Classification for Nasal Aspirator (non powered)

marmotte

Involved In Discussions
#1
Dear all,

I am advising a friend who's marketing a nasal aspirator. This is a "basic" one, where the user sucks on a small tube hoping not to swallow the patient's nasal material :cool:

I found a post already related to this issue - MDD-wise a nasal aspirator seems to be a medical device.http://www.elsmar.com/Forums/showthread.php?t=36554&highlight=nasal+aspirator

I am interested though at this stage at the US/FDA definition.
I am not sure what regulation actually covers it.
Under the ENT rule - which is the first to come to mind - there is nothing appropriate.It seems like only powered devices are even classified!
https://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfcfr/CFRSearch.cfm?CFRPart=874

880.6740 Vacuum-powered body fluid suction apparatus is my next bet...but
WTF classII? Hospital panel ? Special control --- Sounds to me like the “vacuum” means “vacuum from a powered source”.

Any other idea?
I think there may be a case to justify that it is NOT EVEN a medical device, as it does not intend to treat the disease, but alleviate the discomfort of some of the symptoms (i.e. runny nose)
 
Elsmar Forum Sponsor

Pads38

Trusted Information Resource
#2
I can not immediately answer your question, but I was myself searching for an applicable ProCode and found that a complete listing of ProCodes is available, as a spreadsheet, for download. A new one is produced each week here:-

http://www.fda.gov/MedicalDevices/D...nce/Overview/ClassifyYourDevice/ucm051668.htm

Once you have that you can use the 'Filter' controls of your spreadsheet to reduce and sort to allow you to "zoom in" on any possible codes.
 

SteveK

Trusted Information Resource
#3
If it is one of these nasal sprays that just form a barrier like Prevalin; which contains inactive ingredients such as bentonite, xanthan gum etc (which are not absorbed into the body from the nose) – then it is a medical device. Under the MDD it should be a Class I device under rule 5.

Steve
 

Ronen E

Problem Solver
Staff member
Moderator
#6
Seems like there's no existing ProCode that suits.
Doubtfully it is a Medical Device - what disease is it intended to mitigate (if any)?
If it's a Medical Device you could make a 513(g) submission, under which the FDA will classify it for you (a fee applies).

Cheers,
Ronen.
 
Last edited:

marmotte

Involved In Discussions
#7
Thx Ronen.

I agree that indeed it does not treat any disease . Just the uncomfort created by some of the symptoms.
Theree are plethora of competing devices. On amazon, some even claim to be "FDA approved" :lol:...
But i can't find anyhting on any FDADB, including MAUDE and product registrations....
 

Ronen E

Problem Solver
Staff member
Moderator
#8
Thx Ronen.

I agree that indeed it does not treat any disease . Just the uncomfort created by some of the symptoms.
Theree are plethora of competing devices. On amazon, some even claim to be "FDA approved" :lol:...
But i can't find anyhting on any FDADB, including MAUDE and product registrations....
Just to clarify, "mitigate a disease" doesn't equal "treat a disease". It can include your description - "treat the uncomfort created by symptoms". Since "mitigate a disease" appears in the FDA Medical Device definition, claiming downright treatment is not necessary to render a product a Medical Device.

The important question is whether the symptoms you referred to can be attributed to a disease, or they're just the "normal", healthy manifests of the subject's physiology. What really matters in that context is labeling contents, which includes advertising materials.
 

QA-Man

Involved In Discussions
#10
Try these (separated by semi-colon)

PRODUCT CODE; Device; Reg; Class
BZE; heater, breathing system w/wo controller (not humi ... ; 868.527; 2
OGG; humidifier nebulizer kit ; 868.545; 2
KFZ; humidifier, non-direct patient interface (home-use ... ; 868.546; 1
BTT; humidifier, respiratory gas, (direct patient inter ... ; 868.545; 2
KCO; inhaler, nasal ; 874.522; 1
CAF; nebulizer (direct patient interface) ; 868.563; 2
CCQ; nebulizer, medicinal, non-ventilatory (atomizer) ; 868.564; 1
JPW; pump, nebulizer, electrically powered ; 874.522; 1
EPN; pump, nebulizer, manual ; 874.522; 1
CAD; vaporizer, anesthesia, non-heated ; 868.588; 2
 
