventory Control - Class II (IIb for CE) medical device ISO 13485:2016 - Medical Device Quality Management Systems 6
K Post Market Clinical Follow up and CE Marking for Class IIb medical X-Ray products CE Marking (Conformité Européene) / CB Scheme 1
K Calibration of Dose Area Product (DAP) Meter in our Class IIb X-Ray Medical Device General Measurement Device and Calibration Topics 1
H Classification of a Class IIb Medical Device with 3 Components CE Marking (Conformité Européene) / CB Scheme 1
A Technical file requirement for Class IIb medical device CE Marking (Conformité Européene) / CB Scheme 3
MDD_QNA Class I Medical Device into a higher class (IIa or IIb) CE Marking (Conformité Européene) / CB Scheme 7
L What are Accredited laboratories? (Medical Device Class IIb Related) EU Medical Device Regulations 1
E When to submit a Technical File for a Class IIb Medical Device EU Medical Device Regulations 3
J Canada Bulk Packaging and IFU requirements for Class IIa and IIb Medical Devices Canada Medical Device Regulations 2
B Sri Lanka Class IIb Medical Device Registration Other Medical Device Regulations World-Wide 4
L Labeling Requirements in Part A Class IIb Medical Device Technical File CE Marking (Conformité Européene) / CB Scheme 2
C Bulgarian regulatory requirements for export of class IIb medical devices EU Medical Device Regulations 4
M Labeling a Small Class IIb Medical Device EU Medical Device Regulations 8
G Cost for Generating one Class IIb Medical Device GMDN Code Other Medical Device and Orthopedic Related Topics 2
S Medical Device Technical Files (Class IIa & IIb) ? Design Development Documentation EU Medical Device Regulations 7
somashekar Multisite CE certificate - Class IIB medical devices per MDD Annex 2 EU Medical Device Regulations 4
I Medical Device Directive, Conversion from class IIb to class III EU Medical Device Regulations 4
V CE Marking for Power Supply for Medical Device - Class IIb device per MDD, Rule 11 EU Medical Device Regulations 3
P Servicing of Class IIb devices under MDD EU Medical Device Regulations 2
D Shelf life of non sterile class IIb product. ISO 13485:2016 - Medical Device Quality Management Systems 1
A Class IIb implantable-TD sampling plan EU Medical Device Regulations 0
L Class IIb devices marked EC according to the 93/42/CE - Portugal CE Marking (Conformité Européene) / CB Scheme 3
S We are looking for EU authorised Representative for both Class I and Class IIb devices EU Medical Device Regulations 7
J Custom-made implants (Class IIb) and pre-commercialization clinical trials in Spain EU Medical Device Regulations 4
B Class IIB Device - IEC 62304 Software Classification IEC 62304 - Medical Device Software Life Cycle Processes 13
A Level of details required for Class IIb Product Verification and Validation CE Marking (Conformité Européene) / CB Scheme 1
mikinnear Evidence - Class IIb Disinfectant Sanitiser EU Medical Device Regulations 2
C Sampling of Class IIb Implantables under EU MDR - Assessments to the EU MDR EU Medical Device Regulations 4
R Risk Analysis of Class IIb Disinfectant ISO 14971 - Medical Device Risk Management 6
A Endoscopy Software - Class I or Class IIa or Class IIb CE Marking (Conformité Européene) / CB Scheme 4
M MDD 93/42/EEC, Annex IX - Class IIa vs IIb: "Potentially Hazardous" definition EU Medical Device Regulations 2
P Valid NB Technical Review Nonconformance? New Class IIb Product EU Medical Device Regulations 4
S GMDN Code (Class IIb) for Male Condom Other Medical Device and Orthopedic Related Topics 4
S Is using Annex V of 93/42/EEC for CE Certification incorrect for a Class IIb Device? EU Medical Device Regulations 7
G New intended use - Validation of a class IIb implantable (non-active) ISO 13485:2016 - Medical Device Quality Management Systems 5
R Class I MDSW CE Mark CE Marking (Conformité Européene) / CB Scheme 6
N Class III PMA shelf life extension US Medical Device Regulations 2
A Class 2b medical device - best pathway to obtain CE mark? (2022, MDR 2017/745) CE Marking (Conformité Européene) / CB Scheme 2
A Class 2a CE Mark Regulatory Pathway CE Marking (Conformité Européene) / CB Scheme 1
normhowe Surprising shoe policies for Class 7 Cleanrooms US Medical Device Regulations 5
L Class I under MDD EU Medical Device Regulations 1
C Legacy device under Class A sterile devices EU Medical Device Regulations 0
N FDA class 1 US Food and Drug Administration (FDA) 6
P Self Certified Class 1 MDR 2017/745 Design Change CE Marking (Conformité Européene) / CB Scheme 2
M UDI IVD Medical Device Class III US Food and Drug Administration (FDA) 1
M Class Exempt Medical Device Recall Question US Food and Drug Administration (FDA) 3
J Clinical Evaluation For Instruments - Class I, Class Ir, Class Im, and Class IIa - EU MDR EU Medical Device Regulations 4
XRAY_3121 Class II heating pad - does the power supply have to have IEC 60601-1 testing?? US Medical Device Regulations 5

Similar threads

Top Bottom