I have an ECG recording device (low cost) that does not comply with the particular standard 601-2-25. Bandwidth is lower, sampling rate is lower etc. How shall I classify it? it? Can I only apply the 601-1-1 with no particulars? It is for home use.
MIREGMGR
Re: Partial compliance with particular standard
One conforms to a performance standard as a means of achieving/demonstrating conformance to a medical device regulatory system's requirements for your device type. In USA, you might be conforming to the performance standard so as to meet the expectations to file for and obtain a 510(k) premarket clearance. In Europe, you might be demonstrating conformance to the Essential Requirements.
Particularly under the MDD, there are three basic conformance paths: conformance to a Harmonized Standard, conformance to some other recognized standard, and a non-standard-based approach. The former approach requires the least evidence and explanation to the Notified Body or other reviewer, the latter approach the most evidence and explanation.
If you're only partly complying to a standard, you're following the third path.
FDA utilizes a similar approach, though not as clearly defined. To obtain a 510(k), you have to convince the examiner that your device is safe and effective. If your standards compliance is partial, you'll have more explaining to do.
Peter Selvey
Re: Classification of Medical Device with Partial Compliance with a particular standa
In this case it would be advisable to comply with all of the standard except for specific points (e.g. frequency response). Then you only have the job of explaining the rationale behind those points only. Also, although you may not meet the full bandwidth, you still need to meet a bandwidth as justified by your application.
If you provide:
- a complete report, showing the non-conformity(s)
- the alternate specifications (e.g. revised bandwidth)
- verification report for the alternate specifications
- justification for the alternate specifications (expected is a clinical discussion why a reduced bandwidth is reasonable, for example, limit to adult population, target for particular disease etc)
Then it is complete and complies with the regulations. An auditor would have to provide objective evidence the justification is not valid, which is unlikely if the justification is properly documented.
If you don't apply the whole standard, it's unlikely to be accepted. There are many other ECG tests besides bandwidth like CMRR, noise, sensitivity accuracy etc etc which are likely to be reasonable to apply. Again, there may be justification for revised specifications for a home use device, but most of these tests are still reasonable.
Roland chung
Hello Peter,
Based on this topic, I would like to ask a question related to particulars of 3rd edition.
IEC 60601-2-25:2011 now includes the contents of the IEC 60601-2-51:2003. For ECG function, as part of patient monitors or defibrillators, which particulars should be applied? IEC 60601-2-25:2011 or IEC 60601-2-27:2011?
Peter Selvey
The scope of IEC 60601-2-25:2011 now clearly states that "cardiac monitors covered by IEC 60601-2-27 [are excluded] where not intended for obtaining ECG REPORTS for diagnostic purposes".
Previously, IEC 60601-2-51 did apply to a patient monitor if there was an arrhythmia interpretation function, which many patient monitors had (e.g. to trigger alarms).
However in the new edition (IEC 60601-2-25, which replaces IEC 60601-2-51), the interpretation section is removed. I don't know why, maybe they plan a separate standard. So without that there is not much justification to apply to a patient monitor which is intended primarily to detect the heart rate.
There are a couple of catches: first is many monitors include a "diagnostic" filter with 0.05Hz ~ 150Hz. If this is included, this could be a reasonable justification for applying IEC 60601-2-25.
Also, some patient monitors include a function to monitor the ST segment. Again, this could trigger the application of IEC 60601-2-25.
The good news is, if there are no automated measurements made and displayed (e.g. QRS amplitudes, durations) then it is really similar to IEC 60601-2-27, just with tighter limits (e.g. 5% for sensitivity, instead of 20%). And that's quite reasonable as well - a limit of 5% makes much more sense than 20% if diagnostic decisions are being made based on the ECG record, and 5% is easily achievable with modern technology.
Roland chung
Thank you, Peter.