Last edited:
Thread starter Similar threads Forum Replies Date
K Biological indicator system classification under MDR EU Medical Device Regulations 0
A FDA Class Classification for a cabinet 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
dinaroxentool Question about FDA Classification of a Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
N What is our product classification? (Does Unclassified classification still exists) Other US Medical Device Regulations 14
nadhar2 Classification of Action Items Misc. Quality Assurance and Business Systems Related Topics 3
A Class medical device (MDD) - Classification help EU Medical Device Regulations 1
D IEC 62304 Risk Classification - With and without hardware control IEC 62304 - Medical Device Software Life Cycle Processes 2
M Classification of Clean Rooms Other Medical Device Related Standards 9
A ISO 10993-7:2008/ Amd.1:2019 - Classification of special populations Other Medical Device Related Standards 3
DuncanGibbons Classification of aerospace parts depending on their risk and criticality etc. Federal Aviation Administration (FAA) Standards and Requirements 3
M Barrier cream classification EU Medical Device Regulations 2
N Medical Device Classification under MDR - Rule 21 EU Medical Device Regulations 11
M Risk Classification For Supplier - Clinical Research Organisation (CRO) Supply Chain Security Management Systems 3
D Classification of Syringe (nozzle) of needle free injection system. EU Medical Device Regulations 8
S CE Mark - Classification Confusion EU Medical Device Regulations 12
S Medical Device Product Classification in South Korea Other Medical Device Regulations World-Wide 1
M Informational US FDA – Ear, Nose, and Throat Devices; Classification of the Self-Fitting Air-Conduction Hearing Aid Medical Device and FDA Regulations and Standards News 0
D MDR 2017/745 Classification Guide EU Medical Device Regulations 8
M Informational EU – Guidance on Qualification and Classification of Software in Regulation (EU) 2017/745 – MDR and Regulation (EU) 2017/746 – IVDR Medical Device and FDA Regulations and Standards News 2
dgrainger Informational MDCG 2019-11: Guidance on Qualification and Classification of Software in Regulation (EU) 2017/745 – MDR and Regulation (EU) 2017/746 – IVDR Medical Device and FDA Regulations and Standards News 0
M Informational EU MDR Classification Rule 11 – what??? Medical Device and FDA Regulations and Standards News 9
M Informational US FDA Final Guidance – Acceptance Review for De Novo Classification Requests Medical Device and FDA Regulations and Standards News 1
A Fire hazard classification - Clause in IEC 60601 about gauze and electricity ISO 14971 - Medical Device Risk Management 0
J MDR Annex VIII, Rule 6 Classification - Implication for lower risk CV products? CE Marking (Conformité Européene) / CB Scheme 3
A Device Classification - India CDCSO - How to classify the device? Other Medical Device Related Standards 1
B ASTM F2924-14 /ASTM F3001 - Room temperature classification criteria Other US Medical Device Regulations 0
M Informational EU – MANUAL ON BORDERLINE AND CLASSIFICATION IN THE COMMUNITY REGULATORY FRAMEWORK FOR MEDICAL DEVICES Version 1.22 (05-2019) Medical Device and FDA Regulations and Standards News 2
E Wristwatch Skin-proximal glucose sensor - Classification under MDR EU Medical Device Regulations 4
K MEDDEV 2.4/1 rev. 9 Classification EU Medical Device Regulations 1
M Informational USFDA – Medical Devices; Anesthesiology Devices; Classification of the Ventilatory Electrical Impedance Tomograph Medical Device and FDA Regulations and Standards News 0
N Medical Device Re-Classification EU Medical Device Regulations 2
B Class IIB Device - IEC 62304 Software Classification IEC 62304 - Medical Device Software Life Cycle Processes 13
C Suppliers re-classification from Critical to Significant or from Significant to Non-critical Supply Chain Security Management Systems 0
J Medical Device re-classification and marketing ISO 13485:2016 - Medical Device Quality Management Systems 13
M APQC PCF (Process Classification Framework) and ISO 9001 - Processes Based Approach ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
M Informational TGA – Several proposed changes to classification to better align with the EU MDR Medical Device and FDA Regulations and Standards News 0
N Medical Device Accessory Classification - Software as a Medical Device Other Medical Device Regulations World-Wide 5
G EU MDR 2017/745 Rule 11 interpretation - Re-classification of a Software as Medical Device Other Medical Device Related Standards 0
M Medical Device News Manual On Borderline And Classification In The Community Regulatory Framework For Medical Devices Medical Device and FDA Regulations and Standards News 0
M FDA News USFDA Final Rule – Medical Device Classification Procedures: Incorporating Food and Drug Administration Safety and Innovation Act Procedures Medical Device and FDA Regulations and Standards News 0
M FDA News The FDA Issues Final Rule on Medical Device Classification Procedures Medical Device and FDA Regulations and Standards News 0
K Biocompatibility classification for Lancet CE Marking (Conformité Européene) / CB Scheme 7
sagai Classification for compiler's runtime libraries IEC 62304 - Medical Device Software Life Cycle Processes 3
K Classification of IVF/ICSI devices (micromanipulator) in Mexico Other Medical Device Regulations World-Wide 0
M Medical Device News Medical Device Borderline and Classification Manual 1.20 - October 2018 EU Medical Device Regulations 0
H Classification of a device that uses an existing battery IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
H Classification of a Class IIb Medical Device with 3 Components CE Marking (Conformité Européene) / CB Scheme 1
L Classification of Medical vs Wellness Devices Canada Medical Device Regulations 2
D Changing software classification via software - IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 3