Do you sum up significant changes between 1st edition and 2rd edition of IEC 60601-2-25?
shimonv Classification of a cloud- base viewer for the output from a medical device US Food and Drug Administration (FDA) 7
A Class medical device (MDD) - Classification help EU Medical Device Regulations 1
N Medical Device Classification under MDR - Rule 21 EU Medical Device Regulations 11
S Medical Device Product Classification in South Korea Other Medical Device Regulations World-Wide 1
N Medical Device Re-Classification EU Medical Device Regulations 2
J Medical Device re-classification and marketing ISO 13485:2016 - Medical Device Quality Management Systems 13
N Medical Device Accessory Classification - Software as a Medical Device Other Medical Device Regulations World-Wide 5
G EU MDR 2017/745 Rule 11 interpretation - Re-classification of a Software as Medical Device Other Medical Device Related Standards 0
M FDA News USFDA Final Rule – Medical Device Classification Procedures: Incorporating Food and Drug Administration Safety and Innovation Act Procedures Medical Device and FDA Regulations and Standards News 0
M FDA News The FDA Issues Final Rule on Medical Device Classification Procedures Medical Device and FDA Regulations and Standards News 0
M Medical Device News Medical Device Borderline and Classification Manual 1.20 - October 2018 EU Medical Device Regulations 0
H Classification of a Class IIb Medical Device with 3 Components CE Marking (Conformité Européene) / CB Scheme 1
R Vertebroplasty - Medical Device Classification EU Medical Device Regulations 0
C Substance based Medical Device Classification under MDR - Rule 21 EU Medical Device Regulations 9
rob73 EU Medical Device Classification Spreadsheet EU Medical Device Regulations 2
S Medical Device Classification - Laser Guided Positioner EU Medical Device Regulations 2
W Determining Medical Device Classification in Mexico Other Medical Device Regulations World-Wide 5
shimonv EU classification for a Medical Device Accessory EU Medical Device Regulations 16
J0anne MDD 93/42/EEC Medical Device Classification EU Medical Device Regulations 3
B CE Classification of a Medical Device CE Marking (Conformité Européene) / CB Scheme 17
K Classification of the Reseller of Stand Alone Medical Device Software ISO 13485:2016 - Medical Device Quality Management Systems 3
T CE Marking Medical Device and classification CE Marking (Conformité Européene) / CB Scheme 6
E Japan - Medical Device Classification for Mass Spectrometry Japan Medical Device Regulations 2
A Classification of Medical Device Accessories 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
T What is Medical Device Classification for diagnostic equipment in ASEAN ? Other Medical Device Regulations World-Wide 2
T CFDA seeks for public comments on newly revised medical device classification rules China Medical Device Regulations 1
S Please explain Medical Device Product Classification Rationale 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
Z Kyrgyzstan Medical Device Classification and Registration Requirements ISO 13485:2016 - Medical Device Quality Management Systems 3
B KOREA MFDS (KFDA) Medical Device Classification Rules Other Medical Device Regulations World-Wide 3
N Software Medical Device Classification Medical Information Technology, Medical Software and Health Informatics 8
P Classification of Medical Device with Sub Code Devices 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
Ajit Basrur Guidance Document - Medical Device Classification Product Codes Other US Medical Device Regulations 1
A Japan - Medical Device Classification for Autoclave Tape Japan Medical Device Regulations 6
D Keeping Same Medical Device Name but Changing Classification Medical Information Technology, Medical Software and Health Informatics 3
T X-Ray Template Medical Device Classification (US) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
R FDA Rules on Medical Device Classification Listings? Other US Medical Device Regulations 3
Z Medical Device Classification of a Needle Holder EU Medical Device Regulations 9
Y FDA Classification for Medical Device Power Adaptors (Power Supplies) Other US Medical Device Regulations 12
D 510(k) Medical Device Classification and Regulatory Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
D Need some brainstorming on the Classification of a Medical Device CE Marking (Conformité Européene) / CB Scheme 8
F Classification of Medical Device Fastener in Canada (EU classification IIa) Canada Medical Device Regulations 5
R Medical Device FDA - Classification of a device used as an interface with a computer 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
M Medical Device Registration and Classification in Peru Registrars and Notified Bodies 8
L Medical Device FDA Product Classification help needed 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
M Medical Device Classification Question - Liquid Nitrogen Freezers EU Medical Device Regulations 10
T Safety Classification in Medical Device Standalone Software IEC 62304 - Medical Device Software Life Cycle Processes 26
S Medical Device Embedded Software Classification IEC 62304 - Medical Device Software Life Cycle Processes 4
W Classification of a new product - Is it a Medical Device Supplier Quality Assurance and other Supplier Issues 7
S Medical Device Battery Powered Equipment with Battery Charger Classification IEC 60601 - Medical Electrical Equipment Safety Standards Series 17
E Classification (Medical Device or Drug) of Blood Bag in Japan Japan Medical Device Regulations 6